AHI number - HELP

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
Streaming Bob
Posts: 18
Joined: Sun Apr 24, 2011 9:14 am
Location: Colorado

AHI number - HELP

Post by Streaming Bob » Wed May 04, 2011 8:20 am

I have thought all along that my AHI number was 40. Yesterday I visited my ENT Dr. and he looked at the sleep study and said my AHI is 18.5.
Here is what my sleep study says
"The patient initially slept for a total of 133 minutes out of time in bed of 203 with a sleep efficiency of 66%. The patient was noted to have one central apnea, 25 obstructive apneas, 15 hypopneas for an apnea hypopneic index of 18.5. He also had significant respiratory effort related arousals of 47 with an index of 21 for an overall respiratory disturbance index of 39.7. These events were disruptive to sleep and triggered cycling oxygen desaturation that went as low as 79%."
On Monday I got the read out from the chip on my new machine and it showed a AHI on Thur. of 18.1, Fri 31.6 Sat 31.6 and Sun of 7.3.

Thur, Fri and Sat. I was sleeping on my back on an incline with pillows and on Sunday I slept on my side.
I think this is why my AHI was so much better on Sunday. Got prescripition for new sleeping pill yesterday and will try them tonight.
I still am feeling tire. Going to see my internal medicine Dr. today. I just start having shortness of breath.
Any suggestion would be welcome.

Mary Z
Posts: 1493
Joined: Wed Oct 28, 2009 9:55 am

Re: AHI number - HELP

Post by Mary Z » Wed May 04, 2011 8:50 am

It can take awhile to feel better- for some of us it takes six months. See you Internal Medicine Doctor about the shortness of breath. I doubt it would be related to CPAP. You may have other medical issues. I see your AHI was much better sleeping on your side- you noted this yourself. Also, during your sleep study the machine does not know whether you're awake, but the EEG does. That's how they can tell how long you slept and in what sleep stages. Good Luck.

_________________
Mask: Wisp Nasal CPAP Mask with Headgear - Fit Pack
Additional Comments: PR System One Remstar BiPap Auto AS Advanced.
Last edited by Mary Z on Wed May 04, 2011 12:00 pm, edited 1 time in total.
Dog is my copilot

User avatar
avi123
Posts: 4509
Joined: Tue Dec 21, 2010 5:39 pm
Location: NC

Re: AHI number - HELP

Post by avi123 » Wed May 04, 2011 9:05 am

At the sleep study they know how much time you slept which was 133 min = 2.2 hr. The total Obstructive and Hypopneas events that you had was 25 + 15 =40

By dividing 2.2 into 40 I get 18.2 as an AHI

This is close enough to the AHI of 18.5

While you are on CPAP the machine does not know if you're asleep. So it divides those events into usage time.

This could explain the diff.

Shortness of breath could be related to the pressure set on your CPAP if it is hi. But you don't give details of it nor your age.

_________________
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments:  S9 Autoset machine; Ruby chinstrap under the mask straps; ResScan 5.6
see my recent set-up and Statistics:
http://i.imgur.com/TewT8G9.png
see my recent ResScan treatment results:
http://i.imgur.com/3oia0EY.png
http://i.imgur.com/QEjvlVY.png

Streaming Bob
Posts: 18
Joined: Sun Apr 24, 2011 9:14 am
Location: Colorado

Re: AHI number - HELP

Post by Streaming Bob » Wed May 04, 2011 12:15 pm

avi123 wrote:
Shortness of breath could be related to the pressure set on your CPAP if it is hi. But you don't give details of it nor your age.
I am on BiPAP with the setting at 8 and 12. I will be 70 in Oct. and have been very healthy up until I started this BiPAP thing. Before I started BiPAP I would sleep 8 to 9 hours only waking once for a potty break. Now I wake up every 2 to 3 hours and only get about 6 hours sleep. Feel tired most of the day and now the shortness of breath has stated. The sleep Dr. now wants me on sleeping pill so I at least sleep 6 hr. without waking. I am starting some new sleeping pills tonight.
Thanks for your help

User avatar
70sSanO
Posts: 264
Joined: Thu Apr 07, 2011 1:44 pm
Location: SoCal

Re: AHI number - HELP

Post by 70sSanO » Wed May 04, 2011 1:50 pm

Something doesn't look right.

At a sleep study without any therapy you have an AHI of 18.5.

Now with therapy on your machine you are getting 18.1, 31.6, and 31.6.

Did you sleep on your side during the sleep study because you side sleep AHI at home is 7.3?

Do you have a breakdown of your AHI numbers?

Does you Dr know about your latest results?

John
AHI: 2.5
Central: 1.7
Obstructive: 0.3
Hypopnea: 0.5
Pressure: 6.0-8.0cm on back with cervical collar.
Compliance: 15 Years

Janknitz
Posts: 8511
Joined: Sat Mar 20, 2010 1:05 pm
Location: Northern California

Re: AHI number - HELP

Post by Janknitz » Wed May 04, 2011 3:15 pm

Another consideration is that you only slept a little over 2 hours in the study, and probably did not reach REM sleep. So your actual AHI may be much higher than indicated by the limited parameters of your study. And what of your titration study--were they able to observe you in different positions and during REM sleep for that???

You do seem to have a positional component--what position were you in during the sleep study posted?

Your machine is not doing a good job of controlling your apneas, especially in supine but even in side lying it's not too good. There are some factors that may contribute--primarily incorrect pressure settings and mask leaks. Can you access your leak data as that would be the first place to look?

Please be sure to post what make and model machine you have so we will be able to tell you how to access the appropriate data. And we need to know what mask you are using to determine leak rates.

Right now the CPAP doesn't sound like it's set correctly for you, so you are having interrupted sleep. Your sleep can and will improve when the therapy is dialed in to what is optimal for you. Hang in there, and let your doctor know about your sleep issues and very high AHI's on the machine.
What you need to know before you meet your DME http://tinyurl.com/2arffqx
Taming the Mirage Quattro http://tinyurl.com/2ft3lh8
Swift FX Fitting Guide http://tinyurl.com/22ur9ts
Don't Pay that Upcharge! http://tinyurl.com/2ck48rm

User avatar
avi123
Posts: 4509
Joined: Tue Dec 21, 2010 5:39 pm
Location: NC

Re: AHI number - HELP

Post by avi123 » Wed May 04, 2011 6:46 pm

Bob, I am also going to see one of my internist next week. I been on CPAP since Nov 2010. For 3 months it helped me to overcome the sleep fragmentation that I had before because of a Peripheral Neuropathy that I come down with suddenly. While CPAPing at home my AHI never came down below 5/hour. They did come down during the sleep study from 6 months ago from 40 with No CPAP to about 3 with CPAP. I don't know how they did it to come so low.

See here my recent averages for AHI while on a plain CPAP:.......................................................................................................


Image

During the last month I have realized that I have a residual daytime sleepiness in spite of my 8 hrs sleep during the night. The sleep is broken into three to go urinate. But I fall back asleep on the CPAP immediately. Two months ago I stopped taking the 10 mg of Zolpidem (for sleep) and now just take 1/2 mg Xanax 4 times daily.

During the last 2 months I have also been realizing some shortness of breath on exertion.

Currently, I have started to use an S9 Autoset, set at 8 to 13 cmH2O and I already see improvement in my AHI.

Hoping it helps you somewhat.

_________________
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments:  S9 Autoset machine; Ruby chinstrap under the mask straps; ResScan 5.6
Last edited by avi123 on Thu May 05, 2011 3:21 pm, edited 4 times in total.
see my recent set-up and Statistics:
http://i.imgur.com/TewT8G9.png
see my recent ResScan treatment results:
http://i.imgur.com/3oia0EY.png
http://i.imgur.com/QEjvlVY.png

ozij
Posts: 10462
Joined: Fri Mar 18, 2005 11:52 pm

Re: AHI number - HELP

Post by ozij » Wed May 04, 2011 9:55 pm

Streaming Bob wrote:I have thought all along that my AHI number was 40. Yesterday I visited my ENT Dr. and he looked at the sleep study and said my AHI is 18.5.
Here is what my sleep study says
"The patient initially slept for a total of 133 minutes out of time in bed of 203 with a sleep efficiency of 66%. The patient was noted to have one central apnea, 25 obstructive apneas, 15 hypopneas for an apnea hypopneic index of 18.5. He also had significant respiratory effort related arousals of 47 with an index of 21 for an overall respiratory disturbance index of 39.7. These events were disruptive to sleep and triggered cycling oxygen desaturation that went as low as 79%."
On Monday I got the read out from the chip on my new machine and it showed a AHI on Thur. of 18.1, Fri 31.6 Sat 31.6 and Sun of 7.3.
The text I marked in red is where the number 40 comes from. You sleep was interrupted by efforts to breathe 40 times an hour. That's your RDI. Your ENT focused on the AHI alone, the sleep report looked at your overall sleep.
If I remember correctly, your readout, on a Respironics machine, is always a 7 day average. This would mean the effect of side sleeping on Sunday was even more dramatic.

_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: Machine: Resmed AirSense10 for Her with Climateline heated hose ; alternating masks.
And now here is my secret, a very simple secret; it is only with the heart that one can see rightly, what is essential is invisible to the eye.
Antoine de Saint-Exupery

Good advice is compromised by missing data
Forum member Dog Slobber Nov. 2023

Streaming Bob
Posts: 18
Joined: Sun Apr 24, 2011 9:14 am
Location: Colorado

Re: AHI number - HELP

Post by Streaming Bob » Thu May 05, 2011 7:31 am

70sSanO wrote:Something doesn't look right.


Now with therapy on your machine you are getting 18.1, 31.6, and 31.6.

For 18.l I had 13.0 mins of large leak - For 31.6 had 9 mins of large leaks and for the other 31.6 had 145 mins of large leaks.
For the 7.3 had 12 mins of large leakss

Did you sleep on your side during the sleep study because you side sleep AHI at home is 7.3?

Yes.

Do you have a breakdown of your AHI numbers? No Where do I see that. I have CA, OA, H, FL, VS, RE. is the the breakdown????

Does you Dr know about your latest results?

Yes he is on vacation and I will talk to him next week.

I get my printouts for my DME. Take my chip to their office

John

Streaming Bob
Posts: 18
Joined: Sun Apr 24, 2011 9:14 am
Location: Colorado

Re: AHI number - HELP

Post by Streaming Bob » Thu May 05, 2011 7:36 am

Janknitz wrote:Another consideration is that you only slept a little over 2 hours in the study, and probably did not reach REM sleep. So your actual AHI may be much higher than indicated by the limited parameters of your study. And what of your titration study--were they able to observe you in different positions and during REM sleep for that???

You do seem to have a positional component--what position were you in during the sleep study posted?

Your machine is not doing a good job of controlling your apneas, especially in supine but even in side lying it's not too good. There are some factors that may contribute--primarily incorrect pressure settings and mask leaks. Can you access your leak data as that would be the first place to look?
I added the leak data in the above message

Please be sure to post what make and model machine you have so we will be able to tell you how to access the appropriate data. And we need to know what mask you are using to determine leak rates.

My model machine is at the bottom of my post.

Right now the CPAP doesn't sound like it's set correctly for you, so you are having interrupted sleep. Your sleep can and will improve when the therapy is dialed in to what is optimal for you. Hang in there, and let your doctor know about your sleep issues and very high AHI's on the machine.

Streaming Bob
Posts: 18
Joined: Sun Apr 24, 2011 9:14 am
Location: Colorado

Re: AHI number - HELP

Post by Streaming Bob » Fri May 06, 2011 8:20 am

Can't get this quote thing figured out. I did add the leak in one of the above messages I sent.


I saw the Internal Dr. Wed. and he thought the shortness of breath was due to the fact that I had not gotten much excerise during the winter. Now that we are getting nice weather I will be more active. Suggested that I also lose some weight so my wife and myself our now on a diet. Great no more ice cream.

Dr. said to try and keep going with the BiPAP machine but he knew that some people just can't do it. Didn't know if he could do it.
Bob

ozij
Posts: 10462
Joined: Fri Mar 18, 2005 11:52 pm

Re: AHI number - HELP

Post by ozij » Fri May 06, 2011 9:54 am

[quote="Streaming Bob"]Can't get this quote thing figured out. I did add the leak in one of the above messages I sent.


I saw the Internal Dr. Wed. and he thought the shortness of breath was due to the fact that I had not gotten much excerise during the winter. Now that we are getting nice weather I will be more active. Suggested that I also lose some weight so my wife and myself our now on a diet. Great no more ice cream.

Dr. said to try and keep going with the BiPAP machine but he knew that some people just can't do it. Didn't know if he could do it.
Bob[/quote]

Hi Bob,
I intentionally disabled BBCode in the message, so you could see how quote codes look.
[quote] this is the code to start a quote
[/quote] this it the code to end the quote.
You write you own text after the quote ends

You can also copy single sentences, and quote them, like this:
[quote]Can't get this quote thing figured out.[/quote]

If you edit your message and put an end quote i.e. [/quote] where the qoute ends, your own text will come out of the quote box. Delete the [/quote] that comes after your own text, it is unnecessary now.

_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: Machine: Resmed AirSense10 for Her with Climateline heated hose ; alternating masks.