Good afternoon all.
I'm thinking of the quote, "No two nares are created the same".
With both the Swift FX for Her and Aloha, I have not had the difficulty with the pillows as I'm currently having with the Hybrid.
You know what the kicker is? I'm only having issues with the outer edge of the left nare. The right nare is fine as far as I can tell.
If I could just get the left nare under control, I'd be a happy camper If I slightly lift the outer edge of the left pillow, no hissing sound.
However, I can't sleep like that Suggestions would be much appreciated.
Starlette
Hybrid FFM - Pillow Leaks
- chunkyfrog
- Posts: 34545
- Joined: Mon Jul 12, 2010 5:10 pm
- Location: Nowhere special--this year in particular.
Re: Hybrid FFM - Pillow Leaks
Hve you checked this out?
viewtopic.php?t=13882
viewtopic.php?t=13882
_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear |
Additional Comments: Airsense 10 Autoset for Her |
Re: Hybrid FFM - Pillow Leaks
Thank you ChunkyFrog for the link. What Roster said about tightening up the top strap did the trick.
I'll be able to use it tonight.
Starlette
I'll be able to use it tonight.
Starlette