My mask is singing

Post by RocketGirl » Mon Dec 26, 2011 10:56 am

Two months in and everything is going pretty well, but in the past week or so I've awakened a few times with headaches, and my Opus 360 has developed a very high-pitched, intermittent whistle somewhere on the right side, that has awakened me in the night a few times. I thought it had to be a leak and was considering ordering replacement pillows even though my insurance won't pay for them yet. But looking at the data, I'm not sure how to interpret the leaks as shown. (I am new to Sleepyhead.) Are these something that experienced users would worry about? Here are last night's graphs, which are among the higher ones of the past week:

Image
Image
Grateful for any input on these.

Also very, very grateful to JediMark for Sleepyhead - it is an amazing piece of software!

User avatar
SleepingBearDoNtWake
Posts: 114
Joined: Fri Jun 24, 2011 9:38 am

Re: My mask is singing

Post by SleepingBearDoNtWake » Mon Dec 26, 2011 11:48 am

Hi RocketGirl,

I had the swift fx, and it was doing something like yours, and it was driving me crazy. I don't know how your mask works, but if it is simular you might want to look at the part where the hose connects to the pillows. I found if I gently pulled on the hose the nose would stop. For me it was the connection between the two. It might be something to try.

_________________
MaskHumidifier
Additional Comments: Zeo user

jnk
Posts: 5784
Joined: Mon Jun 30, 2008 3:03 pm

Re: My mask is singing

Post by jnk » Mon Dec 26, 2011 12:10 pm

Do you use o-rings?

I've been able to get LT nasal pillows to last for well over a year with the o-ring mod posted by "gl" a while back:

Image

Something similar is done by some here with the Opus.

There is also another o-ring mod for the Opus for holding the pillow stems together, I understand:

viewtopic/t36877/Swift-LT-vs-Opus-360.html#p321170

I mean, anyone named RocketGirl should be using o-rings, one way or the other, I would think.

WARNING: If you are not yet a mask-modifier, and you start now, there is absolutely NO going back!
Last edited by jnk on Mon Dec 26, 2011 12:13 pm, edited 1 time in total.

User avatar
GumbyCT
Posts: 5780
Joined: Fri Sep 14, 2007 6:22 pm
Location: CT
Contact:

Re: My mask is singing

Post by GumbyCT » Mon Dec 26, 2011 12:12 pm

Whistling would indicate a leak. Air moving causing the noise.

I am not familiar with your mask but you could try cleaning out the holes the exhaust goes thru. Sometimes that collects drool or other goober doo.

_________________
Humidifier: HC150 Heated Humidifier With Hose, 2 Chambers and Stand
Additional Comments: New users can't remember they can't remember YET!
BeganCPAP31Jan2007;AHI<0.5
I have no doubt, how I sleep affects every waking moment.
I am making progress-NOW I remember that I can't remember
;)
If this isn’t rocket science why are there so many spaceshots?
Be your own healthcare advocate!

User avatar
chunkyfrog
Posts: 34544
Joined: Mon Jul 12, 2010 5:10 pm
Location: Nowhere special--this year in particular.

Re: My mask is singing

Post by chunkyfrog » Mon Dec 26, 2011 1:13 pm

My Swift FX was whistling at me just this morning.
I wiggled it around until it shut up.

_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear
Additional Comments: Airsense 10 Autoset for Her

jnk
Posts: 5784
Joined: Mon Jun 30, 2008 3:03 pm

Re: My mask is singing

Post by jnk » Mon Dec 26, 2011 1:20 pm

Which is why chunkyfrog's ticklish husband doesn't whistle at her anymore.

User avatar
chunkyfrog
Posts: 34544
Joined: Mon Jul 12, 2010 5:10 pm
Location: Nowhere special--this year in particular.

Re: My mask is singing

Post by chunkyfrog » Mon Dec 26, 2011 1:35 pm

FYI: Never did. He's Ukie--has a thing about whistling indoors.

_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear
Additional Comments: Airsense 10 Autoset for Her

jnk
Posts: 5784
Joined: Mon Jun 30, 2008 3:03 pm

Re: My mask is singing

Post by jnk » Mon Dec 26, 2011 1:42 pm

So THAT's why I'm always broke!



I found that using larger nasal pillows got rid of a whistling sound that only occurred when I breathed out.

It sounded like it was IN my nose instead of outside it. But whatever it was, it went away with the larger pillows, which also happened to be new, so I don't know which really solved it.

DannyPh
Posts: 287
Joined: Tue Oct 16, 2007 6:05 pm
Location: South East Texas

Re: My mask is singing

Post by DannyPh » Mon Dec 26, 2011 5:21 pm

I use the Mirage Liberty FF mask and I was wondering if the O Rings on the Pillows would benefit me on the Mirage Mask ? The problem I am having is that the Pillows are too close together and not getting a good seal until I re position them around. I am using the Large Pillows. i was wondering if the O Rings would keep the Pillows apart some. Guess it is worth a try.

User avatar
RocketGirl
Posts: 266
Joined: Thu Nov 03, 2011 10:48 pm

Re: My mask is singing

Post by RocketGirl » Mon Dec 26, 2011 7:57 pm

jnk wrote:Do you use o-rings?
I do now
jnk wrote: I mean, anyone named RocketGirl should be using o-rings, one way or the other, I would think.
Thanks, jnk, for the ideas and links to the other threads. Interesting stuff! Since I'm so new to this, I've held off modding my mask until I understood it a little better. This definitely seems like a worthwhile thing to try (and I do have a jar of all sizes of o-rings on the workbench). Will give it a try tonight.
jnk wrote:WARNING: If you are not yet a mask-modifier, and you start now, there is absolutely NO going back!
Forewarned is forearmed... I do recognize this as a slippery slope, but things are a lot more interesting that way. Cowabunga!

User avatar
RocketGirl
Posts: 266
Joined: Thu Nov 03, 2011 10:48 pm

Re: My mask is singing

Post by RocketGirl » Mon Dec 26, 2011 8:00 pm

SleepingBearDoNtWake wrote:Hi RocketGirl,

I had the swift fx, and it was doing something like yours, and it was driving me crazy. I don't know how your mask works, but if it is simular you might want to look at the part where the hose connects to the pillows. I found if I gently pulled on the hose the nose would stop. For me it was the connection between the two. It might be something to try.
and
chunkyfrog wrote:My Swift FX was whistling at me just this morning.
I wiggled it around until it shut up.

Thank you chunkyfrog and SleepingBearDoNtWake - I will try these suggestions tonight, too. The Opus has a kind of funky swivelly piece between the hose quick-connect and the mask, so I'll try fiddling with that and the hose to see if that quashes the whistle.

Hose_Head
Posts: 804
Joined: Wed Mar 12, 2008 6:43 pm
Location: Ontario Canada

Re: My mask is singing

Post by Hose_Head » Mon Dec 26, 2011 8:12 pm

RocketGirl wrote:
SleepingBearDoNtWake wrote:Hi RocketGirl,

I had the swift fx, and it was doing something like yours, and it was driving me crazy. I don't know how your mask works, but if it is simular you might want to look at the part where the hose connects to the pillows. I found if I gently pulled on the hose the nose would stop. For me it was the connection between the two. It might be something to try.
and
chunkyfrog wrote:My Swift FX was whistling at me just this morning.
I wiggled it around until it shut up.

Thank you chunkyfrog and SleepingBearDoNtWake - I will try these suggestions tonight, too. The Opus has a kind of funky swivelly piece between the hose quick-connect and the mask, so I'll try fiddling with that and the hose to see if that quashes the whistle.
Something to think about is my experience with an Activa LT mask. This mask is prone to "singing" if you have any rainout, whatsoever. The water collects in the mask's exhaust ports and restricts them in a way that causes the mask to sing. This is unlikely to be your problem, but if you cannot find a solution through the other suggestions, you might want to check this one out.
I'm workin' on it.

User avatar
archangle
Posts: 9293
Joined: Sun Mar 27, 2011 11:55 am

Re: My mask is singing

Post by archangle » Tue Dec 27, 2011 12:14 am

It's possible that a leak big enough to sing won't leak enough liters/minute to show up on the graphs. See if you can catch it singing and figure out where you can poke or wiggle to make it start or stop. Maybe a thorough cleaning for good luck. Disassemble and assemble and check the connection between parts very carefully.

If it starts telling you to do things while it's singing to you, you need to see a different kind of doctor.

_________________
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Also SleepyHead, PRS1 Auto, Respironics Auto M series, Legacy Auto, and Legacy Plus
Please enter your equipment in your profile so we can help you.
Click here for information on the most common alternative to CPAP.
If it's midnight and a DME tells you it's dark outside, go and check for yourself.

Useful Links.

User avatar
red1
Posts: 90
Joined: Sat Oct 22, 2011 12:13 am

Re: My mask is singing

Post by red1 » Tue Dec 27, 2011 1:41 am

I have both the FX & the Lt. They both seem to leak after they've been on awhile. As the night goes on, the areas around both the outside & the entrance to the inside of the nasal passages become oily & have begun to sweat . This will cause a lowered friction between the skin & the silicone, causing it to leak. ITs good to wash your face well, blow your nose, then, using the tissue, dry around the openings of your nasal passages, just before putting on the nasal pillow mask. I find about 5 or so hours later it begins leaking a lot,especially if you happen to subconsciously move your nose, so I go out & repeat the above & I also clean the moisture off the nasal pillow prongs.
Air Curve 10 VAuto
Swift FX Nasal

User avatar
squid13
Posts: 2301
Joined: Fri Jul 15, 2011 3:47 pm
Location: Pensacola, FL

Re: My mask is singing

Post by squid13 » Tue Dec 27, 2011 11:34 am

knk what size O ring do you use for those nasal pillows?

_________________
Machine: ResMed AirCurve 10 ASV Machine with Heated Humidifier
Mask: Evora Full Face Mask - Fitpack
Additional Comments: AirCurve 10 ASV, Oscar V1.0.1-r-1
US Navy Retired 1973,AirCurve 10 ASV, Mode: ASV Auto, Min EPAP: 7.2, Max EPAP: 15.0, Min PS:4.0, Max PS: 15.0, Mask ResMed Airtouch F20, Backup: (2) AirCurve 10 ASV