Need To Adjust Pressure on Older Bipap Pro 2 HELP!!

Post by MrStein » Sat Oct 20, 2012 1:15 pm

Hello All....Newb back for more questions/help.....Last night I finally got my 2nd night of sleep study in!! Other than the fact it took me 1.5hrs to actually fall asleep, I did really well. This morning when I was getting unwired, the tech told me I ended up on a pressure of around 11-12cm . I told him that I have an older Resperonics BiPaP Pro 2 with Biflex that my father gave me before I started sleep study. Of course he said it's not recommended to use it without official RX, but I have already tried it and have no problems. The only thing is....I cant adjust the pressure of course!!...I can't even get in to read what the pressures are. HELP! Does anybody know where I can find a clinicians/setup manual for this machine???....I have googled until my googler wore out...lol....I have found manuals for every machine except this one!!......The tech said it would maybe be 2 weeks or so until I get my machine from my DME. I have the same mask I used last night (from my hospital stay 2 months ago), brand new hose/filters...ect.
I am pretty sure that the current pressure setting is a bit too much but not sure..I am a very tech savvy guy and understand the EPAP/IPAP and flex/rise functions, I just can't get into the menu's to adjust them. I fear sleeping without cpap assistance while I wait on my DME, Can anybody here point me to the setup manual for this machine?
Actually I just need the button combination to bring up the setup menu......help please!!

Here's the make/model/serial of the machine..... Resperonics BiPaP Pro 2 with Biflex model# 1012986 serial# 3688801

and a pic......Image

_________________
Machine: PR System One REMStar 60 Series Auto CPAP Machine
Mask: Amara View Full Face CPAP Mask with Headgear
Additional Comments: 12cm-16cm Auto, Encore Basic 2.0 and SleepyHead
"The future of yesterday is tomorrow's today"

User avatar
greatunclebill
Posts: 1503
Joined: Mon Feb 20, 2012 7:48 pm
Location: L.A. (lower alabama)

Re: Need To Adjust Pressure on Older Bipap Pro 2 HELP!!

Post by greatunclebill » Sat Oct 20, 2012 1:27 pm

did you look down a few posts at this thread? it should help.
viewtopic/t83294/Help-with-quotnewquot- ... r-Pro.html

_________________
Mask: Quattro™ FX Full Face CPAP Mask with Headgear
Additional Comments: myAir, OSCAR. cms-50D+. airsense 10 auto & (2009) remstar plus m series backups
First diagnosed 1990
please don't ask me to try nasal. i'm a full face person.
the avatar is Rocco, my Lhasa Apso. Number one "Bama fan. 18 championships and counting.
Life member VFW Post 4328 Alabama
MSgt USAF (E-7) medic Retired 1968-1990

User avatar
Pugsy
Posts: 65298
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Need To Adjust Pressure on Older Bipap Pro 2 HELP!!

Post by Pugsy » Sat Oct 20, 2012 1:30 pm

This is the only one I have and it is for the Pro which predates the Pro 2.
http://www.apneaboard.com/manuals51/BIPAP-PRO.pdf

If that doesn't help check the link above that greatunclebill posted.
The combination for the cpap Pro2 and the BiPap Pro2 should work.


REMstar Pro 2 Setup

When in the Setup Menu, the humidifier ^ and ramp v buttons operate as up and down keys to change the settings, the left/right user buttons < > allow you to go to the previous/next question or setting, and the pressure start/stop button is used to exit the Setup Menu. Holding the humidifier or ramp buttons down will cause the values to change more quickly.

To enter the Therapy Setup Menu, hold the two top user buttons < > down while plugging in the power cord.
Continue holding the buttons down until the REMstar Pro 2 with C-Flex beeps twice.

Note: The word "setup" will appear on all of the screens indicating that you are in the Therapy Setup Menu.
If you press the Pressure start/stop button, you will exit the Setup Menu.

a. Compliance hours/nights: (recommend leaving alone, but CAN be cleared at this point)
Select next setting with >

b. Therapy mode: (CPAP or CFLE) select with ^ or v
Select next setting with >

c. CPAP or CFLE pressure setting: Select with ^ or v
Select next setting with >

d. C-Flex mode (if you chose CFLE mode): Select setting 1, 2 or 3 with ^
Select next setting with >

e. Ramp time setting: Select with ^ or v
(ramp time will be turned off with a setting of 0:00)
Select next setting with >

f. Ramp pressure setting: Select with ^ or v
Select next setting with >

g. Patient disconnect setting: 1 = on 0 = off Select with ^ or v
Select next setting with >

h. Buttons lights setting: 1 = on 0 = off Select with ^ or v

(Last setting. Use On/Off button to exit Setup)

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
MrStein
Posts: 154
Joined: Mon Oct 15, 2012 2:38 pm
Location: Near Nashville Tn

Re: Need To Adjust Pressure on Older Bipap Pro 2 HELP!!

Post by MrStein » Sat Oct 20, 2012 2:13 pm

Thank you SO much for the info! I will attempt this today, and report back accordingly. This site has been a Godsend for my wife and I. What a great online community! I have found so many answers to questions without even having to post. Last night the tech at the sleep center informed me officially that I am dx'd with severe OSA. Had I not been spending so much time here reading and learning, I probably wouldn't have slept at all. Thanks to whom ever runs this site, and to all who share here!!

_________________
Machine: PR System One REMStar 60 Series Auto CPAP Machine
Mask: Amara View Full Face CPAP Mask with Headgear
Additional Comments: 12cm-16cm Auto, Encore Basic 2.0 and SleepyHead
"The future of yesterday is tomorrow's today"

geofftodds

Re: Need To Adjust Pressure on Older Bipap Pro 2 HELP!!

Post by geofftodds » Wed Nov 06, 2013 9:03 pm

Worked like a charm. Mine didn't say "set up" it said "flex" I believe on the screen. The instructions worked perfectly though, thank you very much.