How to View Pulse-Ox Data in SleepyHead

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
User avatar
Papit
Posts: 844
Joined: Sun Dec 18, 2011 4:47 pm
Location: USA CA

How to View Pulse-Ox Data in SleepyHead

Post by Papit » Mon Apr 23, 2012 2:54 pm

Here's a clarified operating procedure that should help newer users incorporate their CMS-50 pulse oximetry data, time-synchronized among the SleepyHead graphs.

How to View Pulse-Ox Data in SleepyHead

1. Open SleepyHead and click Oximetry tab at the top. Click the button in the lower right to ‘Open .spo/R File’. Select the session file you want to view and click ‘Open’. You should see your red pulse and blue oximetry charts vs. clock time.
2. Click the Save button just left of ‘Open .spo/R File’ and confirm ‘Yes’ that you want to save the data. (Disregard the ‘No Oximetry Data’ screen.)
3. Click ‘Data’ on the top main menu line, then click ‘Rebuild Oximetry Indices’.
4. Close and then restart SleepyHead to view the sleep session graphs including pulse oximetry aligned in the same time scale.
5. Click the Daily tab button in the SleepyHead Navigation bar on the right. The dates that appear in red on the calendar have pulse-oximtry data. Click on a red date to see that session's pulse-ox charts and other graphs for that day. (Steps 1 – 4 above need to have previously been performed to see each date's pulse-ox graphs.)
( If the clock time settings are too dissimilar you will see a notice that they cannot be linked. Resychronize the current time on the CMS-50 so its time setting and that on the xpap machine are the same or very close. )

_________________
Mask: DreamWear Nasal CPAP Mask with Headgear
Additional Comments: Machine: AirCurve 10 ASV (37043), Software:ResScan 5.7.0.9477, SleepyHead V1.00BETA2, Oximeter:CMS-50i

User avatar
Breathing
Posts: 43
Joined: Tue May 01, 2012 6:52 pm

Re: How to View Pulse-Ox Data in SleepyHead

Post by Breathing » Fri Jun 08, 2012 8:15 pm

1. Open SleepyHead and click Oximetry tab at the top. Click the button in the lower right to ‘Open .spo/R File’. Select the session file you want to view and click ‘Open’
There are no sessions. I need to somehow get the data from the device to the computer.

_________________
Mask: Zest Nasal CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments:  SleeyHead software. v0.9.1(beta)
http://www.chinupstrip.com

I have sleep apnea, but I don't snore. Never have. (I am also not male, obese or hypertensive)
I believe the screening criteria for sleep apnea is missing a LOT of people, who have a lot of unnecessary OSA related disease.

User avatar
tigrpal
Posts: 49
Joined: Sat May 05, 2012 4:41 pm
Location: Westchester, IL

Re: How to View Pulse-Ox Data in SleepyHead

Post by tigrpal » Fri Jun 08, 2012 8:28 pm

Make sure your device is on, recording off, and in the menu mode(menu showing on the display).

Also, make sure that when you stop recording in the morning you don't accidentally start recording again. It only stores data from one session. Every time you start recording it erases the previous data.

I learned these lessons the hard way.

User avatar
b360155
Posts: 163
Joined: Sat Oct 22, 2011 7:46 pm
Location: North Richland Hills, TX

Re: How to View Pulse-Ox Data in SleepyHead

Post by b360155 » Fri Jun 08, 2012 8:38 pm

Breathing wrote:
1. Open SleepyHead and click Oximetry tab at the top. Click the button in the lower right to ‘Open .spo/R File’. Select the session file you want to view and click ‘Open’
There are no sessions. I need to somehow get the data from the device to the computer.
Here's some excellent instructions that should get you going:
http://www.apneaboard.com/wiki/index.ph ... SleepyHead

_________________
Mask: Quattro™ FX Full Face CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Using ResScan Version 3.16 Software and Sleepyhead 0.9.2

User avatar
Breathing
Posts: 43
Joined: Tue May 01, 2012 6:52 pm

Re: How to View Pulse-Ox Data in SleepyHead

Post by Breathing » Fri Jun 08, 2012 8:44 pm

That looks like just what I need.... a step by step guide.. Thanks so much. I am going to try again in the morning and see what kind of report I can get.

I appreciate the help.


_________________
Mask: Zest Nasal CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments:  SleeyHead software. v0.9.1(beta)
http://www.chinupstrip.com

I have sleep apnea, but I don't snore. Never have. (I am also not male, obese or hypertensive)
I believe the screening criteria for sleep apnea is missing a LOT of people, who have a lot of unnecessary OSA related disease.

User avatar
Papit
Posts: 844
Joined: Sun Dec 18, 2011 4:47 pm
Location: USA CA

How to View Pulse-Ox Data in SleepyHead

Post by Papit » Sat Jun 09, 2012 5:02 am

The followiing is a much more updated set of instructions I think will help:

Sleepyhead CMS-50 Pulse-Oximetry Graphs - Time Syncronized

Here's the operating procedure I use. I hope it helps new and regular CMS-50E/F Pulse-Oximeter users experience consistent ease of use and convenient access to their pulse-ox graphs both in the provided SpO2 Review software and in SleepyHead.

(Note: If you continue to have difficulty after using this procedure, go to 'Manuals/Downloads' at http://www.pulseoxstore.com/Manuals-Downloads.html to download and reinstall the latest software .)

CMS-50E/F Pulse Oximeter operating procedure
A. Record your Pulse-Ox data over night
1. Turn ON your Pulse-Oximeter after masking up at night.
2. Turn Record to ON.
3. Insert a finger into the sensor and sleep with it all night.
4. Turn Record OFF when you awaken in the morning.

B. 'Import’ your Xpap data to Sleepyhead
1. If you use SleepyHead software, remove the SD memory card from your xpap machine and insert it into the SD card reader on your computer.
2. Open SleepyHead and click ‘Import’ to upload the sleep session data from your SD card.
3. Click ‘My Computer’ on your desktop screen, right click on the drive containing the SD card, click ‘Eject’ and remove the card.
4. Return the SD card to your xpap machine and reinstall it.


C. Upload and view your Pulse-Ox data
1. Use the provided USB cable to connect the Pulse-Oximeter to your computer.
2. Open SpO2 Review software.
3. Click ‘Not Active’ button to open a New Session.
(Click ‘Cancel’ and repeat this Step if ‘New Session’ button on top line is not shown selected.)
4. Enter a file name (e.g., today’s date) and click OK. A message, "Device Connected: waiting for data…," should be displayed.
(If you see a message, "Can’t open COM 6", click OK, disconnect the USB cable from the Pulse-Ox, then reconnect it and go back to Step 3.)
5. If Record is ON, turn Record OFF on the Pulse-Ox.
6. Turn Upload ON on the Pulse-Ox.
7. Wait (up to a minute) while ‘Receiving data, Please wait….’ is displayed.
8. View the Pulse-Ox graphs. Click ‘M’ on bottom line and set the Time scale by changing 'HH: 1’ to the number of whole hours you slept last night rounded up by one. Click OK.

D. View the Pulse-Ox Reports and Graphs in the SpO2 Review CMS-50 software
9. Click the Print Settings button in the lower far right, select a report (e.g.,'OxiMetry Report'), and click ‘Preview’. Print it for your file and/or for your doctor, or just view it. For easier viewing change ‘ % ‘ to 100% or higher (right-most button at top of screen).
10. Press and hold Power off to turn OFF the Pulse-Ox (“Bye Bye!”).
11. Recharge Pulse-Oximeter.

D. View the Pulse Oximetry Reports and Graphs Time-aligned in SleepyHead software
1. Open SleepyHead and click the 'Oximetry' tab at the top. Click the ‘Open .spo/R File’ button in the lower right. Select the sleep file you want to view and click ‘Open’. You should see your red pulse and blue oximetry charts vs. clock time.
2. Click the Save button just left of ‘Open .spo/R File’ button and confirm ‘Yes’ that you want to save the data. (Disregard the ‘No Oximetry Data’ screen.)
3. Click ‘Data’ on the top main menu line, then click ‘Rebuild Oximetry Indices’.
4. Exit out, then restart SleepyHead and click the 'Daily' tab to view the graphs, now including pulse and oximetry, all aligned to the same time scale.
5. Click the 'Daily' button in the SleepyHead Navigation bar on the right. The dates that appear in red on the calendar will display time aligned pulse-oximetry data and graphs. Click on a red date to see the pulse-ox charts for your sleep on that date along with the standard SleepyHead graphs.

(All the steps above need to have previously been performed to see a date's pulse-ox graphs and data in SleepyHead. If the clock time setting on the Pulse-Ox and on your xpap machine do not match closely enough, you will see a notice that the pulse-ox data cannot be linked with SleepyHead. Re-sychronize the current time on the Pulse-Ox so its time setting and that on the xpap machine are the same or very close. Note that fractured sleep sessions are listed in the lower left and each can be toggled ON and OFF.)

_________________
Mask: DreamWear Nasal CPAP Mask with Headgear
Additional Comments: Machine: AirCurve 10 ASV (37043), Software:ResScan 5.7.0.9477, SleepyHead V1.00BETA2, Oximeter:CMS-50i

User avatar
Breathing
Posts: 43
Joined: Tue May 01, 2012 6:52 pm

Re: How to View Pulse-Ox Data in SleepyHead

Post by Breathing » Sat Jun 09, 2012 6:05 am

Code: Select all

C. Upload and view your Pulse-Ox data
1.	Use the provided USB cable to connect the Pulse-Oximeter to your computer. 
2.	Open SpO2 Review software.
3.	Click ‘Not Active’ button to open a New Session. 
(Click ‘Cancel’ and repeat this Step if ‘New Session’ button on top line is not shown selected.)
4.	Enter a file name (e.g., today’s date) and click OK. A message, "Device Connected: waiting for data…," should be displayed.

You say to have it on "menu".... is that a screen that says, "Settings, Alarm, Record off, Exit"?

I've got the "device connected: waiting for data" message.

still waiting for data

_________________
Mask: Zest Nasal CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments:  SleeyHead software. v0.9.1(beta)
http://www.chinupstrip.com

I have sleep apnea, but I don't snore. Never have. (I am also not male, obese or hypertensive)
I believe the screening criteria for sleep apnea is missing a LOT of people, who have a lot of unnecessary OSA related disease.

User avatar
Breathing
Posts: 43
Joined: Tue May 01, 2012 6:52 pm

Re: How to View Pulse-Ox Data in SleepyHead

Post by Breathing » Sat Jun 09, 2012 7:26 am

Got it! The USB cable won't read through a manifold. Plugged it directly in and it's working.


_________________
Mask: Zest Nasal CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments:  SleeyHead software. v0.9.1(beta)
http://www.chinupstrip.com

I have sleep apnea, but I don't snore. Never have. (I am also not male, obese or hypertensive)
I believe the screening criteria for sleep apnea is missing a LOT of people, who have a lot of unnecessary OSA related disease.

User avatar
Papit
Posts: 844
Joined: Sun Dec 18, 2011 4:47 pm
Location: USA CA

Re: How to View Pulse-Ox Data in SleepyHead

Post by Papit » Sat Jun 09, 2012 8:14 am

Way to go! Sleep well.

_________________
Mask: DreamWear Nasal CPAP Mask with Headgear
Additional Comments: Machine: AirCurve 10 ASV (37043), Software:ResScan 5.7.0.9477, SleepyHead V1.00BETA2, Oximeter:CMS-50i

-tim
Posts: 710
Joined: Sat Dec 03, 2011 7:46 pm
Location: Melbourne Australia

Re: How to View Pulse-Ox Data in SleepyHead

Post by -tim » Sat Jun 09, 2012 9:22 am

These directions work for the CMS-50. Not the over priced resmed SpO2 device... There is a way to get it to read the data but I haven't figured it out yet.

_________________
Mask: AirFit™ N20 Nasal CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Resmeds overpriced SpO2

User avatar
b360155
Posts: 163
Joined: Sat Oct 22, 2011 7:46 pm
Location: North Richland Hills, TX

Re: How to View Pulse-Ox Data in SleepyHead

Post by b360155 » Sun Jun 10, 2012 8:15 am

Breathing wrote:Got it! The USB cable won't read through a manifold. Plugged it directly in and it's working.
Good job!
Papit wrote:Way to go! Sleep well.
Just one clarification I wanted to add. The link I posted to the Wiki article that explained how to synchronize the data in your Sleepyhead graphs was specifically for the Contec CMS-50D Plus model pulse oximeter.

This step does not work when trying to synchronize the data with the CMS-50D Plus model.
"3. Click ‘Data’ on the top main menu line, then click ‘Rebuild Oximetry Indices’"

The CMS-50D Plus model does not have a built in time stamp feature like the CMS-50E and above models. It only keeps track of the amount of time that the recording function is on. So, in order to synchronize your graphs in Sleepyhead if you have the CMS-50D Plus, you need to follow the steps in the link in my earlier post.