ResScan - Any way to tell OSA's from CA's?

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
User avatar
thegregster
Posts: 15
Joined: Fri May 07, 2010 4:15 pm

ResScan - Any way to tell OSA's from CA's?

Post by thegregster » Wed Oct 19, 2011 3:32 pm

I recently started dowloading info from my Resmed VPAP-S (similar to Resmed S8 Series) into ResScan 3.14. Is there any way to distinguish central apneas from obstructive apneas? Am I corect in assuming the AI and AHI numbers include both types?

User avatar
Pugsy
Posts: 65411
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: ResScan - Any way to tell OSA's from CA's?

Post by Pugsy » Wed Oct 19, 2011 3:38 pm

S8 machines did not score/record centrals. S8 machines scored obstructive apneas and Hyponea events.

In ResScan and using the S9 machines the AI (is apneas + centrals)...AHI would be the AI plus the hyponeas.
Centrals are flagged separately from the Obstructive apneas by the S9 machine.

To see if you are having any centrals you would need to upgrade to S9 full data machine.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
thegregster
Posts: 15
Joined: Fri May 07, 2010 4:15 pm

Re: ResScan - Any way to tell OSA's from CA's?

Post by thegregster » Wed Oct 19, 2011 5:33 pm

That's good to know. Thanks for the answer.