Programmability of Remstar Plus

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
avgeo
Posts: 6
Joined: Sun Mar 19, 2006 10:26 pm
Location: orange county, Ca.

Programmability of Remstar Plus

Post by avgeo » Fri Apr 14, 2006 5:01 pm

My Remstar Plus was programmed by the local vendor. The person providing instructions said the only way to change settings was by doctor's orders. I received no literature with the unit. Is this thing programmable?
Avgeo

Guest

Post by Guest » Fri Apr 14, 2006 5:42 pm

The flow rate and the ramping (the triangle symbol, which controls the beginning rate of the flow) are 'programmable'. Those are the only features on that machine, in addition to turning on/off (if available, an integrated heated humidifier). The settings are set by the equipment supplier according to directions from your doctor. You can request the booklet for your machine from your supplier - after all, it contains the directions for cleaning the equipment, etc.


User avatar
Wulfman
Posts: 12316
Joined: Thu Jul 07, 2005 3:43 pm
Location: Nearest fishing spot

Re: Programmability of Remstar Plus

Post by Wulfman » Fri Apr 14, 2006 10:22 pm

avgeo wrote:My Remstar Plus was programmed by the local vendor. The person providing instructions said the only way to change settings was by doctor's orders. I received no literature with the unit. Is this thing programmable?
avgeo,

I don't know if this is what you're looking for.....???
Sometime back, someone else (a "guest") posted these instructions for doing setup programming on a REMstar Plus. I made a copy of them and here they are.

Best wishes,

Den


REMstar Plus CPAP.

1. With the machine plugged into a wall outlet, unplug the cord from the back of the CPAP. Then, press and hold down the two user buttons below the display screen while plugging the cord back into the CPAP, and continue to hold the buttons down until the CPAP beeps. This will indicate that the CPAP is in the Therapy Setup Menu.

2. The display will first show the number of hours the machine has been used. To erase this total and begin again at zero, press and hold the ramp or humidifier button. An “X” will appear on the display. Hold the button down until the time change to “0” and the “X” disappears.

To skip this setting or continue, press the right user button.

3. The elevation setting will appear next. To change this setting, press the ramp or humidifier button until the correct setting appears. Here are the elevation settings: 1 = less than 2,500 ft.; 2 = 2,500 to 5,000 ft.; 3 = 5,001 to 7,500 ft.

To skip this setting or continue, press the right user button.

4. The therapy mode will appear, either CPAP or CFLEX. To change this setting, press the ramp or humidifier button until the correct setting appears.

To skip this setting or continue, press the right user button.

5. The CPAP pressure will appear. To change this setting, press the ramp or humidifier button until the correct setting appears.

To skip this setting or continue, press the right user button.

6. The CPAP fine adjustment setting will appear. This setting allows you to calibrate your CPAP pressure to a manometer. We suggest you do not try to adjust this setting.

To skip this setting or continue, press the right user button.

7. The CFLEX setting will appear if you selected the CFLEX mode in number 4 above. A setting of 1 offers the least drop in pressure, 3 offers the most. To change this setting, press the ramp or humidifier button until the correct setting appears.

To skip this setting or continue, press the right user button.

8. The ramp time will appear next. It may be set for 0 to 45 minutes in 5 minute increments. To change this setting, press the ramp or humidifier button until the correct setting appears.

To skip this setting or continue, press the right user button.

9. The ramp starting pressure will appear. To change this setting, press the ramp or humidifier button until the correct setting appears.

To skip this setting or continue, press the right user button.

10. The patient disconnect setting will appear. This sounds an alarm when there is a mask leak and turns the airflow off. 1 = On; 2 = Off. To change this setting, press the ramp or humidifier button until the correct setting appears.


There are no other settings available. Press the pressure start/stop button to exit the menu or press the right user button to scroll through the settings again.

(5) REMstar Autos w/C-Flex & (6) REMstar Pro 2 CPAPs w/C-Flex - Pressure Setting = 14 cm.
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05

sleepless in colorado

Re: Programmability of Remstar Plus

Post by sleepless in colorado » Thu Nov 11, 2010 11:05 pm

Wanted to thank Den for posting those instructions for the REMstar plus C-Flex! Had been searching all over the internet for those. Was just about to give up, (and use the more 'plentiful' service manual for the 'M'-series...Not quite the same!), when I came upon your post.

Thanks so much! And it really works too! -Marc

Janknitz
Posts: 8518
Joined: Sat Mar 20, 2010 1:05 pm
Location: Northern California

Re: Programmability of Remstar Plus

Post by Janknitz » Fri Nov 12, 2010 12:04 am

You might want to consider taking the machine back because it has no efficacy data.

You can't determine what the proper settings are to "program" with no data to back it up.
What you need to know before you meet your DME http://tinyurl.com/2arffqx
Taming the Mirage Quattro http://tinyurl.com/2ft3lh8
Swift FX Fitting Guide http://tinyurl.com/22ur9ts
Don't Pay that Upcharge! http://tinyurl.com/2ck48rm

User avatar
diboja
Posts: 100
Joined: Sat Oct 17, 2009 10:44 am
Location: Winnipeg CANADA

Re: Programmability of Remstar Plus

Post by diboja » Fri Nov 12, 2010 11:59 am

sleepless in colorado wrote:Wanted to thank Den for posting those instructions for the REMstar plus C-Flex! Had been searching all over the internet for those. Was just about to give up, (and use the more 'plentiful' service manual for the 'M'-series...Not quite the same!), when I came upon your post.

Thanks so much! And it really works too! -Marc

You can print off the manual here: http://www.apneaboard.com/CPAP%20Adjustment.htm

FYI