Ramp in cpap Mode

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
sol
Posts: 457
Joined: Mon Dec 20, 2010 3:52 am

Ramp in cpap Mode

Post by sol » Wed Apr 13, 2011 12:14 pm

I have been trying the S9 autoset in cpap mode with Ramp at 10 mins ,when start the machine it goes from the min pressure to the set pressure instantly ,it seem that Ramp don,t apply in cpap mode .

User avatar
jabman
Posts: 549
Joined: Wed Jun 23, 2010 2:14 pm
Location: Minnesota

Re: Ramp in cpap Mode

Post by jabman » Wed Apr 13, 2011 12:21 pm

Hi,
What is the starting pressure of the ramp set at?
I used the ramp feature on my s9 (in cpap mode) up untel last month. My pressure is 11 but the starting ramp pressure was at 5.

_________________
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Machine is set on CPAP mode/ Pressure is set at 12.
Psalm 150: 6
Let everything that has breath praise the LORD.
Praise the LORD.


"If God does not exist, one will lose nothing by believing in him, while if he does exist, one will lose everything by not believing. " - Blaise Pascal.

sol
Posts: 457
Joined: Mon Dec 20, 2010 3:52 am

Re: Ramp in cpap Mode

Post by sol » Wed Apr 13, 2011 1:31 pm

Thanks : start pressure 8 and the set pressure 11.8

User avatar
Pugsy
Posts: 65250
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Ramp in cpap Mode

Post by Pugsy » Wed Apr 13, 2011 5:14 pm

I wonder if there is a setting somewhere that is affecting ramp. Turned off by mistake. I have such a setting on my APAP. Different brand but it actually has more than one choice pertaining to ramp but if I have the main one turned off the others don't matter.
Look closely for all your choices pertaining to ramp in the clinical menu.

Do you have the clinician manual? If not forum member Uncle_Bob has it available in his signature.

Otherwise look close in your setup menu and make sure all choices seem to be proper.
http://www.apneaboard.com/ResMed/ResMed ... Setup.html

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

sol
Posts: 457
Joined: Mon Dec 20, 2010 3:52 am

Re: Ramp in cpap Mode

Post by sol » Wed Apr 13, 2011 11:51 pm

Quote : "I wonder if there is a setting somewhere that is affecting ramp. Turned off by mistake."
Thanks pugsy : the Ramp max setting in the clinical menu set at 30 mins and in the patient menu at 10 mins ,as far as i know there is no other setting to turn or off .

User avatar
Pugsy
Posts: 65250
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Ramp in cpap Mode

Post by Pugsy » Thu Apr 14, 2011 7:13 am

I am sorry, that was my only idea. Since I don't have your type of machine I can't play around to see if I can duplicate your problem.

How long have you had the machine? Have you considered taking it to DME to see if they have an answer? Have you tried the ramp in straight CPAP mode just to see if it works as it should there? Did you get the clinician manual to see if there are any hints in there?

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

BernieRay
Posts: 390
Joined: Tue Nov 30, 2010 8:21 am

Re: Ramp in cpap Mode

Post by BernieRay » Thu Apr 14, 2011 7:58 am

Ramp Settings for the S9:

Home Menu - patient ramp time selection
Setup (Clinician) Menu - max ramp time, start pressure

That's it as far as ramp settings go. But...

In CPAP mode EPR Inhale can be set to medium or fast. If you're using EPR and it's set to fast, that might be causing a fast ramp start, but I would expect it to drop according to your EPR setting once you take a breath.

And, although it probably doesn't have a role to play in this, EPR can also be set to ramp only. I'm only mentioning it to include everything ramp-related in the clinician's manual.

With your start and CPAP pressures not being very far apart, I wonder how easy it would be to tell them apart?
Ray
Diagnosed in 1997