What does it say on the left of the daily panel under details and statistics where it says Channel Min Med, etc.?
Is there a leak number line there?
Why I don't have a leak graph in Sleepyhead?
Re: Why I don't have a leak graph in Sleepyhead?
_________________
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear |
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control |
Additional Comments: Also SleepyHead, PRS1 Auto, Respironics Auto M series, Legacy Auto, and Legacy Plus |
Please enter your equipment in your profile so we can help you.
Click here for information on the most common alternative to CPAP.
If it's midnight and a DME tells you it's dark outside, go and check for yourself.
Useful Links.
Click here for information on the most common alternative to CPAP.
If it's midnight and a DME tells you it's dark outside, go and check for yourself.
Useful Links.
-
- Posts: 122
- Joined: Mon Jun 27, 2011 11:32 am
Re: Why I don't have a leak graph in Sleepyhead?
I am going to try this later today.deltadave wrote:Can you run the data on ResScan, or upload the data files from the SD card to Google docs?
And simultaneously, have someone send YOU a known "good" file?
I don't believe it is possible...deltadave wrote:Or perhaps, the value for the whole night is 0.00.CatherineF wrote:When only leak graph is checked there is flying cube with a sheep where daily graph are supposed to be, means no data
No it isn't there.archangle wrote:What does it say on the left of the daily panel under details and statistics where it says Channel Min Med, etc.?
Is there a leak number line there?
_________________
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear |
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control |
Additional Comments: Quattro FX and Mirage Quattro as backup masks |
Re: Why I don't have a leak graph in Sleepyhead?
Boy, this is a stumper!
Leak is part of a file (*PLD.edf) that includes

so if you have all the other parameters, it would seem unlikely that the file itself is corrupt.
I think seeing if it appears in ResScan will narrow the search.
Leak is part of a file (*PLD.edf) that includes

so if you have all the other parameters, it would seem unlikely that the file itself is corrupt.
I think seeing if it appears in ResScan will narrow the search.
...other than food...
Re: Why I don't have a leak graph in Sleepyhead?
Yo Catherine!
In a review of bugs that Jedimark was working on, one of them appeared to relate to leaks with BiPAP:
http://sourceforge.net/tracker/index.ph ... id=2089181
I think the proof would be to take a look at your data in ResScan.
In a review of bugs that Jedimark was working on, one of them appeared to relate to leaks with BiPAP:
http://sourceforge.net/tracker/index.ph ... id=2089181
I think the proof would be to take a look at your data in ResScan.
...other than food...
Re: Why I don't have a leak graph in Sleepyhead?
I don't believe it is possible...CatherineF wrote:deltadave wrote:Or perhaps, the value for the whole night is 0.00.CatherineF wrote:When only leak graph is checked there is flying cube with a sheep where daily graph are supposed to be, means no data
I had a night with no leaks, 0 for the hole night.
_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
Mask: Brevida™ Nasal Pillow CPAP Mask with Headgear |
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control |
Additional Comments: S9 VPAP Auto Bilevel Machine |