ResScan and S9 VPAP Adapt

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
MAXDRI
Posts: 21
Joined: Sun Sep 09, 2012 5:10 pm

Re: ResScan and S9 VPAP Adapt

Post by MAXDRI » Wed Oct 03, 2012 3:43 pm

Pugsy,
THANK YOU, THANK YOU. I followed your directions, and the detailed graphs appeared. I have printed the manual & now I will have to study it. This forum is fantastic! I really appreciate all the advice & help. Question: Which of the graphs should I first study?
MAXDRI

_________________
MaskHumidifier

User avatar
archangle
Posts: 9293
Joined: Sun Mar 27, 2011 11:55 am

Re: ResScan and S9 VPAP Adapt

Post by archangle » Wed Oct 03, 2012 4:41 pm

avi123 wrote:So far no one has shown, on this discussion board, ResScan graphs from Autoset CS, Autoset CS2, VPAP Adapt, and VPAP AdaptSV, of CAs in the Events graph. Not even DreamDiver himself who has either CompSAS (called also CPAP-emergent Central Sleep Apnea), or CSAS. So MAXDRI is still left unanswered.
Well, someone with an S9 VPAP Adapt, put it on, try and fake a few centrals, and publish the results.

To fake a central, you have to stop breathing, but not close off your airway. It's a little difficult to do, but I've done it on my S9 and PRS1 Autos.

If you get centrals recorded, it would show the machine records them. Unfortunately, if you don't get any centrals, it wouldn't prove it never records centrals, just that you couldn't fool it.

_________________
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Also SleepyHead, PRS1 Auto, Respironics Auto M series, Legacy Auto, and Legacy Plus
Please enter your equipment in your profile so we can help you.
Click here for information on the most common alternative to CPAP.
If it's midnight and a DME tells you it's dark outside, go and check for yourself.

Useful Links.

MAXDRI
Posts: 21
Joined: Sun Sep 09, 2012 5:10 pm

Re: ResScan and S9 VPAP Adapt

Post by MAXDRI » Wed Oct 03, 2012 5:07 pm

Thank you Pugsy, with your directions, I was able to downlaod detailed graphs.
Henry Jr. said "The ResMed rep said when S9VPAP Adapt is operating in ASV mode, it will never identify a CA because the ASV algorithm should always adjust/increase pressure to prevent a Clear Airway even from ever happening.." Maybe that is why, when I findally was able to download the detailed graphs, the area for CA was empty?

_________________
MaskHumidifier

User avatar
avi123
Posts: 4509
Joined: Tue Dec 21, 2010 5:39 pm
Location: NC

Re: ResScan and S9 VPAP Adapt

Post by avi123 » Wed Oct 03, 2012 7:00 pm

Nathe who has compSAS provided these graphs by S9 VPAP Adapt and ResScan:


Image

Notice the poor control of leaks that Nathe had, most likely b/c of the Aloha Nasal Pillow CPAP Mask.

Image

Notice: no CAs are shown.

See it here:


viewtopic.php?f=1&t=73197&p=674948&hili ... ph#p674948

_________________
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments:  S9 Autoset machine; Ruby chinstrap under the mask straps; ResScan 5.6
Last edited by avi123 on Thu Oct 04, 2012 10:35 am, edited 2 times in total.

User avatar
Pugsy
Posts: 65388
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: ResScan and S9 VPAP Adapt

Post by Pugsy » Wed Oct 03, 2012 7:05 pm

MAXDRI wrote:Henry Jr. said "The ResMed rep said when S9VPAP Adapt is operating in ASV mode, it will never identify a CA because the ASV algorithm should always adjust/increase pressure to prevent a Clear Airway even from ever happening.." Maybe that is why, when I findally was able to download the detailed graphs, the area for CA was empty?
That is my understanding also. Centrals aren't flagged at all.
There was a discussion a while back about this little bit of detail when one of the forum members was asking the same thing and in his case his doctor specifically wanted centrals to be flagged and after verifying with ResMed that centrals weren't flagged then they switched RX over to the Respironics ASV machine which does flag centrals.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
Xney
Posts: 842
Joined: Fri Jul 20, 2012 6:53 pm

Re: ResScan and S9 VPAP Adapt

Post by Xney » Thu Oct 04, 2012 12:18 am

Hmm just checked my ASV; never have seen a central. Somebody want to try and hold their breath?

_________________
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Actually a S9 VPAP Adapt, and Respironics M Series Auto BiPAP

User avatar
archangle
Posts: 9293
Joined: Sun Mar 27, 2011 11:55 am

Re: ResScan and S9 VPAP Adapt

Post by archangle » Fri Oct 05, 2012 7:42 pm

Someone send me a free machine and I'll try it.

_________________
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Also SleepyHead, PRS1 Auto, Respironics Auto M series, Legacy Auto, and Legacy Plus
Please enter your equipment in your profile so we can help you.
Click here for information on the most common alternative to CPAP.
If it's midnight and a DME tells you it's dark outside, go and check for yourself.

Useful Links.

User avatar
NateS
Posts: 1716
Joined: Wed Dec 14, 2011 8:53 pm
Location: Kaatskill Mts-Washington Irving

Re: ResScan and S9 VPAP Adapt

Post by NateS » Tue Oct 09, 2012 1:12 pm

avi123 wrote:Nathe who has compSAS provided these graphs by S9 VPAP Adapt and ResScan:


Image

Notice the poor control of leaks that Nathe had, most likely b/c of the Aloha Nasal Pillow CPAP Mask.
Sorry for tuning in late - note the date on that graph of mine you posted - it was in January, which was the first month I was on therapy. I was not using the Aloha then. In fact I don't believe the Aloha was even released for sale back then. I was using a Swift FX.

Also, I have much better control over my leak rate now and it is usually below the red line. I bought the Aloha when it came out and I am still using an Aloha, and like it a lot.

As for the issue about my model of machine not distinguishing between centrals and obstructives in ResScan reports, I believe this was studied at length, at my inquiry, by another more technically-trained member, whose name I don't immediately recall. He did a great job with the data I sent him at his request, and figured it all out. His avatar used to look like a squirrel although maybe he has changed it since.

I think if you search under "algorithm" + "central" + "ResMed" you will quickly find his research, conclusions and explanation.

I like my machine. I think it has done a great job for me. But I have never used the ASV Phillips makes/sells so I cannot compare the two.

Regards, Nate

_________________
Mask: DreamWear Nasal CPAP Mask with Headgear
Additional Comments: ResMed AirCurve 10 ASV; Dreamwear Nasal Mask Original; CPAPMax Pillow; ResScan & SleepyHead
Central sleep apnea AHI 62.6 pre-VPAP. Now 0 to 1.3
Present Rx: EPAP: 8; IPAPlo:11; IPAPHi: 23; PSMin: 3; PSMax: 15
"I've had a perfectly wonderful evening, but this wasn't it." —Groucho Marx

User avatar
avi123
Posts: 4509
Joined: Tue Dec 21, 2010 5:39 pm
Location: NC

Re: ResScan and S9 VPAP Adapt

Post by avi123 » Tue Oct 09, 2012 9:04 pm

Hi Nate, it's nice to see you again.

A question, you mentioned that you have ComplexSAS but I don't remember if you have it as a result of having emergent central apneas after usage of a CPAP?
My other question is if you have selected the Aloha pillows over the ResMed Swift™ FX CPAP Nasal Pillows, for any reason?

_________________
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments:  S9 Autoset machine; Ruby chinstrap under the mask straps; ResScan 5.6

User avatar
NateS
Posts: 1716
Joined: Wed Dec 14, 2011 8:53 pm
Location: Kaatskill Mts-Washington Irving

Re: ResScan and S9 VPAP Adapt

Post by NateS » Wed Oct 10, 2012 11:21 am

avi123 wrote:Hi Nate, it's nice to see you again.

A question, you mentioned that you have ComplexSAS but I don't remember if you have it as a result of having emergent central apneas after usage of a CPAP?
My other question is if you have selected the Aloha pillows over the ResMed Swift™ FX CPAP Nasal Pillows, for any reason?
My diagnosis was as follows:

1) Central Sleep Apnea complicating CPAP treatment of Obstructive Sleep Apnea (Central Sleep Apnea Due to Medical Condition, Not Cheyne Stokes (327.27))
2) Obstructive Sleep Apnea (327.23) (AHI=29.0 events/hour)
3) Periodic Limb Movement Disorder (327.51) (PLM Index=15.9 events/hour)
4) Physiological Hypersomnia (327.10) — ESS=16/24)
5. Nocturnal Premature Ventricular Contractions (427.69)
6. Nocturnal Premature Atrial Contractions (427.61)

Comments: …The degree of sleep fragmentation and disruption of sleep architecture places the patient at risk for cognitive impairments and metabolic sequelae including hypertension, cardiovascular disease, diabetes, and the metabolic syndrome.


After reading it and describing to my doctor how absolutely miserable I was during and after the constant pressure of the CPAP sleep test, I was able to persuade my doctor to skip the 6 week CPAP home trial and give me a third sleep study whose results qualified me for going directly to the ASV made by ResMed as described above and below.

As for the Aloha, if you will type NateS and Aloha into our search box, you will find detailed descriptions I posted regarding my experiences with the Aloha and why I think it is superior to the Swift FX, one advantage out of many being the articulating angle by which the pillows can approach and enter the nostrils. Also many useful posts written by other Aloha users.

Regards, Nate

_________________
Mask: DreamWear Nasal CPAP Mask with Headgear
Additional Comments: ResMed AirCurve 10 ASV; Dreamwear Nasal Mask Original; CPAPMax Pillow; ResScan & SleepyHead
Central sleep apnea AHI 62.6 pre-VPAP. Now 0 to 1.3
Present Rx: EPAP: 8; IPAPlo:11; IPAPHi: 23; PSMin: 3; PSMax: 15
"I've had a perfectly wonderful evening, but this wasn't it." —Groucho Marx

User avatar
NateS
Posts: 1716
Joined: Wed Dec 14, 2011 8:53 pm
Location: Kaatskill Mts-Washington Irving

Re: ResScan and S9 VPAP Adapt

Post by NateS » Wed Oct 10, 2012 11:53 am

PS - In addition, the cheek straps or "hockey sticks" on the Aloha are rigid in the vertical plane, which tends to support the placement of the nasal pillows directly below one's nostrils, in most cases preventing the pillows from slipping down away from the bottom of the nose, which was a constant problem for me with the Swift FX.

Nate

_________________
Mask: DreamWear Nasal CPAP Mask with Headgear
Additional Comments: ResMed AirCurve 10 ASV; Dreamwear Nasal Mask Original; CPAPMax Pillow; ResScan & SleepyHead
Central sleep apnea AHI 62.6 pre-VPAP. Now 0 to 1.3
Present Rx: EPAP: 8; IPAPlo:11; IPAPHi: 23; PSMin: 3; PSMax: 15
"I've had a perfectly wonderful evening, but this wasn't it." —Groucho Marx

User avatar
avi123
Posts: 4509
Joined: Tue Dec 21, 2010 5:39 pm
Location: NC

Re: ResScan and S9 VPAP Adapt

Post by avi123 » Wed Oct 10, 2012 1:47 pm

NateS wrote:PS - In addition, the cheek straps or "hockey sticks" on the Aloha are rigid in the vertical plane, which tends to support the placement of the nasal pillows directly below one's nostrils, in most cases preventing the pillows from slipping down away from the bottom of the nose, which was a constant problem for me with the Swift FX.

Nate

Reply,

Thanks Nate. I am going to try the Swift FX for 30 days. IMO, the most important issue is the double silicone flaps on the pillows, that Resmed offers in the Swift FX. All the rest, be it the wire clip per Papit or fixing the headgear, is marginal. I assume that you are wearing the chinstrap with it. BTW, I have yet to see a decent leak graph from you and Papit while on those nasal pillows.

_________________
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments:  S9 Autoset machine; Ruby chinstrap under the mask straps; ResScan 5.6
Last edited by avi123 on Wed Oct 10, 2012 3:38 pm, edited 2 times in total.

User avatar
avi123
Posts: 4509
Joined: Tue Dec 21, 2010 5:39 pm
Location: NC

Re: ResScan and S9 VPAP Adapt

Post by avi123 » Wed Oct 10, 2012 2:04 pm

NateS wrote:
avi123 wrote:Hi Nate, it's nice to see you again.

A question, you mentioned that you have ComplexSAS but I don't remember if you have it as a result of having emergent central apneas after usage of a CPAP?
My other question is if you have selected the Aloha pillows over the ResMed Swift™ FX CPAP Nasal Pillows, for any reason?
My diagnosis was as follows:

1) Central Sleep Apnea complicating CPAP treatment of Obstructive Sleep Apnea (Central Sleep Apnea Due to Medical Condition, Not Cheyne Stokes (327.27))
2) Obstructive Sleep Apnea (327.23) (AHI=29.0 events/hour)
3) Periodic Limb Movement Disorder (327.51) (PLM Index=15.9 events/hour)
4) Physiological Hypersomnia (327.10) — ESS=16/24)
5. Nocturnal Premature Ventricular Contractions (427.69)
6. Nocturnal Premature Atrial Contractions (427.61)

Comments: …The degree of sleep fragmentation and disruption of sleep architecture places the patient at risk for cognitive impairments and metabolic sequelae including hypertension, cardiovascular disease, diabetes, and the metabolic syndrome.


After reading it and describing to my doctor how absolutely miserable I was during and after the constant pressure of the CPAP sleep test, I was able to persuade my doctor to skip the 6 week CPAP home trial and give me a third sleep study whose results qualified me for going directly to the ASV made by ResMed as described above and below.

As for the Aloha, if you will type NateS and Aloha into our search box, you will find detailed descriptions I posted regarding my experiences with the Aloha and why I think it is superior to the Swift FX, one advantage out of many being the articulating angle by which the pillows can approach and enter the nostrils. Also many useful posts written by other Aloha users.

Regards, Nate

Comment,

What stands out from your diagnosis is that you have CSAS related to CHF.

_________________
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments:  S9 Autoset machine; Ruby chinstrap under the mask straps; ResScan 5.6

MAXDRI
Posts: 21
Joined: Sun Sep 09, 2012 5:10 pm

Re: ResScan and S9 VPAP Adapt

Post by MAXDRI » Wed Oct 10, 2012 6:21 pm

Thanks to all of you.
I have determined that I am using the correct version of ResScan for my ResMed S9 VPAP Adapt.
I am also able to download all the detailed reports.
I haven't done that for several days - due to urgent events in my life.
However, in the daily report from my machine, it states that my leaks are in the 15-25 range. YEA!
It also says No AHI and NO AI.
Does that mean my treatment is working? And preventing the AHI and AI from happening?
I guess I can answer my own question by downloading the detailed data. If the detailed data confirms that I have No AHI or AI, then this ASV machine is amazing.

_________________
MaskHumidifier

User avatar
JohnBFisher
Posts: 3821
Joined: Wed Oct 14, 2009 6:33 am

Re: ResScan and S9 VPAP Adapt

Post by JohnBFisher » Wed Oct 10, 2012 8:39 pm

MAXDRI wrote:... It also says No AHI and NO AI. ... Does that mean my treatment is working? And preventing the AHI and AI from happening? ...
Yup. And yes, it is amazing.

_________________
Mask: Quattro™ FX Full Face CPAP Mask with Headgear
Additional Comments: User of xPAP therapy for over 20 yrs. Resmed & Respironics ASV units with EEP=9cm-14cm H2O; PSmin=4cm H2O; PSmax=15cm H2O; Max=25cm H2O
"I get up. I walk. I fall down. Meanwhile, I keep dancing” from Rabbi Hillel
"I wish to paint in such a manner as if I were photographing dreams." from Zdzisław Beksiński