Pressure Question

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
User avatar
sleepingatlast
Posts: 29
Joined: Mon Jan 03, 2011 8:48 pm

Pressure Question

Post by sleepingatlast » Wed Feb 01, 2012 7:17 pm

Intellipap Auto. Last 30 days, 95th percentile is 8.5, 90th percentile is 8.0. AHI is 7. Machine minimum pressure set to 4 and max set to 12. I have read that standard rule is 2 above and below 95th percentile. Is that accurate? Should I adjust me machine to 6.5 and 10.5?

User avatar
Pugsy
Posts: 65204
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Pressure Question

Post by Pugsy » Wed Feb 01, 2012 7:33 pm

Sounds like a good plan. The maximum doesn't really matter so much because the machine doesn't go there unless it thinks it needs to. For sure the minimum needs to go up though. Give the machine a bit more of a head start.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
Slartybartfast
Posts: 1633
Joined: Wed Sep 01, 2010 12:34 pm

Re: Pressure Question

Post by Slartybartfast » Wed Feb 01, 2012 8:45 pm

If you haven't done so already, you might want to take a look at how your machine is set up. The EPR and I-rounding and E-rounding settings are important to have set right or you will be fighting the machine. Take a look at what this person did:

viewtopic/t73873/First-Night-with-IntelliPAP.html