Can't get ResScan to see SD Card

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
BernieRay
Posts: 390
Joined: Tue Nov 30, 2010 8:21 am

Re: Can't get ResScan to see SD Card

Post by BernieRay » Fri Feb 25, 2011 8:46 pm

avi123 wrote:SD card from a camera would NOT work with my ResScan 3.12. The SD card needs to formatted per Resmed. One such card came with my machine . The word Resmed is printed on it. If you don't have such cards I would go to a DME and buy them there.
Any SD card can work. It can even be formatted by a computer or a camera. But, if it has any non-Resmed files on it when it is inserted into an S9, the S9 will either complain, or it will initialize it for use in the S9. An S9 initialization (which the S9 calls "format") is very different from what format means in the computer world.
Ray
Diagnosed in 1997

User avatar
avi123
Posts: 4509
Joined: Tue Dec 21, 2010 5:39 pm
Location: NC

Re: Can't get ResScan to see SD Card

Post by avi123 » Fri Feb 25, 2011 9:27 pm

360, have you checked if there is a little slider lock switch on the side of your SD card?

If there is, it might just be in the "lock mode" which may prevent it from recording new data.

As to formatting if the card had other stuff on it from before then the S9 would ask you if it's ok to erase the info on the card and format it.

_________________
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments:  S9 Autoset machine; Ruby chinstrap under the mask straps; ResScan 5.6
see my recent set-up and Statistics:
http://i.imgur.com/TewT8G9.png
see my recent ResScan treatment results:
http://i.imgur.com/3oia0EY.png
http://i.imgur.com/QEjvlVY.png

ozij
Posts: 10492
Joined: Fri Mar 18, 2005 11:52 pm

Re: Can't get ResScan to see SD Card

Post by ozij » Fri Feb 25, 2011 11:54 pm

BernieRay wrote:
avi123 wrote:SD card from a camera would NOT work with my ResScan 3.12. The SD card needs to formatted per Resmed. One such card came with my machine . The word Resmed is printed on it. If you don't have such cards I would go to a DME and buy them there.
Any SD card can work. It can even be formatted by a computer or a camera. But, if it has any non-Resmed files on it when it is inserted into an S9, the S9 will either complain, or it will initialize it for use in the S9. An S9 initialization (which the S9 calls "format") is very different from what format means in the computer world.
Ray is right.

_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: Machine: Resmed AirSense10 for Her with Climateline heated hose ; alternating masks.
And now here is my secret, a very simple secret; it is only with the heart that one can see rightly, what is essential is invisible to the eye.
Antoine de Saint-Exupery

Good advice is compromised by missing data
Forum member Dog Slobber Nov. 2023

360
Posts: 60
Joined: Sun Jan 16, 2011 11:01 am
Location: MO, USA

Re: Can't get ResScan to see SD Card

Post by 360 » Sat Feb 26, 2011 11:28 pm

The plot thickens. I get a SD USB reader, plug it in, all is going well, select detailed data, download starts and then the following pops up.

Error occurred during download
The current download was interrupted by a download fault

An error occurred during a download. the last session that was downloading will be incomplete. You may choose to either keep or discard the incomplete data.

Click,

Keep to keep the data
Discard to discard the data

Any suggestions?

I am beyond frustrated. I've got a huge leak issue and from what I can gather there is no way to pinpoint what I need to do before reading the charts.

User avatar
idamtnboy
Posts: 2186
Joined: Mon Nov 01, 2010 2:12 pm
Location: Idaho

Re: Can't get ResScan to see SD Card

Post by idamtnboy » Sun Feb 27, 2011 12:03 am

360 wrote:The plot thickens. I get a SD USB reader, plug it in, all is going well, select detailed data, download starts and then the following pops up.

Error occurred during download
The current download was interrupted by a download fault

An error occurred during a download. the last session that was downloading will be incomplete. You may choose to either keep or discard the incomplete data.

Click,

Keep to keep the data
Discard to discard the data

Any suggestions?

I am beyond frustrated. I've got a huge leak issue and from what I can gather there is no way to pinpoint what I need to do before reading the charts.
At this point the best I can suggest is try reading your SD card on a different computer. If you have problems there you may have a bad card. If all works good on the other PC, then that points to your PC having some sort of hardware problem. Have you tried a different SD card yet?

_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Hose management - rubber band tied to casement window crank handle! Hey, it works! S/W is 3.13, not 3.7

BernieRay
Posts: 390
Joined: Tue Nov 30, 2010 8:21 am

Re: Can't get ResScan to see SD Card

Post by BernieRay » Sun Feb 27, 2011 12:47 am

I've seen that problem a few timess, 350. In each case, I simply shut ResScan down, removed the card, re-inserted the card, and tried again. In each case, it worked.

If the problem persists, you might try removing any other USB removable memory devices you might have connected. The ResScan manual mentions that some other devices can cause problems when connected at the same time. I think it was specific to Resmed adapters, but who knows. It might work.
Ray
Diagnosed in 1997

ozij
Posts: 10492
Joined: Fri Mar 18, 2005 11:52 pm

Re: Can't get ResScan to see SD Card

Post by ozij » Sun Feb 27, 2011 2:53 am

Do you have any other device connected to a USB port in this computer?
I have clashes when my pulse oxymeter is connected --

_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: Machine: Resmed AirSense10 for Her with Climateline heated hose ; alternating masks.
And now here is my secret, a very simple secret; it is only with the heart that one can see rightly, what is essential is invisible to the eye.
Antoine de Saint-Exupery

Good advice is compromised by missing data
Forum member Dog Slobber Nov. 2023

360
Posts: 60
Joined: Sun Jan 16, 2011 11:01 am
Location: MO, USA

Re: Can't get ResScan to see SD Card

Post by 360 » Sun Feb 27, 2011 7:30 am

We have a winner. Tried another computer of mine and that had less things attached via USB and it worked perfectly. I'll post the results in another post right after I figure out how to post the PDF doc.

Thanks a bunch to everyone who helped me out.