Wierd air feeling coming from hose?

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
jjlady
Posts: 84
Joined: Wed Jul 18, 2012 8:22 pm
Location: Las Vegas, Nevada

Wierd air feeling coming from hose?

Post by jjlady » Thu Aug 09, 2012 10:36 am

Last night the air on the inhale felt wierd, I am not sure how to describe it.....felt lie a bubble, or it was going over a bump, not sure what else to say about it? It felt like it was right at the intersection of where the air comes into the mask.

This didn't happen alot maybe 6-10 times, each one was very fast. Also it didn't happen right away, after quite a few hours on the mask. Once I noticed it I couldn't sleep because I kept waiting for it to happen. Between that and a horrible leg cramp I just threw the mask off and slept 2 hours without it.

I also tried a pad a cheek anti leak strap last night for the first time but I can't see how that would affect the air inside or how it comes from the hose to the mask, I made sure the strap was over the intake thing like they said to do (and just like the picture on there site). I even checked the hose to see if there were any kinks or bends and there wasn't any.

Has anybody had this before??

_________________
Mask: Mirage Quattro™ Full Face CPAP Mask with Headgear
Additional Comments: Sleep test showed AHI 41, SETTING IS 15

User avatar
Conrad
Posts: 698
Joined: Mon Jul 30, 2012 9:10 am
Location: Northern Illinois

Re: Wierd air feeling coming from hose?

Post by Conrad » Thu Aug 09, 2012 10:40 am

I'm very new to all this but it sounds like you're talking about a normal pulse coming from your machine. I'm not exactly sure what these pulses are for but mine does it from time to time as well.
ResMed AirCurve 10 Vauto Swift FX

Do not regret growing older. It is a privilege denied to many...

It takes less time to do a thing right than to explain why you did it wrong.

jjlady
Posts: 84
Joined: Wed Jul 18, 2012 8:22 pm
Location: Las Vegas, Nevada

Re: Wierd air feeling coming from hose?

Post by jjlady » Thu Aug 09, 2012 10:54 am

I have had that pulse, the pulse is sort of like the mask breathing or going up and down a bit, this was more concentrated like a ball of air bouncing around in a small area right as it enters the mask.

_________________
Mask: Mirage Quattro™ Full Face CPAP Mask with Headgear
Additional Comments: Sleep test showed AHI 41, SETTING IS 15

User avatar
chunkyfrog
Posts: 34544
Joined: Mon Jul 12, 2010 5:10 pm
Location: Nowhere special--this year in particular.

Re: Wierd air feeling coming from hose?

Post by chunkyfrog » Thu Aug 09, 2012 11:18 am

aliens?

_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear
Additional Comments: Airsense 10 Autoset for Her

User avatar
MaxDarkside
Posts: 1199
Joined: Sun Dec 18, 2011 4:21 pm
Location: Minneapolis, MN

Re: Wierd air feeling coming from hose?

Post by MaxDarkside » Thu Aug 09, 2012 12:27 pm

I can think of two things that might be like you describe: 1) the flapper valve in the elbow coming into the mask was flipper-flapping a bit. Might happen at lower pressures. 2) I don't have a PR system, but a ResMed S9 AutoSet and I can feel the predictive aspect of flow coming from the machine, that is, it seems the machine prepares airflow in anticipation of the next inhale and so sometimes I feel that little push, particularly if I slow my breathing.

Or 3), ALIENS !


_________________
Mask: Quattro™ FX Full Face CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: CMS-50E Oximeter, Zeo Bedside, Software: Comm'l grade AI analytics server & tools & SleepyHead
Do or Die... Sleep Apnea killed me, but I came back. Click for my story
Please visit my My Apnea Analytics blog. Maybe we can help each other.
54 yrs, 6' 1", 160->172 lbs

jjlady
Posts: 84
Joined: Wed Jul 18, 2012 8:22 pm
Location: Las Vegas, Nevada

Re: Wierd air feeling coming from hose?

Post by jjlady » Thu Aug 09, 2012 12:35 pm

It is just like me to have an unexplainable thing happen, nothing in my life is "normal".

_________________
Mask: Mirage Quattro™ Full Face CPAP Mask with Headgear
Additional Comments: Sleep test showed AHI 41, SETTING IS 15

User avatar
MaxDarkside
Posts: 1199
Joined: Sun Dec 18, 2011 4:21 pm
Location: Minneapolis, MN

Re: Wierd air feeling coming from hose?

Post by MaxDarkside » Thu Aug 09, 2012 12:46 pm

jjlady wrote:It is just like me to have an unexplainable thing happen, nothing in my life is "normal".
One must surrender normalcy to be special Those we honor with greatness were abnormal, by definition. Carry on!

What's your starting pressure? Down low like 4-6 cm H2O?

_________________
Mask: Quattro™ FX Full Face CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: CMS-50E Oximeter, Zeo Bedside, Software: Comm'l grade AI analytics server & tools & SleepyHead
Do or Die... Sleep Apnea killed me, but I came back. Click for my story
Please visit my My Apnea Analytics blog. Maybe we can help each other.
54 yrs, 6' 1", 160->172 lbs

jjlady
Posts: 84
Joined: Wed Jul 18, 2012 8:22 pm
Location: Las Vegas, Nevada

Re: Wierd air feeling coming from hose?

Post by jjlady » Thu Aug 09, 2012 1:05 pm

It is fixed at 15 but I ramp from 4 for 30 minutes. Maybe when I put the anti leak strap over the hole and up against the mask something got moved out of position, but it only started happening at least 2 hours (guessing) into my sleep or trying to sleep. It kind of felt like as the air came into the mask it tumbled around or formed a ball and I inhaled it...I know that sounds wierd but that is the best way I can describe it!!

_________________
Mask: Mirage Quattro™ Full Face CPAP Mask with Headgear
Additional Comments: Sleep test showed AHI 41, SETTING IS 15

User avatar
MaxDarkside
Posts: 1199
Joined: Sun Dec 18, 2011 4:21 pm
Location: Minneapolis, MN

Re: Wierd air feeling coming from hose?

Post by MaxDarkside » Thu Aug 09, 2012 1:09 pm

I'm guessing it has something to do with that flapper valve. If you block the output of that valve, it may not want to properly "slap shut". It's the differential pressure, 4 cm vs. bedroom pressure that keeps it shut, and 4 cm is not a lot of "umpfh" so maybe it's moving in there a bit giving you the sensation. At 2 hours (guessing) into your sleep your breathing may be subdued enough that it somehow it's getting rather a bit "flippy flappy" (technical term )

_________________
Mask: Quattro™ FX Full Face CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: CMS-50E Oximeter, Zeo Bedside, Software: Comm'l grade AI analytics server & tools & SleepyHead
Do or Die... Sleep Apnea killed me, but I came back. Click for my story
Please visit my My Apnea Analytics blog. Maybe we can help each other.
54 yrs, 6' 1", 160->172 lbs

jjlady
Posts: 84
Joined: Wed Jul 18, 2012 8:22 pm
Location: Las Vegas, Nevada

Re: Wierd air feeling coming from hose?

Post by jjlady » Thu Aug 09, 2012 1:25 pm

MaxDarkside wrote:I'm guessing it has something to do with that flapper valve. If you block the output of that valve, it may not want to properly "slap shut". It's the differential pressure, 4 cm vs. bedroom pressure that keeps it shut, and 4 cm is not a lot of "umpfh" so maybe it's moving in there a bit giving you the sensation. At 2 hours (guessing) into your sleep your breathing may be subdued enough that it somehow it's getting rather a bit "flippy flappy" (technical term )

That sounds like a possibility, also to be candid, yesterday I was very anxious so I took about 3/4 of an ativan to try and relax (I rarely do this, maybe 3x a year), so maybe my breathing was off (although I took it early at 530pm).

_________________
Mask: Mirage Quattro™ Full Face CPAP Mask with Headgear
Additional Comments: Sleep test showed AHI 41, SETTING IS 15

User avatar
Conrad
Posts: 698
Joined: Mon Jul 30, 2012 9:10 am
Location: Northern Illinois

Re: Wierd air feeling coming from hose?

Post by Conrad » Thu Aug 09, 2012 2:55 pm

Just to be clear, your mask doesn't look like this right?

Image
ResMed AirCurve 10 Vauto Swift FX

Do not regret growing older. It is a privilege denied to many...

It takes less time to do a thing right than to explain why you did it wrong.

jjlady
Posts: 84
Joined: Wed Jul 18, 2012 8:22 pm
Location: Las Vegas, Nevada

Re: Wierd air feeling coming from hose?

Post by jjlady » Thu Aug 09, 2012 6:40 pm

I came home and looked at my data....interesting, periodic breathing last night was 14% while the other nites/ or last week was 3%. Not sure exactly what that means but I bet it has something to do with that...and I bet the ativan caused the periodic breathing as I heard it can affect respiration. Although I took the ativan at least 7 hrs prior.

_________________
Mask: Mirage Quattro™ Full Face CPAP Mask with Headgear
Additional Comments: Sleep test showed AHI 41, SETTING IS 15

User avatar
MaxDarkside
Posts: 1199
Joined: Sun Dec 18, 2011 4:21 pm
Location: Minneapolis, MN

Re: Wierd air feeling coming from hose?

Post by MaxDarkside » Thu Aug 09, 2012 7:00 pm

jjlady wrote:I took the ativan at least 7 hrs prior.
Ativan half-life is 9–16 hours (source: wiki). That means that dose you took 7 hrs earlier... more than half of it was still swirling about your system. If we take the 9 hrs half-life, after 18 hrs a quarter of that dose is still swirling about.

_________________
Mask: Quattro™ FX Full Face CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: CMS-50E Oximeter, Zeo Bedside, Software: Comm'l grade AI analytics server & tools & SleepyHead
Do or Die... Sleep Apnea killed me, but I came back. Click for my story
Please visit my My Apnea Analytics blog. Maybe we can help each other.
54 yrs, 6' 1", 160->172 lbs

jjlady
Posts: 84
Joined: Wed Jul 18, 2012 8:22 pm
Location: Las Vegas, Nevada

Re: Wierd air feeling coming from hose?

Post by jjlady » Thu Aug 09, 2012 7:12 pm

MaxDarkside wrote:
jjlady wrote:I took the ativan at least 7 hrs prior.
Ativan half-life is 9–16 hours (source: wiki). That means that dose you took 7 hrs earlier... more than half of it was still swirling about your system. If we take the 9 hrs half-life, after 18 hrs a quarter of that dose is still swirling about.
My math isn't good, but I think that means it is still in my system now and will be tonight. No wonder I was tired all day today!

_________________
Mask: Mirage Quattro™ Full Face CPAP Mask with Headgear
Additional Comments: Sleep test showed AHI 41, SETTING IS 15

User avatar
MaxDarkside
Posts: 1199
Joined: Sun Dec 18, 2011 4:21 pm
Location: Minneapolis, MN

Re: Wierd air feeling coming from hose?

Post by MaxDarkside » Thu Aug 09, 2012 7:29 pm

jjlady wrote:
MaxDarkside wrote:
jjlady wrote:I took the ativan at least 7 hrs prior.
Ativan half-life is 9–16 hours (source: wiki). That means that dose you took 7 hrs earlier... more than half of it was still swirling about your system. If we take the 9 hrs half-life, after 18 hrs a quarter of that dose is still swirling about.
My math isn't good, but I think that means it is still in my system now and will be tonight. No wonder I was tired all day today!
The amount in your body goes in half every 9-16 hrs. So, if you take some at 10 PM, half of it is swirling about at 7 AM. One quarter of it is swirling around at 4 PM the next day.

_________________
Mask: Quattro™ FX Full Face CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: CMS-50E Oximeter, Zeo Bedside, Software: Comm'l grade AI analytics server & tools & SleepyHead
Do or Die... Sleep Apnea killed me, but I came back. Click for my story
Please visit my My Apnea Analytics blog. Maybe we can help each other.
54 yrs, 6' 1", 160->172 lbs