Someone please help this Newbie!

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
User avatar
SleepyT
Posts: 775
Joined: Wed Jul 08, 2009 10:50 am

Re: Someone please help this Newbie!

Post by SleepyT » Thu Feb 24, 2011 2:49 pm

10-4, AJ. Just try to only change one thing at a time...so when you get results...you will know what caused them! Keep us posted!
"Knowledge is power."

User avatar
robysue
Posts: 7520
Joined: Sat Sep 18, 2010 2:30 pm
Location: Buffalo, NY
Contact:

Re: Someone please help this Newbie!

Post by robysue » Thu Feb 24, 2011 3:07 pm

AJimen05 wrote: His tech called me back and I gave him all the figures. Sounded confused.
Confused about what? That your AHI is still way too high? Or confused that you are capable of reading data off the machine's LCD? Or confused that you are asking for a new machine?
I told him that I wanted another machine, an APAP...S9 Autoset. He said he would get back to me because he thinks I MIGHT NEED AN ADJUSTMENT TO MY PRESSURE! He's going to talk to the Doctor and get back to me. I'm tired of me...anyone care to take over control? J/k.
If he gets back with just another pressure figure for you to try, politely ask what evidence was used to come up with the new pressure level and politely, but firmly tell him that you really would prefer to do a week or two of autotitration at home in an effort to get the pressure setting correct instead of relying on guesswork.

Also since you don't yet have the software, bring the chip into the DME and asking them to print a report of all of your data for you and also fax a copy of the all the data to the doctor's office. And see if the DME would be willing to add the word "URGENT" to the fax's cover letter. Or bring the chip directly to the sleep doctor's office and ask the receptionist to print off a copy of all the data---not just the compliance data for you and make sure the doctor sees a copy of the data too.

_________________
Machine: DreamStation BiPAP® Auto Machine
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear
Additional Comments: PR System DreamStation and Humidifier. Max IPAP = 9, Min EPAP=4, Rise time setting = 3, minPS = 3, maxPS=5

User avatar
AJimen05
Posts: 64
Joined: Wed Feb 23, 2011 1:30 pm
Location: South Florida

Re: Someone please help this Newbie!

Post by AJimen05 » Thu Feb 24, 2011 3:24 pm

robysue wrote:
AJimen05 wrote: His tech called me back and I gave him all the figures. Sounded confused.
Confused about what? That your AHI is still way too high? Or confused that you are capable of reading data off the machine's LCD? Or confused that you are asking for a new machine?
I told him that I wanted another machine, an APAP...S9 Autoset. He said he would get back to me because he thinks I MIGHT NEED AN ADJUSTMENT TO MY PRESSURE! He's going to talk to the Doctor and get back to me. I'm tired of me...anyone care to take over control? J/k.
If he gets back with just another pressure figure for you to try, politely ask what evidence was used to come up with the new pressure level and politely, but firmly tell him that you really would prefer to do a week or two of autotitration at home in an effort to get the pressure setting correct instead of relying on guesswork.

Also since you don't yet have the software, bring the chip into the DME and asking them to print a report of all of your data for you and also fax a copy of the all the data to the doctor's office. And see if the DME would be willing to add the word "URGENT" to the fax's cover letter. Or bring the chip directly to the sleep doctor's office and ask the receptionist to print off a copy of all the data---not just the compliance data for you and make sure the doctor sees a copy of the data too.

LOL, I think he was confused by both actually...I called Apria so that I can get them involved with this. I'll be seeing them today with the SD card so hopefully there would be a resolution to this soon.

_________________
Mask: Quattro™ FX Full Face CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Sleep Study Titrated to 11 cm H2O

User avatar
NancyAZ
Posts: 59
Joined: Wed Jan 26, 2011 11:25 pm
Location: Scottsdale, AZ

Re: Someone please help this Newbie!

Post by NancyAZ » Thu Feb 24, 2011 3:51 pm

Good luck with this. I am in the process of getting my S9 Elite replaced with a S9 Autoset and it took a few phone calls but I am supposed to get a new machine on Monday. Stick with it because at first I thought, "Oh well, I already have this machine," but the more I read on here about the importance of an APAP machine, I realized that I would be so much better off if I just insisted on the change now. Most insurance companies won't buy another machine for five years so it is important to do it now. Good luck! Your doctor is the one you need to convince. Once my doctor faxed a new Rx to the DME, I received a call that day saying the machine would be replaced.

Nancy

_________________
Mask: Swift™ FX For Her Nasal Pillow CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: ResScan 3.13; Pressure 8-12; started CPAP Jan. 26, 2011

User avatar
AJimen05
Posts: 64
Joined: Wed Feb 23, 2011 1:30 pm
Location: South Florida

Re: Someone please help this Newbie!

Post by AJimen05 » Thu Feb 24, 2011 5:26 pm

Pugsy wrote:Off hand without seeing the report graphs it sounds like the 11 cm pressure is not sufficient to ward off events
BUT
given the severity of what is going on with you, I would suggest that you post a screen shot of the report with the graphs and get some better informed ideas.

From what you say with numerical data, we can't blame it on centrals and can't blame it on leak, that pretty much leaves pressure...

Maybe the EPR of 3 is affecting things. It does relate to a 3 cm drop in pressure during exhale. Effectively lowering the pressure by 3 for the amount of time during exhale...

Do you use the ramp? If so how long?

I would want to see the report in graph form to be able to say with more certainty but sure sounds like the pressure isn't warding off the events.

Pugsy,

Stupid question....How do I attach the screenshot I took with my snipping tool onto here?

_________________
Mask: Quattro™ FX Full Face CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Sleep Study Titrated to 11 cm H2O

User avatar
Pugsy
Posts: 65367
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Someone please help this Newbie!

Post by Pugsy » Thu Feb 24, 2011 5:30 pm

AJimen05 wrote:
Pugsy,

Stupid question....How do I attach the screenshot I took with my snipping tool onto here?
Not a stupid question at all, I guess you were supposed to read my mind.
This is what I do.
Use the snipping tool to crop the image that you want.
Save it in jpg format to location you can get to easily. I use my desktop.
You will need to have a photo hosting site like Photobucket.
Upload the file to the photo host site
Then copy the entire IMG address (be sure to include the IMG part) then paste that into a post here.
Use the preview button to see if you see the image.
If not try again, if you can't see the image, we can't.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
Emilia
Posts: 1873
Joined: Mon Jul 12, 2010 8:56 am
Location: Florida, USA

Re: Someone please help this Newbie!

Post by Emilia » Thu Feb 24, 2011 5:31 pm

You need to post the picture to a photo sharing site like Flickr, Picasa, or Photobucket (to name but a few); capture the image url on that site; put that url in a post here; HIGHLIGHT url and then click on the IMG button above the edit box; click Preview to see that it worked; click Submit to post.
Yes, that blue eyed beauty is my cat! He is a seal point, bi-color Ragdoll. I adopted him in '08 from folks who could no longer care for him. He is a joy and makes me smile each and every day.

User avatar
AJimen05
Posts: 64
Joined: Wed Feb 23, 2011 1:30 pm
Location: South Florida

Re: Someone please help this Newbie!

Post by AJimen05 » Thu Feb 24, 2011 5:36 pm

Pugsy wrote:
AJimen05 wrote:
Pugsy,

Stupid question....How do I attach the screenshot I took with my snipping tool onto here?
Not a stupid question at all, I guess you were supposed to read my mind.
This is what I do.
Use the snipping tool to crop the image that you want.
Save it in jpg format to location you can get to easily. I use my desktop.
You will need to have a photo hosting site like Photobucket.
Upload the file to the photo host site
Then copy the entire IMG address (be sure to include the IMG part) then paste that into a post here.
Use the preview button to see if you see the image.
If not try again, if you can't see the image, we can't.

Image

_________________
Mask: Quattro™ FX Full Face CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Sleep Study Titrated to 11 cm H2O

User avatar
AJimen05
Posts: 64
Joined: Wed Feb 23, 2011 1:30 pm
Location: South Florida

Re: Someone please help this Newbie!

Post by AJimen05 » Thu Feb 24, 2011 5:37 pm

I did it! Not sure what all the bars mean but I'll learn....lol

_________________
Mask: Quattro™ FX Full Face CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Sleep Study Titrated to 11 cm H2O

User avatar
Pugsy
Posts: 65367
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Someone please help this Newbie!

Post by Pugsy » Thu Feb 24, 2011 5:44 pm

Good job on the image, this is the summary page.
now one more task, can you get me one night of detailed data?
The page that shows the graphs with the actual leak line and pressure line?

But I already see what I expected. The pressure isn't effective except on those nights that have short red bars. There were only 2 nights with AHI of 5 or so. Looks like the 10 and 11th.

Can you get me an example of one of those and one of your bad nights? Might not be able to get a good night to show since it is over 7 days old. For sure you can get individual night that was recent.

Odd that you had a couple of good nights though.....

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
Lizistired
Posts: 2835
Joined: Tue Dec 14, 2010 10:47 pm
Location: Indiana

Re: Someone please help this Newbie!

Post by Lizistired » Thu Feb 24, 2011 5:49 pm

What did you do different on those 3 or 4 days where the red bars are lower?
Do you have a kink in your hose?!

_________________
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Swift FX sometimes, CMS-50F, Cervical collar sometimes, White noise, Zeo... I'm not well, but I'm better.

User avatar
AJimen05
Posts: 64
Joined: Wed Feb 23, 2011 1:30 pm
Location: South Florida

Re: Someone please help this Newbie!

Post by AJimen05 » Thu Feb 24, 2011 5:52 pm

Pugsy wrote:Good job on the image, this is the summary page.
now one more task, can you get me one night of detailed data?
The page that shows the graphs with the actual leak line and pressure line?

But I already see what I expected. The pressure isn't effective except on those nights that have short red bars. There were only 2 nights with AHI of 5 or so. Looks like the 10 and 11th.

Can you get me an example of one of those and one of your bad nights? Might not be able to get a good night to show since it is over 7 days old. For sure you can get individual night that was recent.

Odd that you had a couple of good nights though.....
I'm sorry Pugsy...but i'm not sure how to pull that data. Does anyone know how to use REscan 3.12?
For the life of me i'm not sure what happened those days though...

_________________
Mask: Quattro™ FX Full Face CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Sleep Study Titrated to 11 cm H2O

User avatar
Lizistired
Posts: 2835
Joined: Tue Dec 14, 2010 10:47 pm
Location: Indiana

Re: Someone please help this Newbie!

Post by Lizistired » Thu Feb 24, 2011 5:56 pm

To the left of that chart, you should have a column of dates. Click on a recent one to highlight it. Then click on the detail graphs tab to the right of the summary tab from the chart you just showed.

_________________
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Swift FX sometimes, CMS-50F, Cervical collar sometimes, White noise, Zeo... I'm not well, but I'm better.

User avatar
AJimen05
Posts: 64
Joined: Wed Feb 23, 2011 1:30 pm
Location: South Florida

Re: Someone please help this Newbie!

Post by AJimen05 » Thu Feb 24, 2011 5:58 pm

Lizistired wrote:To the left of that chart, you should have a column of dates. Click on a recent one to highlight it. Then click on the detail graphs tab to the right of the summary tab from the chart you just showed.
I did that. I clicked the date and then clicked on the details graphs tab and it says No Detailed Data Available.

_________________
Mask: Quattro™ FX Full Face CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Sleep Study Titrated to 11 cm H2O

User avatar
Lizistired
Posts: 2835
Joined: Tue Dec 14, 2010 10:47 pm
Location: Indiana

Re: Someone please help this Newbie!

Post by Lizistired » Thu Feb 24, 2011 6:01 pm

When you download your card, click select data, then select ALL and check the little box under that that says something about detailed session. Then click download

_________________
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Swift FX sometimes, CMS-50F, Cervical collar sometimes, White noise, Zeo... I'm not well, but I'm better.