AHI is rising HELP

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
csroy

AHI is rising HELP

Post by csroy » Tue Aug 16, 2011 2:13 pm

i Have been using APAP for a few months now. Just had UPPP tonsillectomy and had Septoplasty, etc a year ago. My AHI is recently rising and would like some feedback on some of my data. Was wondering if i should try to go back to CPAP (where i started). When i was titrated they never found an ideal pressure for me. I started at 14, which blew me up like a balloon, went down to 12 which helped the bloating (at that time had no data from the cpap). Now i have been on apap and can see the results. Again should i try cpap at 10 or 11 given my data???? Any ideas as i am quite tired

Summary of Daily Events Per Hour
8/11/2011 - 8/15/2011
Total AHI: 7.2
P 4 5 6 7 8 9 10 .11 12 13 14 15 16 17 18 19 20
MaP 0.0 0.0 0.0 0.0 885.2 657.9 477.9 287.0 25.6 0.0 0.0 0.0 0.0 0.0 0.0 0.0 0.0
% 0.0 0.0 0.0 0.0 37.9 28.2 20.5 12.3 1.1 0.0 0.0 0.0 0.0 0.0 0.0 0.0 0.0
FL 0.0 0.0 0.0 0.0 0.9 0.3 0.1 0.2 0.0 0.0 0.0 0.0 0.0 0.0 0.0 0.0 0.0
VS 0.0 0.0 0.0 0.0 0.9 0.5 0.1 0.2 2.3 0.0 0.0 0.0 0.0 0.0 0.0 0.0 0.0
OA 0.0 0.0 0.0 0.0 0.7 0.4 1.3 0.6 0.0 0.0 0.0 0.0 0.0 0.0 0.0 0.0 0.0
CA 0.0 0.0 0.0 0.0 1.4 2.0 2.1 4.8 11.7 0.0 0.0 0.0 0.0 0.0 0.0 0.0 0.0
H 0.0 0.0 0.0 0.0 5.2 4.6 3.0 3.3 4.7 0.0 0.0 0.0 0.0 0.0 0.0 0.0 0.0
RE 0.0 0.0 0.0 0.0 1.7 1.8 2.8 3.8 0.0 0.0 0.0 0.0 0.0 0.0 0.0 0.0 0.0
AHI 0.0 0.0 0.0 0.0 7.3 7.0 6.4 8.7 16.4 0.0 0.0 0.0 0.0 0.0 0.0 0.0 0.0
90%
P - Pressure, MaP - Minutes at Pressure, % - Percent of Night, FL - Flow Limitation, VS - Vibratory Snore, H - Hypopnea,
OA - Obstructed Airway Apnea, CA - Clear Airway Apnea, RE - RERA, AHI - Apnea/Hypopnea Index Legend
8/11/2011 - 8/15/2011
Auto CPAP Time at Pressure

i dont know how to attach the file so i cut and pasted the 5 day summary info. How would i attach a file in my message so the data is more clear.

Thanks,

Guest

Re: AHI is rising HELP

Post by Guest » Tue Aug 16, 2011 2:37 pm

Ok Maybe this info will be easier to read?

Summary of Daily Events Per Hour
8/11/2011 - 8/15/2011
Total AHI: 7.2

p 8 9 10 11 12 13 14 15 16
Map 885.2 657.9477.9287 25.6
% 37.9 28.2 20.5 12.3 1.1
FL 0.9 0.3 0.1 0.2 0
VS 0.9 0.5 0.1 0.2 2.3
OA 0.7 0.4 1.3 0.6 0
CA 1.4 2 2.1 4.8 11.7
H 5.2 4.6 3 3.3 4.7
RE 1.7 1.8 2.8 3.8 0
AHI 7.3 7 6.4 8.7 16.4
90%
P - Pressure, MaP - Minutes at Pressure, % - Percent of Night, FL - Flow Limitation, VS - Vibratory Snore, H - Hypopnea,
OA - Obstructed Airway Apnea, CA - Clear Airway Apnea, RE - RERA, AHI - Apnea/Hypopnea Index Legend
8/11/2011 - 8/15/2011
Auto CPAP Time at Pressure

User avatar
Julie
Posts: 20052
Joined: Tue Feb 28, 2006 12:58 pm

Re: AHI is rising HELP

Post by Julie » Tue Aug 16, 2011 2:51 pm

If you go to User Ctl. Panel under the logo on the main page, click on Profile and then using text vs icons add the full names of your machine and mask, as well as your pressure range, etc. it'll automatically appear under your postings so we can see what you're using... makes it much easier to help you.

csroy8014
Posts: 3
Joined: Tue Aug 16, 2011 2:04 pm

Re: AHI is rising HELP

Post by csroy8014 » Tue Aug 16, 2011 5:31 pm

I have spaced the info a little better and i tried to change my profile to reveal my equipment. I am on a remstar system one apap with a quattro mirage full face mask. Apap is set from 8-16. When i was had a sleep study they could not find an ideal setting for me. They got to 12 during the titration but that was not ideal, the Dr. recommended i start at 14 (which blew my up like a ballon). They then lowered me to 12 which helped a little. I then went to an apap which cured the gas issue but recently my ahi and software results are all over. Should i think about going back to cpap? Maybe try 10? Is the apap contributing to my hypopneas? I have not clue how to get this ahi under control. I am up for an ideas?

Thanks,

Summary of Daily Events Per Hour
8/11/2011 - 8/15/2011
Total AHI: 7.2

p 8 9 10 11 12 13 14 15 16
Map 885.2 657.9 477.9 287 25.6
% 37.9 28.2 20.5 12.3 1.1
FL 0.9 0.3 0.1 0.2 0
VS 0.9 0.5 0.1 0.2 2.3
OA 0.7 0.4 1.3 0.6 0
CA 1.4 2 2.1 4.8 11.7
H 5.2 4.6 3 3.3 4.7
RE 1.7 1.8 2.8 3.8 0
AHI 7.3 7 6.4 8.7 16.4
90%
P - Pressure, MaP - Minutes at Pressure, % - Percent of Night, FL - Flow Limitation, VS - Vibratory Snore, H - Hypopnea,
OA - Obstructed Airway Apnea, CA - Clear Airway Apnea, RE - RERA, AHI - Apnea/Hypopnea Index Legend
8/11/2011 - 8/15/2011
Auto CPAP Time at Pressure

User avatar
Julie
Posts: 20052
Joined: Tue Feb 28, 2006 12:58 pm

Re: AHI is rising HELP

Post by Julie » Tue Aug 16, 2011 5:54 pm

Hi - Can you get the summary from your sleep study - the full report, not just the 1-2 pg summary, because that will tell you more about some of your results, such as whether or not you had centrals, or mostly hypopneas vs obstructive events, etc.?

I also wonder about your having had all that surgery... did you research any of it (e.g. the UPPP) prior to signing on? It can change the picture quite a bit for some people and, unfortunately, may seem to be a good thing for a while, but too often you can end up (within a year, not immediately) back where you were before having it done.

User avatar
Pugsy
Posts: 65130
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: AHI is rising HELP

Post by Pugsy » Tue Aug 16, 2011 6:45 pm

Off hand I think that perhaps the minimum pressure is a bit to low but it would be best to see what one night looks like. When the minimum is too low it takes too long for the machine to raise the pressure to where it needs to be to prevent the events effectively. Need to give it a bit more of a head start. I started out with 8 cm APAP and was having around 10 AHI. When I worked my way to 10 cm minimum (max was 20) my AHI was reduced to 3 to 4.

It would be better if we could see one a single night detailed daily report. Just pick one that seems to be pretty much what you usually see.

Easiest way is to do a screen shot of the report.
Here is how I do it.

Open the image to full size so it is easily read.
I use Vista snipping tool to create a screen shot and crop the image at the same time.
Prt/scr key will also take a screen shot if using XP. If laptop is used sometimes the Fn key has to be pushed at the same time as the prt/scr key
I think windows 7 Home Premium has the snipping tool, Basic may not.
Once the screen shot is created save it in jpg format.
Upload the image to a host site. I use Photobucket it is free, there are others.
Once the image is uploaded then copy the ENTIRE IMG address. Be sure to include the opening and closing IMG in brackets. Paste that copied address into a post here.
Use the preview button. If you can't see the image try again because if you can't see it we can't.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

Guest

Re: AHI is rising HELP

Post by Guest » Tue Aug 16, 2011 7:24 pm

Thanks for the info its a great help. I will try to see if i have the ability to complete the technological feet and post a single night of data.

As for now i may try to set my min at 9 instead of 8 and see if that helps at all. I imaging i must try that for several nights to get a real idea? My 90% pressure is usually 10-11, sometimes a little less or a little more but most commonly 10-11.

Good Idea

User avatar
Pugsy
Posts: 65130
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: AHI is rising HELP

Post by Pugsy » Tue Aug 16, 2011 7:37 pm

Guest wrote:As for now i may try to set my min at 9 instead of 8 and see if that helps at all. I imaging i must try that for several nights to get a real idea? My 90% pressure is usually 10-11, sometimes a little less or a little more but most commonly 10-11.
Yes. You can give 9 cm minimum as trial. Watch your reports (daily reports) you will know in 4 or 5 nights if it is helping. Sometimes just a seemingly small change makes a huge difference.

Please do try to post the report. It helps us when we can see what you are seeing.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

csroy8014
Posts: 3
Joined: Tue Aug 16, 2011 2:04 pm

Re: AHI is rising HELP

Post by csroy8014 » Wed Aug 17, 2011 5:53 am

I am not sure how to do the img start and stop thing. I just created one album on photobucket with the info you requested. I believe i can give you a link to it?

Lets give it a try

http://photobucket/csroy8014

I hope that will show you 5 days of reports and the summary. They may be a little out of order but the info is there. If it is easier to do the img thing let me know. I just dont understand what to type into the two sets of brackets?

Thanks a million!!

User avatar
Pugsy
Posts: 65130
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: AHI is rising HELP

Post by Pugsy » Wed Aug 17, 2011 7:24 am

When at photobucket you will see this box when you hover over one image
Image

See the last line in that box? The line that starts with IMG brackets? Put mouse cursor over it and copy that entire address which will include the IMG box. Then right click mouse to paste it in a post here. Then the image shows up. When doing it this way you don't have to use the IMG or URL brackets on the forum editor. The IMG brackets are already included in the copy.

Image

Above is one of your reports after I finally got the link to open. BTW yes, just a little more minimum pressure and see if it reduces those events.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

Guest

Re: AHI is rising HELP

Post by Guest » Wed Aug 17, 2011 8:07 am

I sincerely appreciate all you help learning to use the message board and also with the APAP. I appears the wealth of knowledge here far exceeds the reactionary info i receive from my ENT DR. I know how to make changes on my machine so the trips to the DME seem senseless ( as long as i dont make drastic changes). I appreciate the feedback and will get back to you on the results of raising my min pressure from 8-9.

Thanks again.

csroy8014
Posts: 3
Joined: Tue Aug 16, 2011 2:04 pm

Re: AHI is rising HELP

Post by csroy8014 » Wed Aug 17, 2011 12:47 pm

I had my first night changing my minimum pressure from 8 to 9. Seemed a little better, do i have to give it several days at 9 before trying say 9.5 or 10 as a min?

Here it is below.

Thanks in advance

Image