Wireless CPAP Modem Causing Internet to Cut Out?

Post by napstress » Sat Jun 11, 2011 11:34 am

Is it possible that the wireless modem on my CPAP machine is causing our Internet to cut out intermittently? If so, could that happen even if the CPAP machine is disconnected from a power source? When I called Verizon about the Internet cutting out, tech support asked if we have a house alarm. We don't. I didn't think of it at the time, but we did start having connectivity problems around the time I brought the CPAP machine home. We lost Internet connection yesterday but the machine was unplugged from the wall.
Epworth Sleepiness Scale: 14
Diagnostic study: overall AHI: 0.2 events/hour; overall RDI: 45 events/hour
Titration study: AHI: 6.1; RDI: 27; CPAP pressures: 5-8cm

Not-tired behind my eyes and with a clear, cool head!

User avatar
Goofproof
Posts: 16087
Joined: Mon Dec 05, 2005 3:16 pm
Location: Central Indiana, USA

Re: Wireless CPAP Modem Causing Internet to Cut Out?

Post by Goofproof » Sat Jun 11, 2011 11:40 am

Do you have it plugged into their DSL interface as ALL phone jacks in the house should have. You should have gotten plenty for all lines, I think five came with mine. Jim

Trying to get phone help from Verizon is useless, might as well be talking to a bowl of Curry.
Use data to optimize your xPAP treatment!

"The art of medicine consists in amusing the patient while nature cures the disease." Voltaire

User avatar
GumbyCT
Posts: 5778
Joined: Fri Sep 14, 2007 6:22 pm
Location: CT
Contact:

Re: Wireless CPAP Modem Causing Internet to Cut Out?

Post by GumbyCT » Sat Jun 11, 2011 11:52 am

Our Dish satellite box also likes to phone home.

_________________
Humidifier: HC150 Heated Humidifier With Hose, 2 Chambers and Stand
Additional Comments: New users can't remember they can't remember YET!
BeganCPAP31Jan2007;AHI<0.5
I have no doubt, how I sleep affects every waking moment.
I am making progress-NOW I remember that I can't remember
;)
If this isn’t rocket science why are there so many spaceshots?
Be your own healthcare advocate!

User avatar
archangle
Posts: 9293
Joined: Sun Mar 27, 2011 11:55 am

Re: Wireless CPAP Modem Causing Internet to Cut Out?

Post by archangle » Sat Jun 11, 2011 1:12 pm

I believe you said earlier that your modem doesn't plug into the phone line or the internet. There's no wire coming out from the modem to the phone line or internet, and you don't have a second box that plugs into the phone or internet. If that's true, it won't affect your internet.

If the machine is unplugged, it is even less likely.

I'm assuming your Verizon internet comes into your house on some sort of cable, not an antenna.

Don't tell Verizon about the CPAP, unless they specifically ask you about a CPAP machine. They're just looking for an excuse to blame you and close the service call and make their numbers look good.

Be tough with them. Don't let them close the service call until it's fixed. Keep calling them until it's fixed. At some point, they'll get tired of having to come back and try to actually find the problem.

_________________
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Also SleepyHead, PRS1 Auto, Respironics Auto M series, Legacy Auto, and Legacy Plus
Please enter your equipment in your profile so we can help you.
Click here for information on the most common alternative to CPAP.
If it's midnight and a DME tells you it's dark outside, go and check for yourself.

Useful Links.

mebeingme

Re: Wireless CPAP Modem Causing Internet to Cut Out?

Post by mebeingme » Sat Jun 11, 2011 8:15 pm

the wireless modems run off of cell signal.

User avatar
GumbyCT
Posts: 5778
Joined: Fri Sep 14, 2007 6:22 pm
Location: CT
Contact:

Re: Wireless CPAP Modem Causing Internet to Cut Out?

Post by GumbyCT » Sat Jun 11, 2011 8:55 pm

mebeingme wrote:the wireless modems run off of cell signal.
and who pays for that?

_________________
Humidifier: HC150 Heated Humidifier With Hose, 2 Chambers and Stand
Additional Comments: New users can't remember they can't remember YET!
BeganCPAP31Jan2007;AHI<0.5
I have no doubt, how I sleep affects every waking moment.
I am making progress-NOW I remember that I can't remember
;)
If this isn’t rocket science why are there so many spaceshots?
Be your own healthcare advocate!

User avatar
idamtnboy
Posts: 2186
Joined: Mon Nov 01, 2010 2:12 pm
Location: Idaho

Re: Wireless CPAP Modem Causing Internet to Cut Out?

Post by idamtnboy » Sat Jun 11, 2011 10:26 pm

GumbyCT wrote:
mebeingme wrote:the wireless modems run off of cell signal.
and who pays for that?
If it's like the Kindle, it's built into the cost of service. The Kindle cell phone cost is part of the cost of the books when you download them.

_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Hose management - rubber band tied to casement window crank handle! Hey, it works! S/W is 3.13, not 3.7

User avatar
napstress
Posts: 462
Joined: Sat Apr 09, 2011 6:59 am

Re: Wireless CPAP Modem Causing Internet to Cut Out?

Post by napstress » Sun Jun 12, 2011 1:06 pm

archangle wrote:I believe you said earlier that your modem doesn't plug into the phone line or the internet. There's no wire coming out from the modem to the phone line or internet, and you don't have a second box that plugs into the phone or internet. If that's true, it won't affect your internet.

If the machine is unplugged, it is even less likely.

I'm assuming your Verizon internet comes into your house on some sort of cable, not an antenna.

Don't tell Verizon about the CPAP, unless they specifically ask you about a CPAP machine. They're just looking for an excuse to blame you and close the service call and make their numbers look good.

Be tough with them. Don't let them close the service call until it's fixed. Keep calling them until it's fixed. At some point, they'll get tired of having to come back and try to actually find the problem.
Thanks for answering my question, Archangle. I believe our Internet comes in through cable.

Thanks also for the heads up about not telling Verizon about the CPAP machine and also how to handle them. So far, they are testing the computer modem remotely and have not mentioned making a house call.
Epworth Sleepiness Scale: 14
Diagnostic study: overall AHI: 0.2 events/hour; overall RDI: 45 events/hour
Titration study: AHI: 6.1; RDI: 27; CPAP pressures: 5-8cm

Not-tired behind my eyes and with a clear, cool head!

User avatar
GumbyCT
Posts: 5778
Joined: Fri Sep 14, 2007 6:22 pm
Location: CT
Contact:

Re: Wireless CPAP Modem Causing Internet to Cut Out?

Post by GumbyCT » Mon Jun 13, 2011 7:01 am

idamtnboy wrote:
GumbyCT wrote:
mebeingme wrote:the wireless modems run off of cell signal.
and who pays for that?
If it's like the Kindle, it's built into the cost of service. The Kindle cell phone cost is part of the cost of the books when you download them.
It's not a kindle. Naptress do you know who pays for the cell service to your modem?

_________________
Humidifier: HC150 Heated Humidifier With Hose, 2 Chambers and Stand
Additional Comments: New users can't remember they can't remember YET!
BeganCPAP31Jan2007;AHI<0.5
I have no doubt, how I sleep affects every waking moment.
I am making progress-NOW I remember that I can't remember
;)
If this isn’t rocket science why are there so many spaceshots?
Be your own healthcare advocate!

User avatar
archangle
Posts: 9293
Joined: Sun Mar 27, 2011 11:55 am

Re: Wireless CPAP Modem Causing Internet to Cut Out?

Post by archangle » Mon Jun 13, 2011 2:32 pm

GumbyCT wrote:It's not a kindle. Naptress do you know who pays for the cell service to your modem?
I believe there are commercial services that provide this service to the DMEs. They provide the modem, wireless service, servers, and a web site that monitors the modem to the DME. I presume the service subcontracts wireless access with the existing cell phone network providers. The DME resells the monitoring service and/or bundles it in with their other services.

I got the impression that some of these services are run by the CPAP manufacturers, or by someone with close ties. I think I saw this service listed on one of the CPAP manufacturer's website. There may be some independent companies that do this, or some of the bigger DME groups may run their own. I doubt "Aunt Mary's power tool and respiratory equipment company" handles any of the technical details.

Napstress, do you get a separate bill for the modem and monitoring, or just one bill from the DME and/or doctor?

_________________
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Also SleepyHead, PRS1 Auto, Respironics Auto M series, Legacy Auto, and Legacy Plus
Please enter your equipment in your profile so we can help you.
Click here for information on the most common alternative to CPAP.
If it's midnight and a DME tells you it's dark outside, go and check for yourself.

Useful Links.

User avatar
napstress
Posts: 462
Joined: Sat Apr 09, 2011 6:59 am

Re: Wireless CPAP Modem Causing Internet to Cut Out?

Post by napstress » Tue Jun 14, 2011 3:23 am

GumbyCT wrote:do you know who pays for the cell service to your modem?
archangle wrote:do you get a separate bill for the modem and monitoring, or just one bill from the DME and/or doctor?
I'm assuming you mean the modem on my CPAP machine, right? I do not know who pays for the cell service to that modem. I only get one bill from the DME, who supplies the CPAP machine. What are you guys thinking, here? That I should call them to see if their modem can interfere with my Internet connectivity?
Epworth Sleepiness Scale: 14
Diagnostic study: overall AHI: 0.2 events/hour; overall RDI: 45 events/hour
Titration study: AHI: 6.1; RDI: 27; CPAP pressures: 5-8cm

Not-tired behind my eyes and with a clear, cool head!

User avatar
archangle
Posts: 9293
Joined: Sun Mar 27, 2011 11:55 am

Re: Wireless CPAP Modem Causing Internet to Cut Out?

Post by archangle » Tue Jun 14, 2011 8:13 am

napstress wrote:What are you guys thinking, here? That I should call them to see if their modem can interfere with my Internet connectivity?
No. It's not interfering with your internet. Especially since you said you unplugged the CPAP and still had internet problems.

It's extremely unlikely it would interfere with your internet anyway.

Is your internet still messed up?

_________________
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Also SleepyHead, PRS1 Auto, Respironics Auto M series, Legacy Auto, and Legacy Plus
Please enter your equipment in your profile so we can help you.
Click here for information on the most common alternative to CPAP.
If it's midnight and a DME tells you it's dark outside, go and check for yourself.

Useful Links.