using sleepyhead software: data not importing

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
User avatar
freelsjd
Posts: 7
Joined: Sat May 12, 2012 12:38 pm
Location: Clinton, TN

using sleepyhead software: data not importing

Post by freelsjd » Sat May 12, 2012 12:43 pm

I have sleepyhead installed on my computer and it is not reading my data.

Any advice ?

_________________
Machine: PR System One REMStar 60 Series Auto + Heated Tube Humidifier Core
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear
Additional Comments: Model 560P, started April 27, 2012

User avatar
Burkebang
Posts: 297
Joined: Fri Dec 09, 2011 4:30 pm
Location: Oslo, Norway

Re: using sleepyhead software: data not importing

Post by Burkebang » Sat May 12, 2012 3:17 pm

Are you sure the importer is pointed to the root of your memorycard? If not, check it and try again.

_________________
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear
Additional Comments: Also use Mirage FX nasal mask a lot. Contec CMS-50D+ Pulseoximeter and Zeo Mobile tracks the quality of my therapy.

User avatar
freelsjd
Posts: 7
Joined: Sat May 12, 2012 12:38 pm
Location: Clinton, TN

Re: using sleepyhead software: data not importing

Post by freelsjd » Sat May 12, 2012 3:39 pm

Thanks for responding Burkebang. Yes, I am fairly certain it is at the right directory because it picked up some of the information such as brand name, model number, and serial number. I did not enter that information, so it did read that from the root directory. But, it does not seem to read any thing else. I know the data is there because the binary files have all the dates I have used the machine since I started with CPAP two weeks ago. I wonder if the data is somehow password protected ?

_________________
Machine: PR System One REMStar 60 Series Auto + Heated Tube Humidifier Core
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear
Additional Comments: Model 560P, started April 27, 2012

User avatar
NateS
Posts: 1716
Joined: Wed Dec 14, 2011 8:53 pm
Location: Kaatskill Mts-Washington Irving

Re: using sleepyhead software: data not importing

Post by NateS » Sat May 12, 2012 5:42 pm

freelsjd wrote:Thanks for responding Burkebang. Yes, I am fairly certain it is at the right directory because it picked up some of the information such as brand name, model number, and serial number. I did not enter that information, so it did read that from the root directory. But, it does not seem to read any thing else. I know the data is there because the binary files have all the dates I have used the machine since I started with CPAP two weeks ago. I wonder if the data is somehow password protected ?
You haven't yet mentioned the make and model of the xPAP from which your data emanates, nor the version of Sleepyhead you are using, nor the operating system of your computer.

Nate

_________________
Mask: DreamWear Nasal CPAP Mask with Headgear
Additional Comments: ResMed AirCurve 10 ASV; Dreamwear Nasal Mask Original; CPAPMax Pillow; ResScan & SleepyHead
Central sleep apnea AHI 62.6 pre-VPAP. Now 0 to 1.3
Present Rx: EPAP: 8; IPAPlo:11; IPAPHi: 23; PSMin: 3; PSMax: 15
"I've had a perfectly wonderful evening, but this wasn't it." —Groucho Marx

User avatar
avi123
Posts: 4509
Joined: Tue Dec 21, 2010 5:39 pm
Location: NC

Re: using sleepyhead software: data not importing

Post by avi123 » Sat May 12, 2012 6:03 pm

freelsjd wrote:I have sleepyhead installed on my computer and it is not reading my data.

Any advice ?
If you stick around untill Pugsy shows up you're sure to get a reply. BTW, do you have the card inside the machine while the data is collected and you'r a sleep?

I also have problems with SH and even gave up on it. I use ResScan.

_________________
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments:  S9 Autoset machine; Ruby chinstrap under the mask straps; ResScan 5.6
see my recent set-up and Statistics:
http://i.imgur.com/TewT8G9.png
see my recent ResScan treatment results:
http://i.imgur.com/3oia0EY.png
http://i.imgur.com/QEjvlVY.png

User avatar
freelsjd
Posts: 7
Joined: Sat May 12, 2012 12:38 pm
Location: Clinton, TN

Re: using sleepyhead software: data not importing

Post by freelsjd » Sat May 12, 2012 7:44 pm

I am new to this forum and have only used my CPAP machine about two weeks now.

I just edited my profile to include my machine information.

I have installed v0.9.2-2 beta of Sleephead just today from source on my Linux/Debian/Squeeze amd64 computer.

I copied the data from the SD card onto my hard drive. I can see the data does exist each day since I have had it.

I would like to try SleepyHead since it appears to be compatible with Respironics.

_________________
Machine: PR System One REMStar 60 Series Auto + Heated Tube Humidifier Core
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear
Additional Comments: Model 560P, started April 27, 2012

User avatar
RandyJ
Posts: 1673
Joined: Thu Apr 21, 2011 7:22 pm
Location: Connecticut, USA

Re: using sleepyhead software: data not importing

Post by RandyJ » Sat May 12, 2012 8:12 pm

Sleepyhead is not yet compatible with the new 60 series machines, but he is working on it. Encore will work, though.

Edit: not sure if Encore will work with your OS, but perhaps you also have access to a computer running Windows...

_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: Backup & Travel Machines: PR System One Bipap Auto, S9 VPAP Auto, S9 Autoset, Oximeter CMS-50E
Last edited by RandyJ on Sat May 12, 2012 9:57 pm, edited 1 time in total.
Diagnosed March 2011, using APAP 14 - 16.5 cm, AFlex+ 2
Alt masks Swift FX pillows, Mirage FX nasal mask, Mirage Quattro full face mask

User avatar
NateS
Posts: 1716
Joined: Wed Dec 14, 2011 8:53 pm
Location: Kaatskill Mts-Washington Irving

Re: using sleepyhead software: data not importing

Post by NateS » Sat May 12, 2012 8:53 pm

freelsjd wrote:I am new to this forum and have only used my CPAP machine about two weeks now.

I just edited my profile to include my machine information.

I have installed v0.9.2-2 beta of Sleephead just today from source on my Linux/Debian/Squeeze amd64 computer.

I copied the data from the SD card onto my hard drive. I can see the data does exist each day since I have had it.

I would like to try SleepyHead since it appears to be compatible with Respironics.

Welcome!

Now it seems we know what the problem is, and those more familiar with your equipment will be able to chime in and make pertinent suggestions.

Best wishes,

Nate

_________________
Mask: DreamWear Nasal CPAP Mask with Headgear
Additional Comments: ResMed AirCurve 10 ASV; Dreamwear Nasal Mask Original; CPAPMax Pillow; ResScan & SleepyHead
Central sleep apnea AHI 62.6 pre-VPAP. Now 0 to 1.3
Present Rx: EPAP: 8; IPAPlo:11; IPAPHi: 23; PSMin: 3; PSMax: 15
"I've had a perfectly wonderful evening, but this wasn't it." —Groucho Marx

User avatar
Pugsy
Posts: 65255
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: using sleepyhead software: data not importing

Post by Pugsy » Sun May 13, 2012 8:09 am

As you have found out....SleepyHead is not yet compatible with the new 60 series machines.

Check your private message box..

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
freelsjd
Posts: 7
Joined: Sat May 12, 2012 12:38 pm
Location: Clinton, TN

Re: using sleepyhead software: data not importing

Post by freelsjd » Mon May 14, 2012 8:00 am

I added model 560P to my "additional comments" profile. I am assuming this fits the machine category for "60 series" ?

_________________
Machine: PR System One REMStar 60 Series Auto + Heated Tube Humidifier Core
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear
Additional Comments: Model 560P, started April 27, 2012

User avatar
Pugsy
Posts: 65255
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: using sleepyhead software: data not importing

Post by Pugsy » Mon May 14, 2012 9:57 am

freelsjd wrote:I added model 560P to my "additional comments" profile. I am assuming this fits the machine category for "60 series" ?
In which software did you add this bit of information? EncoreBasic doesn't care but it hurts nothing to have it there and SleepyHead still won't work.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

fredchu
Posts: 1
Joined: Mon Oct 15, 2012 5:06 pm

Re: using sleepyhead software: data not importing

Post by fredchu » Mon Oct 15, 2012 5:08 pm

10/13/12. Hello & thanks to Mark for SleepyHead. I found it today, but can't get it to read data from the SD chip or hard-drive folder. I successfully upgraded to v0.92 tonight, too. I pointed the Import button to E: for the SD card. It appears to read my machine info: Respironics 460p, serial etc. But first Use and Last Use are blank.

And just above, however, is a warning: No CPAP Machine Data Imported.

I would sincerely appreciate your advice. I'm sure I'm doing something
dumb.

Here is the file structure on the SD card.

E:
P-Series
Clear0 (folder)
e (folder)
000.004
p0 (folder)
0000000048.001
0000000168.002
properties.txt
P0657704xxxxC (folder - which is serial number of machine)
p0 (folder)
0000000001.001
0000000001.002
0000000062.001
0000000062.002
0000000062.005
plus 114 other files ending in either .001, .002, .005
properties.txt
last

User avatar
Pugsy
Posts: 65255
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: using sleepyhead software: data not importing

Post by Pugsy » Mon Oct 15, 2012 8:33 pm

fredchu wrote: It appears to read my machine info: Respironics 460p,
Your machine is one of the new 60 series machines and they came out after SleepyHead so SleepyHead won't work with that model machine just yet.
I will send you a private message in a little bit with some information about a software that will work with your machine. So check your private message box.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

Cubic302

Re: using sleepyhead software: data not importing

Post by Cubic302 » Wed Mar 13, 2013 7:25 pm

Respironics PR System One REMstar Pro with C-Flex+ (460P)

Hi All..
I was hoping some one could assist me. I am new to the forum, and to CPAP.. Just received it, used twice, and the health provider, (believe or not) pulls the data via wifi.. My CPAP has a wifi adapter, which connects to a cell service..

Anyway, I have the same issue with Sleepyhead. I use a Macbook Pro by Apple, and installed the MAC version of Sleepyhead v0.9.2

I can see the data.. many files ######.001, and ######.002, ect.. I initially thought the data was encrypted, since it goes over the network, and should be secure..(HIPA)... But I ready the the forum, and see that some members sent private messages with info about another software that may be helpful..

Can any one assist?

My CPAP = Respironics PR System One REMstar Pro with C-Flex+ (460P)

User avatar
jdm2857
Posts: 2982
Joined: Wed Aug 20, 2008 8:29 pm
Location: South Jersey

Re: using sleepyhead software: data not importing

Post by jdm2857 » Wed Mar 13, 2013 7:46 pm

I don't have the specifics, but there is a patch for SleepyHead that makes it accomodate the 60 series machines.

Someone with the link to the patch will be by shortly.

And if I remember correctly, when installing the patch, do not double click on the icon to run it.

Instead, right click and choose "Run as Administrator".
jeff