Resmed S9 autoset: Wrong sleep log

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
kinggrayskull
Posts: 6
Joined: Sun Mar 06, 2011 7:28 pm

Resmed S9 autoset: Wrong sleep log

Post by kinggrayskull » Sun Mar 06, 2011 7:32 pm

Hi Cpaptalk!

I just recently bought a Remsed S9 Autoset. For the first time, I used it together with my Mirage Quattro Full Face Mask last night and actually slept for 6 hours. However, the screen shows that I only slept for 1.2 hours.

How do I fix this?

Thanks!

User avatar
scrapper
Posts: 776
Joined: Mon Nov 16, 2009 1:32 pm
Location: USA

Re: Resmed S9 autoset: Wrong sleep log

Post by scrapper » Sun Mar 06, 2011 7:46 pm

I suspect that the time is not set correctly and you slept over the 12 noon hour according to the clock. At 12 noon, the machine zeros out the day's values, and moves to the next 24 hour period. Check the clock setting first.

_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: ResScan software 3.13, Pressure 21/15
“Life is 10% what happens to you, and 90% what you make of it.” Charles Swindoll

kinggrayskull
Posts: 6
Joined: Sun Mar 06, 2011 7:28 pm

Re: Resmed S9 autoset: Wrong sleep log

Post by kinggrayskull » Sun Mar 06, 2011 7:47 pm

Thanks Scrapper! Will try this out. Will let you know what happens.

kinggrayskull
Posts: 6
Joined: Sun Mar 06, 2011 7:28 pm

Re: Resmed S9 autoset: Wrong sleep log

Post by kinggrayskull » Sun Mar 06, 2011 8:47 pm

Hi Scrapper.

I can't seem to find the clock settings on my S9 machine.

I am now at View Settings and there isn't anything for time or clock.

Help! Thanks.

mstevens
Posts: 137
Joined: Sun Oct 15, 2006 9:15 pm

Re: Resmed S9 autoset: Wrong sleep log

Post by mstevens » Sun Mar 06, 2011 9:31 pm

kinggrayskull wrote:Hi Scrapper.

I can't seem to find the clock settings on my S9 machine.

I am now at View Settings and there isn't anything for time or clock.

Help! Thanks.
Weird - I was willing to swear one of the first things I did when mine arrived was to set the time. I just went through the patient menus and could not find this, nor does anything about it appear in the manual.

You may need to enter the clinical menu to set the clock. If so, that's really dumb. What about someone who travels halfway around the world? Having the clock off by 12 hours would really screw up the data, including compliance measures.

_________________
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Don't assume physicians have an easier time getting information or resources from equipment manufacturers!
Untreated AHI: 86
Treated AHI: 0.4

User avatar
scrapper
Posts: 776
Joined: Mon Nov 16, 2009 1:32 pm
Location: USA

Re: Resmed S9 autoset: Wrong sleep log

Post by scrapper » Sun Mar 06, 2011 9:57 pm

I no longer have my S9 or the manual, but have found ResMed machines to be set up the same........so try pressing the big round button and the down arrow at the same time. Look for Options--the time and date are in that section.

_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: ResScan software 3.13, Pressure 21/15
“Life is 10% what happens to you, and 90% what you make of it.” Charles Swindoll

kinggrayskull
Posts: 6
Joined: Sun Mar 06, 2011 7:28 pm

Re: Resmed S9 autoset: Wrong sleep log

Post by kinggrayskull » Sun Mar 06, 2011 10:32 pm

Thank you Scrapper!
I got the Clock Setting right.

User avatar
AMUW
Posts: 85
Joined: Sat Dec 11, 2010 4:11 pm

Re: Resmed S9 autoset: Wrong sleep log

Post by AMUW » Sun Mar 06, 2011 11:08 pm

The issue becomes more apparent if you have the sw and are able to compare the time scale to your own clock.
You may still not have fixed your problem in the long run.
My S9 Autoset was delivered in the winter, but still set to daylight savings time. I found and modified the clock in the machine's clinical settings. But one day later it back ... that's because I took the SD card to the PC and back ... SD card settings apparently take precedence, and can only be modified in the ResScan program.
Moderate-severe OSA, ResMed S9 AutoSet EPR + H5i Humidifier, ResMed Masks: trying Swift FX Nasal Pillow, Mirage Nasal, Mirage Quatro or Quattro FX Full Face
ResMed SD card & USB adaptor, ResScan 3.10

User avatar
scrapper
Posts: 776
Joined: Mon Nov 16, 2009 1:32 pm
Location: USA

Re: Resmed S9 autoset: Wrong sleep log

Post by scrapper » Mon Mar 07, 2011 8:12 am

sw????

_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: ResScan software 3.13, Pressure 21/15
“Life is 10% what happens to you, and 90% what you make of it.” Charles Swindoll

User avatar
Pugsy
Posts: 65139
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Resmed S9 autoset: Wrong sleep log

Post by Pugsy » Mon Mar 07, 2011 10:02 am

scrapper wrote:sw????
I think software.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
idamtnboy
Posts: 2186
Joined: Mon Nov 01, 2010 2:12 pm
Location: Idaho

Re: Resmed S9 autoset: Wrong sleep log

Post by idamtnboy » Mon Mar 07, 2011 10:03 am

AMUW wrote:I found and modified the clock in the machine's clinical settings. But one day later it back ... that's because I took the SD card to the PC and back ... SD card settings apparently take precedence, and can only be modified in the ResScan program.
Did you press the center knob each time after changing the hour and minute? You have to explicitly make sure each individual setting gets changed before leaving the menu. Putting the SD card into your PC has nothing to do with time on the S9.

_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Hose management - rubber band tied to casement window crank handle! Hey, it works! S/W is 3.13, not 3.7