Flow Limitations and Pressure, Runaway S9?

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
User avatar
BigLou
Posts: 63
Joined: Thu Jun 28, 2012 9:05 pm

Flow Limitations and Pressure, Runaway S9?

Post by BigLou » Mon Oct 08, 2012 4:32 pm

Image

Why do I have increased flow limitations with the increase in pressures? Is this normal? Can it cause sleep fragmentations or am I barking up the wrong tree!?
Please also notice the rattier looking flow pattern during these times. My EPR is set at 2. Am afraid to increase the top range due to aerophgia and what not. No events usually slip thru but am just concerned about the sleep quality. I feel OK in the a.m. but not GREAT even with such low AHI's. Was titrated @ 10 straight CPAP 4 months ago. Decided myself to experiment into APAP. AHI's have been lower since.
Last edited by BigLou on Mon Oct 15, 2012 5:36 pm, edited 2 times in total.
:) :) :)

User avatar
Pugsy
Posts: 65148
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Flow Limitations and Pressure

Post by Pugsy » Mon Oct 08, 2012 4:38 pm

The increase in pressure is in response to the flow limitations. The machine is trying to do its job and having a problem it looks like.
Seems to want to use a little more pressure and you have it maxed at 12 cm. For some reason that is not enough.
Can't see if any events happened during that time. Did you have any obstructive events during the time frame with the flow limitations and higher pressure?
My first suspects for the apparent need for more pressure would be either supine sleeping or REM stage sleep or both.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
BigLou
Posts: 63
Joined: Thu Jun 28, 2012 9:05 pm

Re: Flow Limitations and Pressure

Post by BigLou » Mon Oct 08, 2012 4:46 pm

Thanks Pugsy, only had 1 event this night (hypopnea @ 0200). No events are slipping by at the higher pressure (usually) but the S9 keeps wanting to go higher and I'm just not going to let it unless letting it would decrease the flow limitations and have a positive result on my sleep quality. From my limited knowledge, it seems that my leaks would increase (for sure) and FL would increase also? Still trying to figure out the correlations in the data.
:) :) :)

User avatar
Pugsy
Posts: 65148
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Flow Limitations and Pressure

Post by Pugsy » Mon Oct 08, 2012 4:54 pm

The leaks might increase with more pressure...guess it depends on how much more pressure is needed. Doesn't appear to need a lot though. You are barely skirting that max.
If no events to speak of (and with only 1 hyponea that is no big deal), no wake ups and you are feeling decent...it may not be a big deal to have those flow limitations show up. The increase in pressure might create more of an issue than just letting them happen would create.
If your sleep study pointed to your OSA being worse when sleeping on your back or REM stage sleep...then you likely have the culprit identified.

If that was my report and I had a truckload of events along with the flow limitations I would let the machine fight a little harder with more pressure but if you aren't having any problems then I wouldn't worry about them.
Now if I was still extremely fatigued or having wake ups....stuff like that then I probably would try to limit those flow limitations somewhat to see if it would help me feel better.
It might be more about the minimum pressure than the maximum though.
A little more minimum might do a better job of warding off the limitations...prevention vs reduction after they have materialized.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
BigLou
Posts: 63
Joined: Thu Jun 28, 2012 9:05 pm

Re: Flow Limitations and Pressure

Post by BigLou » Mon Oct 08, 2012 5:21 pm

Thanks again Pugsy......my sleep study did reveal a strong positional component (AHI 23) to my Dx of OSA, supine of course leading to most of my apneas.
I do try to start sleeping on my side (Swift LT helps with this) but I'm not sure where I am during the night or where I end up at the end of the sleep session. Would something like a ZEO clear this up( as per REM phases ) or are they just too unreliable? Or should I just leave the pressure ranges where they are for a week (just raised the maximum from 11.4 to 11.8 last night) and raise the minimum to about 10.2 from 9.8? I have been using a 5 minute ramp lately, I feel it eases my anxiety after masking up.
:) :) :)

User avatar
Pugsy
Posts: 65148
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Flow Limitations and Pressure

Post by Pugsy » Mon Oct 08, 2012 6:46 pm

BigLou wrote: Would something like a ZEO clear this up( as per REM phases ) or are they just too unreliable?
It is probably close enough to give you an idea but then what do you do with the information? I have mixed feelings about the Zeo and not due to its accuracy or lack of or whatever. My mixed feelings are what do we do about what we might see...like it your situation. What would having that information actually change or maybe make you do something different? Are you going to change REM stages?
I happen to be documented much worse in REM sleep....like 5 times more events in REM sleep than in non REM sleep. To me it doesn't really matter when or why something occurs other than maybe explaining it...it still needs to be dealt with. That's why I have an auto adjusting pressure machine. I expect it to do its job and I don't worry about sleep stages or sleep positions. I have enough to deal with without adding some other restriction. I do try to help the machine do its job with more efficiency though and that is where the minimum and maximums come in.

In your situation we see the pressures go up and kiss the max line and no real reduction in that chunk of flow limitations as best I can see from the report even though they stay up there for quite a while. This is where I like the PR S1 Auto machine better...it actually flags the Flow limitations and counts them so it gives an easier to see image of what is going on. So I am not so sure that increasing the max will do much...it doesn't seem to want to go a whole lot higher and the flow limitations are still pretty significant. I would think that we treat those flow limitations sort of like we would a cluster of full grown obstructive events of some sort and that would be with more minimum pressure...Prevention being the goal and not fixing after the fact.

If it were me and I wanted to see if I can reduce that chunk of flow limitations I would probably just try a little more minimum and I think your 0.4 increase would be a good choice. Sometimes it takes only tiny changes to bring about positive results when we are at this level of our therapy where things are really pretty good and we just want to tweak things a little more. I know that when I used the Respironics APAP at 10 cm minimum I did really well with it and didn't have any REM clusters to speak of and my Flow limitations were pretty much nonexistent but when I dropped the minimum to 9.5 cm I saw some clustering (not huge but some small) of both the obstructive events and the Flow limitations. I was mainly interested in addressing the obstructive events and in my case the flow limitation reduction was a by product of the increase to 10 min. Respironics machines only allow adjustments in the 0.5 range..nothing smaller.

So if it were me I wouldn't worry so much about why the flow limitations show up at the time they do...if REM related we don't/can't limit REM. Is is worth the cost of the Zeo? I can't answer that for you but ask yourself what you would do with the information when you got it assuming it was accurate? Sometimes that curiosity itch needs to be scratched and if you are that type of person and it doesn't take food off the table to buy it then go for it. I did get a chance to try a Zeo a while back. One of the forum members was kind enough to offer hers for a trial. I used it 3 nights and my curiosity itch was totally scratched.
Pretty much confirmed what I already suspected. Some people like the reassurance that the Zeo data offers. That is just fine.
Whatever it takes for someone to feel like they have they have all the tools they need to satisfy whatever needs they feel they have.

You could also choose to experiment with really trying to stay on your side if you are so inclined. I always start the night on my side and I actually prefer my side because sleeping on my back causes me to have increased back pain. So I would rather not sleep on my back but the effort to try to force me to stay on my side causes more issues than just a minor wake up to realize I am on my back and my back hurts (which is probably what woke me up) and I turn back on to my side and go back to sleep.
Trying to prevent supine sleeping actually disrupts my sleep more than just that little wake up and turn on my side and go back to sleep. The cure causes more problems than the disease...if you know what I mean. If trying to stay on your side happens to cause more problems than it fixes...I don't see why people want to do it. If you could rig up some way to know for sure that you don't ever get supine then that might at least answer the question about the Flow limitations being related to sleep position if they totally went away and the pressure line stayed nice and flat.

Me...I am a lazy sot and the reason I have the machine I have is so it can do the job it was designed to do. I do my job by setting it close enough to where it might need or want to go so it can get there in a timely fashion and do its job the best it can.
Also sometimes when the minimum is set maybe just a little bit "better" head start then the maximum sometimes actually never gets touched because prevention is easier than fixing sometimes.
A while back someone posted this about their new APAP. I think you will find it interesting to read. Different machine and different software but the principle is the same no matter what machine is being used.
viewtopic.php?f=1&t=39869&st=0&sk=t&sd=a

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
avi123
Posts: 4509
Joined: Tue Dec 21, 2010 5:39 pm
Location: NC

Re: Flow Limitations and Pressure

Post by avi123 » Mon Oct 08, 2012 7:07 pm

I would raise the max pressure by 2 cm and see if the pressure graph still "kisses" the top limit line. It should not. The leak graph that you have is typical to using nasal pillows. Those pillows vibrate in the nostrils and causing the jagged graph. The inverted U in the graph at 3:30, 5 and 7 indicate mouth leaks. But all the leaks are below the red line which is good.

Here are a few graphs which were posted by users of nasal pillows masks, in this forum, attesting that these masks are causing jagged (not smooth) leak graphs. This is besides the affect of nasal pillows masks on causing Flow Limitation (which includes UAR and RERA) to rise:



Both of the following by Papit on Mon May 07, 2012 6:15 pm

Machine: S9 AutoSet™ CPAP Machine Mask: Aloha Nasal Pillow CPAP Mask with Headgear Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control Additional Comments: Software:ResScan 4.1, SH 0.9.2. CMS-50F Oximeter. Neck Roll 2.5" :


Image


Image

By NateS:

Mask: Aloha Nasal Pillow CPAP Mask with Headgear

Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control

Additional Comments: S9 ResMed VPAP Adapt; Swift Bella Loops; Aloha Mask; CPAPMax Pillow; Ruby Chin Strap - Size Large; ResScan & SleepyHead:


Image



by BigLou on Wed Oct 10, 2012 8:30 pm:

Machine: S9 AutoSet™ CPAP Machine Mask: Swift™ LT Nasal Pillow CPAP Mask with Headgear Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control Additional Comments: Also using SleepyHead SW and ResScan:


Image

_________________
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments:  S9 Autoset machine; Ruby chinstrap under the mask straps; ResScan 5.6
Last edited by avi123 on Mon Oct 15, 2012 12:49 pm, edited 2 times in total.
see my recent set-up and Statistics:
http://i.imgur.com/TewT8G9.png
see my recent ResScan treatment results:
http://i.imgur.com/3oia0EY.png
http://i.imgur.com/QEjvlVY.png

User avatar
Pugsy
Posts: 65148
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Flow Limitations and Pressure

Post by Pugsy » Mon Oct 08, 2012 7:16 pm

avi123 wrote:I would raise the max pressure by 2 cm and see if the pressure graph still "kisses" the top limit line. It should not. The leak graph that you have is typical to using nasal pillows. Those pillows vibrate in the nostrils and causing the jagged graph.
Get off the nasal pillows causing vibration and jagged graphs. You have no idea what you are talking about again.
When was the last time you used a nasal pillow mask?
They do NOT vibrate in the nostrils. Please quit dissing the nasal pillow masks. Despite your personal feeling about them they do a darn good job for a lot of people.
How come you only tell the men this garbage and not the women? Misogynist?

2 cm increase in minimum pressure is not warranted and even if it was that is way too much of an increase way too fast.
There is zero sense in making such a huge jump in the situation. Totally unneeded since the obstructive events aren't an issue.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
BigLou
Posts: 63
Joined: Thu Jun 28, 2012 9:05 pm

Re: Flow Limitations and Pressure

Post by BigLou » Mon Oct 08, 2012 7:54 pm

Thank you both for your inputs. I have archived in ResScan and have found no real basis AVI for this vibratory effect from my nasal pillows. I have used a nasal mask, hybrid, FFM and have still seen these spiky leaks with all throughout the night. Just find the Swift LT mask to be the best all around for me.

Pugsy, maybe raising the minimum pressure to 10.2 and leaving the max alone for now may somewhat decrease the magnitude of the FL's ? I don't know but it seems like whenever that pressure rises so do the FL's so not exactly sure how raising the minimum could be preventative except for maybe a more appropriate lower threshold pressure level. (I did read your explanation). I think this will be my next experimental move though. Should I keep my EPR at 2 or raise it to 3....or would this be counterproductive by letting the pressure go down to 7.2 on exhalation?

AVI, raising the maximum even higher makes less sense to me because of the side effects and the fact that I'm not really having any events when the pressure kisses the max line, usually for <20% of the night. I do not want to use any more pressure than I absolutely need to. That's why I keep my machine APAP range so tight.
:) :) :)

User avatar
Todzo
Posts: 2014
Joined: Tue Apr 24, 2012 8:51 pm
Location: Washington State U.S.A.

Re: Flow Limitations and Pressure

Post by Todzo » Mon Oct 08, 2012 8:28 pm

Hi BigLou!

I have some thoughts.
BigLou wrote:Why do I have increased flow limitations with the increase in pressures?
You seem to use a lot of air (your flow waveform is wide). From my experience moving a lot of air irritates the airway passage (dries, cools, less CO2 frustrates circulation and metabolism). So now, if you raise pressure you will tend to use more air, further irritate the airway passage which will cause more flow limitations which will raise the pressure (+loop . . .).
BigLou wrote:Can it cause sleep fragmentations or am I barking up the wrong tree!?
In your case, apparently so.

As I look at the entire night you have a few bumps to start with a fairly major change at around 2:20a.m. The little narrowing of the waveform makes me think that the event eventually may have produced an arousal or awakening. Things quite down, indeed are fairly quiet but by 5:10 or so your breathing appears to take off and become unstable. I think the increased flow limitations are consistent with unstable breathing in this context. And, indeed, it would appear they resulted in a full awakening (probably a bathroom break for you). The time to the next batch of unstable breathing is reduced in your second set on the machine.

Our breathing control appears to take the inputs from our O2 and CO2 sensors and then adds in a factor from our level of central nervous system activation which then drives our muscles to breath. It is intended, so they think, to be a negative feedback loop. Obstructive Sleep Apnea (OSA) upsets this system in ways we do not yet understand. What we do know is that the process is upset. If the gain of the system goes above one, you have unstable breathing. It can go the range of simply breathing levels considerably above what is necessary for good metabolism, to periodic breathing where the breathing level oscillates as the system tries to find stability, to full fledged hypocapnic central apnea.

It appears that your high breathing levels result in arousals – which increase central nervous system activation, which results in heaver breathing - . . . - until the breathing becomes fully destabilized and fully awakens you.
BigLou wrote:Am afraid to increase the top range due to aerophgia and what not.
After reading up on unstable breathing I am more afraid of what it does to me than I am of OSA!! I would not use APAP if it were me. You might even consider trying less pressure on CPAP.
BigLou wrote:No events usually slip thru but am just concerned about the sleep quality. I feel OK in the a.m. but not GREAT even with such low AHI's. Was titrated @ 10 straight CPAP 4 months ago. Decided myself to experiment into APAP. AHI's have been lower since.
My pressure, proscribed about nine years ago, has been bumped up twice. Recently weight loss and other changes brought me to the point where I could see unstable breathing creeping into my data. My AHI remained quite reasonable (under two) but I know what unstable breathing does to us (starves our cells, moves the brain into starve Peter to keep Paul going mode, it is not pretty). Unable to connect with medical help I decided to self titrate.

I tried Auto-PAP set to the extremes of the proscribed numbers (13 to 15). Snoring kicked it to 15 and I know that snoring is one of the signs of unstable breathing. When my nose stuffs up anymore I take it as a sign to breath less – which consistently clears it up. No tissue required. Auto-PAP would not work for me.

I then started at 13 cm/H2O and looked at my data for at least a week each time – deciding to lower another 1 cm/H2O – until – some two months later I raised it from 7 back to 8 cm/H2O were it will stay until I see good reason to move it.

As I reduced the pressure the average AHI (over a week of data) stayed under two. Since I use techniques to help repair some of the damage to my breathing system which cause me to breath less and slower sometimes the events are false and simply the result of breathing less often. So that is even better than it looks.

I decided to go with 8 because at 7 I did see very real looking apnea flow waveforms and because we are going into a time and season where I expect to see more inflammation (and possible weight gain) in my body so the higher pressure – well – I have considerable “head room” before unstable breathing I know from the process of finding where OSA occurs, pressure wise.

If I were you I would:

1. Reduce Stress!
2. Go back to using CPAP.
3. Raise the head of my bed a couple of inches. Concrete bricks from a building suppy store do this nicely weighing in at less than five dollars for the pair. This should lessen your need for pressure and my help with the tendency for aerophgia as well.
4. If the above yields AHI less than 2 you might consider going lower in small steps as I mention above. Note that I spent 16 days trying out 13 and average a bit over nine days a try for each pressure as I moved pressures. Move slow looking at the data with care.
5. I do regular aerobic and interval training exercise in an effort to help my breathing reflexes work better. I really do believe it helps. I tried for about three months to connect with good medical care all that time at pressure which was much higher than it should have been. The exercise helped me keep my wits and saved my health!

Have a great week!

Todzo
May any shills trolls sockpuppets or astroturfers at cpaptalk.com be like chaff before the wind!

User avatar
Pugsy
Posts: 65148
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Flow Limitations and Pressure

Post by Pugsy » Mon Oct 08, 2012 8:40 pm

BigLou wrote:Pugsy, maybe raising the minimum pressure to 10.2 and leaving the max alone for now may somewhat decrease the magnitude of the FL's ? I don't know but it seems like whenever that pressure rises so do the FL's so not exactly sure how raising the minimum could be preventative except for maybe a more appropriate lower threshold pressure level. (I did read your explanation). I think this will be my next experimental move though. Should I keep my EPR at 2 or raise it to 3....or would this be counterproductive by letting the pressure go down to 7.2 on exhalation?
I don't know if raising the minimum will reduce the flow limitations or not but it wouldn't hurt to try as long as you don't have any problem with the increase (like aerophagia). It was an idea or a theory that might make an impact on things and may not.
I wouldn't go changing EPR to 3 though because that is going to reduce your pressure even more. At least not while evaluating the minimum pressure changes. Remember Science experiments 101...limit your variables as best you can so that you can evaluate the results and have a greater chance to attribute any changes to just the one variable.

Whenever pressures rise...those pressures are responding to the flow limitation's presence. The pressure increases are NOT the cause of the flow limitations though. In this case the flow limitations materialize and then the pressure increases in an effort to prevent the limitation. Flow limitations and snores are the obstructive sleep apnea event precursors or warning signs. The machine increases the pressure in an attempt to stop the warning signs. In your case the increase seems to deal with the actual full grown obstructive events very well but doesn't seem to be able to deal with the flow limitation part.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
avi123
Posts: 4509
Joined: Tue Dec 21, 2010 5:39 pm
Location: NC

Re: Flow Limitations and Pressure

Post by avi123 » Mon Oct 08, 2012 9:06 pm

Hi Flow Limitations are wose than obstructive apneas so said the sleep expert Dr Barry Krakow (Kraiko).
BigLou wrote:Thank you both for your inputs. I have archived in ResScan and have found no real basis AVI for this vibratory effect from my nasal pillows. I have used a nasal mask, hybrid, FFM and have still seen these spiky leaks with all throughout the night. Just find the Swift LT mask to be the best all around for me.

Pugsy, maybe raising the minimum pressure to 10.2 and leaving the max alone for now may somewhat decrease the magnitude of the FL's ? I don't know but it seems like whenever that pressure rises so do the FL's so not exactly sure how raising the minimum could be preventative except for maybe a more appropriate lower threshold pressure level. (I did read your explanation). I think this will be my next experimental move though. Should I keep my EPR at 2 or raise it to 3....or would this be counterproductive by letting the pressure go down to 7.2 on exhalation?

AVI, raising the maximum even higher makes less sense to me because of the side effects and the fact that I'm not really having any events when the pressure kisses the max line, usually for <20% of the night. I do not want to use any more pressure than I absolutely need to. That's why I keep my machine APAP range so tight.
Comment,

Hi Flow Limitations are worse than obstructive apneas, so said the sleep expert Dr Barry Krakow (Kraiko)( and notice: not Pugsy). This b/c you get suffocated by it. You need to find out what causing it. Could be deviated septum, excessive UARs or RERAs, or something else at the upper airway. The machine needs to raise the pressure b/c of the F/L. If you continue to have it the situation may get worse, the APAP will continue to raise pressure and you may need to switch to a BiLevel.
As to your pressure on the S9 APAP even if you raise your max pressure to 12 cm then the 95 percentile would be around 10 cm.

_________________
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments:  S9 Autoset machine; Ruby chinstrap under the mask straps; ResScan 5.6
Last edited by avi123 on Thu Oct 11, 2012 3:19 pm, edited 1 time in total.
see my recent set-up and Statistics:
http://i.imgur.com/TewT8G9.png
see my recent ResScan treatment results:
http://i.imgur.com/3oia0EY.png
http://i.imgur.com/QEjvlVY.png

User avatar
Lizistired
Posts: 2835
Joined: Tue Dec 14, 2010 10:47 pm
Location: Indiana

Re: Flow Limitations and Pressure

Post by Lizistired » Mon Oct 08, 2012 9:46 pm

Biglou, Please stick to the screenshots. AVI is the only one who posts the report format.

_________________
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Swift FX sometimes, CMS-50F, Cervical collar sometimes, White noise, Zeo... I'm not well, but I'm better.

User avatar
BigLou
Posts: 63
Joined: Thu Jun 28, 2012 9:05 pm

Re: Flow Limitations and Pressure

Post by BigLou » Tue Oct 09, 2012 5:50 pm

Image

Last nights report. Did not sleep for long. Pressure did not raise as much as night before but still w/ flow limitations (although less).
:) :) :)

User avatar
Lizistired
Posts: 2835
Joined: Tue Dec 14, 2010 10:47 pm
Location: Indiana

Re: Flow Limitations and Pressure

Post by Lizistired » Tue Oct 09, 2012 6:15 pm

BigLou, That "rattier" looking flow line is what mine looks like during REM. I also have increased flow limitations during REM. ResScan has those 2 apneas flagged as CA's and they are short, so I wouldn't be too concerned about it.

_________________
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Swift FX sometimes, CMS-50F, Cervical collar sometimes, White noise, Zeo... I'm not well, but I'm better.