Rescan software help please

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
User avatar
Merryj
Posts: 82
Joined: Wed Jun 22, 2011 6:23 pm
Location: Central New York

Rescan software help please

Post by Merryj » Fri Sep 09, 2011 6:37 pm

I just downloaded the rescan software to my computer and downloaded my statistics. For some reason it does not show me leak data? It show usage, AHI and pressure and that is all. I have looked at the manuals and can't figure out why I can't get the leak data to show. Can anyone help?

_________________
Mask: Zest Nasal CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Trying to use the Opus 360 Nasal Pillow CPAP Mask with Headgear

User avatar
apneawho
Posts: 609
Joined: Wed Jul 27, 2011 10:09 am
Location: Atlanta, GA, USA

Re: Rescan software help please

Post by apneawho » Fri Sep 09, 2011 6:50 pm

Did you select detailed data as well as summary data?

_________________
Mask: Swift™ FX For Her Nasal Pillow CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: ResMed Climate line tubing, Hose cozy, PurSleep Clear aroma therapy, Using latest version Sleepyhead sofware-thanks Jedimark!

User avatar
Merryj
Posts: 82
Joined: Wed Jun 22, 2011 6:23 pm
Location: Central New York

Re: Rescan software help please

Post by Merryj » Fri Sep 09, 2011 6:58 pm

When it asked I had it download all available data - but these are the only reports that show up...

_________________
Mask: Zest Nasal CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Trying to use the Opus 360 Nasal Pillow CPAP Mask with Headgear

User avatar
Pugsy
Posts: 65327
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Rescan software help please

Post by Pugsy » Fri Sep 09, 2011 6:59 pm

Do you have AHI/pressure, flow...daily detailed graphs for each different night?
Isn't there a choice in ResScan as to what data to show? Make sure leak data is chosen.
If you have detailed data for all the other stuff and make sure the check mark is chosen to show leak, then unplug your machine and plug it back in forcing it to reboot and see if leak data shows up tomorrow.
Some people have had some weird stuff go on like AHI of 0.0 for 2 weeks straight or Leak stuck on 15 for a week and never change. May or may not be related to your problem but sure wouldn't hurt to try.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
apneawho
Posts: 609
Joined: Wed Jul 27, 2011 10:09 am
Location: Atlanta, GA, USA

Re: Rescan software help please

Post by apneawho » Fri Sep 09, 2011 7:00 pm

Do you see tab at the top that say summary graphs? You should be able to see leak graphed there.

_________________
Mask: Swift™ FX For Her Nasal Pillow CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: ResMed Climate line tubing, Hose cozy, PurSleep Clear aroma therapy, Using latest version Sleepyhead sofware-thanks Jedimark!

User avatar
apneawho
Posts: 609
Joined: Wed Jul 27, 2011 10:09 am
Location: Atlanta, GA, USA

Re: Rescan software help please

Post by apneawho » Fri Sep 09, 2011 7:10 pm

I just went into my software to see if I can tell what might be happening. When I first load my card, it automatically downloads summary info only. At the top of the page, there are tabs and if you select download and then select detailed data you should be able to get all data which will include leak. Summary graphs show everything for all days and detailed graph will show a single session in detail.

_________________
Mask: Swift™ FX For Her Nasal Pillow CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: ResMed Climate line tubing, Hose cozy, PurSleep Clear aroma therapy, Using latest version Sleepyhead sofware-thanks Jedimark!

User avatar
Merryj
Posts: 82
Joined: Wed Jun 22, 2011 6:23 pm
Location: Central New York

Re: Rescan software help please

Post by Merryj » Sat Sep 10, 2011 7:40 am

Thanks - I will try and download the information again and see what happens.

_________________
Mask: Zest Nasal CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Trying to use the Opus 360 Nasal Pillow CPAP Mask with Headgear

User avatar
Lizistired
Posts: 2835
Joined: Tue Dec 14, 2010 10:47 pm
Location: Indiana

Re: Rescan software help please

Post by Lizistired » Sat Sep 10, 2011 8:20 am

Here is the link to the tutorial.
http://montfordhouse.com/cpap/resscan_tutorial/

If you are downloading "all" and checking high rate data sessions and it still isn't coming up,
there is an icon of a graph with a check mark on it in the upper right area of the detailed window. Click there and select which info you want graphed and how many to display.
But really, watch the tutorial, it will make things alot easier.

_________________
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Swift FX sometimes, CMS-50F, Cervical collar sometimes, White noise, Zeo... I'm not well, but I'm better.

User avatar
apneawho
Posts: 609
Joined: Wed Jul 27, 2011 10:09 am
Location: Atlanta, GA, USA

Re: Rescan software help please

Post by apneawho » Sat Sep 10, 2011 11:39 am

[quote="Lizistired"]Here is the link to the tutorial.
http://montfordhouse.com/cpap/resscan_tutorial/

Thank you for posting that tutorial. It was extremely well done and helpful.

_________________
Mask: Swift™ FX For Her Nasal Pillow CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: ResMed Climate line tubing, Hose cozy, PurSleep Clear aroma therapy, Using latest version Sleepyhead sofware-thanks Jedimark!

User avatar
cherylann
Posts: 169
Joined: Sun Sep 04, 2011 9:51 pm
Location: Tulsa, OK

Re: Rescan software help please

Post by cherylann » Sat Sep 10, 2011 5:50 pm

Lizistired wrote:Here is the link to the tutorial.
http://montfordhouse.com/cpap/resscan_tutorial/

If you are downloading "all" and checking high rate data sessions and it still isn't coming up,
there is an icon of a graph with a check mark on it in the upper right area of the detailed window. Click there and select which info you want graphed and how many to display.
But really, watch the tutorial, it will make things alot easier.
Thanks so much for the tutorial. It's been a great help to me as well.

_________________
MachineMask
My priorities in life are God, family, and baseball. Starting in April, I do change the order around a bit.

User avatar
Merryj
Posts: 82
Joined: Wed Jun 22, 2011 6:23 pm
Location: Central New York

Re: Rescan software help please

Post by Merryj » Sat Sep 10, 2011 6:40 pm

Thanks for the tutorial. I may have to try and download and try asking for the high rate data sessions.

Could it be that my machine is not set up to track leak rate etc? Maybe I have to call my dme..

_________________
Mask: Zest Nasal CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Trying to use the Opus 360 Nasal Pillow CPAP Mask with Headgear

User avatar
apneawho
Posts: 609
Joined: Wed Jul 27, 2011 10:09 am
Location: Atlanta, GA, USA

Re: Rescan software help please

Post by apneawho » Sat Sep 10, 2011 7:13 pm

Merryj wrote:Thanks for the tutorial. I may have to try and download and try asking for the high rate data sessions.

Could it be that my machine is not set up to track leak rate etc? Maybe I have to call my dme..
I think it is probably not your machine, but that you have to look at the parameters on your software graphs. Did you see that part of the tutorial where you can select and deselect graph parameters?

_________________
Mask: Swift™ FX For Her Nasal Pillow CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: ResMed Climate line tubing, Hose cozy, PurSleep Clear aroma therapy, Using latest version Sleepyhead sofware-thanks Jedimark!

User avatar
Merryj
Posts: 82
Joined: Wed Jun 22, 2011 6:23 pm
Location: Central New York

Re: Rescan software help please

Post by Merryj » Sun Sep 11, 2011 6:15 am

I do see that, but there is nothing there but these three items AHI, Pressure, Usage. Under detailed graphs it just says no data available.

I am frustrated that I can't figure this out.

_________________
Mask: Zest Nasal CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Trying to use the Opus 360 Nasal Pillow CPAP Mask with Headgear

User avatar
AmIawake
Posts: 44
Joined: Fri Jul 08, 2011 5:41 pm

Re: Rescan software help please

Post by AmIawake » Sun Sep 11, 2011 9:28 am

Make sure you have updated your download options to include detailed data and high rate data.

Download your card again. Be sure to select OVERWRITE when prompted.

ResScan is buggy. If it didn't download correctly the first time and you don't overwrite on the next try, you will still have the same incomplete data/report staring back at you. Over the past week, I have had to download the data with overwrite 4 times. That's not a good track record for software used in patient monitoring. Wonder what a DME would do? No I guess I don't wonder.

_________________
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: ResScan 4.30
When do I get the t-shirt?