New user with S9 question

Post by kessa3 » Sat May 28, 2011 3:28 pm

I just got my S9 CPAP a little less than 2 weeks ago (not sure what subtype it is) and I did have an issue with it turning the pressure down. I took it in the have the pressure measure and it was way down. Turned out the Flex was turned on, so he turned it off. I also don't use the ramp because I don't find it necessary.
Today I woke up again feeling like the pressure was less. I have a good mask fit (Mirage Quattro full face with headgear) so I know its not leaking. Plus the onscreen report didn't show any leaks. I just turned on the AHI reporting too, but I don't remember what the number was and it doesn't show last night's numbers now.
My doctor and the DME said you do get used to the pressure, so is that all I'm feeling now? Do you wake up in the morning feeling like the pressure is lower even though it isn't, or do I still have a problem with the machine?

Joy in WI
Pressure 8
S9 with humidifier
Joy

User avatar
scrapper
Posts: 776
Joined: Mon Nov 16, 2009 1:32 pm
Location: USA

Re: New user with S9 question

Post by scrapper » Sat May 28, 2011 3:37 pm

I agree with your provider.......you do get used to the pressure, and don't notice it, especially in the am when you've been wearing it all night. You can trust the numbers.

The numbers roll into the average after noon each day..........so the numbers you viewed in the morning, are not available in the afternoon or evening. Most machines gather data noon to noon.

_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: ResScan software 3.13, Pressure 21/15
“Life is 10% what happens to you, and 90% what you make of it.” Charles Swindoll

User avatar
LSAT
Posts: 13411
Joined: Sun Nov 16, 2008 10:11 am
Location: SE Wisconsin

Re: New user with S9 question

Post by LSAT » Sat May 28, 2011 3:46 pm

I don't feel the pressure at all (11). You get used to it rather quickly if you use it regularly. You can feel the force of the exhaust and...you can unhook the mask and feel the pressure coming directly from the hose,

User avatar
kessa3
Posts: 3
Joined: Sat May 28, 2011 3:12 pm
Location: Wisconsin Rapids, WI
Contact:

Re: New user with S9 question

Post by kessa3 » Sat May 28, 2011 4:09 pm

Thank you so much! I do definitely feel and hear the air flow when I take the mask off. And its also good to know about the times the data is collected.

My computer has an S-card reader (several sizes), but I'm afraid to pop it in there for fear I'll wreck the card. Can I do that as long as I have the software to interpret the data on the card?

Thanks again,

Joy
Joy

User avatar
scrapper
Posts: 776
Joined: Mon Nov 16, 2009 1:32 pm
Location: USA

Re: New user with S9 question

Post by scrapper » Sat May 28, 2011 4:18 pm

Absolutely plug the card in and download the data when you have time.........I do it all of the time. Having the standardized card is one of the pro's of the S9 series......

_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: ResScan software 3.13, Pressure 21/15
“Life is 10% what happens to you, and 90% what you make of it.” Charles Swindoll

User avatar
Bright Choice
Posts: 596
Joined: Fri Jan 14, 2011 5:17 pm

Re: New user with S9 question

Post by Bright Choice » Sat May 28, 2011 5:24 pm

You will need Resmed software which is called Rescan. You can find it and download it here. Look for posts by "uncle bob". The link is under his name. Good luck!

_________________
Mask: Mirage™ FX For Her Nasal CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: S9 VPAP Adapt, CompSA, RLS/PLMD, Insomnia, started 12/30/10 Rescan 3.14

User avatar
kessa3
Posts: 3
Joined: Sat May 28, 2011 3:12 pm
Location: Wisconsin Rapids, WI
Contact:

Re: New user with S9 question

Post by kessa3 » Sat May 28, 2011 5:48 pm

It is almost downloaded (93%) right now, so I'm on my way. I'm a nerd who can't wait to see my data!

Joy
Joy

Jacob8515
Posts: 3
Joined: Sat Jul 23, 2011 3:55 pm

Re: New user with S9 question

Post by Jacob8515 » Sat Jul 23, 2011 4:08 pm

I just took possession of my S9 elite and mirage mask yesterday. I had a really good night sleep for the first time in memory. My wife said she checked me 2x during the night to see if I was still breathing because I was so quite. She can get rid of the earplugs. Where can I find a link to download the latest rescan software to start managing my data?

User avatar
Bright Choice
Posts: 596
Joined: Fri Jan 14, 2011 5:17 pm

Re: New user with S9 question

Post by Bright Choice » Sat Jul 23, 2011 5:07 pm

Jacob8515 wrote:I just took possession of my S9 elite and mirage mask yesterday. I had a really good night sleep for the first time in memory. My wife said she checked me 2x during the night to see if I was still breathing because I was so quite. She can get rid of the earplugs. Where can I find a link to download the latest rescan software to start managing my data?
Wow that is impressive. Congratulations!

_________________
Mask: Mirage™ FX For Her Nasal CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: S9 VPAP Adapt, CompSA, RLS/PLMD, Insomnia, started 12/30/10 Rescan 3.14

sleepnationtv
Posts: 283
Joined: Mon Jul 19, 2010 7:23 pm

Re: New user with S9 question

Post by sleepnationtv » Sun Jul 24, 2011 1:35 pm

With the EPR Feature when the provider is measuring the pressure it thinks it is in the breathing out cycle so it lowers the pressure. So you can turn the feature back on because when you actually use it, it will be at the right pressure. That is a common issue that most DME's don't even know about.

Also you can check the AHI for different time periods right on the machine. You have to go into the advanced information screen. Hold down the two silver buttons together.

Here is a video that will help with that.
http://www.cpaplibrary.com/machines.html

Dane
Dane Schapper, Founder cpaplibrary.com
Check out the cpaplibrary video bloghttp://www.cpaplibrary.com

User avatar
robysue
Posts: 7520
Joined: Sat Sep 18, 2010 2:30 pm
Location: Buffalo, NY
Contact:

Re: New user with S9 question

Post by robysue » Mon Jul 25, 2011 12:20 pm

sleepnationtv wrote:With the EPR Feature when the provider is measuring the pressure it thinks it is in the breathing out cycle so it lowers the pressure. So you can turn the feature back on because when you actually use it, it will be at the right pressure. That is a common issue that most DME's don't even know about.
I'm lost. Can you clarify what you mean here?

For example: Suppose you've got an S9 AutoSet that is running in straight CPAP mode with the pressure set at 9cm and EPR set to 3.

Are you saying the pressure is 9cm on the EXHALE and then it goes up to 12cm when you INHALE?

Or are you saying the pressure is 9cm on the INHALE and then it goes down to 6cm when you Exhale?

And out of curiosity, where does one find this information documented?

Thanks

_________________
Machine: DreamStation BiPAP® Auto Machine
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear
Additional Comments: PR System DreamStation and Humidifier. Max IPAP = 9, Min EPAP=4, Rise time setting = 3, minPS = 3, maxPS=5

User avatar
archangle
Posts: 9293
Joined: Sun Mar 27, 2011 11:55 am

Re: New user with S9 question

Post by archangle » Mon Jul 25, 2011 12:28 pm

Kessa, next to the power button on your S9, it should say "escape, escape auto, elite, or AutoSet."

I use a pressure of around 12. When I've had my mask on for for a while, I have to open my mouth and let some air escape or lift up the edge of my mask sometimes to convince myself the machine is on because I can't feel the pressure any more. You can also put your hand in the air stream from your mask exhaust and get used to how that feels when the pressure is right.

_________________
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Also SleepyHead, PRS1 Auto, Respironics Auto M series, Legacy Auto, and Legacy Plus
Please enter your equipment in your profile so we can help you.
Click here for information on the most common alternative to CPAP.
If it's midnight and a DME tells you it's dark outside, go and check for yourself.

Useful Links.

User avatar
Alshain
Posts: 424
Joined: Thu Jan 13, 2011 4:27 pm
Location: East Texas

Re: New user with S9 question

Post by Alshain » Mon Jul 25, 2011 2:23 pm

robysue wrote:Or are you saying the pressure is 9cm on the INHALE and then it goes down to 6cm when you Exhale?
This, the prescription pressure will be maintained during the inhale and pressure decreases to allow an easier exhale.
robysue wrote:And out of curiosity, where does one find this information documented?

Thanks
It's not, at least not for ResMed, even in the clinical manual. But if you look up C-Flex for Respironics, it's a little better documented and it's the same thing.

This is really my own opinion/experience but you shouldn't need this feature unless your pressure is at least 10. I'm sure people will disagree because it really depends on the user and how easily they can exhale against the machine but, that's my thought.

_________________
Mask: Quattro™ FX Full Face CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: ClimateLine, 6cm H2O
But they that wait upon the LORD shall renew their strength; they shall mount up with wings as eagles; they shall run, and not be weary; and they shall walk, and not faint. - Isaiah 40:31 (KJV)

User avatar
robysue
Posts: 7520
Joined: Sat Sep 18, 2010 2:30 pm
Location: Buffalo, NY
Contact:

Re: New user with S9 question

Post by robysue » Tue Jul 26, 2011 11:50 am

Alshain wrote:
robysue wrote:Or are you saying the pressure is 9cm on the INHALE and then it goes down to 6cm when you Exhale?
This, the prescription pressure will be maintained during the inhale and pressure decreases to allow an easier exhale.p
Alshain, this is, of course, exactly what I thought and what my RT said when she set the S9 up back last Septemeber. But sleepnationtv's English seems to imply otherwise:
sleepnationtv wrote:With the EPR Feature when the provider is measuring the pressure it thinks it is in the breathing out cycle so it lowers the pressure. So you can turn the feature back on because when you actually use it, it will be at the right pressure. That is a common issue that most DME's don't even know about.
Re-read that part that I emphasized: Is sleepnationtv saying that the (set) pressure is measured on the breathing OUT (EXHALE) cycle when you use EPR? Because if that's what he is saying, then his statement definitely runs counter to what DME's and RT's (and even posters here) commonly tell S9 users about how EPR works. Because most DMEs and most RT's do think EPR lowers the pressure from the set pressure on exhale (hence exhale relief) and then raises the pressure back up to the set pressure on inhale. Which is why there is some concern by some DMEs and RTs (and even a few users here) that turning on EPR might adversely affect the quality of treatment. And that raises the question: What does sleep nationtv mean by that statement : That is a common issue that most DME's don't even know about.

And if he's not meaning to say the set pressure is measured on EXHALE when using EPR (which is what it sounds to me like he's saying), then I'd like him to clean up his statement so it's no longer ambiguous.

_________________
Machine: DreamStation BiPAP® Auto Machine
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear
Additional Comments: PR System DreamStation and Humidifier. Max IPAP = 9, Min EPAP=4, Rise time setting = 3, minPS = 3, maxPS=5