Central Apneas appear with the start of BiPAP Treatment

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
Tom W
Posts: 223
Joined: Thu Sep 01, 2011 4:59 am
Location: West Central Florida

Central Apneas appear with the start of BiPAP Treatment

Post by Tom W » Tue May 01, 2012 9:01 am

I’m not sure where I am at this point and not sure in which direction to go but I do know that I’m extremely frustrated, exhausted, etc.

I’ve spent 11 months or so with a System One CPAP and APAP trying to reduce my AHI without much success.

My AHI seems to be uncorrelated to pressure but seems to be inversely correlated to periodic breathing (my AHI seems to be lower when my PB is higher).

My original doctor of 13 years adjusted my pressure and looked at my results and kept promising that we would do something more radical ‘next’ time but in the end it was, “Lets try another pressure change and see what happens”.

Out of frustration I saw another doctor and he suggested that I just crank my pressure up from 12 to 18 to see what happens. I explained that I couldn’t tolerate the higher pressures due to ear pain and air entering my stomach. He prescribed a BiPAP machine with Ipress of 18 and Epress of 12. While I understand the increase in the number of Centrals my AHI due to the higher pressure my AHI due to Obstructive Apneas is still close to 10 (I understand I’ve only been on this machine for just over a week but it didn’t seem to do anything for my Obstructives and just added a significant number of Centrals).

I will admit that I seem slightly less exhausted with BiPAP but it’s marginal and perhaps just the placebo effect.

I go back in 2 weeks and even if he prescribes an ASV machine to reduce/eliminate the Centrals I’ll still wind up with an AHI close to or just over 10. Is this the best I can expect? Is my glottis reacting to the higher pressures by closing and negating the effects of the higher pressure? Is xPAP just not that effective with some people?

CPAP/APAP:

Image

BiPAP:

Image

Any Suggestions?
Last edited by Tom W on Thu May 03, 2012 11:26 am, edited 4 times in total.

User avatar
chunkyfrog
Posts: 34545
Joined: Mon Jul 12, 2010 5:10 pm
Location: Nowhere special--this year in particular.

Re: Lost...

Post by chunkyfrog » Tue May 01, 2012 9:09 am

If you add BIPAP to your title line, the bipappers here will certainly chime in.
They are a most helpful bunch.
You are in good hands.

_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear
Additional Comments: Airsense 10 Autoset for Her

User avatar
chunkyfrog
Posts: 34545
Joined: Mon Jul 12, 2010 5:10 pm
Location: Nowhere special--this year in particular.

Re: Lost On CPAP - Now Trying BiPAP

Post by chunkyfrog » Tue May 01, 2012 11:38 am

New post; to bump this up to the top where the right folks will see it.

_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear
Additional Comments: Airsense 10 Autoset for Her

User avatar
Pugsy
Posts: 65127
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Lost On CPAP - Now Trying BiPAP

Post by Pugsy » Tue May 01, 2012 12:11 pm

Another bump.. I have read it but I can't formulate anything constructive to offer.
I certainly don't feel comfortable undermining the doctor's 2 week plan. So won't go there but I don't know where they came up with IPAP 18.....but I guess there was a reason somewhere.

Calling JohnBFisher...for his thoughts maybe...I am at a loss.

How about editing the subject...remove "lost on cpap" and add Centrals in its place...get John's attention.
He doesn't read all thread and normally zeroes in on "central" threads because that is where is knowledge is more valuable.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

-SWS
Posts: 5301
Joined: Tue Jan 11, 2005 7:06 pm

Re: Central Apneas appear with the start of BiPAP Treatment

Post by -SWS » Tue May 01, 2012 12:45 pm

Tom W wrote: I go back in 2 weeks and even if he prescribes an ASV machine to reduce/eliminate the Centrals I’ll still wind up with an AHI close to or just over 10. Is this the best I can expect? Is my glottis reacting to the higher pressures by closing and negating the effects of the higher pressure? Is xPAP just not that effective with some people?
Let's see... You present obstructive apneas at CPAP and lower pressures. However, you additionally present periodic breathing and central apneas (aka "unstable ventilation") using BiLevel at higher pressures. That's not exactly rare. An ASV trial certainly sounds in order. However, the purpose of ASV in cases like yours is to treat BOTH the obstructive and central components. So it is not a forgone conclusion that ASV will leave you with an AHI close to or just over 10. If you have yet to try ASV, then you do not yet know what ASV will do for your obstructive and central components.

ASV's EPAP and minimum IPAP settings will endeavor to address your obstructive component---while ASV's fluctuating IPAP (that's the adaptive part of servo ventilation) will ideally address the undershoot/overshoot part of your central component. Additionally, the machine's backup rate should help with those "treatment emergent" central apneas. That said, ASV works well for some users while not working so well for others. Below are two links in which Gilmartin et al and Rapoport discuss ventilatory-instability's suspected hypocapnia mechanism----as well as an experimental adjunct method of countering hypocapnia's CO2 depletion:

http://www.ncbi.nlm.nih.gov/pmc/articles/PMC3014237/

http://www.ncbi.nlm.nih.gov/pmc/articles/PMC3014238/

Good luck with the ASV trial. But don't rule out ASV's potential until you have tried it...

-SWS
Posts: 5301
Joined: Tue Jan 11, 2005 7:06 pm

Re: Central Apneas appear with the start of BiPAP Treatment

Post by -SWS » Tue May 01, 2012 6:57 pm

Tom W wrote: My AHI seems to be uncorrelated to pressure but seems to be inversely correlated to periodic breathing (my AHI seems to be lower when my PB is higher).
It's not uncommon for AHI to be more severe in REM. Nor is it uncommon for PB to be largely NREM based. So that inverse correlation just may be a reflection of how much time you spend in NREM versus REM.

Tom W wrote: Is my glottis reacting to the higher pressures by closing and negating the effects of the higher pressure?
Dr. Rapoport may have described an important characteristic of your obstructive component in that second link presented above:
Discussing ventilatory instability, David M. Rapoport, M.D. wrote: Instability, we are now told, contributes importantly to the underlying pathophysiology of obstructive as well as central sleep apnea, at least in some patients.
http://www.ncbi.nlm.nih.gov/pmc/articles/PMC3014238/


bump for others

User avatar
JohnBFisher
Posts: 3821
Joined: Wed Oct 14, 2009 6:33 am

Re: Central Apneas appear with the start of BiPAP Treatment

Post by JohnBFisher » Wed May 02, 2012 12:08 pm

First, let me offer my apologies for my delay in responding to this. Pugsy pointed it out to me ... but I've had a pretty busy period at home. We recently adopted a new rescue dog. He's about one year old and keeping us busy - with training and adjusting to one another.

The good news is that it appears he will work out and become a service dog for me .. at least at home .. and possibly everywhere. He's VERY friendly and meets people well. Plus he seems to be driven to try to please.

That's a blessing because it's possible I will need it sooner than I would like. I appear to be developing vocal cord paralysis. All it takes is a busy day and suddenly my vocal cords seem to lack all normal vibration (it feels as if someone gave me something to numb my vocal cords). My speech just has the sibilants, no tone. Sigh! If it's not one thing it's another.

Add to that, I seem to get "weak" .. feel as if I must REALLY concentrate .. if I have a busy day. It's almost as if I've had too much to drink, needing to do that concentration to get things done .. but without the fuzzy headed feeling. Yuck!

So, my apologies on my delay. I've had more fun that I would like.

Now, let's tackle your post.

First, Tom W, as you might know, Pugsy and -SWS are two of the best members of the forum. They are offering good information for you. I'm glad to see you were in very capable hands.

Next, let's tackle the BiPAP versus CPAP versus ASV treatment. As you've noted, as your pressure increases the central apneas tend to increase. You are reacting to the pressure in a negative manner. This type of apnea is known as Complex Sleep Apnea Syndrome (ComplexSA). Until recently it was not well understood or accepted by most doctors. Most now accept that it exists (based on plenty of studies). And most also recognize the gold standard for treating ComplexSA is the use of an ASV unit to provide therapeutic Positive Airway Pressure. It beats out older techniques, such as the Spontaneous/Timed BiPAP unit (BiPAP S/T) and the addition of supplemental oxygen to the xPAP therapy.

You note:
Tom W wrote:... I go back in 2 weeks and even if he prescribes an ASV machine to reduce/eliminate the Centrals I’ll still wind up with an AHI close to or just over 10. Is this the best I can expect? ...
The fact that you then go on to note that your glottis seems to close at the higher pressure may well interfere with your xPAP therapy (regardless of the type). However, it may not be all that bad. It's possible that is mostly positional (on your back).

So, you might want to explore if there is a position where that is not as much of a problem.

You should also question the use of just BiPAP to address the central sleep apnea.

Do NOT settle for "this is the best I can expect"!!!! Untreated central sleep apnea is just as deadly (literally) as obstructive sleep apnea. I had a hard time getting my doctors to listen to it. It took me breaking down in tears before I could get a doctor to listen that I had HORRIBLE problems with sleep onset central sleep apnea. (I would stop breathing so long - over a minute - that my BiPAP unit would think I was no longer attached and turn itself off !!! But because it was sleep onset, it was not "scored" ... At more than a minute, WHO GIVES A TINKERS DAMN when it occurs !!!.) Unfortunately, by that time the uncontrolled high blood pressure had damaged my kidneys. I now have to be fairly careful about my diet and medications. Gee THANKS!

So, I repeat. Do NOT settle for "this is the best I can expect"!!!! Untreated central sleep apnea is deadly.

Now, I don't want you to panic about this. There are always tweaks that can be made to the therapy that will make it work. For example, as if the glottis closing is positional (you might have noted that as well). So, you might need to pin the tennis balls to the back of a T-Shirt to help stay off your back.

Also, if you find the BiPAP does NOT address the central apneas, be CERTAIN to INSIST on a trial of ASV. This will require an ASV titration study. Essentially, they will titrate to eliminate the obstructive sleep apnea and make certain the high pressure is not too much for you.

You might be wondering if the high pressure is bad with Complex Sleep Apnea. In fact, it is not. Because the unit is NOT attempting to clear an obstruction. Instead it is helping to continue your respiration even when you fail to breathe. You see, Central sleep apnea (which is a part of Complex Sleep Apnea) results from an overshoot / undershoot cycle. That is for whatever reason (in your case the Positive Airway Pressure itself) you stop breathing as you should. This is the initial UNDERSHOOT side of the cycle. When that happens, the CO2 builds up in the blood stream until your body finally kicks in and restarts respiration. However, since the CO2 has built up abnormally high you start to slightly hyperventilate and blow off too much CO2. This is the OVERSHOOT side of the cycle. And unfortunately, it sets the stage for another UNDERSHOOT side, since you often will blow off too much CO2. This also leads to your periodic breathing (which your sleep studies and equipment have observed).

To break this repetitive cycle, the ASV quickly increases the pressure until it helps ascertain respiration even when you are not breathing as you should. This respiration allows your body to blow off enough CO2 so that you do not kick into the OVERSHOOT side of the cycle. Because the unit adapts to your breathing style and includes the timed response, it is known as Adaptive Servo Ventilation (ASV). For what it's worth, I note in my ICE app on my phone and in my MedicAlert emergency health record that I use a ventilator at night. You can not assume that I will breathe as I should when I am not conscious.

Yes, it can take a bit of tweaking to get ASV to work well with ComplexSA, but it does work. You are definitely not alone here. So, don't loose hope.

Anyway, I hope that helps.

_________________
Mask: Quattro™ FX Full Face CPAP Mask with Headgear
Additional Comments: User of xPAP therapy for over 20 yrs. Resmed & Respironics ASV units with EEP=9cm-14cm H2O; PSmin=4cm H2O; PSmax=15cm H2O; Max=25cm H2O
"I get up. I walk. I fall down. Meanwhile, I keep dancing” from Rabbi Hillel
"I wish to paint in such a manner as if I were photographing dreams." from Zdzisław Beksiński

Tom W
Posts: 223
Joined: Thu Sep 01, 2011 4:59 am
Location: West Central Florida

Re: Central Apneas appear with the start of BiPAP Treatment

Post by Tom W » Wed May 02, 2012 12:38 pm

Thank You Everyone!

I have several questions but want to take some additional time to review the information provided first.

apnea2142
Posts: 38
Joined: Sat Nov 26, 2011 2:38 am

Re: Central Apneas appear with the start of BiPAP Treatment

Post by apnea2142 » Thu May 03, 2012 4:42 am

I've read on forums that higher pressures help treat centrals and technicians will raise the pressure on a titration study in response to centrals (I think it's IPAP)
anyway sometimes when I set my asv max PS to 10 I wake up feeling terrible instead of 15

Tom W
Posts: 223
Joined: Thu Sep 01, 2011 4:59 am
Location: West Central Florida

Re: Central Apneas appear with the start of BiPAP Treatment

Post by Tom W » Thu May 03, 2012 7:39 am

JohnBFisher wrote:
The good news is that it appears he will work out and become a service dog for me .. at least at home .. and possibly everywhere. He's VERY friendly and meets people well. Plus he seems to be driven to try to please.
Best of luck with the new dog. He seems to be a good one.
JohnBFisher wrote:
You see, Central sleep apnea (which is a part of Complex Sleep Apnea) results from an overshoot / undershoot cycle. That is for whatever reason (in your case the Positive Airway Pressure itself) you stop breathing as you should. This is the initial UNDERSHOOT side of the cycle. When that happens, the CO2 builds up in the blood stream until your body finally kicks in and restarts respiration. However, since the CO2 has built up abnormally high you start to slightly hyperventilate and blow off too much CO2. This is the OVERSHOOT side of the cycle. And unfortunately, it sets the stage for another UNDERSHOOT side, since you often will blow off too much CO2. This also leads to your periodic breathing (which your sleep studies and equipment have observed).
My number of centrals has increased quite a bit since being on BiPAP but other than short periods on APAP 18 cm/H20 is the highest pressure that I've been prescribed.

Are these what you would consider examples of undershoot / overshoot cycles?

Image

Image

Image
Last edited by Tom W on Thu May 03, 2012 11:21 am, edited 4 times in total.

User avatar
JohnBFisher
Posts: 3821
Joined: Wed Oct 14, 2009 6:33 am

Re: Central Apneas appear with the start of BiPAP Treatment

Post by JohnBFisher » Thu May 03, 2012 7:49 am

apnea2142 wrote:I've read on forums that higher pressures help treat centrals and technicians will raise the pressure on a titration study in response to centrals (I think it's IPAP) anyway sometimes when I set my asv max PS to 10 I wake up feeling terrible instead of 15
Tom W, unlike almost all of the ASV users, this poster has been unable to attain good therapy due to many changes without acting as if it were a controlled experiment. I recommend for most folks that they work with their doctor if they want to change their pressure. There may be information of which your doctor is aware that would make him want to shy away from making the change. Even with over 20 years of experience, I do *exactly* that. And if I do any experimentation with pressures, I change only one thing .. monitor the data for a while .. record the results .. and then I discuss it with my doctor.

In response to your post:
Tom W wrote:Are these what you would consider examples of undershoot / overshoot cycles?
Yes. You can see the waxing and waning of your breathing followed by the central apnea. In fact, it's almost textbook perfect example of periodic breathing.

I am not attempting to alarm you, but you will find when you look for more information on periodic breathing that there are a lot of references to Cheyne-Stokes Respiration (CSR), which is indicative of heart issues. If you have been given a good bill of health in that respect, then this is just how you respond to the higher pressure. But if you have other factors (weight, age, diabetes, high blood pressure, etc) that might make you want to check for heart issues, then please do discuss this with your doctors. However, it is not at all unusual for folks with Central Sleep Apneas (or ComplexSA) to exhibit periodic breathing throughout the night.

Hope that helps.

_________________
Mask: Quattro™ FX Full Face CPAP Mask with Headgear
Additional Comments: User of xPAP therapy for over 20 yrs. Resmed & Respironics ASV units with EEP=9cm-14cm H2O; PSmin=4cm H2O; PSmax=15cm H2O; Max=25cm H2O
"I get up. I walk. I fall down. Meanwhile, I keep dancing” from Rabbi Hillel
"I wish to paint in such a manner as if I were photographing dreams." from Zdzisław Beksiński

User avatar
JohnBFisher
Posts: 3821
Joined: Wed Oct 14, 2009 6:33 am

Re: Central Apneas appear with the start of BiPAP Treatment

Post by JohnBFisher » Thu May 03, 2012 8:24 am

By the way, here's an example of my periodic breathing. I do not have it occur as often or as pronounced as you do. My periodic breathing occurs less than 2% of the night. As you can see, it is a lot more "ragged". Maybe the computer can tell it's periodic, but it's not all that obvious to the rest of us:

Image

Hope that helps.

_________________
Mask: Quattro™ FX Full Face CPAP Mask with Headgear
Additional Comments: User of xPAP therapy for over 20 yrs. Resmed & Respironics ASV units with EEP=9cm-14cm H2O; PSmin=4cm H2O; PSmax=15cm H2O; Max=25cm H2O
"I get up. I walk. I fall down. Meanwhile, I keep dancing” from Rabbi Hillel
"I wish to paint in such a manner as if I were photographing dreams." from Zdzisław Beksiński

Tom W
Posts: 223
Joined: Thu Sep 01, 2011 4:59 am
Location: West Central Florida

Re: Central Apneas appear with the start of BiPAP Treatment

Post by Tom W » Fri May 04, 2012 6:56 am

-SWS wrote:
Tom W wrote: My AHI seems to be uncorrelated to pressure but seems to be inversely correlated to periodic breathing (my AHI seems to be lower when my PB is higher).
It's not uncommon for AHI to be more severe in REM. Nor is it uncommon for PB to be largely NREM based. So that inverse correlation just may be a reflection of how much time you spend in NREM versus REM.
I might have stumbled upon this relationship by chance. I've found on nights that I often wake for mask leaks my AHI tends to be lower. I assume the constant arousals reduce the amount of REM sleep I experience during the night. I do notice that on those nights with very low (for me) AHI I usually feel like a train wreck the next day.

[
-SWS wrote:
Tom W wrote: Is my glottis reacting to the higher pressures by closing and negating the effects of the higher pressure?
Dr. Rapoport may have described an important characteristic of your obstructive component in that second link presented above:
Discussing ventilatory instability, David M. Rapoport, M.D. wrote: Instability, we are now told, contributes importantly to the underlying pathophysiology of obstructive as well as central sleep apnea, at least in some patients.


http://www.ncbi.nlm.nih.gov/pmc/articles/PMC3014238/

bump for others
I did stumble across this article a few months back:

http://jap.physiology.org/content/79/1/186.abstract

In the following flow patterns you can see both OA's and CA's that look pretty similarly shaped. I assume from the response of the pressure pulses that sometimes the airway is clear and sometimes it is closed. I'm beginning to think that perhaps the only difference is the position of my glottis. I don't really have any experience in this field besides trying to improve my own sleep but in these examples even the OA's seem to be central in nature. The apparent (to me) deliberate starting and stopping resembles a pattern of breath holding that I occasionally do during the day but obviously I can't hold my breath that long/often when awake. Just throwing some thoughts out for discussion. Feel free to tell me I'm wrong, crazy, etc..

Image

Image

-SWS
Posts: 5301
Joined: Tue Jan 11, 2005 7:06 pm

Re: Central Apneas appear with the start of BiPAP Treatment

Post by -SWS » Fri May 04, 2012 10:37 pm

Tom W wrote: I did stumble across this article a few months back:

http://jap.physiology.org/content/79/1/186.abstract
The article suggests that PAP treatment induces glottis narrowing in some patients. They don't say what the epiglottis is doing. Regardless, there you have a proposed mechanism that might explain your obstructive apneas at higher pressures.

Here's another document suggesting that unstable ventilation causes or triggers obstructive events in some patients:
David P. White, M.D. wrote: hypoxia-induced periodic breathing can lead to obstructive hypopneas and apneas in individuals in whom only snoring was observed under normoxic conditions
http://ajrccm.atsjournals.org/content/172/11/1363.full
Tom W wrote: My AHI seems to be uncorrelated to pressure
Combine your observations with an established link between lacking ventilatory control and secondary obstruction, and I think you have incentive to pursue treatment that will also stabilize your ventilatory control. You just might have a primary obstructive component at atmospheric pressure---but another obstructive component emerging as secondary to poor ventilatory control at higher CPAP pressures. This proposed mechanism of compounded obstructive dyscontrol might explain at least some cases in which treatment pressures will not inversely correlate with obstructive AHI. Perhaps straightening out your ventilatory control at moderate-to-high pressures will also eliminate this proposed secondary/emergent obstructive component.
Tom W wrote: In the following flow patterns you can see both OA's and CA's that look pretty similarly shaped. I assume from the response of the pressure pulses that sometimes the airway is clear and sometimes it is closed. I'm beginning to think that perhaps the only difference is the position of my glottis. I don't really have any experience in this field besides trying to improve my own sleep but in these examples even the OA's seem to be central in nature. The apparent (to me) deliberate starting and stopping resembles a pattern of breath holding that I occasionally do during the day but obviously I can't hold my breath that long/often when awake. Just throwing some thoughts out for discussion. Feel free to tell me I'm wrong, crazy, etc..
Well, your own observations and analysis might prove to be more valuable than your doctors. At this point one good avenue of treatment-exploration seems to be that of actively addressing BOTH the central and obstructive components. Good luck to you my friend!

Tom W
Posts: 223
Joined: Thu Sep 01, 2011 4:59 am
Location: West Central Florida

Requirements for ASV Tritation Needed

Post by Tom W » Thu May 10, 2012 8:23 am

As of today my 20 day average AHI is 22.12 on BiPAP at my current settings (18/12) and I'm scheduled to see my new doctor who prescribed this machine/settings this coming Monday.

From what I've seen online I don't qualify for an ASV machine since my Centrals only make up 43 percent of my total apneas and it needs to be greater than 50 percent to meet the requirements for Complex Sleep Apnea (CSA). I do however meet all the other requirements for CSA.

I thought I remember seeing something (which I can 't find now that I need it) which stated that an ASV machine could/would be approved on a trial basis if an ASV tritation was perfomed and it showed significant improvement over current CPAP/BiPAP usage results.

Are my current stats 'bad' enough to be considered a failure on BiPAP?

Since I don't meet the CSA criteria is there any other way to be considered for an ASV machine?

Thanks again to everyone....