My CPAP Machine will not fully start. Keeps turning ON/OFF:(

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
User avatar
GumbyCT
Posts: 5778
Joined: Fri Sep 14, 2007 6:22 pm
Location: CT
Contact:

Re: My CPAP Machine will not fully start. Keeps turning ON/OFF:(

Post by GumbyCT » Sat Sep 24, 2011 12:28 pm

Pugsy wrote:Contact cpap.com. Philips Respironics will just tell you to work through the supplier.

Are you talking about the "brick" portion being bad? Should be easy to get a replacement for it to see for sure.
This is where having a DC cable would be helpful.

_________________
Humidifier: HC150 Heated Humidifier With Hose, 2 Chambers and Stand
Additional Comments: New users can't remember they can't remember YET!
BeganCPAP31Jan2007;AHI<0.5
I have no doubt, how I sleep affects every waking moment.
I am making progress-NOW I remember that I can't remember
;)
If this isn’t rocket science why are there so many spaceshots?
Be your own healthcare advocate!

User avatar
archangle
Posts: 9293
Joined: Sun Mar 27, 2011 11:55 am

Re: My CPAP Machine will not fully start. Keeps turning ON/OFF:(

Post by archangle » Sat Sep 24, 2011 12:47 pm

sanjay0864 wrote:Looks like the power adapter may have totally failed. Which in a way may be a good thing, for it at least makes clear what the problem is.
Are you familiar enough with voltmeters to use one? It's pretty easy to probe the DC connector on the end of the cable coming from the brick. I got 12.2V or so on the output of my PRS1 brick.

_________________
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Also SleepyHead, PRS1 Auto, Respironics Auto M series, Legacy Auto, and Legacy Plus
Please enter your equipment in your profile so we can help you.
Click here for information on the most common alternative to CPAP.
If it's midnight and a DME tells you it's dark outside, go and check for yourself.

Useful Links.

User avatar
Pugsy
Posts: 65122
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: My CPAP Machine will not fully start. Keeps turning ON/OFF:(

Post by Pugsy » Sat Sep 24, 2011 1:04 pm

GumbyCT wrote:This is where having a DC cable would be helpful.

And this is where you totally lose me. Have no clue what you are talking about but I assume it has to do with not having to use that darn brick. I guess the brick does something but have no clue what. Complete dufus when it comes to this stuff. Now give me a computer or software problem... I can hold my own.
Sigh.. I am not well rounded in some areas. Electrical issues being one of them.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
GumbyCT
Posts: 5778
Joined: Fri Sep 14, 2007 6:22 pm
Location: CT
Contact:

Re: My CPAP Machine will not fully start. Keeps turning ON/OFF:(

Post by GumbyCT » Sat Sep 24, 2011 1:35 pm

Pugsy wrote:
GumbyCT wrote:This is where having a DC cable would be helpful.

And this is where you totally lose me. Have no clue what you are talking about but I assume it has to do with not having to use that darn brick. I guess the brick does something but have no clue what. Complete dufus when it comes to this stuff. Now give me a computer or software problem... I can hold my own.
Sigh.. I am not well rounded in some areas. Electrical issues being one of them.
Well yes, IF you eliminate the brick and it works correctly then you can be reasonably certain the brick is the problem.

_________________
Humidifier: HC150 Heated Humidifier With Hose, 2 Chambers and Stand
Additional Comments: New users can't remember they can't remember YET!
BeganCPAP31Jan2007;AHI<0.5
I have no doubt, how I sleep affects every waking moment.
I am making progress-NOW I remember that I can't remember
;)
If this isn’t rocket science why are there so many spaceshots?
Be your own healthcare advocate!

User avatar
archangle
Posts: 9293
Joined: Sun Mar 27, 2011 11:55 am

Re: My CPAP Machine will not fully start. Keeps turning ON/OFF:(

Post by archangle » Sat Sep 24, 2011 1:53 pm

Pugsy wrote:
GumbyCT wrote:This is where having a DC cable would be helpful.

And this is where you totally lose me. Have no clue what you are talking about but I assume it has to do with not having to use that darn brick.
If you had the DC cable, you could try plugging the CPAP machine into the cigarette lighter jack on your car and eliminate the brick completely.

_________________
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Also SleepyHead, PRS1 Auto, Respironics Auto M series, Legacy Auto, and Legacy Plus
Please enter your equipment in your profile so we can help you.
Click here for information on the most common alternative to CPAP.
If it's midnight and a DME tells you it's dark outside, go and check for yourself.

Useful Links.

User avatar
Pugsy
Posts: 65122
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: My CPAP Machine will not fully start. Keeps turning ON/OFF:(

Post by Pugsy » Sat Sep 24, 2011 3:03 pm

archangle wrote:If you had the DC cable, you could try plugging the CPAP machine into the cigarette lighter jack on your car and eliminate the brick completely
Gottcha. A possible way to see if the brick was defective. Good idea. I would need a different truck. Cigarette lighter quit working 60 K miles ago but I understand the principle. Husband explained DC and AC. These machines will work with DC I guess. Obviously my idea of camping is a camper with full electric, shower etc. That is why I avoid the "how to use machine in the woods threads". I used to do a lot of primitive camping back when I did the horseback endurance racing. I can't ride anymore like that so haven't been camping in years. I will leave the electrical stuff to you guys who are much more comfortable with it than I am.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

rkmontg342

Re: My CPAP Machine will not fully start. Keeps turning ON/OFF:(

Post by rkmontg342 » Sat Apr 28, 2012 5:08 pm

Okay here's another problem....I have Auto ON set to ON for coming ON automatically when I put ON the mask. It was turning itself on for a month (got this machine in February 2012), then stopped doing it. A few nights ago, it decided to turn itself on but only for that one night. Any suggestions? btw... it is: Respironics Remstar Pro C-flex. Thanks.

User avatar
pats
Posts: 289
Joined: Sun Apr 22, 2012 5:52 pm

Re: My CPAP Machine will not fully start. Keeps turning ON/OFF:(

Post by pats » Sat Apr 28, 2012 5:24 pm

sanjay0864 wrote:I think I may have not fully or correctly explained the problem I am facing with my CPAP machine. The problem is not that it won't start, because it does at times. It also works fine at times and I can go thru a whole night without any issues. But then when it decides to have problems, the machine will simply go into a loop, wherein it will start to boot and then shut off and then start to boot and shut off. If I unplug the power adapter from the wall socket and wait a couple of minutes and then plug it in again, the machine then starts properly. Strangely enough, if unplug the power adapter from the back of the machine, it does not stop the loop of starting - shutting off. I am now leaning to believe that the problem may lie with the power adapter and not the machine itself. The latest update is that now the machine is not powering up at all. Looks like the power adapter may have totally failed. Which in a way may be a good thing, for it at least makes clear what the problem is.
I've seen similar problems, in other electro-mechanical equipment, caused by a failing capacitor in the power supply. The capacitor should be maintaining the right voltage during the power up, when a lot of power is used to get things started. As it weakens, more and more attempts to power up fail, leading to increased cycling. Unplugging, waiting long enough for all capacitors in the power supply to discharge, and then plugging in again gives it a clean start. During this phase, the machine will go on working after a successful power-up, unless something puts an unusual demand on the power supply. Finally, it stops being able to even try.

If I saw the early stages of this type of behavior in something important, I would work on getting it replaced ASAP.

_________________
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear
Additional Comments: Software: SleepyHead.

User avatar
Julie
Posts: 20051
Joined: Tue Feb 28, 2006 12:58 pm

Re: My CPAP Machine will not fully start. Keeps turning ON/OFF:(

Post by Julie » Sat Apr 28, 2012 8:22 pm

Do you realize you're replying to a six month old post?

User avatar
GumbyCT
Posts: 5778
Joined: Fri Sep 14, 2007 6:22 pm
Location: CT
Contact:

Re: My CPAP Machine will not fully start. Keeps turning ON/OFF:(

Post by GumbyCT » Sat Apr 28, 2012 8:56 pm

rkmontg342 wrote:Okay here's another problem....I have Auto ON set to ON for coming ON automatically when I put ON the mask. It was turning itself on for a month (got this machine in February 2012), then stopped doing it. A few nights ago, it decided to turn itself on but only for that one night. Any suggestions? btw... it is: Respironics Remstar Pro C-flex. Thanks.
It sounds like you have a mask sealing problem. When the mask doesn't seal tight the machine doesn't sense you breathing and therefore does NOT turn On. No clue given as to mask make or model.

_________________
Humidifier: HC150 Heated Humidifier With Hose, 2 Chambers and Stand
Additional Comments: New users can't remember they can't remember YET!
BeganCPAP31Jan2007;AHI<0.5
I have no doubt, how I sleep affects every waking moment.
I am making progress-NOW I remember that I can't remember
;)
If this isn’t rocket science why are there so many spaceshots?
Be your own healthcare advocate!

Rich rich rich

Re: My CPAP Machine will not fully start. Keeps turning ON/OFF:(

Post by Rich rich rich » Tue May 24, 2016 10:55 pm

The fix i found is to put an extra screw on the back right of the printed circuit board.
Remove knob by pulling straight up. Then you need to use a torx driver T15 to open the casing. Then the lid lifts off. Find a short screw more than 4mm thick to insert in the hole already in the board. There is a hollow tube for the screw to attach to. This keeps the pressure on the seals of the air pressure sensors.

kimberly@yahoo.com

Re: My CPAP Machine will not fully start. Keeps turning ON/OFF:(

Post by kimberly@yahoo.com » Sat Oct 21, 2017 2:07 pm

The red light on my cpap machine keeps beeping and waking me up. I got a different mask that just covers my nose, which I prefer, but this is when the beeping started.
I had the full mask covering nose and mouth before. I move around a lot in my sleep. I'm a very restless sleeper and sometimes end up on my face. I am so sick and feel horrible when I wake up. My doctor told me I have the worst case of central sleep apnea he's ever seen. I'm so frustrated, and sick and tired of being sick and tired

User avatar
Goofproof
Posts: 16087
Joined: Mon Dec 05, 2005 3:16 pm
Location: Central Indiana, USA

Re: My CPAP Machine will not fully start. Keeps turning ON/OFF:(

Post by Goofproof » Sat Oct 21, 2017 2:48 pm

Sounds like your XPAP is telling you by switching from a FF to a nasal mask, you are killing your treatment by air escaping thru your mouth. If you used software to monitor your treatment, it could show you. Jim
Use data to optimize your xPAP treatment!

"The art of medicine consists in amusing the patient while nature cures the disease." Voltaire

User avatar
Okie bipap
Posts: 3567
Joined: Thu Oct 15, 2015 4:14 pm
Location: Central Oklahoma

Re: My CPAP Machine will not fully start. Keeps turning ON/OFF:(

Post by Okie bipap » Sat Oct 21, 2017 3:15 pm

You did not give enough information to be helped. Please register and fill out the information on your machine and mask, In the mean time, what machine are you using, and what mask are you currently using. It would also be nice ti know what full face mask you were using before the new mask. Could it be a large leak alarm going off on your machine? If you were using a full face mask in the future and did not get this red light and beeping, it could be caused by air escaping through your mouth when you are sleeping.

You will be better off starting a new post of your own instead of resurrecting an old post.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Mask: Evora Full Face Mask - Fitpack
Additional Comments: IPAP 20-25, ps 4, OSCAR software
Growing old is mandatory, but growing up is optional.