Just got Resmed S9 and we are so LOST

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
cabo
Posts: 13
Joined: Sat Feb 19, 2011 9:18 am

Just got Resmed S9 and we are so LOST

Post by cabo » Sat Feb 19, 2011 10:06 am

OK so the sleep study was done . The prescription written. The Resmed s9 with Humi bought. Don't know anything about this machine. EPR? My husband put it on last night and snored all night. He said this morning,"I don't think anything was happening" It was turned on we did the mask fit thing,the ramp of 5 mins and apparently there is more to it. We are lost. People talk on this forum about settings 6-10 4-7 etc. ???His script was written for "9" and that's apparently what it was set at.HELP!!!!!!!!!!!!!!!! Boughtn the machine on line sleep store. (No Help) paid cash. Insurance No. Please help to get us started.

User avatar
Pugsy
Posts: 65115
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Just got Resmed S9 and we are so LOST

Post by Pugsy » Sat Feb 19, 2011 10:10 am

First question.
Exactly which S9 machine did you get? There are multiple models and responders to your questions will need to know which model.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
scrapper
Posts: 776
Joined: Mon Nov 16, 2009 1:32 pm
Location: USA

Re: Just got Resmed S9 and we are so LOST

Post by scrapper » Sat Feb 19, 2011 10:12 am

You're not lost............you have us.

Answer the question by Pugsy and we'll all help--probably more than you wish.

_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: ResScan software 3.13, Pressure 21/15
“Life is 10% what happens to you, and 90% what you make of it.” Charles Swindoll

User avatar
robysue
Posts: 7520
Joined: Sat Sep 18, 2010 2:30 pm
Location: Buffalo, NY
Contact:

Re: Just got Resmed S9 and we are so LOST

Post by robysue » Sat Feb 19, 2011 10:13 am

Second question: Did you read the entire owner's manual carefully?

Third question: Do you have the provider's manual (also known as the clinical manual)? If so, did you read it carefully?

_________________
Machine: DreamStation BiPAP® Auto Machine
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear
Additional Comments: PR System DreamStation and Humidifier. Max IPAP = 9, Min EPAP=4, Rise time setting = 3, minPS = 3, maxPS=5

cabo
Posts: 13
Joined: Sat Feb 19, 2011 9:18 am

Re: Just got Resmed S9 and we are so LOST

Post by cabo » Sat Feb 19, 2011 10:19 am

Wow a reply already. TY. It is the Resmed autoset & Elite with H5i. I was sent a Welcome guide. Less than informative.Only 12 small pages(mostly pictures)Settings for tubing,mask,then mask fit, Adjujsting humidity and ramp time.That's it.

User avatar
Pugsy
Posts: 65115
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Just got Resmed S9 and we are so LOST

Post by Pugsy » Sat Feb 19, 2011 10:24 am

It will help if your will go to the User Control Panel in the upper left portion of the page under the banner and go to your Profile, then to the equipment page and fill in text please, what your equipment is. The S9 Auto
and also the type of mask being used. This will show everyone at a glance what you use and save time and questions.

So with that, what type of mask is being use? Brand and model?

I don't use the ResMed machine so all I can do is ask questions and I will let those that do use the ResMed machine help you with those details.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

cabo
Posts: 13
Joined: Sat Feb 19, 2011 9:18 am

Re: Just got Resmed S9 and we are so LOST

Post by cabo » Sat Feb 19, 2011 10:32 am

Quattro FX full face mask ResMed

User avatar
scrapper
Posts: 776
Joined: Mon Nov 16, 2009 1:32 pm
Location: USA

Re: Just got Resmed S9 and we are so LOST

Post by scrapper » Sat Feb 19, 2011 10:52 am

Do you like the Quattro? How are your leaks?

Finding the right mask is almost as hard as finding a good spouse.

_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: ResScan software 3.13, Pressure 21/15
“Life is 10% what happens to you, and 90% what you make of it.” Charles Swindoll

cabo
Posts: 13
Joined: Sat Feb 19, 2011 9:18 am

Re: Just got Resmed S9 and we are so LOST

Post by cabo » Sat Feb 19, 2011 10:58 am

Only had the whole thing one night.Nothing worked. I'm telling you we are LOST. I am one to read manuals cover to cover. This stuff came with a 5 step guide on how to make it work. Plug it in,Turn it on,Put face mask on, Lie down,Turn it off in the morning. WHAT????

User avatar
DoriC
Posts: 5214
Joined: Sat Sep 13, 2008 9:28 pm
Location: NJ

Re: Just got Resmed S9 and we are so LOST

Post by DoriC » Sat Feb 19, 2011 11:00 am

I'm sorry if I'm confusing you, but if your machine is an Elite then it is a cpap and can only be set to one pressure, 9cm in your case, if it's an auto you can set it to a range of pressures(8-12cms)for example. Let us know for sure. Good luck and glad you found us.

_________________
Mask: Mirage Quattro™ Full Face CPAP Mask with Headgear
Additional Comments: 14/8.4,PS=4, UMFF, 02@2L,
"Do or Do Not-There Is No Try"-"Yoda"
"We are what we repeatedly do,so excellence
is not an act but a habit"-"Aristotle"
DEAR HUBBY BEGAN CPAP 9/2/08

User avatar
Jayjonbeach
Posts: 133
Joined: Sun Aug 15, 2010 9:04 am
Location: Mississauga

Re: Just got Resmed S9 and we are so LOST

Post by Jayjonbeach » Sat Feb 19, 2011 11:02 am

cabo wrote:Wow a reply already. TY. It is the Resmed autoset & Elite with H5i. I was sent a Welcome guide. Less than informative.Only 12 small pages(mostly pictures)Settings for tubing,mask,then mask fit, Adjujsting humidity and ramp time.That's it.
It can NOT be an Elite and Autoset model, it is one or the other. Whatever it says on the actual machine will tell you what it is.

If it is an Autoset, find and download the Clinic Manual. With a titration pressure of 9, I would set the low pressure at 9 and the high pressure at 14 just to start.

Snoring usually indicates the pressure is too low. Note that one night in a sleep lab is not always the best setting to do a titration in the first place so having the machine you own set on APAP therapy will help you judge what is a good pressure (though you will need to get the software and do LOTS of reading on this forum to figure this all out).

If it is the Elite model, then you will be limited to only having one pressure. If it is on 9, I would now try 10 since he is snoring with 9.

Changing the pressure is very easy on either machine, there are two buttons you have to hold down for 3 seconds which give you access to the Clinical menu, from there you can change all the settings, the Clinical Manual tells you all this.

Good luck, note the above is just my own opinion based on experience and info I have learned over time.

_________________
Mask: Quattro™ FX Full Face CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Software is actually 3.11. Curious about Oxygen and whether I might need it or just want it.
Tired of being tired for 20 years running, hoping this is the answer...

User avatar
rested gal
Posts: 12881
Joined: Thu Sep 09, 2004 10:14 pm
Location: Tennessee

Re: Just got Resmed S9 and we are so LOST

Post by rested gal » Sat Feb 19, 2011 11:10 am

Jayjonbeach wrote:It can NOT be an Elite and Autoset model, it is one or the other. Whatever it says on the actual machine will tell you what it is.
Right. The confusion is probably because the clinical manual is a combined manual for both machines. The clinical manual will say both Autoset and Elite on the cover. Sounds like the User manual (or the "Welcome Guide", anyway) is also a combo.

As Jay mentioned, the specific name of the machine will be written out on top of the machine itself. The word Autoset or the word Elite (not both) will be written on the machine.
ResMed S9 VPAP Auto (ASV)
Humidifier: Integrated + Climate Control hose
Mask: Aeiomed Headrest (deconstructed, with homemade straps
3M painters tape over mouth
ALL LINKS by rested gal:
viewtopic.php?t=17435

User avatar
NancyAZ
Posts: 59
Joined: Wed Jan 26, 2011 11:25 pm
Location: Scottsdale, AZ

Re: Just got Resmed S9 and we are so LOST

Post by NancyAZ » Sat Feb 19, 2011 5:51 pm

I have the ResMed Elite and the book that came with the machine was almost worthless. It does say both Elite and Autoset on the booklet. My machine has the name "Elite" above the button on the lefthand side. Yours might say Autoset. Whatever, the name will be on the top of the machine. Any information that I got about my machine I got from people on this website. I downloaded the software, the manual for the software, and the manual for the machine all from links given here. Post an email asking for those items and you will receive help. I can't help you with that but someone here will help you (Yes he will!). There is also a video that gives lots of information about the S9 on http://www.sleepnation.com.

Good luck. You will be fine. The beginning is always overwhelming for everyone, even the lucky ones who have few problems. Read everything that you can read on this forum, ask questions and share any problems that you have. I have never been on a forum with a more caring bunch of people set on helping others.
Nancy

_________________
Mask: Swift™ FX For Her Nasal Pillow CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: ResScan 3.13; Pressure 8-12; started CPAP Jan. 26, 2011

Alicia41964
Posts: 43
Joined: Fri Feb 04, 2011 9:48 pm
Location: Rochester, New York

Re: Just got Resmed S9 and we are so LOST

Post by Alicia41964 » Sat Feb 19, 2011 10:19 pm

Cabo, the beginning was a challenge for me also! Saw my provider @ 30 days and they told me I was non-compliant!

I used it each and every night, but through my own ignorance, it wasn't doing anything. I've sinced learned much, mostly through trial and error. No, I did not get a manual with mine, although it was brand spanking new. By the way, whatever you do, please don't email Res Med looking for info on the software. I am able to see all my numbers/charts/graphs, etc... However, I cannot print them. I get an error message and my program shuts right down. Silly me, I emailed their customer support and was told that since I am not a "clinician", I have no business downloading anything or trying to utilize this software. Boy, was I mad. I was only trying to be an educated patient! Keep looking to the people on this site as they seem to know the tricks of the trade! One can learn a lot on here. If neccessary, each night, try one thing differently on your machine and pretty soon, you should have it all set. Best of luck and keep plugging away!

_________________
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Humidification set @ 2.0 with 4 Liters of Oxygen bled in at night BTW - I'm looking for who I used to be..... Has anyone seen her? LOL

Kylejbarton
Posts: 2
Joined: Mon Feb 21, 2011 9:23 am

Re: Just got Resmed S9 and we are so LOST

Post by Kylejbarton » Mon Feb 21, 2011 9:45 am

I have been using my machine since December and finally decided to research it a bit. I've read through various posts here and now know how to change the settings and access the sleep reports. Thank you. I was prescribed a setting of 7. My AHI is 0.4 and leak 10.7. I just changed my setting to to 7.4. I need to find out how to download the software and clinical manual please.
Resmed Autoset S9 with h5i humidifier, CPAP mode, 7.4 setting, slimline swift tube with nasal pillow, humidity 3.0, EPR 0. Severe obstructive apnea.