Cpap Data 11.11.2011

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
User avatar
wmcanally
Posts: 133
Joined: Sat Nov 05, 2011 8:35 am
Location: Jacksonville Texas

Cpap Data 11.11.2011

Post by wmcanally » Mon Nov 14, 2011 3:07 pm

Please let me know if changes are needed. Im new and learning.

http://i1227.photobucket.com/albums/ee4 ... 112011.jpg

_________________
Mask

User avatar
bdp522
Posts: 3378
Joined: Tue Jun 27, 2006 5:13 pm

Re: Cpap Data 11.11.2011

Post by bdp522 » Mon Nov 14, 2011 3:36 pm

Before you do anything else you need to get the leaks under control. Most of the night(90%) you were at a pressure of 9.57 or less. The leak for your mask at 9.57 is about 31. The leak you had was 74.33. The rest of the night you were between 9.57 and 13.7 for pressure and your leak was 124.00, but the accepted leak for the mask at those pressures is about 37. With leaks that high you can't trust the data as far as events(AHI). So first you have to control the leaks, THEN you can dial in the pressure.

Brenda

_________________
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Love my papillow, Aussie heated hose and PAD-A-CHEEKS! Also use Optilife, UMFF(with PADACHEEK gasket), and Headrest masks Pressure; 10.5

User avatar
Pugsy
Posts: 65354
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Cpap Data 11.11.2011

Post by Pugsy » Mon Nov 14, 2011 3:38 pm

This one your leak was a little more and drove your pressure up a bit in one spot. Still a good AHI and depending on how things continue...maybe a slight increase in minimum pressure might be considered but would need a lot more nights like this one. Don't make changes based on one or 2 nights of data when you are first starting out. Things are likely to just be unstable at first. Especially leak which messes with everything else.

Image

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.