Hello,
I am selling my Fisher Paykel Icon Primo cpap and would like to know how to reset this cpap back to factory settings and erase all of my information. Does anyone know how to do this?
Thanks!
Reset Icon Primo
Re: Reset Icon Primo
http://www.apneaboard.com/CPAP%20Adjustment.htm
The Clinical/Provider manual should explain how to reset therapy data to zero. If you don't have it you can request it from above link. Total machine blower hours cannot usually be changed on any machine.
The Clinical/Provider manual should explain how to reset therapy data to zero. If you don't have it you can request it from above link. Total machine blower hours cannot usually be changed on any machine.
I may have to RISE but I refuse to SHINE.
-
palexis60
Re: Reset Icon Primo
Reset Icon Auto+; Primo and others icon machines:
1. Go to the doctors menu: push center and clockwise turn dial 180 and then counterclockwise one click back by holding middle push button. the doctors menu will be open and you will see an additional dots on left side of dial. Choose menu # 2.
2. Turn dial to info dot. (looks like a step 5 o'clock on dial and push a center dial. Move one click from smily face to next dot clockwise and push dial again.
3. Dial left to fined Auto or CPAP and push for 3-5 sec. until IFORMATION (560hours for exemple) starts flashing. Push center again to zeroed. All Info will be resset.
Alex
1. Go to the doctors menu: push center and clockwise turn dial 180 and then counterclockwise one click back by holding middle push button. the doctors menu will be open and you will see an additional dots on left side of dial. Choose menu # 2.
2. Turn dial to info dot. (looks like a step 5 o'clock on dial and push a center dial. Move one click from smily face to next dot clockwise and push dial again.
3. Dial left to fined Auto or CPAP and push for 3-5 sec. until IFORMATION (560hours for exemple) starts flashing. Push center again to zeroed. All Info will be resset.
Alex