Did I miss something?

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
CJNAZ
Posts: 15
Joined: Fri May 13, 2011 9:54 am
Location: Sonoran Desert, Arizona

Did I miss something?

Post by CJNAZ » Tue May 24, 2011 9:40 am

Hello to all of you. I am quite new to CPAP therapy. I have had my machine for 12 days now and am adjusting OK. I do experience leaks and am trying all of the tricks that I have been reading about on this forum. By the way, this forum is absolutely great. I've learned so much since I did my sleep study.
Here's the question. I downloaded the data from my SD card into Encore Viewer yesterday and the only data that I see on the report is the hours of use and the pressure setting. I've looked thru a bunch of topics and cannot seem to find if I have a brick for a machine or if I need to go into the setting and change the setup so it records all the data that the software says it is capable of capturing. Any quick advice?

_________________
Mask: Quattro™ FX Full Face CPAP Mask with Headgear
Additional Comments: Pressure set at 11

CJNAZ
Posts: 15
Joined: Fri May 13, 2011 9:54 am
Location: Sonoran Desert, Arizona

Re: Did I miss something?

Post by CJNAZ » Tue May 24, 2011 9:51 am

I am guessing I do not have a data capable machine but how do I find out exactly? Or did I already when I downloaded the SD card and there was very little data available.

_________________
Mask: Quattro™ FX Full Face CPAP Mask with Headgear
Additional Comments: Pressure set at 11

User avatar
Pugsy
Posts: 65397
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Did I miss something?

Post by Pugsy » Tue May 24, 2011 10:33 am

The machine that you are showing in your profile (if you have chosen the correct one) does record full data and you should be able to see all the data on the software reports.

Not sure what viewing choices show in Encore Viewer 2.0 but in Encore Pro it defaults to summary data which is just hours of use and pressure but when we choose "Detail" data there are more pages in the report. Look in the software to see if there are more options for viewing the reports.

If you don't have the "Pro" model and instead have the PR System One Remstar Plus that machine is hours of use only.

You could also use Onkor online analyzer to see if it shows the additional data. If you can see more than hours of use on that report then it is just a matter of finding the full reports in the software.
Instructions are at the website.
http://www.onkor.net/

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
archangle
Posts: 9293
Joined: Sun Mar 27, 2011 11:55 am

Re: Did I miss something?

Post by archangle » Tue May 24, 2011 10:52 am

Right above the display on the blower unit for your CPAP, it probably says "REMstar" followed by something else like "Plus," "Pro," etc.

What does the text say after REMstar on your unit?

_________________
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Also SleepyHead, PRS1 Auto, Respironics Auto M series, Legacy Auto, and Legacy Plus
Please enter your equipment in your profile so we can help you.
Click here for information on the most common alternative to CPAP.
If it's midnight and a DME tells you it's dark outside, go and check for yourself.

Useful Links.

User avatar
Pugsy
Posts: 65397
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Did I miss something?

Post by Pugsy » Tue May 24, 2011 11:16 am

Using Encore Viewer 2.0 now I don't see any choices for viewing other than what it shows. I am assuming then that since your machine is only showing pressure and hours of use that the machine in use is likely the PR System One Remstar Plus and does not offer any more data. You can confirm by the number of pages in the report that comes up. If you scroll through all the pages (I had 16) and you don't see the daily details, then there aren't any available other than what you see.

Confirm what the machine says it is as archangle has suggested. If it is the PR System One Remstar Plus it won't ever offer anything more than hours of use and pressure. You would need to at least upgrade to the machine that you currently show in your Profile the PR System One Pro Cpap with CFlex Plus.

Since new to therapy there could be some options available to you depending on DME and insurance.
If you confirm it is the non data capable machine then I would suggest that you start by reading Janknitz blog about what to do if you discover you have been given a brick titled "Help, I have been give a brick"
http://maskarrayed.wordpress.com/

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
LoQ
Posts: 1475
Joined: Sun Apr 06, 2008 1:59 pm
Location: America

Re: Did I miss something?

Post by LoQ » Tue May 24, 2011 11:17 am

With that title, I was sure this was another Rapture thread. Heh.

Please let us know what your machine says on it. You'll surely get help once you post that.

CJNAZ
Posts: 15
Joined: Fri May 13, 2011 9:54 am
Location: Sonoran Desert, Arizona

Re: Did I miss something?

Post by CJNAZ » Tue May 24, 2011 11:37 am

WOW!!! Thanks for the quick replies. I called my DME and told them that I wanted to know what was up with my machine and why I wasn't seeing data. You should have heard the "silence" on the other end when I rattled off all kinds of info and knowledge (thanks to this forum), and told them what I was looking for. Guess he hasn't heard of someone being educated about their therapy choices.The guy tried to suggest that I call Respironics and was about to give me their number. I stopped him and said "as my DME provider I expect YOU to have the info I need and that I should not be calling the equipment manufacturer". He called back just now and "politely" said that they had given me the wrong machine versus what my Dr prescribed. They gave me a Respironics model# DS250 instead of a DS450!! 12 days of lost data. Oh well, at least I am going to get what I need. I will go pickup the new one tomorrow and start from there. Thank you to everyone who inputs info on this forum!!! You just saved me a ton of time and frustration, and possible a life. If someone were to get so frustrated they stopped therapy the consequences could be very bad. Thank you again and I look forward to getting to know all of you.

_________________
Mask: Quattro™ FX Full Face CPAP Mask with Headgear
Additional Comments: Pressure set at 11

User avatar
OutaSync
Posts: 2048
Joined: Sun Sep 23, 2007 8:49 am
Location: Virginia

Re: Did I miss something?

Post by OutaSync » Tue May 24, 2011 11:42 am

You saved yourself by coming here and educating yourself. Great job!

We must have the same DME Mine tried to pawn off a brick, too.
Diagnosed 9/4/07
Sleep Study Titrated to 19 cm H2O
Rotating between Activa and Softgel
11/2/07 RemStar M Series Auto with AFlex 14-17
10/17/08 BiPAP Auto SV 13/13-23, BPM Auto, AHI avg <1

User avatar
archangle
Posts: 9293
Joined: Sun Mar 27, 2011 11:55 am

Re: Did I miss something?

Post by archangle » Tue May 24, 2011 11:58 am

Be sure they make a written note somehow that you have 12 days of the required compliance for insurance purposes. You don't want them to lose the data from the old machine and report back to the insurance company that you didn't use the machine for 12 days and run into an insurance payment problem because you aren't compliant.

Remember, they've already made a <cough><cough> "mistake" so they might make another one when doing insurance compliance reports.

And check the new machine to be sure it says "REMstar PRO" above the display on the new blower unit.

And check that the "REMstar PRO" part isn't a sticker they just took and stuck over the label on your old machine.

_________________
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Also SleepyHead, PRS1 Auto, Respironics Auto M series, Legacy Auto, and Legacy Plus
Please enter your equipment in your profile so we can help you.
Click here for information on the most common alternative to CPAP.
If it's midnight and a DME tells you it's dark outside, go and check for yourself.

Useful Links.

User avatar
Pugsy
Posts: 65397
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Did I miss something?

Post by Pugsy » Tue May 24, 2011 12:03 pm

Yeah "cough cough" mistake dispensing machine my @ss. They knew exactly what they were doing.

Good job on standing up to them and for being proactive in your own behalf.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
2flamingos
Posts: 439
Joined: Mon Apr 27, 2009 7:26 pm
Location: L.A. (Lower Alabama)

Re: Did I miss something?

Post by 2flamingos » Tue May 24, 2011 12:08 pm

archangle wrote:Be sure they make a written note somehow that you have 12 days of the required compliance for insurance purposes. You don't want them to lose the data from the old machine and report back to the insurance company that you didn't use the machine for 12 days and run into an insurance payment problem because you aren't compliant.

Remember, they've already made a <cough><cough> "mistake" so they might make another one when doing insurance compliance reports.

And check the new machine to be sure it says "REMstar PRO" above the display on the new blower unit.

And check that the "REMstar PRO" part isn't a sticker they just took and stuck over the label on your old machine.

I would also be sure to keep a copy of the report that you did run - you should be able to save it as a pdf, jpeg, or document writer file. We all know that the insurance folks can be as bad as the the DME folks.

I agree with Pugsy - they knew what they were doing - they were giving you a less expensive model, receiving the same payment from your insurance, and pocketing the rest.

_________________
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Pressure - 12-16 ClimateLine Hose. Zeo Mobile, Chin Strap
Old enough to know better


Karma is a Wonderful Thing.

Janknitz
Posts: 8518
Joined: Sat Mar 20, 2010 1:05 pm
Location: Northern California

Re: Did I miss something?

Post by Janknitz » Tue May 24, 2011 12:41 pm

Ha, that's a great story! I'm so proud of you for being firm with them and getting them to admit their error and, more importantly, FIX it! You rock!

After you get your pro and confirm you have the right machine, I'd love to see you document this in writing. Just report the conversation as you have here with the date, time, and who you spoke to. Start the letter with something like: "This is just to confirm our conversation of [date} where you stated I had been given the wrong machine. I should have received the DS450 but your company provided a DS250 instead. Thank you for correcting the error and sending me the DS450 requested (by my physician)." Then CC it to your insurer and your doctor. The insurer so you don't get double billed, and your doctor so he/she knows what kinds of games this DME plays. It will also put this DME on notice that they can't play games with YOU anymore.
What you need to know before you meet your DME http://tinyurl.com/2arffqx
Taming the Mirage Quattro http://tinyurl.com/2ft3lh8
Swift FX Fitting Guide http://tinyurl.com/22ur9ts
Don't Pay that Upcharge! http://tinyurl.com/2ck48rm

Kilgore Trout
Posts: 109
Joined: Thu May 05, 2011 8:57 pm
Location: Baltimore, MD
Contact:

Re: Did I miss something?

Post by Kilgore Trout » Tue May 24, 2011 9:44 pm

CJNAZ wrote:He called back just now and "politely" said that they had given me the wrong machine versus what my Dr prescribed.
Woohoo! Congrats on the big win

CJNAZ
Posts: 15
Joined: Fri May 13, 2011 9:54 am
Location: Sonoran Desert, Arizona

UPDATE!!! Re: Did I miss something? UPDATE!!!

Post by CJNAZ » Thu May 26, 2011 8:58 am

Well, I received the correct machine yesterday. The Respironics DS450, like I was suppose to have in the beginning, not the DS250. I am totally and completely disturbed at this point! This machine is SO LOUD compared to the other. I can really really hear myself breathe. The machine makes a noticeable "rushing air" sound when I inhale. The other machine was quieter than a mouse fart. My wife heard my breaths, thru the machine noise, all night long. Not good. I can't imagine that there is a difference between the two models and I thought the only difference was the algorithm for reading efficacy. What did I do, I can't sleep with that sound. I don't think I received a used machine but maybe I did. I wonder if it is defective. Is this a normal sound for a machine that seems identical except the computer board capability within? I am going to contact my DME this morning but am hoping I can get some insight from any of you out there first. I'd like to have a bit of knowledge before lowering the hammer on this place again. Please tell your thoughts about this. Thanks!!!!!

_________________
Mask: Quattro™ FX Full Face CPAP Mask with Headgear
Additional Comments: Pressure set at 11

User avatar
OutaSync
Posts: 2048
Joined: Sun Sep 23, 2007 8:49 am
Location: Virginia

Re: Did I miss something?

Post by OutaSync » Thu May 26, 2011 9:22 am

Turn the CFlex off, if the noise bothers you. If that doesn't help, it may be a defective machine.

My machine was so noisey when I first got it that I had to put a styrofoam cooler over it to muffle the noise. After 5 days, I noticed that it was much quieter and I quit using the cover. I don't know it it actually got quieter as it broke in, or if I got used to it, but it doesn't bother me now.
Diagnosed 9/4/07
Sleep Study Titrated to 19 cm H2O
Rotating between Activa and Softgel
11/2/07 RemStar M Series Auto with AFlex 14-17
10/17/08 BiPAP Auto SV 13/13-23, BPM Auto, AHI avg <1