S8 Mask Setting

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
woodmizer5
Posts: 15
Joined: Tue Oct 18, 2011 4:41 pm

S8 Mask Setting

Post by woodmizer5 » Sun Dec 18, 2011 5:06 pm

Can anyone tell me what mask setting to put for the opus 360, on an S8 vpap? I have the swift displayed, but it is showing to much leakage to register.
Thank You.

User avatar
Pugsy
Posts: 65354
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: S8 Mask Setting

Post by Pugsy » Sun Dec 18, 2011 5:16 pm

On the S8 machines I believe ResMed says for non ResMed machines to just choose "standard" or whichever it is that doesn't point to a specific ResMed model. Swift nasal pillows are of course similar to the Opus 360 and the vent rate will be very similar. It will get you very close.
If you are having too much leak to register then the problem isn't the mask selection choice. The problem is in the leak itself. Mask selection feature only makes minor adjustments to the vent rate for mask and pressure before reporting the excess leak.
If mask isn't leaking perhaps a hole in the hose or something like that.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
RandyJ
Posts: 1673
Joined: Thu Apr 21, 2011 7:22 pm
Location: Connecticut, USA

Re: S8 Mask Setting

Post by RandyJ » Sun Dec 18, 2011 6:48 pm

There is a useful pdf you can download here: http://www.resmed.com/us/assets/html/pr ... c=patients

It shows Resmed masks only and the settings for the various Resmed machines, but if you can look up the vent rate for your non-Resmed mask at your pressure, you can find which Resmed mask is most similar to yours and go from there. (Usually the vent rate chart is in the user info that comes with your mask; if not you can find it online at the manufacturer's website.)

Edit: From what I can see, "Swift" is the correct setting, but "Standard" would also work. Vent rates for Opus 360 are almost the same as for Swift LT at same pressures. Pugsy was right on without even seeing the charts.

_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: Backup & Travel Machines: PR System One Bipap Auto, S9 VPAP Auto, S9 Autoset, Oximeter CMS-50E
Diagnosed March 2011, using APAP 14 - 16.5 cm, AFlex+ 2
Alt masks Swift FX pillows, Mirage FX nasal mask, Mirage Quattro full face mask