Air Pressure question with Apap

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
User avatar
Daisy004
Posts: 88
Joined: Tue Sep 11, 2012 7:41 pm

Air Pressure question with Apap

Post by Daisy004 » Fri Nov 23, 2012 10:39 am

Ok so the lab originally prescibed my air pressure number as a 7. I got the Apap machine and it was set for 6-9. Lately I noticed it was maxing out at 9 so my hubby moved it up to number 10. Last night it maxed out at 10! My leak was good at 7ml, AHI 1.4 as it usually is, can it be possible that sleep lab got it wrong and I actually need this higher pressure? Can I trust the Apap machine? Would welcome your comments, thanks.
"I'm not sleeping, I'm just checking my eyelids for pinholes!" - My Dad

User avatar
greatunclebill
Posts: 1503
Joined: Mon Feb 20, 2012 7:48 pm
Location: L.A. (lower alabama)

Re: Air Pressure question with Apap

Post by greatunclebill » Fri Nov 23, 2012 10:49 am

what program are you using? what does maxed out mean? how long? how many times? a screen shot from your program would help.

_________________
Mask: Quattro™ FX Full Face CPAP Mask with Headgear
Additional Comments: myAir, OSCAR. cms-50D+. airsense 10 auto & (2009) remstar plus m series backups
First diagnosed 1990
please don't ask me to try nasal. i'm a full face person.
the avatar is Rocco, my Lhasa Apso. Number one "Bama fan. 18 championships and counting.
Life member VFW Post 4328 Alabama
MSgt USAF (E-7) medic Retired 1968-1990

User avatar
Daisy004
Posts: 88
Joined: Tue Sep 11, 2012 7:41 pm

Re: Air Pressure question with Apap

Post by Daisy004 » Fri Nov 23, 2012 10:56 am

Im sorry I have no clue how to do that screenshot thingy or how to read sleepyhead, just not that computer savvy.
"I'm not sleeping, I'm just checking my eyelids for pinholes!" - My Dad

User avatar
Pugsy
Posts: 65359
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Air Pressure question with Apap

Post by Pugsy » Fri Nov 23, 2012 11:19 am

Screen shots of your reports are how we see what you are seeing.
We discussed how to get them and how to post them here in this thread.
viewtopic.php?f=1&t=81072&p=737779#p737779

APAP machines respond to snores and flow limitations and it is possible that the "maxing" out is just a normal response.
It is common to have varying pressure needs...sleeping position maybe, like supine sleeping or maybe REM related events causing an increase in pressure needs. We don't sleep the same each night so a set pressure doesn't always eliminate every possible snore or flow limitation. If you max out at 10 it might be just a brief kissing of that 10 cm line or it might be a prolonged time. It might mean that you would need a little more pressure during that particular time asleep.

We really can't offer anything more than general comments without seeing what you are seeing on the reports.
Is it possible that your sleep lab got it wrong? Yeah, sure is but it is also possible that normal things are causing the need for more pressure. My RX was for 8 cm but in REM sleep I often need a LOT more pressure but during the titration part of my sleep study I only got 6 minutes of REM. So it comes as no surprised that they didn't get it spot on. They were close though and likely for a good bit of the night I probably could use 8 cm and do well...I just need more pressure sometimes.
The sleep lab had one night in a foreign environment to try to come up with a pressure. It's not surprising that you need a slightly different pressure at home.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
Daisy004
Posts: 88
Joined: Tue Sep 11, 2012 7:41 pm

Re: Air Pressure question with Apap

Post by Daisy004 » Sat Nov 24, 2012 9:57 am

Thank you guys for the replys...I guess I am just scared Im getting too much pressure, silly I know, but it is an autoset so I will trust it. It just seems the higher I set the end pressure, the machine reaches it. Kinda freaked me out a bit. If I do manage to get into sleepyhead, how do I take a screen caption? Thanks
"I'm not sleeping, I'm just checking my eyelids for pinholes!" - My Dad

User avatar
Sheriff Buford
Posts: 4111
Joined: Mon Aug 09, 2010 8:01 am
Location: Kingwood, Texas

Re: Air Pressure question with Apap

Post by Sheriff Buford » Sat Nov 24, 2012 10:02 am

I would "bump" it up to 11. The autoset won't go up to a pressure unless it needs it. Sheriff

User avatar
RandyJ
Posts: 1673
Joined: Thu Apr 21, 2011 7:22 pm
Location: Connecticut, USA

Re: Air Pressure question with Apap

Post by RandyJ » Sat Nov 24, 2012 10:14 am

An auto won't go to the higher pressures if it doesn't need to. As was said above, looking at your reports would help a lot, to see if the forays into higher pressure are brief or if you stay there for most of the night.

Another option is to set the machine for a range like 7-18 or 7-20 (or wherever you feel comfortable setting a maximum) and sleep with those settings for a week. After 7 days, Sleepyhead or ResScan will report a 95% pressure number and a median pressure number that will be helpful in determining your true pressure needs.

You could find that higher pressures bother you, or perhaps not. But the machine won't go where it doesn't need to.

_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: Backup & Travel Machines: PR System One Bipap Auto, S9 VPAP Auto, S9 Autoset, Oximeter CMS-50E
Diagnosed March 2011, using APAP 14 - 16.5 cm, AFlex+ 2
Alt masks Swift FX pillows, Mirage FX nasal mask, Mirage Quattro full face mask

User avatar
Daisy004
Posts: 88
Joined: Tue Sep 11, 2012 7:41 pm

Re: Air Pressure question with Apap

Post by Daisy004 » Sat Nov 24, 2012 10:15 am

Thank You Sherriff I will set it for 11. Im in Houston, I attempted to drive through Kingwood yesterday omg..crazy traffic!! I think they were all headed to the mall for the sales. We were headed to Lake Livingston taking the boys fishing. Anyhoo, I wil up the pressure to 11 tonight and see what happens. thanks.
"I'm not sleeping, I'm just checking my eyelids for pinholes!" - My Dad

User avatar
avi123
Posts: 4509
Joined: Tue Dec 21, 2010 5:39 pm
Location: NC

Re: Air Pressure question with Apap

Post by avi123 » Sat Nov 24, 2012 10:44 am

Lack of sufficient OP data.

_________________
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments:  S9 Autoset machine; Ruby chinstrap under the mask straps; ResScan 5.6
Last edited by avi123 on Sat Nov 24, 2012 11:49 am, edited 6 times in total.
see my recent set-up and Statistics:
http://i.imgur.com/TewT8G9.png
see my recent ResScan treatment results:
http://i.imgur.com/3oia0EY.png
http://i.imgur.com/QEjvlVY.png

User avatar
Pugsy
Posts: 65359
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Air Pressure question with Apap

Post by Pugsy » Sat Nov 24, 2012 10:51 am

Daisy004 wrote: If I do manage to get into sleepyhead, how do I take a screen caption?
In my previous post I posted a link to forum thread where we discussed how to obtain screen shots and how to post them on the forum.

If your computer is Windows Vista or Win 7 (unless the basic version) you have the Windows Snipping Tool and using it is super easy. Just look at my above post for that thread. Links for how to use Snipping Tool are in that thread or you can simply Google "Windows Snipping Tool".

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
Burkebang
Posts: 297
Joined: Fri Dec 09, 2011 4:30 pm
Location: Oslo, Norway

Re: Air Pressure question with Apap

Post by Burkebang » Sat Nov 24, 2012 11:14 am

Sheriff Buford wrote:I would "bump" it up to 11. The autoset won't go up to a pressure unless it needs it. Sheriff
I would not trust that to be the case:
viewtopic/t84496/S9-goes-from-515-in-au ... 0-min.html

_________________
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear
Additional Comments: Also use Mirage FX nasal mask a lot. Contec CMS-50D+ Pulseoximeter and Zeo Mobile tracks the quality of my therapy.

User avatar
Pugsy
Posts: 65359
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Air Pressure question with Apap

Post by Pugsy » Sat Nov 24, 2012 11:24 am

It's true the APAP won't increase the pressure unless it thinks it needs to but sometimes it thinks it needs to for something like a snore or flow limitation and the increase in pressure causes more of a problem than the snore or flow limitation would. Problem might be something like aerophagia at higher pressures or simply the changing of the pressure itself being disruptive to sleep. Some people are highly sensitive to pressure changes and some people can sleep through a hurricane.
Sometimes compromises are needed depending on the situation.

The thread that Burkebang mentions the machine is increasing the pressure while the user is awake and that shouldn't happen unless there is some sort of unusual situation or a possible problem with the machine.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
ughwhatname
Posts: 604
Joined: Sun Oct 14, 2012 8:23 am

Re: Air Pressure question with Apap

Post by ughwhatname » Sat Nov 24, 2012 11:31 am

I'd say don't be scared of more pressure. You may find your tiredness improves with better treatment.

Daisy004 wrote:Thank you guys for the replys...I guess I am just scared Im getting too much pressure, silly I know, but it is an autoset so I will trust it. It just seems the higher I set the end pressure, the machine reaches it. Kinda freaked me out a bit. If I do manage to get into sleepyhead, how do I take a screen caption? Thanks

_________________
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Mask is the TAP PAP Nasal Pillow CPAP Mask with Stability Mouthpiece

User avatar
squid13
Posts: 2301
Joined: Fri Jul 15, 2011 3:47 pm
Location: Pensacola, FL

Re: Air Pressure question with Apap

Post by squid13 » Sat Nov 24, 2012 12:48 pm

Some people just put it at 20 and let the machine go to where it needs to go. It will only go to the pressure it needs.

_________________
Machine: ResMed AirCurve 10 ASV Machine with Heated Humidifier
Mask: Evora Full Face Mask - Fitpack
Additional Comments: AirCurve 10 ASV, Oscar V1.0.1-r-1
US Navy Retired 1973,AirCurve 10 ASV, Mode: ASV Auto, Min EPAP: 7.2, Max EPAP: 15.0, Min PS:4.0, Max PS: 15.0, Mask ResMed Airtouch F20, Backup: (2) AirCurve 10 ASV

User avatar
archangle
Posts: 9293
Joined: Sun Mar 27, 2011 11:55 am

Re: Air Pressure question with Apap

Post by archangle » Sat Nov 24, 2012 5:32 pm

The main drawbacks to too much pressure are leaks, discomfort or increased apneas due to pressure induced central apnea.

As long as you're increasing pressure slowly and watching your AHI numbers, there's not a lot of risk of something bad happening. If you do start having central apneas, your AHI number will go up.

Many people are overly afraid of central apnea. It's probably no more harmful than regular obstructive apnea, but it's harder to treat. If you start having too many centrals as you increase the pressure, just turn the pressure back down.

_________________
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Also SleepyHead, PRS1 Auto, Respironics Auto M series, Legacy Auto, and Legacy Plus
Please enter your equipment in your profile so we can help you.
Click here for information on the most common alternative to CPAP.
If it's midnight and a DME tells you it's dark outside, go and check for yourself.

Useful Links.