EPR settings

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
Mark184
Posts: 26
Joined: Sun Jul 31, 2011 9:12 am
Location: Bellevue, WA

EPR settings

Post by Mark184 » Fri Oct 21, 2011 9:54 am

So I'm given the choice of off, 1, 2, and 3. My doctor has set mine to 2. What differences will I feel as I lower the number setting? Increase? Off? The clinician's manual mentions the settings, but doesn't define what the numbers mean? (unless my apnea induced confusion has made me miss this entry!)

pap4life
Posts: 207
Joined: Tue Jan 25, 2011 12:08 pm
Location: Florida

Re: EPR settings

Post by pap4life » Fri Oct 21, 2011 10:19 am

Mark.. the EPR is a comfort setting, to assist you with your exhaling. That, of course, is why it available to you as a user. The Doc/DME or whoever set up your equipment, set it on (2) because thats a middle of the range to help you get started. They have no other means to know what your "comfort" range may be. You may not even care to use the EPR, once you get accostomed to CPAP use.
The setting of (1) gives a small amount of exhale relief, maybe (depending on your equip) drops the constant pressure down about 1 -2cm
(2) middle of the road, drops the pressure down 2-3
and (3) the most EPR relief, drops the pressure down maybe 3-4cm during your exhalation.
Some people don;t even use the EPR.
Strictly comfort!! Play with the settings.. I suggest changing it to another setting, if uncomfortable, after 3 or 4 days of trial. You must at least stay with one setting for several nights, just to get used to the EPR. Trial and error, that all.

Anon

Re: EPR settings

Post by Anon » Fri Oct 21, 2011 10:23 am

I have an Intellipap, and I can change that setting with the machine on. So I put my mask on and sat there and figured out which one felt the best to me. 3 seemed to blow back into my face too much when it came back up to pressure. 1 didn't seem to let me relax enough - I felt the effort to exhale. So, 2 it is!

User avatar
Lizistired
Posts: 2835
Joined: Tue Dec 14, 2010 10:47 pm
Location: Indiana

Re: EPR settings

Post by Lizistired » Fri Oct 21, 2011 10:50 am

Also if you are in CPAP mode, there is an additional setting to adjust how fast the pressure comes back. This is not availablein AUTO.

EPR made me feel like I had to breathe at the machine's pace, so I turned it off and had much better results. It's an individual thing.

_________________
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Swift FX sometimes, CMS-50F, Cervical collar sometimes, White noise, Zeo... I'm not well, but I'm better.

User avatar
RandyJ
Posts: 1673
Joined: Thu Apr 21, 2011 7:22 pm
Location: Connecticut, USA

Re: EPR settings

Post by RandyJ » Fri Oct 21, 2011 11:59 am

Lizistired wrote:Also if you are in CPAP mode, there is an additional setting to adjust how fast the pressure comes back. This is not availablein AUTO.

EPR made me feel like I had to breathe at the machine's pace, so I turned it off and had much better results. It's an individual thing.

I had the same experience as Lizistired. I tried EPR in all different settings and couldn't get used it. I turned it off and don't have any difficulty breathing out against 14 cm and higher.

_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: Backup & Travel Machines: PR System One Bipap Auto, S9 VPAP Auto, S9 Autoset, Oximeter CMS-50E
Diagnosed March 2011, using APAP 14 - 16.5 cm, AFlex+ 2
Alt masks Swift FX pillows, Mirage FX nasal mask, Mirage Quattro full face mask

BrandonB
Posts: 77
Joined: Tue Oct 18, 2011 3:41 pm

Re: EPR settings

Post by BrandonB » Fri Oct 21, 2011 1:30 pm

In my titration study, I had a very hard time exhaling against. At home, I initially had EPS set to 3 with "medium" breaths (only medium and short I think were the choices). As Anon already mentioned, the "blowback" was uncomfortable for me. I've been at 1 the past three nights. Works for me. I'd suggest using the mask fit mode to test or if you have a straight CPAP set on your Autoset, turn it on for a few minutes (lying down) and see which feels more comfortable. Nice thing is you can easily change it as well. As Pap4Life already mentioned, I'd give it at least 3 days though when trying a new setting to make a permanent decision.
Life before my CPAP was really no life at all!

Mark184
Posts: 26
Joined: Sun Jul 31, 2011 9:12 am
Location: Bellevue, WA

Re: EPR settings

Post by Mark184 » Fri Oct 21, 2011 4:02 pm

Thanks for all of these helpful replies. My S9 is set to auto, and my average pressure seems to be about 6.5 to 7.5 most nights. The doc has the auto pressure settings between 5 and 20. So if I understand what I've read from your answers, the S9 is dropping my pressure to between 4.5 and 5.5 on exhale. I'll give a try lowering the EPR setting to 1, since sometimes it barely feels like the cpap is doing anything while I'm using it.

scottjf8
Posts: 128
Joined: Sat Jul 16, 2011 7:52 pm

Re: EPR settings

Post by scottjf8 » Fri Oct 21, 2011 4:11 pm

cpapman1 wrote:Hope this helps a little:

EPR is a reduction in pressure on the exhale. 3 is equal to a drop in pressure of 3 cm H2O. 0 is no reduction in pressure. It means that the pressure will drop up to the numbers of cm H2O you have selected. I always suggest a setting of 3 unless the set pressure is 7 cm H2O or less. At a pressure setting of 7 cm H2O and an EPR setting of 3, the CPAP machine would drop to a pressure of 4 cm H2O on the exhale. That is as low as the machine will go.
I'm curious about this... my S9 (Elite, not autoset) is set at 7cm per my titration... I feel pretty good, but not great. I also have EPR on 3.

Will changing/taking it off help more? I'm using a FFM if that helps...

Let me know... kinda wondering now..

_________________
Mask: Quattro™ FX Full Face CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Started therapy 7/15/2011

dtsm
Posts: 1097
Joined: Sun Mar 08, 2009 8:34 am
Location: CT

Re: EPR settings

Post by dtsm » Fri Oct 21, 2011 4:54 pm

Keep in mind when EPR is used, it effectively reduces cpap pressure during exhalation....therefore in theory one needs to increase the cpap pressure by the same EPR number. For instance, if your cpap setting is at 9 and you decide to go with EPR of 2, you should increase cpap from 9 to 11. If you're on apap, as long as the range is broad enough, no need to adjust.

User avatar
Pugsy
Posts: 64932
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: EPR settings

Post by Pugsy » Fri Oct 21, 2011 5:13 pm

scottjf8 wrote:'m curious about this... my S9 (Elite, not autoset) is set at 7cm per my titration... I feel pretty good, but not great. I also have EPR on 3.

Will changing/taking it off help more?
Give it a try. Some people it helps, some it doesn't. Reducing it a little will effectively raise the overall average a wee bit and it may feel better, work better or do nothing. I have seen reducing EPR or turning it off make huge changes in a person's data and I have also seen where it makes no change at all. There is no hard fast rule on what to expect but it hurts nothing to try. It is a comfort feature. It is there to make exhale easier and more comfortable. Might be greatly appreciated at 14 cm of pressure but at 7 cm not needed nearly so much. It is only important to how YOU perceive it.
Impact, if any, will vary from person to person.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
Lizistired
Posts: 2835
Joined: Tue Dec 14, 2010 10:47 pm
Location: Indiana

Re: EPR settings

Post by Lizistired » Fri Oct 21, 2011 5:26 pm

dtsm wrote:Keep in mind when EPR is used, it effectively reduces cpap pressure during exhalation....therefore in theory one needs to increase the cpap pressure by the same EPR number. For instance, if your cpap setting is at 9 and you decide to go with EPR of 2, you should increase cpap from 9 to 11. If you're on apap, as long as the range is broad enough, no need to adjust.
I don't agree with this. If you increase your pressure by your epr, say to 11, you are exhaling against 9 again. There's no point in that and now your inhale pressure is higher than needed. I can see the comfort feature to get used to the machine but I don't see the sense in raising your inhale pressure to exhale against the same pressure.

_________________
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Swift FX sometimes, CMS-50F, Cervical collar sometimes, White noise, Zeo... I'm not well, but I'm better.

pap4life
Posts: 207
Joined: Tue Jan 25, 2011 12:08 pm
Location: Florida

Re: EPR settings

Post by pap4life » Fri Oct 21, 2011 7:46 pm

Well said, Liz.

The intent of EPR is for ""comfort"" relief. It makes no sense whatever (theory???) to increase your inhale (IPR) because of a intended drop of a couple of cm/H2O in the EPR for comfort. Where is the comfort going to come from in that manuevour? If that were the case, then why did the manufacturers put the EPR settings on the machines?. Hit "0" and turn the EPR off, would serve the same results, (no comfort).

User avatar
archangle
Posts: 9293
Joined: Sun Mar 27, 2011 11:55 am

Re: EPR settings

Post by archangle » Fri Oct 21, 2011 8:31 pm

Mark184 wrote:I'll give a try lowering the EPR setting to 1, since sometimes it barely feels like the cpap is doing anything while I'm using it.
You tend to notice the pressure less and less as you get used to it. If I put mine on, I sometimes can't feel the pressure at all and sometimes I have to lift up the edge of the mask to convince myself the machine is on.

Once you get used to it, you might want to turn EPR off and see if it feels comfortable.

_________________
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Also SleepyHead, PRS1 Auto, Respironics Auto M series, Legacy Auto, and Legacy Plus
Please enter your equipment in your profile so we can help you.
Click here for information on the most common alternative to CPAP.
If it's midnight and a DME tells you it's dark outside, go and check for yourself.

Useful Links.

User avatar
ozij
Posts: 10432
Joined: Fri Mar 18, 2005 11:52 pm

Re: EPR settings

Post by ozij » Fri Oct 21, 2011 11:25 pm

In humans, exhalation is the result of muscular relaxation, and not of muscular effort. Therefore, it's the difference between inhale pressure and exhale pressure that makes exhalation easier. And that's why as far as exhale relief is concerned, it does make sense to increase your pressure by your EPR. There may be other reasons not to increase Inhale pressure - but if you can set your exhale pressure to something lower, exhalation relief is not one of them.

Encyclopaedia Brittanica: The mechanics of breathing

_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: Machine: Resmed AirSense10 for Her with Climateline heated hose ; alternating masks.
And now here is my secret, a very simple secret; it is only with the heart that one can see rightly, what is essential is invisible to the eye.
Antoine de Saint-Exupery

Good advice is compromised by missing data
Forum member Dog Slobber Nov. 2023

scottjf8
Posts: 128
Joined: Sat Jul 16, 2011 7:52 pm

Re: EPR settings

Post by scottjf8 » Sat Oct 22, 2011 7:49 am

I moved my EPR from 3 to 1 last night. Slept pretty good overall.

_________________
Mask: Quattro™ FX Full Face CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Started therapy 7/15/2011