No Answers Yet For Problems With ASV

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
Guest

Re: No Answers Yet For Problems With ASV

Post by Guest » Thu Jul 14, 2011 7:35 pm

NotMuffy wrote: The body needs to build up CO2 and then a regular breathing pattern resumes. This phenomenon occurs following an "arousal" - a sleep disturbance that causes a deep breath, or "sigh" followed by a compensatory pause.

............

Those need to be left alone.
Should machine-induced centrals also be left alone?

Exactly how does a machine-induced central develop?

User avatar
dsm
Posts: 6996
Joined: Mon Jun 20, 2005 6:53 am
Location: Near the coast.

Re: No Answers Yet For Problems With ASV

Post by dsm » Thu Jul 14, 2011 8:19 pm

Guest wrote:
NotMuffy wrote: The body needs to build up CO2 and then a regular breathing pattern resumes. This phenomenon occurs following an "arousal" - a sleep disturbance that causes a deep breath, or "sigh" followed by a compensatory pause.

............

Those need to be left alone.
Should machine-induced centrals also be left alone?

Exactly how does a machine-induced central develop?
Guest, Here is but one scenario (as I understand it) for that situation (in very simple terms) ...

If your xPAP machine is set incorrectly (cpap/auto with pressure too high for what is needed, bipap with badly set risetime and/or epap/ipap gap) and an event (e.g. sleep arousal) results in you breathing faster/deeper than you would normally then the result stemming from the machine's intervention can cause your breathing to drop your blood CO2 level to below your normal blood CO2 threshold, in which case your respiratory sensing (chemoreceptors) will signal to your breathing to slow or pause until the sensing system sees your blood CO2 level back in range.

I believe there are several factors that can affect the extent/depth of the reaction.

Should machine induced centrals be responded to - of course but that may involve changing the machine's set up. If the machine is causing central dysfunction, it needs attention.

Cheers

DSM
xPAP and Quattro std mask (plus a pad-a-cheek anti-leak strap)

User avatar
Paper_Nanny
Posts: 430
Joined: Fri Apr 15, 2011 3:52 pm
Location: Southern Oregon

Re: No Answers Yet For Problems With ASV

Post by Paper_Nanny » Thu Jul 14, 2011 9:20 pm

NotMuffy wrote:Anyway, you only breathe when you need to (insert 47 exceptions here later).
I thought two more exceptions were added last year, bringing the total to 49, with the exception of the Czech Republic, where it is still only 35.
NotMuffy wrote:This phenomenon occurs following an "arousal" - a sleep disturbance that causes a deep breath, or "sigh" followed by a compensatory pause.
Are arousals the only time this occurs?
NotMuffy wrote:Anyway, cross out all those centrals in the waveforms and recalculate your AHI.
I will say 6.1. It is my final answer.

Deborah

_________________
Mask: FitLife Total Face CPAP Mask with Headgear
Humidifier: DreamStation Heated Humidifier
Additional Comments: DreamSTation DSX500
PR System One BiPAP Auto SV Advanced

StillAnotherGuess
Posts: 132
Joined: Sat Apr 23, 2011 4:26 pm

Re: No Answers Yet For Problems With ASV

Post by StillAnotherGuess » Thu Jul 14, 2011 11:17 pm

Paper_Nanny wrote: I will say 6.1. It is my final answer.
Which may or may not make you smarter than a fifth grader, so lets say the real answer is 5. That would put you in a unique position of being a solution in search of a problem. The problem would still appear to be what is the problem?

_________________
Mask: Quattro™ FX Full Face CPAP Mask with Headgear
Additional Comments: S9 Adapt for Home and Travel, On-Board Firmware, Std. Tubing. EEP 9.8, Min PS 4.6, Max Pressure 21

User avatar
Paper_Nanny
Posts: 430
Joined: Fri Apr 15, 2011 3:52 pm
Location: Southern Oregon

Re: No Answers Yet For Problems With ASV

Post by Paper_Nanny » Fri Jul 15, 2011 12:23 am

StillAnotherGuess wrote:Which may or may not make you smarter than a fifth grader, so lets say the real answer is 5. That would put you in a unique position of being a solution in search of a problem. The problem would still appear to be what is the problem?
Oh, Silly Guess, that by no means puts me in a unique position. It puts me in exactly the same position as all human beings. We are all solutions seeking a problem. We all have our own special blends of talents and gifts. Part of the human condition is a search to find ways to use those for the benefit of others, while at the same time, finding those whose talents and gifts will be of benefit to us.

It is with that in mind that I ponder your most recent posts to this thread. I am trying to understand why you feel drawn to me and my situation. Why do you feel the need to say these things to me? Are you here to give help me or are you looking for me to give something to you? Or looking for something from someone else here? I ponder, but come to no conclusions. Ponder and ponder... Ponder and wander, meander and contemplate, knit and purl, pearl and pear as I go from here to there.

As for being smarter than a fifth grader, the way I look at it is I don't have to be smarter because I'm bigger. And I am allowed to cuss and they're not.

Deborah

_________________
Mask: FitLife Total Face CPAP Mask with Headgear
Humidifier: DreamStation Heated Humidifier
Additional Comments: DreamSTation DSX500
PR System One BiPAP Auto SV Advanced

ozij
Posts: 10469
Joined: Fri Mar 18, 2005 11:52 pm

Re: No Answers Yet For Problems With ASV

Post by ozij » Fri Jul 15, 2011 12:40 am


Deborah, your posts are giving me so much pleasure, and I don't mean just because of your incredible wit. It's the way you embroider the tapestry of humanity that has me awed and touched and chuckling all at once.

A thread full of wonder -- and like the best threads on this forum, teaching us about people at the same time as it teaches us about machines and breathing.

Thank you for continuing it.

_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: Machine: Resmed AirSense10 for Her with Climateline heated hose ; alternating masks.
And now here is my secret, a very simple secret; it is only with the heart that one can see rightly, what is essential is invisible to the eye.
Antoine de Saint-Exupery

Good advice is compromised by missing data
Forum member Dog Slobber Nov. 2023

User avatar
rested gal
Posts: 12880
Joined: Thu Sep 09, 2004 10:14 pm
Location: Tennessee

Re: No Answers Yet For Problems With ASV

Post by rested gal » Fri Jul 15, 2011 12:48 am

ozij wrote::lol:
Deborah, your posts are giving me so much pleasure, and I don't mean just because of your incredible wit. It's the way you embroider the tapestry of humanity that has me awed and touched and chuckling all at once.

A thread full of wonder -- and like the best threads on this forum, teaching us about people at the same time as it teaches us about machines and breathing.

Thank you for continuing it.
You're absolutely right, ozij... on all counts about Deborah ~~~ and her thread.
ResMed S9 VPAP Auto (ASV)
Humidifier: Integrated + Climate Control hose
Mask: Aeiomed Headrest (deconstructed, with homemade straps
3M painters tape over mouth
ALL LINKS by rested gal:
viewtopic.php?t=17435

User avatar
Mr Bill
Posts: 532
Joined: Fri Feb 18, 2011 11:56 pm
Location: Grand Junction, CO

Re: No Answers Yet For Problems With ASV

Post by Mr Bill » Fri Jul 15, 2011 1:50 am

NotMuffy wrote:If that time span was 10 seconds or more, that was a central apnea (not really, you'd need to be asleep, but pretend).

Anyway, you only breathe when you need to (insert 47 exceptions here later). The response to the large breath was normal. The body needs to build up CO2 and then a regular breathing pattern resumes. This phenomenon occurs following an "arousal" - a sleep disturbance that causes a deep breath, or "sigh" followed by a compensatory pause. On the waveforms, it looks like this:

Image

Those need to be left alone. I would imagine if you tried to attack it with an ASV, the already low pCO2 would be driven lower no matter how aggressive the attack is, and a run of chain centrals could occur (you would probably need to have some of that "increased chemoresponsiveness" for that to really perpetuate. Normal chemoresponsiveness is like a shock absorber. Hit a bump and you only bounce once).

Anyway, cross out all those centrals in the waveforms and recalculate your AHI.
I thought those sighs could be discounted! Cool. IIRC somebody noted that the Respironics ASV algorithm never intervenes on the first missed or abnormal breath but progressively intervenes on the ones following.

I see many pressure pulses marked with the red block but also pressure pulses with no red blocks. So it appears to me that the marked red pulses are marking pulses used to check for CA. Whereas the other unmarked pulses are interventions to keep things moving along.

What about the data around 5:37 AM and 5:43 AM? There are shallower but faster breaths and there is a pattern of pressure pulses that rises to a peak and then falls off repeated several times in the middle of those breaths. Is this an example of where pressure was used to intervene and keep all the breaths the same magnitude? In that case is this evidence that the ASV is doing something beneficial right there that simple CPAP or BIPAP would not accomplish?
EPAP min=6, EPAP max=15, PS min=3, PS max=12, Max Pressure=30, Backup Rate=8 bpm, Flex=0, Rise Time=1,
90% EPAP=7.0, Avg PS=4.0, Avg bpm 18.3, Avg Min vent 9.2 Lpm, Avg CA/OA/H/AHI = 0.1/0.1/2.1/2.3 ... updated 02/17/12

User avatar
NotMuffy
Posts: 962
Joined: Wed Sep 16, 2009 6:56 am
Location: Dunno. GPS is dead.

Re: No Answers Yet For Problems With ASV

Post by NotMuffy » Fri Jul 15, 2011 4:46 am

Paper_Nanny wrote:
NotMuffy wrote:This phenomenon occurs following an "arousal" - a sleep disturbance that causes a deep breath, or "sigh" followed by a compensatory pause.
Are arousals the only time this occurs?
As demonstrated, you can obviously do this in wake. The inability to discern actual sleep is a huge stumbling block in attempting to analyze single-channel data.

In addition, your frequency of awakenings (16 in the titration study, probably representative of your baseline) may also generate a lot of stuff that should be exluded (wake events are not scored during NPSG. Single waveform automated analysis obviously does not have that ability).
"Don't Blame Me...You Took the Red Pill..."

StillAnotherGuess
Posts: 132
Joined: Sat Apr 23, 2011 4:26 pm

Re: No Answers Yet For Problems With ASV

Post by StillAnotherGuess » Fri Jul 15, 2011 9:48 am

Paper_Nanny wrote:[. I am trying to understand why you feel drawn to me and my situation. Why do you feel the need to say these things to me? Are you here to give help me or are you looking for me to give something to you? Or looking for something from someone else here?
I'm drawn to you because I'm trying to understand what your situation is, currently, that you have not articulated very well. I also enjoy your humor and your wit. Your original problem of PTB is mitigated. You have a small residual AHI issue. CPAP will not cure fatigue, MS, or pills. What is your current CPAP problem?

_________________
Mask: Quattro™ FX Full Face CPAP Mask with Headgear
Additional Comments: S9 Adapt for Home and Travel, On-Board Firmware, Std. Tubing. EEP 9.8, Min PS 4.6, Max Pressure 21

User avatar
Paper_Nanny
Posts: 430
Joined: Fri Apr 15, 2011 3:52 pm
Location: Southern Oregon

Re: No Answers Yet For Problems With ASV

Post by Paper_Nanny » Fri Jul 15, 2011 11:05 am

Mr Bill wrote:In that case is this evidence that the ASV is doing something beneficial right there that simple CPAP or BIPAP would not accomplish?
Is my machine doing things like an ASV? I thought with the current settings, it was acting strictly as a BiPAP would. Settings: BPM=0, Pres Max=12, EPAP Max= 12, EPAP Min=8, PS Max= 4, PS Min=0.

Or is it still doing something ASVish because I wasn't able to set PS min= 4?

Deborah

_________________
Mask: FitLife Total Face CPAP Mask with Headgear
Humidifier: DreamStation Heated Humidifier
Additional Comments: DreamSTation DSX500
PR System One BiPAP Auto SV Advanced

User avatar
Paper_Nanny
Posts: 430
Joined: Fri Apr 15, 2011 3:52 pm
Location: Southern Oregon

Re: No Answers Yet For Problems With ASV

Post by Paper_Nanny » Fri Jul 15, 2011 11:13 am

NotMuffy wrote:
Paper_Nanny wrote:
NotMuffy wrote:This phenomenon occurs following an "arousal" - a sleep disturbance that causes a deep breath, or "sigh" followed by a compensatory pause.
Are arousals the only time this occurs?
As demonstrated, you can obviously do this in wake.
In sleep, are there reasons not associated with arousals that one would sigh?

Given that we cannot tell from my wave form data whether or not my sighs were post- arousal, I am asking more for the purpose of satisfying a curiosity rather than looking for practically applicable information.

Deborah

_________________
Mask: FitLife Total Face CPAP Mask with Headgear
Humidifier: DreamStation Heated Humidifier
Additional Comments: DreamSTation DSX500
PR System One BiPAP Auto SV Advanced

StillAnotherGuess
Posts: 132
Joined: Sat Apr 23, 2011 4:26 pm

Re: No Answers Yet For Problems With ASV

Post by StillAnotherGuess » Fri Jul 15, 2011 12:29 pm

Paper_Nanny wrote:Is my machine doing things like an ASV?
Yes. It is not only doing things like an ASV. It is currently set up in ASV mode. PS Max > 0 = ASV Mode.
Paper_Nanny wrote:I thought with the current settings, it was acting strictly as a BiPAP would.
To act as a BiPAP, you would need to set PS Min > 0.
Paper_Nanny wrote:Settings: BPM=0, Pres Max=12, EPAP Max= 12, EPAP Min=8, PS Max= 4, PS Min=0.
Typical ASV settings
Paper_Nanny wrote:Or is it still doing something ASVish because I wasn't able to set PS min= 4?
No. Currently PS Min=0 is sitting on your EPAP value. Think CPAP.

I'm not your enemy.

_________________
Mask: Quattro™ FX Full Face CPAP Mask with Headgear
Additional Comments: S9 Adapt for Home and Travel, On-Board Firmware, Std. Tubing. EEP 9.8, Min PS 4.6, Max Pressure 21
Last edited by StillAnotherGuess on Fri Jul 15, 2011 2:00 pm, edited 5 times in total.

StillAnotherGuess
Posts: 132
Joined: Sat Apr 23, 2011 4:26 pm

Re: No Answers Yet For Problems With ASV

Post by StillAnotherGuess » Fri Jul 15, 2011 1:17 pm

To state it more correctly, your machine is currently set up in CPAP mode with ASV assist.

_________________
Mask: Quattro™ FX Full Face CPAP Mask with Headgear
Additional Comments: S9 Adapt for Home and Travel, On-Board Firmware, Std. Tubing. EEP 9.8, Min PS 4.6, Max Pressure 21

StillAnotherGuess
Posts: 132
Joined: Sat Apr 23, 2011 4:26 pm

Re: No Answers Yet For Problems With ASV

Post by StillAnotherGuess » Fri Jul 15, 2011 1:34 pm

Which brings to mind that ASV assist may not be doing you any favors, as Bob and NM previously stated, so you could try setting PS Max=0.

_________________
Mask: Quattro™ FX Full Face CPAP Mask with Headgear
Additional Comments: S9 Adapt for Home and Travel, On-Board Firmware, Std. Tubing. EEP 9.8, Min PS 4.6, Max Pressure 21