Philips Respironics System One problem

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
Cathy2903
Posts: 4
Joined: Sat Jun 11, 2011 6:15 am

Philips Respironics System One problem

Post by Cathy2903 » Sat Nov 26, 2011 10:12 am

I have had nothing but trouble with my Philips Respironics Sysyem One Bipap. My DME just switched out the unit for the third time and the same problem keeps occurring. The machine shuts off while I am sleeping and I wake up because I am literally being suffocated and there is a steady beep. There is nothing that can be done at this point but to unplug the unit and then plug it back in and sometimes I can get up to an hours sleep before it happens again. The DME is acting like there is nothing wrong and it takes a couple days for them to come out each time and I have a hard time sleeping with no bipap at all. Has anyone run into this problem, I am at the end of my rope with this and even tried calling Philips which was a huge waste of time because they won't help you and tell you to call your provider.

User avatar
DoriC
Posts: 5214
Joined: Sat Sep 13, 2008 9:28 pm
Location: NJ

Re: Philips Respironics System One problem

Post by DoriC » Sat Nov 26, 2011 10:30 am

Is it possible you have Split Night set to ON?? If so, turn if OFF, that's the only thing I can come up with. Also check if Mask Alert is set to ON, you might try it turned OFF. Keep us posted.

_________________
Mask: Mirage Quattro™ Full Face CPAP Mask with Headgear
Additional Comments: 14/8.4,PS=4, UMFF, 02@2L,
"Do or Do Not-There Is No Try"-"Yoda"
"We are what we repeatedly do,so excellence
is not an act but a habit"-"Aristotle"
DEAR HUBBY BEGAN CPAP 9/2/08

Cathy2903
Posts: 4
Joined: Sat Jun 11, 2011 6:15 am

Re: Philips Respironics System One problem

Post by Cathy2903 » Sat Nov 26, 2011 2:19 pm

Don't have a split sleep as far as I know. Just tried to take a nap and it is shutting off every 5 minutes again. I am thinking maybe there is something wrong with the power cord since they haven't switched that out only the unit.

Janknitz
Posts: 8513
Joined: Sat Mar 20, 2010 1:05 pm
Location: Northern California

Re: Philips Respironics System One problem

Post by Janknitz » Sat Nov 26, 2011 2:24 pm

Check to make sure they aren't just giving you the same machine back--write down the serial number--don't trust the DME.

Check the build date (it's somewhere on the bottom of the machine-I'm sure someone here can tell you where to look).

Also see if you can find the firmware build and try to get a machine with a different version.

If your DME can't won't help you, contact PR directly. I've seen other complaints about this, and I'm guessing it's a known manufacturers defect so they should replace it under warranty, but you may need to get PR to light a fire under the DME.
What you need to know before you meet your DME http://tinyurl.com/2arffqx
Taming the Mirage Quattro http://tinyurl.com/2ft3lh8
Swift FX Fitting Guide http://tinyurl.com/22ur9ts
Don't Pay that Upcharge! http://tinyurl.com/2ck48rm

User avatar
Slinky
Posts: 11372
Joined: Wed Nov 01, 2006 3:43 pm
Location: Mid-Michigan

Re: Philips Respironics System One problem

Post by Slinky » Sat Nov 26, 2011 4:49 pm

Cathy said she has already called PR and their only response was to go back to her local DME provider. Given the KNOWN problem and the many replacements for this problem by PR that certainly does NOT speak well for PR!!!!!!

Given the situation, now is the time to make some NOISE! A tactful but strong letter of complaint to PR sent Certified, Return Receipt Requested. The same to the local DME provider. Yet another to your state business licensing board. Try to see if your sleep doctor or sleep lab will assist you in this matter.

_________________
Mask: Quattro™ FX Full Face CPAP Mask with Headgear
Additional Comments: PR SystemOne BPAP Auto w/Bi-Flex & Humidifier - EncorePro 2.2 Software - Contec CMS-50D+ Oximeter - Respironics EverFlo Q Concentrator
Women are Angels. And when someone breaks our wings, we simply continue to fly.....on a broomstick. We are flexible like that.
My computer says I need to upgrade my brain to be compatible with its new software.

User avatar
Pugsy
Posts: 65148
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Philips Respironics System One problem

Post by Pugsy » Sat Nov 26, 2011 4:56 pm

Ask the DME to swap out power supply (both cords including the brick) since you say that hasn't been changed.
Perhaps a bad connector or other internal component in the power supply. Multiple machines with same problem is very odd.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
Slinky
Posts: 11372
Joined: Wed Nov 01, 2006 3:43 pm
Location: Mid-Michigan

Re: Philips Respironics System One problem

Post by Slinky » Sat Nov 26, 2011 5:11 pm

Cathy,

1] Check the bottom plate on your PR S1 for the manufacture date (and serial number).
2] When unplugging and plugging your S1 back in keep a close eye on the LED screen for the software version number, it appears and disappears quickly.
3] Go thru the entire Clinicians Menu and make sure that Split Night, SmartStart and SmartStop (or AutoStart/AutoStop) are both turned OFF.
4] As Pugsy suggested ask that both power supply cords (w/power brick) be exchanged. (Unobtrusively mark them w/finger nail polish or some such so you can tell if they really have been exchanged).

It does sound like you are being taking advantage of by an uncaring or less than honest local DME supplier. At the very least they should be checking these things out themselves in your presence.

_________________
Mask: Quattro™ FX Full Face CPAP Mask with Headgear
Additional Comments: PR SystemOne BPAP Auto w/Bi-Flex & Humidifier - EncorePro 2.2 Software - Contec CMS-50D+ Oximeter - Respironics EverFlo Q Concentrator
Women are Angels. And when someone breaks our wings, we simply continue to fly.....on a broomstick. We are flexible like that.
My computer says I need to upgrade my brain to be compatible with its new software.

User avatar
RocketGirl
Posts: 266
Joined: Thu Nov 03, 2011 10:48 pm

Re: Philips Respironics System One problem

Post by RocketGirl » Sat Nov 26, 2011 6:51 pm

Just a shot in the dark, but on the chance that the DME actually did swap out your machine three times - is it at all possible that the problem is not in the machine, but is in your electrical outlet instead?

Some sort of electrical fault - low voltage, a short-circuit, reversed polarity, strong RF interference, or another device with a heavy draw on the same circuit - in other words, something that kicks in other than the machine itself, that the machine might be reacting to?

Just a thought.

User avatar
archangle
Posts: 9293
Joined: Sun Mar 27, 2011 11:55 am

Re: Philips Respironics System One problem

Post by archangle » Sun Nov 27, 2011 9:46 pm

I have a PRS1 machine where the plug on the back of the machine doesn't seem to fit right. If I jiggle the plug, the machine will turn off and back on. Check to see if you have this problem.

_________________
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Also SleepyHead, PRS1 Auto, Respironics Auto M series, Legacy Auto, and Legacy Plus
Please enter your equipment in your profile so we can help you.
Click here for information on the most common alternative to CPAP.
If it's midnight and a DME tells you it's dark outside, go and check for yourself.

Useful Links.

Mattbyz

Re: Philips Respironics System One problem

Post by Mattbyz » Mon Mar 21, 2016 5:22 am

I didn't check to see how old this post is. If you haven't had it fixed yet. I'm working to fix same issue. Don't let anyone tell you it's anything but the sensor for the heating element. You can get the machine to work without the heating element on after you unplug it, flip it to zero before it turns pressure on. It will work fine...only problem is your without steam. Turns out it's super important.

Stokes
Posts: 1
Joined: Wed Feb 28, 2018 1:40 pm

Re: Philips Respironics System One problem

Post by Stokes » Wed Feb 28, 2018 1:49 pm

Cathy2903 wrote:
Sat Nov 26, 2011 2:19 pm
Don't have a split sleep as far as I know. Just tried to take a nap and it is shutting off every 5 minutes again. I am thinking maybe there is something wrong with the power cord since they haven't switched that out only the unit.
Yes, I had the same problem. It was the power supply not the CPAP unit.
Tested the power supply (12V 6.6A) with a dummy load (50W light bulb ~ 4A) and the power supply voltage dropped to 5.5V..... fail!
I used another (good) power supply and all good, no problems.

I have tested 20 to 30 brand new Respironics power supplies at the local electrical junk market and found about 40% didn't meet the specifications stated on the label.

I have since purchased 2 spare power supplies from eBay for AU$12.00 each. Make sure you get ones with the correct plug for your CPAP machine.