Can't Handle the Pressure!!

Post by TiredMominCanada » Fri Oct 12, 2012 4:36 pm

Hello all!
Had my titration study last night and my visit to the DME today to get my machine. Two things:

1 My new prescription is 15cm (have been on a trial of 9 for the last 3 weeks)

2 The loaner machine that they gave me (S9 Autoset), I cannot have - long storty short - script is for CPAP only, to get an auto I would have to pay completely out of pocket. So I'm stuck with straight CPAP which I'll have to deal with ( I got the S9 Elite).

So the sleep tech said ("I had the pressure lower and it stopped your events, but couldn't stop your snoring until I put it at 15") Frankly I don't give a crap if I'm snoring as long as I'm breathing!

I tried my machine with 15cm pressure and I couldn't take it! It was like my lungs were filling up and it couldn't get rid of the air (even with the EPR on). Soooo, I figured I'm going to set the machine at a lower pressure (like 11 or 12) and see what my AHI is. I figure if it is under 5, I feel rested and I can handle the pressure it's good enough.

Thoughts???

_________________
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Basic Resmed Nasal Mask - so far...

123.Shawn T.W.
Posts: 748
Joined: Sun Jun 24, 2012 12:39 pm
Location: Cochise County AZ

Re: Can't Handle the Pressure!!

Post by 123.Shawn T.W. » Fri Oct 12, 2012 4:53 pm

I'd go with 10, and set your ramp to start at 7 and run for 5-10 min, give it a few day then increase...

Do you have software to monitor you results?
"I am a man of peace, but if war comes to my door it will find me home." - Winston Churchill

User avatar
Xney
Posts: 842
Joined: Fri Jul 20, 2012 6:53 pm

Re: Can't Handle the Pressure!!

Post by Xney » Fri Oct 12, 2012 4:54 pm

Can you turn on C-Flex / EPR? Try setting it to 2 or 3.

however, I should note that 15 is kind of the range where people start considering Bi-Level instead of CPAP. I've been okay up to about 15cm with cpap, but then it started to get too much.

The reason the techs want to eliminate snoring is that it's a precursor (often) to obstructive events. However, you may want to take the tact that you'll watch out for events happening rather than trying to eliminate them, rather than suffering at that pressure level. I think that's reasonable if you have the ability to look at your data.

_________________
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Actually a S9 VPAP Adapt, and Respironics M Series Auto BiPAP

User avatar
greatunclebill
Posts: 1503
Joined: Mon Feb 20, 2012 7:48 pm
Location: L.A. (lower alabama)

Re: Can't Handle the Pressure!!

Post by greatunclebill » Fri Oct 12, 2012 5:07 pm

keep returning to the doc. tell him over abnd over you are trying but you absolutely can't breathe against the pressure. he'll have no choice but to prescribe an automatic machine. even if you have to pay out of pocket, they are not that expensive, new, on line and even cheaper lightly used. whatever you do, do not stop treatment.

_________________
Mask: Quattro™ FX Full Face CPAP Mask with Headgear
Additional Comments: myAir, OSCAR. cms-50D+. airsense 10 auto & (2009) remstar plus m series backups
First diagnosed 1990
please don't ask me to try nasal. i'm a full face person.
the avatar is Rocco, my Lhasa Apso. Number one "Bama fan. 18 championships and counting.
Life member VFW Post 4328 Alabama
MSgt USAF (E-7) medic Retired 1968-1990

User avatar
Todzo
Posts: 2014
Joined: Tue Apr 24, 2012 8:51 pm
Location: Washington State U.S.A.

Re: Can't Handle the Pressure!!

Post by Todzo » Fri Oct 12, 2012 5:16 pm

TiredMominCanada wrote:Hello all!
Had my titration study last night and my visit to the DME today to get my machine. Two things:

1 My new prescription is 15cm (have been on a trial of 9 for the last 3 weeks)

2 The loaner machine that they gave me (S9 Autoset), I cannot have - long storty short - script is for CPAP only, to get an auto I would have to pay completely out of pocket. So I'm stuck with straight CPAP which I'll have to deal with ( I got the S9 Elite).

So the sleep tech said ("I had the pressure lower and it stopped your events, but couldn't stop your snoring until I put it at 15") Frankly I don't give a crap if I'm snoring as long as I'm breathing!

I tried my machine with 15cm pressure and I couldn't take it! It was like my lungs were filling up and it couldn't get rid of the air (even with the EPR on). Soooo, I figured I'm going to set the machine at a lower pressure (like 11 or 12) and see what my AHI is. I figure if it is under 5, I feel rested and I can handle the pressure it's good enough.

Thoughts???
Hi TiredMominCanada!

In my experiance it helps to spend quality time with the machine and/or new setting during the day. The change may be little in terms of absolute pressure but is hugh in terms of the changes that need to take place in our breathing reflexes to handle it.

Part of the time should be in bed learning to breath quietly with the machine. For me it helps a lot to put a pause between the breath out - - breath in. Think two slow easly counts in, two counts out, two counts pause.

With the practice in bed then with a book, music, or light TV. Some kind of distraction.

I think I would set aside a half hour each for a couple of days and the half for the distraction time (or full hour for a full TV program) to complete a week.

When you start to use the CPAP at night, use the practice to breath quietly.

Rigging to use CPAP data is strongly advised.

Have a great week, and may you find good sleep!

Todzo
May any shills trolls sockpuppets or astroturfers at cpaptalk.com be like chaff before the wind!

User avatar
LSAT
Posts: 13405
Joined: Sun Nov 16, 2008 10:11 am
Location: SE Wisconsin

Re: Can't Handle the Pressure!!

Post by LSAT » Fri Oct 12, 2012 5:36 pm

If your insurance will pay for the Elite...they should pay for the Autoset. They have the same insurance code. It may be the DME that is refusing to give you the more expensive unit. See if you can get your doctor to specify on the prescription that you need an auto titrating unit. There is nothing wrong with the Elite ...it is a good machine that gives you data.

User avatar
Pugsy
Posts: 65337
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Can't Handle the Pressure!!

Post by Pugsy » Fri Oct 12, 2012 5:45 pm

A jump from 9 to 15 is huge for some people even with EPR maxed at 3. It will take some time and gradual adjustment but it can be done.
In the meantime start with the 11 or 12 if you can tolerate that well. Then give it a week or so then try a little more (like 0.4) and give that a little while. All the time watching your software reports. Keep inching upwards with the pressure but go at it in small increments and you can get a lot closer to 15 that way. Plus it will be much easier on you getting there.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
Slartybartfast
Posts: 1633
Joined: Wed Sep 01, 2010 12:34 pm

Re: Can't Handle the Pressure!!

Post by Slartybartfast » Fri Oct 12, 2012 6:12 pm

LSAT wrote:If your insurance will pay for the Elite...they should pay for the Autoset. They have the same insurance code. It may be the DME that is refusing to give you the more expensive unit. See if you can get your doctor to specify on the prescription that you need an auto titrating unit. There is nothing wrong with the Elite ...it is a good machine that gives you data.
She's in Canadia, not the U.S.

User avatar
Xney
Posts: 842
Joined: Fri Jul 20, 2012 6:53 pm

Re: Can't Handle the Pressure!!

Post by Xney » Fri Oct 12, 2012 6:31 pm

It's a big jump.

When I started, 8cm seemed crazy. I got CPAP all the way up to 15cm by going up bit by bit. You actually can get used to higher pressures, though everybody reaches their own personal point of too much.

I would definitely consider self-titrating up (in consultation with your doctor) slowly or an auto machine!

It's funny but now like 13cm or 14cm is no big deal to me.

_________________
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Actually a S9 VPAP Adapt, and Respironics M Series Auto BiPAP

User avatar
archangle
Posts: 9293
Joined: Sun Mar 27, 2011 11:55 am

Re: Can't Handle the Pressure!!

Post by archangle » Sat Oct 13, 2012 4:18 am

If it were me, and I'd been using 9, I'd set it at 9 or something comfortable, then creep it up a little each night.

Most of us find that we can exhale against any pressure after a little adjustment.

The medical mafia may treat you like some sort of heretic if they find out you changed the pressure on your own. They may sic the CPope on you.

The AutoSet is better, but if you are willing to look at the data and tinker wisely, you can probably get pretty good therapy with the Elite.

_________________
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Also SleepyHead, PRS1 Auto, Respironics Auto M series, Legacy Auto, and Legacy Plus
Please enter your equipment in your profile so we can help you.
Click here for information on the most common alternative to CPAP.
If it's midnight and a DME tells you it's dark outside, go and check for yourself.

Useful Links.

User avatar
TiredMominCanada
Posts: 34
Joined: Fri Sep 14, 2012 5:50 pm
Location: Northern Ontario, Canada

Re: Can't Handle the Pressure!!

Post by TiredMominCanada » Sat Oct 13, 2012 6:18 am

I set the pressure to 11.6cm last night (EPR at 3), Ramp from 9cm for 20 minutes, seemed okay (although I had a dream that I was filling up with air!!). My AHI was 4.4 this morning so a big improvement. I will keep it at this pressure for a few days and keep an eye on the number and inch up slowly if I need to. Thanks so much for all of the advice and support!!

_________________
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Basic Resmed Nasal Mask - so far...

TedVPAP
Posts: 975
Joined: Sat Jun 12, 2010 10:29 am

Re: Can't Handle the Pressure!!

Post by TedVPAP » Sat Oct 13, 2012 8:59 am

I do not use the ramp. I like to make sure I have a good seal and comfortable/relaxed breathing before I fall asleep.

_________________
Machine: DreamStation Auto CPAP Machine
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Humidifier: DreamStation Heated Humidifier
Additional Comments: AutoPAP 16-20, Ultimate Chin Strap http://sleepapneasolutionsinc.com/

User avatar
Denial Dave
Posts: 1334
Joined: Wed Sep 19, 2012 6:45 am
Location: Connecticut

Re: Can't Handle the Pressure!!

Post by Denial Dave » Sat Oct 13, 2012 12:18 pm

I feel your pain

during my sleep study I thought 4 was unbearable....

I was prescribed with bi-pap of 21 /15.......!!


Good Luck

_________________
Mask: Mirage Quattro™ Full Face CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Started at VPAP pressure setting of 20/14.4. I've survived Stage 1 cancer and lost 80+ lbs. Pressure is now 14.5 / 11
if you don't know where you are going... any road will take you there.... George Harrison

User avatar
KEQ5
Posts: 93
Joined: Sat Jul 21, 2012 7:38 pm

Re: Can't Handle the Pressure!!

Post by KEQ5 » Sat Oct 13, 2012 4:14 pm

Ramp is tool. Sometimes it's useful, sometimes not. When I started CPAP in August, I used the ramp with a 5 minute setting. Then I ran without it for a while, now that I've upped my pressures again, I'm using it at 5 minutes again for a bit.

However, I think that when using the ramp feature, you should set the "start" pressure for something not much lower then your prescribed pressure. So if your pressure is set to 12, starting the ramp at 4 is too much. You might not leak at 4 but start leaking at 8 or 9 on the way to 12. In which case, a starting pressure of 8 or 9, then a ramp up to 12 over 5 or 10 minutes would make more sense as you'd be more likely to have a good mask fit.

User avatar
archangle
Posts: 9293
Joined: Sun Mar 27, 2011 11:55 am

Re: Can't Handle the Pressure!!

Post by archangle » Sun Oct 14, 2012 1:11 am

Getting rid of the ramp, gradually if necessary may be a good next step. Either do without or try increasing the ramp pressure until you don't need it.

_________________
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control