Losing data on S9 machine

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
User avatar
JeffH
Posts: 2151
Joined: Mon Dec 04, 2006 3:44 pm

Losing data on S9 machine

Post by JeffH » Mon Mar 12, 2012 8:02 am

After a little over 300 days of use I'm starting to lose data on my S9 ASV machine. I took a pencil eraser and tried cleaning the contacts on the SD card this morning. Any other ideas?

JeffH

_________________
Mask
Additional Comments: Equipment isn't correct, S9 ASV w/H5i

User avatar
NateS
Posts: 1716
Joined: Wed Dec 14, 2011 8:53 pm
Location: Kaatskill Mts-Washington Irving

Re: Losing data on S9 machine

Post by NateS » Mon Mar 12, 2012 11:52 am

JeffH wrote:After a little over 300 days of use I'm starting to lose data on my S9 ASV machine. I took a pencil eraser and tried cleaning the contacts on the SD card this morning. Any other ideas?

JeffH
These cards are so inexpensive now on Amazon, why not play it safe and just buy a new card and copy the data from your existing card before it becomes completely unreliable?

Nate

_________________
Mask: DreamWear Nasal CPAP Mask with Headgear
Additional Comments: ResMed AirCurve 10 ASV; Dreamwear Nasal Mask Original; CPAPMax Pillow; ResScan & SleepyHead
Central sleep apnea AHI 62.6 pre-VPAP. Now 0 to 1.3
Present Rx: EPAP: 8; IPAPlo:11; IPAPHi: 23; PSMin: 3; PSMax: 15
"I've had a perfectly wonderful evening, but this wasn't it." —Groucho Marx

User avatar
Lizistired
Posts: 2835
Joined: Tue Dec 14, 2010 10:47 pm
Location: Indiana

Re: Losing data on S9 machine

Post by Lizistired » Mon Mar 12, 2012 12:17 pm

Hi Jeff, What do you mean by "losing data"? One night, all night?...suddenly?

_________________
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Swift FX sometimes, CMS-50F, Cervical collar sometimes, White noise, Zeo... I'm not well, but I'm better.

User avatar
robysue
Posts: 7520
Joined: Sat Sep 18, 2010 2:30 pm
Location: Buffalo, NY
Contact:

Re: Losing data on S9 machine

Post by robysue » Mon Mar 12, 2012 1:36 pm

JeffH wrote:After a little over 300 days of use I'm starting to lose data on my S9 ASV machine. I took a pencil eraser and tried cleaning the contacts on the SD card this morning.
JeffH:

Asking for clarification:

You say you are using a is using a ResMed S8 series ASV. But your signature shows a Resmed S8 machine. Can you clarify which is correct? If you signature is wrong, can you fix it?

Which machine you have is important:

The older S8 machine uses a proprietary card (and card reader). And replacing the card will be difficult and probably a bit expensive.

The newer S9 machine uses a standard 1 or 2 Gig SD card. You can find replacement cards for cheap at places like your local Walmart or Target. You just need to be sure the card is an SD card and not one of the newer SDHC cards.

_________________
Machine: DreamStation BiPAP® Auto Machine
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear
Additional Comments: PR System DreamStation and Humidifier. Max IPAP = 9, Min EPAP=4, Rise time setting = 3, minPS = 3, maxPS=5

User avatar
Pugsy
Posts: 65326
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Losing data on S9 machine

Post by Pugsy » Mon Mar 12, 2012 2:08 pm

Robysue...look in Jeff's comments. He has a S9 machine. I don't think that cpap.com carries the one he has so there is no way to link to it so people with this machine just add it to the comments. It probably would be clearer if people would then just omit the machine link and that would force us to look in comments section. Lessen confusion that way.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
JeffH
Posts: 2151
Joined: Mon Dec 04, 2006 3:44 pm

Re: Losing data on S9 machine

Post by JeffH » Mon Mar 12, 2012 6:03 pm

Lizistired wrote:Hi Jeff, What do you mean by "losing data"? One night, all night?...suddenly?
Losing it randomly. One line of data for one item will only show up for half the night. I almost never gets the whole night in the summary data, but lately has been losing data in the detailed part of the data as well.

JeffH

_________________
Mask
Additional Comments: Equipment isn't correct, S9 ASV w/H5i

User avatar
JeffH
Posts: 2151
Joined: Mon Dec 04, 2006 3:44 pm

Re: Losing data on S9 machine

Post by JeffH » Mon Mar 12, 2012 6:04 pm

NateS wrote:
JeffH wrote:After a little over 300 days of use I'm starting to lose data on my S9 ASV machine. I took a pencil eraser and tried cleaning the contacts on the SD card this morning. Any other ideas?

JeffH
These cards are so inexpensive now on Amazon, why not play it safe and just buy a new card and copy the data from your existing card before it becomes completely unreliable?

Nate
First happened when we were on vacation last year. I remember going and buying a card then. I'll have to see if I can find it laying around here.

Thanks

JeffH

_________________
Mask
Additional Comments: Equipment isn't correct, S9 ASV w/H5i

User avatar
robysue
Posts: 7520
Joined: Sat Sep 18, 2010 2:30 pm
Location: Buffalo, NY
Contact:

Re: Losing data on S9 machine

Post by robysue » Mon Mar 12, 2012 7:01 pm

Pugsy wrote:Robysue...look in Jeff's comments. He has a S9 machine. I don't think that cpap.com carries the one he has so there is no way to link to it so people with this machine just add it to the comments. It probably would be clearer if people would then just omit the machine link and that would force us to look in comments section. Lessen confusion that way.
Yes, Jeff says he's using an S9, but his signature shows an S8. That's why I'm confused about what machine he has.

Also he has 300+ days of data. And the S9 VPAP only came out about a year ago at most. And the AVS version was later yet, if I recall correctly.

So Jeff, change your signature if you're using an S9 machine please?

_________________
Machine: DreamStation BiPAP® Auto Machine
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear
Additional Comments: PR System DreamStation and Humidifier. Max IPAP = 9, Min EPAP=4, Rise time setting = 3, minPS = 3, maxPS=5

User avatar
JeffH
Posts: 2151
Joined: Mon Dec 04, 2006 3:44 pm

Re: Losing data on S9 machine

Post by JeffH » Mon Mar 12, 2012 9:47 pm

robysue wrote:
Pugsy wrote:Robysue...look in Jeff's comments. He has a S9 machine. I don't think that cpap.com carries the one he has so there is no way to link to it so people with this machine just add it to the comments. It probably would be clearer if people would then just omit the machine link and that would force us to look in comments section. Lessen confusion that way.
Yes, Jeff says he's using an S9, but his signature shows an S8. That's why I'm confused about what machine he has.

Also he has 300+ days of data. And the S9 VPAP only came out about a year ago at most. And the AVS version was later yet, if I recall correctly.

So Jeff, change your signature if you're using an S9 machine please?
Its an S9 and when I filled it out CPAP.com didn't sell those, as was mentioned by previous poster.

_________________
Mask
Additional Comments: Equipment isn't correct, S9 ASV w/H5i

TimWK
Posts: 43
Joined: Wed Jan 12, 2011 8:25 pm
Location: AZ

Re: Losing data on S9 machine

Post by TimWK » Mon Mar 12, 2012 10:03 pm

Last month I lost 1 weeks data in a strange way. While reading the data it showed i had no events but all the other data was there. The next day the same, in fact it was the same for a week. Upon looking at the detailed data you could see that there were events but no flags. I removed the power from my S9 and rebooted it the next morning the event flags were back. I now reboot my S9 weekly.
Before this happened my S9 was not unplugged for 7 months.
Best Regards

User avatar
Lizistired
Posts: 2835
Joined: Tue Dec 14, 2010 10:47 pm
Location: Indiana

Re: Losing data on S9 machine

Post by Lizistired » Mon Mar 12, 2012 10:08 pm

I think I would reboot, and/or a new card and see if it resolves. If not, RESMED is on facebook. Ask Gil on there.

_________________
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Swift FX sometimes, CMS-50F, Cervical collar sometimes, White noise, Zeo... I'm not well, but I'm better.

User avatar
JeffH
Posts: 2151
Joined: Mon Dec 04, 2006 3:44 pm

Re: Losing data on S9 machine

Post by JeffH » Wed Mar 14, 2012 7:02 am

New card worked all night. I also slept thru the night for the first time in a long time. Who knew that a new SD card would help you sleep better...LOL

JeffH

_________________
Mask
Additional Comments: Equipment isn't correct, S9 ASV w/H5i