You have a Respironics machine and are seeing dual leak lines.birdygirl38 wrote:So, Is the large leak line on the top of the sleepyhead leak graph and the bottom line the one I should be comparing to it?
Top line is the total leak line (mask vent rate plus excess leak) and the bottom line is excess leak only.
We don't have a set value to keep an eye on when looking at excess leak only line but we do have a value for the top line which is total leak....keep that line under 80 to 90 L/min and your leak won't be a problem.
If you get into the 100 L/min range the machine can't work properly....it won't/can't sense events properly and it won't respond properly.
Mask vent rates will vary with type of mask used and the pressures used...more pressure the greater the vent rate.
If you are using APAP varying pressures your leak line will vary and that is entirely normal.
Full face masks have a higher vent rate than nasal pillow masks even at the same pressure.
If you look in your mask paperwork there should be a graph that shows the intentional leak/vent rate for your mask ....just look for your pressure on the graph. If using varying pressures it can't be exact but you can use the average to get you close. It doesn't have to perfect. There's always a little plus or minus variation anyway.