Machine is recording AHI's while I am awake

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
spiral_15

Machine is recording AHI's while I am awake

Post by spiral_15 » Fri Jan 27, 2012 9:38 am

Good day all,
I have been struggling with acclimatizing to wearing a full face mask so I decided to wear it while awake last night about an hour before I normally go to sleep. I am currently using a Resmed S9 VPAP S machine. After an hour and 10 minutes I looked at the data on my machine and it had registered 18 AHI's while I was awake. Is this even possible? How can I have apneas while awake? I have been on treatment now for 2 weeks and the lowest AHI numbers I have been able to produce were 18 in a 6 hour period. Any idea what may be happening?

User avatar
LSAT
Posts: 13328
Joined: Sun Nov 16, 2008 10:11 am
Location: SE Wisconsin

Re: Machine is recording AHI's while I am awake

Post by LSAT » Fri Jan 27, 2012 9:41 am

The machine doesn't know if you are asleep or awake. It just looks a the gaps in breathing. You must have some unusual breathing patterns while awake.

Mary Z
Posts: 1493
Joined: Wed Oct 28, 2009 9:55 am

Re: Machine is recording AHI's while I am awake

Post by Mary Z » Fri Jan 27, 2012 9:56 am

For the past two weeks have you been using the machine all night, every night? An AHI is still high, are you going in for a check up anytime soon? Your pressure may need to be adjusted up or down. Are leaks under control? If not your data is probably faulty.

_________________
Mask: Wisp Nasal CPAP Mask with Headgear - Fit Pack
Additional Comments: PR System One Remstar BiPap Auto AS Advanced.
Dog is my copilot

User avatar
Slartybartfast
Posts: 1633
Joined: Wed Sep 01, 2010 12:34 pm

Re: Machine is recording AHI's while I am awake

Post by Slartybartfast » Fri Jan 27, 2012 10:14 am

Everyone's breathing while awake is noticeably irregular. Your VPAP is expecting to be monitoring the steady, regular breathing of someone who is asleep. There's no way for the machine to know whether you're asleep or awake. A sleep lab can determine that, but a small machine can't do so reliably. So don't worry about those readings. I usually read a few pages of a book while I'm masked up, and my data clearly shows more anomalies during that time than after I put the book down and turn out the light.

User avatar
archangle
Posts: 9293
Joined: Sun Mar 27, 2011 11:55 am

Re: Machine is recording AHI's while I am awake

Post by archangle » Fri Jan 27, 2012 10:17 am

Apneas while awake don't "count." Unfortunately, the machine doesn't know you're awake, so it will count them anyway. In a sleep lab, they use EEG electrodes to disregard anything that happens while you're awake.

The machines try to detect apneas by observing your breathing patterns in your sleep. Unfortunately, your breathing pattern while awake may look like something that would indicate an apnea while you're awake.

Get the free SleepyHead software, and you can look at the graphs on your computer, and it will be easier to figure out when you were awake and when you were asleep and ignore the events while you were awake.

Please sign up for an ID on the board and put your equipment in your profile. There's a link in my signature on how to fill it in.

_________________
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Also SleepyHead, PRS1 Auto, Respironics Auto M series, Legacy Auto, and Legacy Plus
Please enter your equipment in your profile so we can help you.
Click here for information on the most common alternative to CPAP.
If it's midnight and a DME tells you it's dark outside, go and check for yourself.

Useful Links.

User avatar
Kody
Posts: 336
Joined: Mon Sep 05, 2011 5:18 pm
Contact:

Re: Machine is recording AHI's while I am awake

Post by Kody » Fri Jan 27, 2012 10:24 am

I'm still in the closely monitoring my data stage. So sometimes I'll wake up too early and think there's a good chance I won't be able to go back to sleep. So I will pop out the card, but still keep using the machine in case I do fall back to sleep. If not, I don't have the "awake events" scored into my data. It's just a little easier than using a lot of extra Math to filter out events etc.
Complex Sleep Apnea

Carmen
Posts: 55
Joined: Mon Dec 19, 2011 5:04 pm

Re: Machine is recording AHI's while I am awake

Post by Carmen » Fri Jan 27, 2012 10:28 am

archangle wrote:Apneas while awake don't "count." Unfortunately, the machine doesn't know you're awake, so it will count them anyway. In a sleep lab, they use EEG electrodes to disregard anything that happens while you're awake.

The machines try to detect apneas by observing your breathing patterns in your sleep. Unfortunately, your breathing pattern while awake may look like something that would indicate an apnea while you're awake.

Get the free SleepyHead software, and you can look at the graphs on your computer, and it will be easier to figure out when you were awake and when you were asleep and ignore the events while you were awake.

Please sign up for an ID on the board and put your equipment in your profile. There's a link in my signature on how to fill it in.
Can you sort of tell from the Sleepyhead software when you're asleep or awake? I've been trying to track how many times I wake up at night. I can only tell from the graphs how many times I was out of bed but I'm awake more often than I'm out of bed.

nanwilson
Posts: 3463
Joined: Tue Apr 13, 2010 10:35 am
Location: Southern Alberta

Re: Machine is recording AHI's while I am awake

Post by nanwilson » Fri Jan 27, 2012 12:06 pm

Your machine does not know if you are awake or asleep nor does sleepyhead or any program to read your data. The only way to tell if you are awake or asleep is by YOU or an EEG.
Started cpap in 2010.. still at it with great results.

User avatar
Pugsy
Posts: 65055
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Machine is recording AHI's while I am awake

Post by Pugsy » Fri Jan 27, 2012 12:24 pm

Carmen wrote:Can you sort of tell from the Sleepyhead software when you're asleep or awake?
Not unless you wake up and turn the machine off so that you have a gap in the therapy line each and every time you wake up and then restart the machine. As others have mentioned, the machine just scores what it senses and it has no way of knowing if you are awake or not. The software can only report what the machine scores.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
idamtnboy
Posts: 2186
Joined: Mon Nov 01, 2010 2:12 pm
Location: Idaho

Re: Machine is recording AHI's while I am awake

Post by idamtnboy » Fri Jan 27, 2012 2:28 pm

Carmen wrote:Can you sort of tell from the Sleepyhead software when you're asleep or awake? I've been trying to track how many times I wake up at night. I can only tell from the graphs how many times I was out of bed but I'm awake more often than I'm out of bed.
Depends on how smooth your breathing pattern is while sleeping vs. being awake. In my case I can tell from the flow graph simply because there is real difference in the flow pattern. Awake breathing is much more erratic than sleep breathing. Study your graphs closely and take particular note of the flow pattern when you know you were asleep and awake. I'll bet you can soon spot most of the awake periods, especially if they last more than, say, half a minute.

_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Hose management - rubber band tied to casement window crank handle! Hey, it works! S/W is 3.13, not 3.7

User avatar
archangle
Posts: 9293
Joined: Sun Mar 27, 2011 11:55 am

Re: Machine is recording AHI's while I am awake

Post by archangle » Fri Jan 27, 2012 2:38 pm

Carmen wrote: Can you sort of tell from the Sleepyhead software when you're asleep or awake?
If you remember not being able to get to sleep for a long time, or lying in bed for a long time in the morning before getting up, you can discount that time.

After you look at the data for several nights, making note of how you breathe when awake and when sleeping, you may be able to get a good idea of when you're awake and when you're sleeping.

You could also deliberately mark your breathing when you wake up in the middle of night if you have the presence of mind. Inhale, hold your breath for 5 seconds, exhale, hold for 5, repeat 3 times (or something similar) and you can see that on the graph if you look in detail.

_________________
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Also SleepyHead, PRS1 Auto, Respironics Auto M series, Legacy Auto, and Legacy Plus
Please enter your equipment in your profile so we can help you.
Click here for information on the most common alternative to CPAP.
If it's midnight and a DME tells you it's dark outside, go and check for yourself.

Useful Links.

derek70
Posts: 99
Joined: Sat Dec 24, 2011 9:55 am

Re: Machine is recording AHI's while I am awake

Post by derek70 » Fri Jan 27, 2012 3:03 pm

It's the reason I only wear my mask just about to sleep.
I started using settling (ramp) too, because when in settling machine don't record any ahi.

_________________
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear
Additional Comments: Diagnosed AHI ~40 / Titrated @ 10 / CPAP8 EPR 0

cpaptex
Posts: 118
Joined: Tue Nov 15, 2011 4:18 pm

Re: Machine is recording AHI's while I am awake

Post by cpaptex » Fri Jan 27, 2012 8:37 pm

If you were wide awake they proably don't count. But, I was recently in the ER with an Oximeter attached to my finger. Each time that I nodded off an alarm would sound because my O2 Sat had dropped of to 85 or less. I was sleepy, and it happened over and over. Was it an Apnea? Dunno
I'm getting better day by day. Few if any OA Events, mostly CAs and Hyponeas.

I have Interstital Lung Disease

paulkellyclarke

Re: Machine is recording AHI's while I am awake

Post by paulkellyclarke » Sat May 18, 2013 11:51 am

same problem loads when awake in bef but vary from 0 to 7when asleep also air not mask material cauae bad face itch and of sleep on side mask come of after time.
aldo definate rise in number of centeral incidents Resmed auto 9 also when pressure shoot up wake me
Please excuse any errors done on mobile phone with small touch pad

User avatar
zoocrewphoto
Posts: 3732
Joined: Mon Apr 30, 2012 10:34 pm
Location: Seatac, WA

Re: Machine is recording AHI's while I am awake

Post by zoocrewphoto » Sat May 18, 2013 4:56 pm

paulkellyclarke wrote:same problem loads when awake in bef but vary from 0 to 7when asleep also air not mask material cauae bad face itch and of sleep on side mask come of after time.
aldo definate rise in number of centeral incidents Resmed auto 9 also when pressure shoot up wake me
Please excuse any errors done on mobile phone with small touch pad

You'll get better responses if you start a new topic as this topic was an old one from a coupe years ago. Also, sign up and list your equipment and tell is more about the problems you are having.

But basically, the machine can't tell when you are awake or asleep.We tend to hold our breath when we move around, so it is easy for the machine to consider that an apnea. When we are new, we tend to react to the pressure and mask more, and we have trouble falling asleep. Once we get used to it, the ahi tends to go down as we sleep more and move around less.

_________________
Mask: Quattro™ FX Full Face CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Resmed S9 autoset pressure range 11-17
Who would have thought it would be this challenging to sleep and breathe at the same time?