System One Not Recording on SD card

Post by biged46us » Tue Jun 26, 2012 8:27 pm

I got my System One at the beginning of June. After some research I learned of the existence of SleepyHead. I downloaded it and when I tried to read my SD card, there was no data to be read...this is after nearly 2 month's use. I tried to read the root directory on the SD card as directed. Is my System One properly set up?


_________________
Mask

User avatar
greatunclebill
Posts: 1503
Joined: Mon Feb 20, 2012 7:48 pm
Location: L.A. (lower alabama)

Re: System One Not Recording on SD card

Post by greatunclebill » Tue Jun 26, 2012 8:45 pm

do you have a system one 60 series? if so, it does not work with sleepyhead. have you looked at your card in the computer, not using sleepyhead, like you would look at any cd or dvd? is anything there? is the card unlocked so the tab on the side is pointing toward the end with the electical contacts?

_________________
Mask: Quattro™ FX Full Face CPAP Mask with Headgear
Additional Comments: myAir, OSCAR. cms-50D+. airsense 10 auto & (2009) remstar plus m series backups
First diagnosed 1990
please don't ask me to try nasal. i'm a full face person.
the avatar is Rocco, my Lhasa Apso. Number one "Bama fan. 18 championships and counting.
Life member VFW Post 4328 Alabama
MSgt USAF (E-7) medic Retired 1968-1990

User avatar
Pugsy
Posts: 65362
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: System One Not Recording on SD card

Post by Pugsy » Tue Jun 26, 2012 9:00 pm

Need to know the exact model name that shows up on the machine AND the model number off the bottom of the blower unit.

DS something or HS something.
150, 250, 450, 460, 550, 560 one of those...

The 60 models...SleepyHead won't work with those machines but something else will.
150 and 250 models...software is useless because the machines only collect hours of use on that SD card.

So need to know exactly which machine...then we can figure out why no data.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
biged46us
Posts: 5
Joined: Sat May 19, 2012 8:20 pm
Location: Columbus, OH
Contact:

Re: System One Not Recording on SD card

Post by biged46us » Fri Jun 29, 2012 8:38 pm

There are a bunch of labels on the bottom of the blower unit, but I think this is the information you need"

REMeter Auto DOM
560P

_________________
Mask

User avatar
Pugsy
Posts: 65362
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: System One Not Recording on SD card

Post by Pugsy » Fri Jun 29, 2012 8:51 pm

Let's get your profile fixed. So please edit your profile equipment tab to avoid confusion later...so we know for sure what you are using.
On the equipment page for machines...look for PR System One Remstar 60 Series Auto CPAP....that is what the 560 model is.
In layman's terms...you have an APAP and not the Pro you have chosen right now which is a straight cpap

For humidifier choose....System One 60 Series heated tube humidifier with heated tube....it is the next to the last one on the list of humidifiers..you have to scroll down.

SleepyHead won't work with your machine just yet but Mark is working on it.
For now you can use EncoreBasic which will give you tons of information. Check your private message box because I have sent you some information on how to get EncoreBasic.

You do have data on the machine...just need different software right now to view the data.
It is Windows based software though. No Mac version. If you have a Mac you have to set up some sort of Windows environment to run the software.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

Kerr
Posts: 22
Joined: Tue Jun 05, 2012 12:22 pm
Location: Knoxville, TN

Re: System One Not Recording on SD card

Post by Kerr » Fri Jun 29, 2012 8:58 pm

I have the same machine, can you send me the information of where to find it too?

User avatar
Pugsy
Posts: 65362
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: System One Not Recording on SD card

Post by Pugsy » Fri Jun 29, 2012 9:07 pm

Kerr wrote:I have the same machine
check your PM box

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

Kerr
Posts: 22
Joined: Tue Jun 05, 2012 12:22 pm
Location: Knoxville, TN

Re: System One Not Recording on SD card

Post by Kerr » Fri Jun 29, 2012 9:16 pm

I get 'media does not belong to an EncoreBasic Patient'

Any thoughts?

lazer
Posts: 1377
Joined: Fri Apr 20, 2012 1:56 pm
Location: Hermitage, PA
Contact:

Re: System One Not Recording on SD card

Post by lazer » Fri Jun 29, 2012 9:20 pm

Kerr wrote:I get 'media does not belong to an EncoreBasic Patient'

Any thoughts?
Just ignore that msg

_________________
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear
Additional Comments: SleepyHead & Encore Basic Software & a Zeo
Image.....................................................ImagePress ESC if the animations BUG you!.....................................................Image

Kerr
Posts: 22
Joined: Tue Jun 05, 2012 12:22 pm
Location: Knoxville, TN

Re: System One Not Recording on SD card

Post by Kerr » Fri Jun 29, 2012 9:22 pm

Ah, fixed it. I had to use the "Assign Card" tab and assign it to me before it would download anything.

rexter
Posts: 2
Joined: Fri Jul 06, 2012 8:22 am

Re: System One Not Recording on SD card

Post by rexter » Fri Jul 06, 2012 8:30 am

>> You do have data on the machine...just need different software right now to view the data.

I can't get my SD card to be read by EncoreViewer or EncorePro v2.7 or SleepyHead 0.9.1. Can you point me to EncoreBasic to see if that will work, please?

User avatar
Pugsy
Posts: 65362
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: System One Not Recording on SD card

Post by Pugsy » Fri Jul 06, 2012 9:14 am

rexter wrote:I can't get my SD card to be read by EncoreViewer or EncorePro v2.7 or SleepyHead 0.9.1. Can you point me to EncoreBasic to see if that will work, please?
Right now the only software that works with the new 60 series machines is EncoreBasic and Encore Pro 2.8...the others aren't supported.
Check your private message box for information on how to get EncoreBasic.
Basic offers the exact same reports as Pro does without being the PITA that Pro is.
If you run into trouble post here at the forum and someone will help you. I won't be available for the next 3 days or so.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

rexter
Posts: 2
Joined: Fri Jul 06, 2012 8:22 am

Re: System One Not Recording on SD card

Post by rexter » Sun Jul 08, 2012 2:27 am

Thanks a million, Pugsy! After two weeks of phoning unhelpful support lines and the purchase of both EncorePro and EncoreViewer, I finally have some overview of the data. Now I must just figure out the details of this software.

I read somewhere that SleepyHead was being updated to possibly cover 60 series machines as well?

Appreciate the help,

R.