What is that sound & How do I fix it?

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
cpap_user_29

What is that sound & How do I fix it?

Post by cpap_user_29 » Mon Nov 28, 2011 12:37 pm

Sometimes during the night, my hose and mask make a very loud knocking or "clunking sound." It sounds like someone is knocking on a door. I think it may be that the hose is filled with too many water droplets and they are hitting the mask. I usually take off the mask, shake out the hose and when I reapply it - it still does the same thing. Can I get some advice on how to prevent this - and what is happening? I cant sleep b/c of the noise - and it wakes up my husband as well.
Thanks in advance!

Mary Z
Posts: 1493
Joined: Wed Oct 28, 2009 9:55 am

Re: What is that sound & How do I fix it?

Post by Mary Z » Mon Nov 28, 2011 12:57 pm

Sounds like you may have your humidifier turned up too high. That can make a clunking or gurgling noise and cause water in the tubing. A hose cozy ( cover for your hose) might help if you don't want to turn the humidity down.
Also, make sure you're not filling the humidifier too high.

_________________
Mask: Wisp Nasal CPAP Mask with Headgear - Fit Pack
Additional Comments: PR System One Remstar BiPap Auto AS Advanced.
Dog is my copilot

User avatar
2 B Sleeping Soundly
Posts: 822
Joined: Sat Nov 05, 2011 2:24 pm
Location: Southern California

Re: What is that sound & How do I fix it?

Post by 2 B Sleeping Soundly » Mon Nov 28, 2011 1:20 pm

cpap_user_29,

Welcome!

I agree with Mary Z’s assessment. I had a similar experience with a System One Humidifier because I filled the water reservoir too high. I also used to get a lot of condensation in the hose because of our house being cooler than the air coming out of the humidifier (our winters are mild and we don’t use a heater), but I cured that with the hose cozy she suggested. I have been using the Snuggle Hose and it is great, but the zippered Pad A Cheek looks like it would do a good job and also make getting the cover very easy to get on and off.

I case you don't know where to look I put the two sources I know of below:)

SnuggleHose
http://www.snugglehose.com/index.html

Pad A Cheek
http://padacheek.com/PAC_Hosecover.html

Hope it helps,

John

_________________
Machine: PR System One REMStar 60 Series Auto CPAP Machine
Additional Comments: 11-09-11 Start PAP therapy. Current settings: APAP 13cmH2O - 17cmH2O / Ramp, off / A-Flex, off / Respironics Premium Chin Strap / Sleepyhead Software
One time a cop pulled me over for running a stop sign. He said, "Didn't you see the stop sign?" I said, "Yeah, but I don't believe everything I read." -- Steven Wright

User avatar
apneawho
Posts: 609
Joined: Wed Jul 27, 2011 10:09 am
Location: Atlanta, GA, USA

Re: What is that sound & How do I fix it?

Post by apneawho » Mon Nov 28, 2011 1:37 pm

Yep. Sounds like condensation. Don't know what equipment you are using. If you could click on user control and add the equipment to your profile that may help us.
I have a ResMed machine that has the option of climate line tubing. I have never had condensation problems with the climate line. Keeping the machine lower than your tubing is helpful. You can hang the tubing with a hose hanger system or place the machine lower than your head. The hose snuggy or cozy keeps the tubing warmer. There are heated hoses you may look into also. I am not sure what your humidifier is set at right now, but you may try experimenting to see if you can drop the setting a tad. That may reduce condensation, but make sure your nasal and oral passages don't get too dry. Sometimes there is a fine line between comfort and reducing condensation or rain out. Warming the tubing up under the covers with you may help also.

_________________
Mask: Swift™ FX For Her Nasal Pillow CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: ResMed Climate line tubing, Hose cozy, PurSleep Clear aroma therapy, Using latest version Sleepyhead sofware-thanks Jedimark!

cpap_user_29

Re: What is that sound & How do I fix it?

Post by cpap_user_29 » Mon Nov 28, 2011 8:57 pm

Thank you for all the replies, and helpful links! This website is the best!

User avatar
archangle
Posts: 9293
Joined: Sun Mar 27, 2011 11:55 am

Re: What is that sound & How do I fix it?

Post by archangle » Wed Nov 30, 2011 1:14 pm

Please do look at the link at the bottom of this post, and sign up for an ID and fill in your equipment.

The next time this happens, stop your CPAP. Disconnect the hose from the machine, and hold the end of the hose up. Disconnect from your mask and do the same and make a "U" shape with your hose. Now, see if there's water in your hose at the bottom of the U. You may be able to see it if your hose is transparent. If you can't see it, see if you can get the water to run downhill and pour out one end of the hose.

If so, you have the dreaded rainout many of us deal with.

Once you tell us your equipment, we may be able to give you specific advice. The simplest solution is if you can arrange the hose so any water formed runs back in to the machine. In the middle of the night, you may be able to lift the hose and make it run back down into the tank as a temporary fix.

_________________
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Also SleepyHead, PRS1 Auto, Respironics Auto M series, Legacy Auto, and Legacy Plus
Please enter your equipment in your profile so we can help you.
Click here for information on the most common alternative to CPAP.
If it's midnight and a DME tells you it's dark outside, go and check for yourself.

Useful Links.

User avatar
avi123
Posts: 4509
Joined: Tue Dec 21, 2010 5:39 pm
Location: NC

Re: What is that sound & How do I fix it?

Post by avi123 » Wed Nov 30, 2011 1:26 pm

It could be the anti asphyxia (a.k.a. non rebreathing) valve. Usually, the valve sounds like that if the machine's flow is stopped and the valve is OK. Also, when the the leaves inside the valve were pushed thru to the other side and need be pushed back.

_________________
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments:  S9 Autoset machine; Ruby chinstrap under the mask straps; ResScan 5.6
Last edited by avi123 on Fri Dec 02, 2011 2:57 pm, edited 1 time in total.
see my recent set-up and Statistics:
http://i.imgur.com/TewT8G9.png
see my recent ResScan treatment results:
http://i.imgur.com/3oia0EY.png
http://i.imgur.com/QEjvlVY.png

User avatar
Pugsy
Posts: 65326
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: What is that sound & How do I fix it?

Post by Pugsy » Wed Nov 30, 2011 1:35 pm

If it is rain out or condensation in the hose, it doesn't take much moisture to make a whole lot of noise.
I have experienced it and am continually amazed that when I pour out the water from the hose there will be such a tiny amount and I was so sure that there was at least a cupful from the sounds I heard.

This winter I bought a heated hose so I don't have to deal with the noise and snorting water up my nose. Heated hose and barrel cozy doing their job now. No more rain out.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.