Request advice concerning initial APAP setting

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
User avatar
Pugsy
Posts: 65369
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Request advice concerning initial APAP setting

Post by Pugsy » Sun Mar 06, 2011 8:40 am

Before increasing the maximum check your leak rate (I don't see it mentioned anywhere in your comments).
If the pressure is bumping 9 much of the night look to see if the leak rate line is also jumping around. If it is, need to work on the leak before increasing the max. The higher pressure could be coming from the machine trying to "chase the leak" and maintain pressure.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

Fenius

Вопрос

Post by Fenius » Sun Mar 06, 2011 9:26 am

Прошу прощения. Не чайно забрел к вам в тему, теперь и забыл что хотел спросить ))))

jonnybee
Posts: 244
Joined: Sun Nov 28, 2010 9:06 am
Location: Eastern NC

Re: Request advice concerning initial APAP setting

Post by jonnybee » Sun Mar 06, 2011 9:46 am

dtsm wrote:
jonnybee wrote:Charts showed pressure banging on 9.0 often which surprised me since my sleep study titration/prescription was for cpap 5cm. Also showed more centrals than previously.I plan to increase the settings tonight to 5/12 and see what happens. Anyone have suggestions? Is this a reasonable approach?
You don't indicate leaks which 'might' be cause for pressure of 9. I would stay the course for now, not change any settings and monitor data [including leaks].
Leak Median: 0.0 95th 0.0 Max 1.2

Pressure Median: 6.4 95th 8.3 Max 8.8
ResMed S9 Autoset
ResMed AirFit F20 mask with headgear