ResMed S9 >>> Humidifier Problems

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
Txdeputy
Posts: 1
Joined: Sat Jun 06, 2015 9:06 am

Re: ResMed S9 >>> Humidifier Problems

Post by Txdeputy » Sat Jun 06, 2015 10:13 am

True, if the heating plate on the humidifier is off or not working, the air coming through the machine, will be room temperature. If the humidifier is operational, the water in the reservoir, will be heated, thus heating the air as it leaves the machine. That is what happened to me, the humidifier failed, resulting in cold air, coming through the machine into my sinus cavity. I immediately checked the heating plate and it was stone cold. When I previously changed out my humidifier at the VA, they told me that they have a high incidence of the humidifier for the ResMed S9, being bad.

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: ResMed S9 >>> Humidifier Problems

Post by palerider » Sat Jun 06, 2015 5:24 pm

Txdeputy wrote:True, if the heating plate on the humidifier is off or not working, the air coming through the machine, will be room temperature. If the humidifier is operational, the water in the reservoir, will be heated, thus heating the air as it leaves the machine. That is what happened to me, the humidifier failed, resulting in cold air, coming through the machine into my sinus cavity. I immediately checked the heating plate and it was stone cold. When I previously changed out my humidifier at the VA, they told me that they have a high incidence of the humidifier for the ResMed S9, being bad.
the modern machines use smart humidifiers, they have humistats that measure the amount of moisture that is in the room air and adapt to that. if the air in your room wasn't dry, then the humidifer will add very little, or no extra moisture. they adapt on an ongoing basis. unless you measured the heating plate with a thermometer, all you know is that it wasn't hot to the touch, but that could be most any temp from room temp up to near body temp, all will feel cold to you.

many people have mistakenly thought that the humidifier in a modern machine has failed, and been wrong, simply because they didn't realize the humidifiers aren't just dumb "heat water" things anymore.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

danabarb

Re: ResMed S9 >>> Humidifier Problems

Post by danabarb » Fri Oct 23, 2015 4:28 pm

I have had the ResMed 9 and had nothing but problems with it. The water tank leaked, so I notified the place where I bought it and they said they replaced it with a new one. That leaked and the place where I bought it from said the machine checks out fine and if I send it back to ResMed, I will have to pay for the repair. I became weak and my legs got knots in them from lack of oxygen. I bought another machine and within two weeks am much better-not one hundred percent. The sleep place has been giving me the run around; told me I need a new sleep study test! and they cannot find my old records or prescription. I went to another source and the records will be sent within ten days!

Physician
Posts: 703
Joined: Tue Jun 29, 2010 5:42 pm
Location: West Coast USA

Re: ResMed S9 >>> Humidifier Problems

Post by Physician » Fri Oct 23, 2015 8:46 pm

danabarb wrote:I have had the ResMed 9 and had nothing but problems with it. The water tank leaked, so I notified the place where I bought it and they said they replaced it with a new one. That leaked and the place where I bought it from said the machine checks out fine and if I send it back to ResMed, I will have to pay for the repair. I became weak and my legs got knots in them from lack of oxygen. I bought another machine and within two weeks am much better-not one hundred percent. The sleep place has been giving me the run around; told me I need a new sleep study test! and they cannot find my old records or prescription. I went to another source and the records will be sent within ten days!

Strange. Why wasn't the repair covered by ResMed's two year warranty ? From where was it leaking ? Did you show them the leaks via photographs or demonstration ? During the warranty, you or the vendor can send it to ResMed.

Which entity indicated you need a new study ? What evidence do you have that you lacked oxygen ? Records sent within ten days from where to where ?

User avatar
LSAT
Posts: 13399
Joined: Sun Nov 16, 2008 10:11 am
Location: SE Wisconsin

Re: ResMed S9 >>> Humidifier Problems

Post by LSAT » Fri Oct 23, 2015 8:52 pm

danabarb wrote:I have had the ResMed 9 and had nothing but problems with it. The water tank leaked, so I notified the place where I bought it and they said they replaced it with a new one. That leaked and the place where I bought it from said the machine checks out fine and if I send it back to ResMed, I will have to pay for the repair. I became weak and my legs got knots in them from lack of oxygen. I bought another machine and within two weeks am much better-not one hundred percent. The sleep place has been giving me the run around; told me I need a new sleep study test! and they cannot find my old records or prescription. I went to another source and the records will be sent within ten days!
The CPAP machine itself cannot leak. If the humidifier was leaking it was the plastic tub. If for some reason the humidifier itself was returned, you could use the CPAP machine without the humidifier so you will not get leg knots(?)

User avatar
Wulfman...
Posts: 6688
Joined: Mon Sep 01, 2014 6:41 pm
Location: Nearest fishing spot

Re: ResMed S9 >>> Humidifier Problems

Post by Wulfman... » Fri Oct 23, 2015 8:57 pm

Same person, similar post here:

viewtopic.php?f=1&t=57626&start=15


Den

.
(5) REMstar Autos w/C-Flex & (6) REMstar Pro 2 CPAPs w/C-Flex - Pressure Setting = 14 cm.
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05

scottgib

Re: ResMed S9 >>> Humidifier Problems

Post by scottgib » Sun Dec 06, 2015 4:40 pm

If heater fails, the air will will be lower than rooms temperature by some amount up to 20 degrees F depending on the room humidity. If the room humidity were say 35%, then I would guess that the air going to the mask would be something like like 10 degrees lower than room temp. ( I am too lazy to calculate the exact amount). This temp change is due to the evaporation of the water. This is like the cooling effect of sweating.

ps.smile4me@yahoo.com

Re: ResMed S9 >>> Humidifier Problems

Post by ps.smile4me@yahoo.com » Mon May 02, 2016 9:26 pm

I have had the same issues from day one with mine, It starts out to be working then it stops heating and tank is cool and it drys out my nose and mouth and i have to no avail get them to under stand the thing does not work then it wont document that i have used it and they say its because i didnt use it when i know that i did in order for my ins to pay for the thing i had to wear it at least 4 hrs a day and the lady said if i was not going to wear it turn it on a leave it running for 4 hrs for ins to pay for it, What the heck, I told her it was not working and then they sold me a mouth strap and said i must be breathing out my mouth at night so that is why my mouth was dry and nose . But i told her that still did not fix the fact the water take was cold and not warm, Well the strap to hold my mouth closed but i still woke up with dry eyes and nose and mouth. Now i have been seeing that this RES Med S9 has issues and has caused me issues and a lot of others a lot of issues here from reading what others have wrote. The people i got this Res Med from made me feel it was all in my head and nothing was wrong with this Res Med S9 It caused me mush distress and i found my self taking it off through the night and got to wear i could not breath good with it like pressure air flow was not right, They made me feel like i was me and not the Res Med , That nothing was wrong with it , Well i stopped using it and suffered a stroke. Now i have a Res Med S9 i can not use and my ins will not pay for another one. This company should owe up the the fact that they have some that do not work and give back there money are make it right by given them one that works.

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: ResMed S9 >>> Humidifier Problems

Post by palerider » Mon May 02, 2016 10:09 pm

someone call a farmer, we've got a load of fertilizer!

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
WRX03
Posts: 137
Joined: Fri Apr 29, 2016 9:37 am
Location: Roanoke, Virginia

Re: ResMed S9 >>> Humidifier Problems

Post by WRX03 » Wed May 04, 2016 7:05 am

A medical equipment repair tech here. I'm looking at another S9 at the moment with the usual reported problems; cuts off and/or humidifier in-op. This one is reported to "cuts off after a few minutes" and "humidifier not heating" Of course I can't duplicate either problem, it's been running for 24hrs straight, and there isn't anything in the error log to support either issue. The last code is 631 and it was 9 months ago.

I now know how a car mechanic feels when a car acts up but they can't duplicate the problem and there isn't any error codes. They feel bad because they can't fix the problem they can't duplicate. You might as well throw a hammer at it and replace what it hits. Then they cringe when the car come back a month or two later.

At least another popular manufacture's error codes are service tech friendly and tell you what needs replaced. If I have a "cuts off issue" there will be a related error code recorded. 99.99999% of the time it correct and fixes the issue.

_________________
Mask: AirTouch™ F20 Full Face CPAP Mask with Headgear
Additional Comments: Actually PCA is a "For Her". It's DIY piece together from spare parts CPAP. It's multi color! LOL

User avatar
chunkyfrog
Posts: 34544
Joined: Mon Jul 12, 2010 5:10 pm
Location: Nowhere special--this year in particular.

Re: ResMed S9 >>> Humidifier Problems

Post by chunkyfrog » Wed May 04, 2016 7:14 am

I keep a cheap little hygrometer next to my machine.
It tells me the ambient humidity.
Then I know why the tank is still mostly full, or nearly dry.
These things are usually more reliable than we think--or wish, if that's the case.