How does a machine know?

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
Hephaestus
Posts: 29
Joined: Wed Sep 08, 2010 7:11 am

How does a machine know?

Post by Hephaestus » Wed Jan 26, 2011 10:00 am

I'm curious, maybe the board experts can advise...

How does my machine know the difference between types of events? All these events are concerned with a stop of breathing in one way or another: central vs. obstructive vs. Hypopnea. How does it tell the difference between a sinus event to a throat event? And while we're on the subject, how does it measure snoring? Are there sensors in the machine that I'm not aware of?

_________________
Mask: Mirage Quattro™ Full Face CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Actually, I'm using ResScan 03.12.016

User avatar
avi123
Posts: 4509
Joined: Tue Dec 21, 2010 5:39 pm
Location: NC

Re: How does a machine know?

Post by avi123 » Wed Jan 26, 2011 10:14 am

Are you interested in a list of articles and reports that explain all these? But first please reply how could I tell the answer to your question about those sensors that you don't know about?

Overall, yes, there are sensors that measure most of the following parameters, but notice that while in Sleep Labs all of these are usually measured and recorded only those marked with a star are measured in home- CPAPs and an Oximeter:

1) Airflow* and Pressures*
2) Snore*
3) Plethysmography (some time)
4) Heart Rate*
5) Chest Effort
6) Abdominal Effort
7) Body Position
Mask Pressure* (in some)
9) Mask Leak*
10) SpO2*
11) Limb Movements
12) EKG
13) EEG
14) EMG of chin and tibials

As to a comparison between measures and calculations between in-home-CPAPS and those at Sleep Study centers:
(as of June 2010)

Studies that compared CPAP-derived AHI to in-lab PSG, AHI showed marked differences between the two concluding that CPAP’s have insufficient signal quality and could potentially generate” misleading results” especially for mild to moderate OSA patients. These studies demonstrate that while superior to oximetry alone and therefore a potentially better in-home “screening” tool, CPAP’s without supplemental information can overestimate apneas and underestimate hypopneas. The studies also showed susceptibility to mouth leaks and mask motion artifacts.

_________________
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments:  S9 Autoset machine; Ruby chinstrap under the mask straps; ResScan 5.6
Last edited by avi123 on Wed Jan 26, 2011 3:07 pm, edited 5 times in total.
see my recent set-up and Statistics:
http://i.imgur.com/TewT8G9.png
see my recent ResScan treatment results:
http://i.imgur.com/3oia0EY.png
http://i.imgur.com/QEjvlVY.png

User avatar
LinkC
Posts: 3154
Joined: Wed Oct 29, 2008 1:06 pm
Location: Amelia Island, FL

Re: How does a machine know?

Post by LinkC » Wed Jan 26, 2011 10:26 am

The short answer is that each type of event has a different profile.

For example, a snore exhibits a "vibration" of the flow resistance during inhale. Other types of events have signatures that involve different phases of the event. But you should know that your machine doesn't always get it right.

_________________
MachineMaskHumidifier
Additional Comments: 11-14 cmH2O
The OSA patient died quietly in his sleep.
Unlike his passengers who died screaming as the car went over the cliff...

StevenXXXX
Posts: 165
Joined: Tue Jan 04, 2011 4:22 pm

Re: How does a machine know?

Post by StevenXXXX » Wed Jan 26, 2011 10:59 am

LinkC said: your machine doesn't always get it right

Do you have an opinion as to how much better the equipment is that is used in a Sleep Study?

_________________
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: ResMed ResScan Software Version 3.16 Do NOT use either Ramp or EPR

WestCoastCdnGrl
Posts: 140
Joined: Fri Dec 24, 2010 1:02 am
Location: London, England

Re: How does a machine know?

Post by WestCoastCdnGrl » Wed Jan 26, 2011 11:15 am

LinkC wrote: your machine doesn't always get it right
I'd rather that it not always get it right but at least be there to help me through the episode than be 100% all of the time and be hit-or-miss as far as reliability/efficacy are concerned... an episode is an episode is an episode and as long as I don't go back to how I was on the night of December 12, 2010 (the night before the start of therapy), then a miss diagnosis of what type of event I have is really nothing to get bent out of shape about.

Hephaestus
Posts: 29
Joined: Wed Sep 08, 2010 7:11 am

Re: How does a machine know?

Post by Hephaestus » Wed Jan 26, 2011 1:43 pm

All right, thank you for the responses. I understand how sleep centers could get that information because there are sensors all over the place. I was curious about how our home machines might derive that information. For example, I'm using an S9 Auto and notice on my reports that some events are classified differently than others. So was curious as to how the S9 determines which is which...

_________________
Mask: Mirage Quattro™ Full Face CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Actually, I'm using ResScan 03.12.016

BernieRay
Posts: 390
Joined: Tue Nov 30, 2010 8:21 am

Re: How does a machine know?

Post by BernieRay » Wed Jan 26, 2011 2:16 pm

You might want to take a look at the clinician's manual for the S9 (google should find it). It explains how the S9 identifies each event as well as the S9's response to each type of event.
Ray
Diagnosed in 1997

Hephaestus
Posts: 29
Joined: Wed Sep 08, 2010 7:11 am

Re: How does a machine know?

Post by Hephaestus » Wed Jan 26, 2011 3:32 pm

Good tip, thank you!

_________________
Mask: Mirage Quattro™ Full Face CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Actually, I'm using ResScan 03.12.016

StevenXXXX
Posts: 165
Joined: Tue Jan 04, 2011 4:22 pm

Re: How does a machine know?

Post by StevenXXXX » Wed Jan 26, 2011 4:39 pm

You can get the Clinian Manual for the ResMed S9 AutoSet and all of the others in the S9 Series by following the directions below. All of the models in the S9 Series are in the same Clinian Manual.

We have a few CPAP Clinician/Setup Manuals (that are too large to post on this web site, but are available through email.

Please follow these email instructions exactly:

If you would like one of the manuals below, please send an email to: apneaboard@gmail.com and put "Manuals" (without the quotes) in the subject line. Then, let us know in the body of your email which of the following manuals you would like us to email to you.

LIMIT: Two (2) Clinician Manuals per individual.
(If you request more than two manuals, your request will be discarded)


I tried to PM the Clinician Manual to you and even tried to attach it in this reply - but apparently the file was too large.

_________________
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: ResMed ResScan Software Version 3.16 Do NOT use either Ramp or EPR

User avatar
jmelby
Posts: 463
Joined: Sat Nov 14, 2009 7:52 pm
Location: Minneapolis, MN

Re: How does a machine know?

Post by jmelby » Wed Jan 26, 2011 4:50 pm

Resmed's website has some information on their central apnea detection that you might find interesting:
http://www.resmed.com/us/products/s9_se ... nc=dealers

"It uses the forced oscillation technique (FOT) to determine the state of the airway during an apnea. When an apnea is detected, small oscillations are added to the pressure to measure airway patency. The CSA algorithm uses the resulting flow and pressure to measure airway patency and differentiate central and obstructive events."

_________________
Mask: Mirage Activa™ LT Nasal CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: ResScan 3.12, APAP 9 - 13, no EPR, ClimateControl 75F
(yet another Jeff)

User avatar
avi123
Posts: 4509
Joined: Tue Dec 21, 2010 5:39 pm
Location: NC

Re: How does a machine know?

Post by avi123 » Wed Jan 26, 2011 4:56 pm

Hephaestus wrote:All right, thank you for the responses. I understand how sleep centers could get that information because there are sensors all over the place. I was curious about how our home machines might derive that information. For example, I'm using an S9 Auto and notice on my reports that some events are classified differently than others. So was curious as to how the S9 determines which is which...

The S9 has computer chips inside similar to a laptop. It also has lots of tiny sensors to measure the physical parameters.

_________________
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments:  S9 Autoset machine; Ruby chinstrap under the mask straps; ResScan 5.6
see my recent set-up and Statistics:
http://i.imgur.com/TewT8G9.png
see my recent ResScan treatment results:
http://i.imgur.com/3oia0EY.png
http://i.imgur.com/QEjvlVY.png

cpapdork
Posts: 103
Joined: Sat Nov 27, 2010 7:57 am
Contact:

Re: How does a machine know?

Post by cpapdork » Thu Jan 27, 2011 2:57 pm

I believe xPAPs generally have just a single type of sensor which measures only airflow*. As some prior posters mention, by measuring the "baseline" flow and the resulting flow when certain variations are purposely introduced by the machine, the machine is able to classify events.

* Maybe some also measure temp and humidity, but these I think are mainly for comfort, not for diagnosis.
Image

MidnightOwl
Posts: 368
Joined: Thu Feb 22, 2007 11:49 pm

Re: How does a machine know?

Post by MidnightOwl » Thu Jan 27, 2011 3:18 pm

avi123 wrote:
As to a comparison between measures and calculations between in-home-CPAPS and those at Sleep Study centers:
(as of June 2010)

Studies that compared CPAP-derived AHI to in-lab PSG, AHI showed marked differences between the two concluding that CPAP’s have insufficient signal quality and could potentially generate” misleading results” especially for mild to moderate OSA patients. These studies demonstrate that while superior to oximetry alone and therefore a potentially better in-home “screening” tool, CPAP’s without supplemental information can overestimate apneas and underestimate hypopneas. The studies also showed susceptibility to mouth leaks and mask motion artifacts.
avi123,

Could you provide a source for the quote you posted?

Thanks.

Midnightowl

Hephaestus
Posts: 29
Joined: Wed Sep 08, 2010 7:11 am

Re: How does a machine know?

Post by Hephaestus » Thu Jan 27, 2011 3:51 pm

jmelby wrote:Resmed's website has some information on their central apnea detection that you might find interesting:
http://www.resmed.com/us/products/s9_se ... nc=dealers

"It uses the forced oscillation technique (FOT) to determine the state of the airway during an apnea. When an apnea is detected, small oscillations are added to the pressure to measure airway patency. The CSA algorithm uses the resulting flow and pressure to measure airway patency and differentiate central and obstructive events."
Very interesting! Knowing this tells me that our modern machines are much more than just a simple air pump. I wouldn't have known that we could gather this type of information via air pressure/flow measurements. Certainly we couldn't do it without cpu's and software algorithm's to precisely control the air pressure and volume.

I'd always thought it would be a good move to get into the medical equipment field!