Hi, I use an AEIO Everest 2, and I think I might've overfilled the humidifier tank...I went to adjust the humidifier settings just now and they're not even showing up on the screen. The bottom where I slide the tank in is a little wet, so I dried it but it's still not working. I don't have a manual :/ and I've only had the machine for a couple of weeks. any ideas?
By the way, if you know of anywhere online I can find a manual for this machine, I'd be grateful.
AEIO Everest 2 Humidifier Not Working
Re: AEIO Everest 2 Humidifier Not Working
Try removing the entire humidifier assy and battery from machine. Set it all someplace dry with good air circulation(a sunny window). In a day or two try starting it again. This should give time for all the moisture to get out of where ever it got to. It might work...it might not...but it's worth a try.
Don't have a link for an on-line manual, but I'll try to answer specific questions for you.
Brenda
Don't have a link for an on-line manual, but I'll try to answer specific questions for you.
Brenda
Re: AEIO Everest 2 Humidifier Not Working
http://www.apneaboard.com/CPAP%20Adjustment.htm
Try above link for instructions on how to obtain clinician manual via email. Scroll all the way to the bottom and I see these shown.
AEIOmed Everest 2 CPAP (Patient & Clinician Quickstart Guide)
AEIOmed Everest 3 CPAP (Complete Clinician Manual)
Try above link for instructions on how to obtain clinician manual via email. Scroll all the way to the bottom and I see these shown.
AEIOmed Everest 2 CPAP (Patient & Clinician Quickstart Guide)
AEIOmed Everest 3 CPAP (Complete Clinician Manual)
I may have to RISE but I refuse to SHINE.
Re: AEIO Everest 2 Humidifier Not Working
Thanks to both of you! I dried it out more thoroughly and aired it out a bit and the humidifier is working again - I was so worried I'd shorted it out or something.
I will email about the manual right now, thanks.
I will email about the manual right now, thanks.
Re: AEIO Everest 2 Humidifier Not Working
I observed that the heated humidifier for the Everest 3 does not work above 8000 ft elevation! This is a consistent observation as it happened in Chile, China, and Ecuador. The pre-heat mode does work at higher elevations.