Sore Nose Bridge

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
User avatar
BigOffshoreDaddy
Posts: 8
Joined: Sun Sep 20, 2009 9:12 am
Contact:

Sore Nose Bridge

Post by BigOffshoreDaddy » Wed Feb 15, 2012 8:48 am

I know that this has been addressed half a million times, but I haven't seen anything like what I have.

Around November, I got a new mask and forgot to wear a band-aid, thus creating a huge sore. It got to the point that I chose to wear the mask over my own comfort, and I created a monster. It took me almost 6 weeks of constant care before it healed.

Two nights ago was the first night that I slept with my machine in 6 weeks. I went through my regular routine (clean face, apply band-aid for protection), put the mask on, and waited for sleep. When I woke up yesterday morning, there was a wonderful looking hematoma right under the band-aid, and it was as if I woke up back in November looking at it for the first time.

I simply cannot go another 6 weeks without therapy. Any tips, tricks, recommendations, equipment changes, and/or suggestions would be most welcome!

_________________
MaskHumidifier

nanwilson
Posts: 3463
Joined: Tue Apr 13, 2010 10:35 am
Location: Southern Alberta

Re: Sore Nose Bridge

Post by nanwilson » Wed Feb 15, 2012 8:54 am

Change your mask to one that does not go on the top of the nose. Take a look at cpap.com at all the choices for full face masks, you will get a good idea on what others have used and like. I use the Innomed Hybrid, it is considered a full face mask as it has nose pillows and a mouth piece.
Started cpap in 2010.. still at it with great results.

User avatar
Maxie
Posts: 331
Joined: Mon Jan 30, 2012 2:54 pm
Location: Lake Norman, North Carolina

Re: Sore Nose Bridge

Post by Maxie » Wed Feb 15, 2012 9:08 am

I had the nose bridge sores also and I changed masks and no problems at all now. I tried a hybrid but eventually went back to my original nasal pillows and all is well now. There are several masks that don't cause nose bridge problems so good luck!

User avatar
LSAT
Posts: 13401
Joined: Sun Nov 16, 2008 10:11 am
Location: SE Wisconsin

Re: Sore Nose Bridge

Post by LSAT » Wed Feb 15, 2012 9:25 am

You may have the forehead bridge adjustment too tight. It should be very loose with the adjustment turned out almost to the end. The silicone should be very light on the upper part of the nose. Have you tried one of the several nose pads from...

http://www.cpapsupplyusa.com/61915-1-Re ... l-Pad.aspx
http://www.remzzzs.com
http://www.padacheek.com
http://www.cpapseal.com/
https://www.cpap.com/productpage/boomer ... -pack.html

Have you read this?
http://maskarrayed.wordpress.com/taming ... e-quattro/
Last edited by LSAT on Wed Feb 15, 2012 10:32 am, edited 1 time in total.

User avatar
SleepingDVader
Posts: 120
Joined: Fri Dec 09, 2011 10:04 am
Location: S. Central Louisiana

Re: Sore Nose Bridge

Post by SleepingDVader » Wed Feb 15, 2012 9:46 am

I've been having fairly good luck lately using the Mirage Liberty hybrid FFM. It uses nasal pillows and has nothing on the bridge of the nose at all. But my primary mask right now is still the Mirage Quattro. I nearly trashed it because of the nose irritation but bought the Boomerang gel pads and they work really well, at least in my case. The gel pad goes over the bridge of the nose and helps with comfort, but also aids in getting rid of leaks in that area too. So now I have no blister on the bridge of my nose and sleep very well with that mask. Good luck and hope this helps.

_________________
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Rx pressure is 20/16, Use papcap w/chinstrap and tape. Sleep number bed and Sobakawa pillow. Also use Swift FX & Quattro FX as required.

User avatar
BigOffshoreDaddy
Posts: 8
Joined: Sun Sep 20, 2009 9:12 am
Contact:

Re: Sore Nose Bridge

Post by BigOffshoreDaddy » Wed Feb 15, 2012 9:52 am

Thanks, everyone, for your input! I've looked at the links, LSAT, and will definitely try what I've seen so far.

_________________
MaskHumidifier

User avatar
b360155
Posts: 163
Joined: Sat Oct 22, 2011 7:46 pm
Location: North Richland Hills, TX

Re: Sore Nose Bridge

Post by b360155 » Wed Feb 15, 2012 12:51 pm

My primary mask these days is the Quattro FX as I'm able to get very little leakage and it's probably the most comfortable ffm I've ever worn. No nose bridge problems at all. Based on some reviews in other threads, I have been trying out a Mirage Quattro for the past couple of weeks. I've found that I've been able to get an excellent seal without much problem at all, however no matter how I adjust the tension on the straps and position the forehead support bar, I still feel a lot of pressure on the bridge of my nose. Just ordered a ResMed Gecko this morning so if that does the trick for me with this mask, I'll report back.

_________________
Mask: Quattro™ FX Full Face CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Using ResScan Version 3.16 Software and Sleepyhead 0.9.2

User avatar
archangle
Posts: 9293
Joined: Sun Mar 27, 2011 11:55 am

Re: Sore Nose Bridge

Post by archangle » Wed Feb 15, 2012 3:41 pm

Did you try any nasal masks before ending up with a full face mask?

_________________
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Also SleepyHead, PRS1 Auto, Respironics Auto M series, Legacy Auto, and Legacy Plus
Please enter your equipment in your profile so we can help you.
Click here for information on the most common alternative to CPAP.
If it's midnight and a DME tells you it's dark outside, go and check for yourself.

Useful Links.

User avatar
Vader
Posts: 1223
Joined: Fri Aug 18, 2006 4:35 pm
Location: Land Of The Free

Re: Sore Nose Bridge

Post by Vader » Wed Feb 15, 2012 4:03 pm

When I regularly used the Mirage Quattro, I had a sore nose bridge far too many times than I'd like to remember.

Putting a few layers of paper tape on the bridge is helpful, and a Padacheek mask liner will also definitely help, but the problem never went away totally until I switched to the Quattro FX.

_________________
Mask
.









Vader

User avatar
napstress
Posts: 462
Joined: Sat Apr 09, 2011 6:59 am

Re: Sore Nose Bridge

Post by napstress » Wed Feb 15, 2012 5:40 pm

I hope you can find either a suitable mask or cushioning. It's such a PITA to be forced off therapy for such a thing! I still get nose sores even on the Quattro FX with multiple interventions! The best I can offer you is how to make the sore go away relatively quickly. I pierce a vitamin E and rub a bit on the spot, leave it on for a few hours, and wipe off before masking up. I jump on this as soon as the sore opens, so it takes only a couple of days to heal. I do this in the mornings. I don't know about evenings: the vit E might make the sore moist, and therefore, vulnerable when the mask rubs on it (maybe even with band-aid).

Does anyone know why it is advised that we not wash the Boomerang with shampoo but an alcohol-based hand sanitizer is OK? I'm assuming soap would also not be OK, right?

Good luck, BOSD.
Epworth Sleepiness Scale: 14
Diagnostic study: overall AHI: 0.2 events/hour; overall RDI: 45 events/hour
Titration study: AHI: 6.1; RDI: 27; CPAP pressures: 5-8cm

Not-tired behind my eyes and with a clear, cool head!

ozze_dollar
Posts: 611
Joined: Wed Jun 08, 2011 5:20 pm
Location: Sydney,Australia

Re: Sore Nose Bridge

Post by ozze_dollar » Wed Feb 15, 2012 7:37 pm

napstress wrote:I hope you can find either a suitable mask or cushioning. It's such a PITA to be forced off therapy for such a thing! I still get nose sores even on the Quattro FX with multiple interventions! The best I can offer you is how to make the sore go away relatively quickly. I pierce a vitamin E and rub a bit on the spot, leave it on for a few hours, and wipe off before masking up. I jump on this as soon as the sore opens, so it takes only a couple of days to heal. I do this in the mornings. I don't know about evenings: the vit E might make the sore moist, and therefore, vulnerable when the mask rubs on it (maybe even with band-aid).

Does anyone know why it is advised that we not wash the Boomerang with shampoo but an alcohol-based hand sanitizer is OK? I'm assuming soap would also not be OK, right?

Good luck, BOSD.
I use a pure soap to wash my nose pads.

User avatar
máirtín
Posts: 254
Joined: Sat Apr 30, 2011 12:01 pm
Location: California Ó Thuaidh; Montsoreau, 49730

Re: Sore Nose Bridge

Post by máirtín » Wed Feb 15, 2012 7:50 pm

I followed all of the advice here without a good result until I visited the dermatologist.
You are using an ASV which limits the mask choices you have and also the ways you can adjust the straps.
Most of the advice here is from members who don't have these restrictions to work within - it's not bad advice, it simply doesn't apply.
Go see a dermatologist.

_________________
Mask: Mirage Quattro™ Full Face CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: I have a Resmed S9 ASV
"Bheadh sé chomh maith agat a rá," a dúirt an Luch Chodlamáin féin, "gurb ionann 'Codlaím nuair a análaím' agus 'Análaím nuair a chodlaím'!"

User avatar
Gekko
Posts: 31
Joined: Wed Mar 31, 2010 12:16 am
Location: Sydney, Australia

Re: Sore Nose Bridge

Post by Gekko » Thu Feb 16, 2012 7:15 pm

Unfortunately, I think changing masks is the only option. I had a Quattro and it was fine for about a year, then suddenly I started to get nose sores! It must have been because the straps were getting old and I could no longer find that sweet spot between no leaks and comfort. I bought a Quattro FX and everything is now OK!

User avatar
b360155
Posts: 163
Joined: Sat Oct 22, 2011 7:46 pm
Location: North Richland Hills, TX

Re: Sore Nose Bridge

Post by b360155 » Sun Feb 19, 2012 9:12 am

I got my ResMed Gecko in the mail and have tried it out the last 2 nights with my Mirage Quattro since this mask was creating excessive pressure on the bridge of my nose. It seems to protect the nose bridge fine but I can't seem to get the mask to seal if I use it. I've taken it off after just a minute or two each night as I get air leaks from where the face cushion transitions from being on the gecko to being on my skin. Tried adjusting straps but nothing seems to stop the leaking. Anybody use this and had success with it? Any secrets to stop it from causing leaks? I'll probably just go back to the Quattro FX which seems to work the best for both seal and comfort for me. Thanks...


Image

_________________
Mask: Quattro™ FX Full Face CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Using ResScan Version 3.16 Software and Sleepyhead 0.9.2

Mary Z
Posts: 1493
Joined: Wed Oct 28, 2009 9:55 am

Re: Sore Nose Bridge

Post by Mary Z » Sun Feb 19, 2012 9:47 am

I have sensitive skin and went through the same thing with the Quattro. I found that once the skin had broken down and required that I not wear the mask for three weeks while it healed that I could not go back to a mask that rested on that area of my nose.
I wear the FitLife Total Face Mask which completely bypasses the nose and works for me. Someone said you are on ASV- could not tell from the photos of your equipment in which case you would have to check with your Doc or DME about this mask. I was allowed to use it during a lab trial of ASV. Good Luck.

_________________
Mask: Wisp Nasal CPAP Mask with Headgear - Fit Pack
Additional Comments: PR System One Remstar BiPap Auto AS Advanced.
Dog is my copilot