Can't get ResScan to see SD Card

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
360
Posts: 60
Joined: Sun Jan 16, 2011 11:01 am
Location: MO, USA

Can't get ResScan to see SD Card

Post by 360 » Thu Feb 24, 2011 7:20 pm

Anyone else have this problem. I have installed ResScan and inserted the SD card into my SD drive but ResScan says "there are no devices available for downloading data".

User avatar
billbolton
Posts: 2264
Joined: Wed Jun 07, 2006 7:46 pm
Location: Sydney, Australia

Re: Can't get ResScan to see SD Card

Post by billbolton » Thu Feb 24, 2011 7:40 pm

360 wrote:Anyone else have this problem. I have installed ResScan and inserted the SD card into my SD drive but ResScan says "there are no devices available for downloading data".
You need to check whether you can "see" the SD card from Windows. It should show up as a drive in Windows Explorer (Computer, or My Computer Icon on your desktop) and you should be able to open up a directory of the files on the card.

If you can't "see" it from Windows, neither can ResScan

Cheers,

Bill

_________________
MachineMask
Additional Comments: Airmini, Medistrom Pilot 24, CMS 60C Pulse Oximeter, ResScan 6

360
Posts: 60
Joined: Sun Jan 16, 2011 11:01 am
Location: MO, USA

Re: Can't get ResScan to see SD Card

Post by 360 » Thu Feb 24, 2011 7:44 pm

Yes, I can see it in in Windows but ResScan is not recognizing it.

User avatar
KatieW
Posts: 1671
Joined: Sun Sep 27, 2009 11:07 am
Location: southern AZ

Re: Can't get ResScan to see SD Card

Post by KatieW » Thu Feb 24, 2011 9:37 pm

Are you opening ResScan 1st, then plugging your SD card in?

_________________
Mask: Pico Nasal CPAP Mask with Headgear
Additional Comments: Resmed AirCurve 10 ASV and Humidifier, Oscar for Mac
KatieW

TimWK
Posts: 43
Joined: Wed Jan 12, 2011 8:25 pm
Location: AZ

Re: Can't get ResScan to see SD Card

Post by TimWK » Thu Feb 24, 2011 9:49 pm

It shouldn't matter weather you put the SD card in first or not. If your using windows 7, as soon as you insert the card it asks if you want to open the file viewer. if you then open ResScan it automatically opens the download manager.

Best Regards

360
Posts: 60
Joined: Sun Jan 16, 2011 11:01 am
Location: MO, USA

Re: Can't get ResScan to see SD Card

Post by 360 » Thu Feb 24, 2011 11:48 pm

I'm running XP and I tried it both ways, doesn't seem to acknowledge it both ways. I also tried my other PC and which has an SD reader on the printer and again it shows up and I can open the file, just can't get ResScan to see it.

User avatar
jdm2857
Posts: 2982
Joined: Wed Aug 20, 2008 8:29 pm
Location: South Jersey

Re: Can't get ResScan to see SD Card

Post by jdm2857 » Fri Feb 25, 2011 12:38 am

I'm using an S8 AutoSet with the ResLink module, which uses a SmartMedia card instead of an SD card, so I don't know if this applies, but when I insert the card into the PC reader and open ResScan's dowload screen, it sometimes takes a bit for the card to appear in ResScan. I just wait and the "no devices available ..." message changes to show the presence of the card.

SmartMedia, by the way, is an older format of memory card that is rapidly going away. I bought a few extra cards on eBay just in case my card dies.
jeff

User avatar
billbolton
Posts: 2264
Joined: Wed Jun 07, 2006 7:46 pm
Location: Sydney, Australia

Re: Can't get ResScan to see SD Card

Post by billbolton » Fri Feb 25, 2011 1:37 am

360 wrote:I have installed ResScan
What version of ResScan have you installed?

Cheers,

Bill

_________________
MachineMask
Additional Comments: Airmini, Medistrom Pilot 24, CMS 60C Pulse Oximeter, ResScan 6

User avatar
Pugsy
Posts: 65129
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Can't get ResScan to see SD Card

Post by Pugsy » Fri Feb 25, 2011 9:00 am

billbolton wrote: What version of ResScan have you installed?
Bill, I believe he has 3.12 since it was the one I directed him to in Uncle Bob's signature just yesterday.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

360
Posts: 60
Joined: Sun Jan 16, 2011 11:01 am
Location: MO, USA

Re: Can't get ResScan to see SD Card

Post by 360 » Fri Feb 25, 2011 9:46 am

Correct 3.12. One thing I did think about was that I did check the boxes to install the card reader drivers in case I ever used that type of reader. I wonder if ResScan is looking for one of them vs. searching my computer for an SD card?

ohammersmith
Posts: 74
Joined: Fri Dec 10, 2010 7:53 am

Re: Can't get ResScan to see SD Card

Post by ohammersmith » Fri Feb 25, 2011 10:08 am

This happens to me all the time, but I "resolve" it by quitting Rescan. I assume you've tried that.

I notice it happens most often when I didn't have a "patient" file open when I insert the card.

This is what I do to keep that bug at bay.

1. Start ResScan
2. Open my patient file
3. Insert SD card (note I've copied the files off the actual SD card and onto another one, but that seems not to affect things)
4. Wait for ResScan to close my patient file (argh!)
5. Escape out of the "download data" dialog because on windows 7 the buttons to do anything are hidden and there is no default enter or double click action (double argh!)
6. Re-open patient file
7. Wait a minute, ResScan is slow (triple argh!)
8. Click "Download data" start the transfer
9. Wait. (quadruple argh!)
10. Click "Discard" when it tells me there is duplicate (unless there's some extenuating circumstance where I did multiple sessions in a day and got the data for the first one already.)

I find that if I skip #2 it will get into this infinite loop of prompting me to download but not letting me open any patient file and claiming there's no card when I cancel, open a file and initiate download. Immediately after telling me there's no device, ResScan auto-detects the card and closes the file and prompts, again... deja vu all over again.

I hate Windows software.

You'll excuse me while I go write my own scripts to bring EDF+ files into R.

_________________
Humidifier

BernieRay
Posts: 390
Joined: Tue Nov 30, 2010 8:21 am

Re: Can't get ResScan to see SD Card

Post by BernieRay » Fri Feb 25, 2011 10:42 am

360 wrote:Correct 3.12. One thing I did think about was that I did check the boxes to install the card reader drivers in case I ever used that type of reader. I wonder if ResScan is looking for one of them vs. searching my computer for an SD card?
Yes, it most likely is. I recommend going into Control Panel -Add or Remove Programs and uninstall the drivers. You may also need to uninstall ResScan. Re-boot your computer, then try it again.

I don't remember the 3 labels for the drivers, but I'm fairly certain that installation of one of them causes ResScan to not even look for an SD card in another device.

Note: Unless a Resmed device is being used to connect to the machine (either a Resmed direct connection/usb adapter or a Resmed card reader) none of the drivers are needed. If you need one of them later, all you have to do is go to (assuming WinXP) C:\Program Files\Resmed\ResScane\Drivers and run the installer for whichever drive is needed.
Ray
Diagnosed in 1997

User avatar
avi123
Posts: 4509
Joined: Tue Dec 21, 2010 5:39 pm
Location: NC

Re: Can't get ResScan to see SD Card

Post by avi123 » Fri Feb 25, 2011 12:07 pm

360 wrote:Anyone else have this problem. I have installed ResScan and inserted the SD card into my SD drive but ResScan says "there are no devices available for downloading data".

Hi, I have the ResScan program icon both at the Start column and at Desktop.

My laptop runs on Windows XP Internet Explorer #8

I use two SD cards that I got from my DME. They need to be formatted per Resmed. Or be free of any other data.

I take the SD card from the S9 and stick it into this Sony reader from Staples:


http://www.staples.com/Sony-USB-Memory- ... rea=SEARCH

Next, I stick the reader with the SD card, in it, into USB slot in my laptop

Next, I click on a ResScan icon. (if the SD reader is not in the USB then the ResScan asks for data from the flow generator)

(To validate you can click: Start, Computer, Removable disk, that the SD is pluged in)

I never loaded any ResScan drivers, I assume that they are inside the Windows.

I downloaded the ResScan program from here:


viewtopic.php?p=474698

see uncle-bob's links

I gave a fictitious name for myself as a patient

I do overwrite when downloading

I daily save the reports which show up as PDF files on the left side under Browser

(The only problem that I have is how to post here in the forum portions of my report without going to a hoster site)

_________________
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments:  S9 Autoset machine; Ruby chinstrap under the mask straps; ResScan 5.6
Last edited by avi123 on Sat Feb 26, 2011 8:15 am, edited 1 time in total.
see my recent set-up and Statistics:
http://i.imgur.com/TewT8G9.png
see my recent ResScan treatment results:
http://i.imgur.com/3oia0EY.png
http://i.imgur.com/QEjvlVY.png

360
Posts: 60
Joined: Sun Jan 16, 2011 11:01 am
Location: MO, USA

Re: Can't get ResScan to see SD Card

Post by 360 » Fri Feb 25, 2011 12:59 pm

I've unistalled and reinstalled 3.13 with no driver and I still get the "there are NO devices available for downloading data" message. I've tried closing patient files, re-opening, etc. and nothing seems to help. I'm I going to just need to buy a USB reader and go that route?

BernieRay
Posts: 390
Joined: Tue Nov 30, 2010 8:21 am

Re: Can't get ResScan to see SD Card

Post by BernieRay » Fri Feb 25, 2011 1:49 pm

Assuming that you explicitly un-installed the drivers first and re-booted, I'm at a bit of a loss. What OS are you using?
Ray
Diagnosed in 1997