sbullett wrote: I am not sure I understand why you wouldn't want it to be near the max number. Is it based on the occurence of apneas that are continuing now?
Yes, it is because of the current AHI showing. If you were below 5 AHI...wouldn't matter so much that you were maxing out your maximum because a less than 5 AHI would be acceptable.
Since you are double that amount then it points to the pressures not being set optimally.
I suspect that the minimum of 5 is not sufficient when you add the EPR of 3 cm reduction into the mix.
It would be helpful to see the overall graphs and not just this summary section. So that we can see the events in relation to the pressures...like are they occurring a times that the pressure is lower (as in near the minimum) or are they occurring when you are maxed out also.
When they did your titration they didn't factor in a 3 cm drop during exhale when they came up with your 7 cm suggested pressure. You would be surprised how much a little 1 or 2 cm pressure difference will make.
Sometimes the minimum has to be adjusted upwards a bit to make up for the EPR reduction....so not only are you starting out less than 7 cm...when you do get to 7 cm on inhale you are still only using 4 cm on exhale...likely allowing some events to sneak past the defenses.
I may have to RISE but I refuse to SHINE.