Updating the firmware on the PR System One machines

General Discussion on any topic relating to CPAP and/or Sleep Apnea.

Does updating your machines firmware interest you?

Yes
200
93%
No
9
4%
I like to vote in polls.
7
3%
 
Total votes: 216

User avatar
blakepro
Posts: 507
Joined: Thu Oct 08, 2009 1:26 pm
Location: Utah

Updating the firmware on the PR System One machines

Post by blakepro » Thu Jan 20, 2011 3:16 pm

So, I have discovered that it is possible to update the firmware on the PR 1 machines.

Why would you want to do this? Well, one reason is the expanded "classic" humidification options dont appear to be available on older firmwares. There may be other benefits or improvements, but the changes between firmware versions dont seem to be documented anywhere.


So, There are two pieces to this whole process.

1:There is a program called "PR System 1 Flash Upgrade Utility 1.0" that is used to upgrade PR1 devices to software version 1.24

2:There is some sort of cable that you need to connect your machine to the PC with. Its Respironics Part #1064749
I was hoping that the update would be done through the SD card itself, but NOPE, you need this part to do it.

This Part# 1064749 is where I'm currently hung up. Does anyone here have any ideas how I could buy that or at least get some information about what it is or what it looks like? Seems like it may be a "DME ONLY" part or something..

One thing I'm also curious about is how exactly it connects to the PR1. Does it plug into the humidifier port? The SD card slot? Who knows. Until I find one, we may never know.

Any ideas would be welcomed!

Thanks.
Last edited by blakepro on Fri Jan 21, 2011 1:08 pm, edited 1 time in total.

User avatar
6PtStar
Posts: 2659
Joined: Mon Jun 18, 2007 1:58 pm
Location: Texas, Is there any other place?

Re: Updating the firmware on the PR System One machines

Post by 6PtStar » Thu Jan 20, 2011 7:30 pm

'

_________________
Humidifier: HC150 Heated Humidifier With Hose, 2 Chambers and Stand
Additional Comments: 11cm/H2O, Encore Pro 1.8i, Pro Analyzer, Encore Viewer1.0 - 3 Remstar Pro2's, 1 Remstar Auto
Last edited by 6PtStar on Thu Jan 27, 2011 10:49 am, edited 1 time in total.
Life's journey is not to arrive at the grave safely in a well preserved body, but rather to skid in sideways, totally worn out, shouting: "Wow what a ride!"
I still play Cowboys and Bad Guys but now I use real bullets. CAS

User avatar
GumbyCT
Posts: 5779
Joined: Fri Sep 14, 2007 6:22 pm
Location: CT
Contact:

Re: Updating the firmware on the PR System One machines

Post by GumbyCT » Thu Jan 20, 2011 8:00 pm

blakepro wrote:So, I have discovered that it is possible to update the firmware on the PR 1 machines.

Why would you want to do this? Well, one reason is the expanded "classic" humidification options dont appear to be available on older firmwares. There may be other benefits or improvements, but the changes between firmware versions dont seem to be documented anywhere.


So, There are two pieces to this whole process.

1:There is a program called "PR System 1 Flash Upgrade Utility 1.0" that is used to upgrade PR1 devices to software version 1.24

2:There is some sort of cable that you need to connect your machine to the PC with. Its Respironics Part #1064749
I was hoping that the update would be done through the SD card itself, but NOPE, you need this part to do it.

This Part# 1064749 is where I'm currently hung up. Does anyone here have any ideas how I could buy that or at least get some information about what it is or what it looks like? Seems like it may be a "DME ONLY" part or something..

One thing I'm also curious about is how exactly it connects to the PR1. Does it plug into the humidifier port? The SD card slot? Who knows. Until I find one, we may never know.

Any ideas would be welcomed!

Thanks.
I found this - http://www.search.philips.com/search/se ... medical&o=

Not sure where you got the p/n but maybe start there?

Personally, I think it will be easier to pick up another machine with the update? Unless you can find a tech who is willing to help.
I do wish you well.

_________________
Humidifier: HC150 Heated Humidifier With Hose, 2 Chambers and Stand
Additional Comments: New users can't remember they can't remember YET!
BeganCPAP31Jan2007;AHI<0.5
I have no doubt, how I sleep affects every waking moment.
I am making progress-NOW I remember that I can't remember
;)
If this isn’t rocket science why are there so many spaceshots?
Be your own healthcare advocate!

User avatar
blakepro
Posts: 507
Joined: Thu Oct 08, 2009 1:26 pm
Location: Utah

Re: Updating the firmware on the PR System One machines

Post by blakepro » Thu Jan 20, 2011 8:26 pm

GumbyCT wrote:
blakepro wrote:So, I have discovered that it is possible to update the firmware on the PR 1 machines.

Why would you want to do this? Well, one reason is the expanded "classic" humidification options dont appear to be available on older firmwares. There may be other benefits or improvements, but the changes between firmware versions dont seem to be documented anywhere.


So, There are two pieces to this whole process.

1:There is a program called "PR System 1 Flash Upgrade Utility 1.0" that is used to upgrade PR1 devices to software version 1.24

2:There is some sort of cable that you need to connect your machine to the PC with. Its Respironics Part #1064749
I was hoping that the update would be done through the SD card itself, but NOPE, you need this part to do it.

This Part# 1064749 is where I'm currently hung up. Does anyone here have any ideas how I could buy that or at least get some information about what it is or what it looks like? Seems like it may be a "DME ONLY" part or something..

One thing I'm also curious about is how exactly it connects to the PR1. Does it plug into the humidifier port? The SD card slot? Who knows. Until I find one, we may never know.

Any ideas would be welcomed!

Thanks.
I found this - http://www.search.philips.com/search/se ... medical&o=

Not sure where you got the p/n but maybe start there?

Personally, I think it will be easier to pick up another machine with the update? Unless you can find a tech who is willing to help.
I do wish you well.
Thanks for the reply and link, Yeah, I checked that same search area too and it produces zero results. The part number came directly from an inquiry to respironics. They said that part is necessary to do the firmware update.

I have the program installed and working on my PC.. just need the cable. heh... So close, yet so far away.

User avatar
jbn3boys
Posts: 849
Joined: Tue Aug 31, 2010 9:04 am
Location: Wisconsin

Re: Updating the firmware on the PR System One machines

Post by jbn3boys » Thu Jan 20, 2011 8:36 pm

Could a DME help out?

_________________
Mask: Swift™ FX For Her Nasal Pillow CPAP Mask with Headgear
Additional Comments: titration 11
Sleep study (Aug 2010): AHI 16 (On mask AHI 0.2) <-- Now, if I could just attain that "0.2" again!
aPAP for 4 months, Switched to BiPap, 2nd sleep study Feb 2011 Possible PLMD
to quote Madalot..."I'm an enigma"

DadscreepyCPAPmask
Posts: 125
Joined: Thu Dec 30, 2010 3:28 am

Re: Updating the firmware on the PR System One machines

Post by DadscreepyCPAPmask » Thu Jan 20, 2011 8:55 pm

Updating firmware on these machines should not be something that is difficult to do. I update my firmware on all my electronic devices in my home, so this should be no different if Respironics does do periodic updates. The DME's probably do not know much about this though. It will be up to the individual owners of these machines to do their own leg work and find out whats what. I am very interested because firmware updates are usually a very good thing.

_________________
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear
Additional Comments: The Ultimate Chin Strap!!

User avatar
sydneybird
Posts: 122
Joined: Sat Dec 04, 2010 8:42 pm

Re: Updating the firmware on the PR System One machines

Post by sydneybird » Thu Jan 20, 2011 9:44 pm

I'm sure you checked the provider's guide and users manual. I just did and found no reference to any data port besides the SD card slot. I even pried off the cover where the card reader is to see if there was something behind it. Nothing behind it. I wonder why there is even a removable cover there in the first place? Maybe there used to be a phone jack there in older models. If yours has a phone jack that could be used as a data port. I'm sure you also know that classic mode is only accessible through the provider's menu. I provided instructions in a humidifier post. I don't mean to insult, just covering the bases. I don't know anything about the older SW versions.

User avatar
snoozysue
Posts: 287
Joined: Sun Oct 17, 2010 2:32 am
Location: Australia

Re: Updating the firmware on the PR System One machines

Post by snoozysue » Thu Jan 20, 2011 10:50 pm

Hi, I am not technical at all but I did find this:
http://stores.nextag.com/store/4798115/ ... load-Cable
I have no idea if its what you need.

cflame1
Posts: 3311
Joined: Sat Mar 11, 2006 6:55 am
Location: expat Canadian in Kentucky

Re: Updating the firmware on the PR System One machines

Post by cflame1 » Thu Jan 20, 2011 11:09 pm


User avatar
jbn3boys
Posts: 849
Joined: Tue Aug 31, 2010 9:04 am
Location: Wisconsin

Re: Updating the firmware on the PR System One machines

Post by jbn3boys » Fri Jan 21, 2011 8:33 am

blakepro wrote:So, I have discovered that it is possible to update the firmware on the PR 1 machines.
So where did you "discover" this information? How would you know if your machine needs the update, or if it already has it? How old is the firmware update? When was it released?

I agree firmware/software updates are both important and easy to do on home electronics. I'm not sure if the same is true of medical equipment. While I agree they are most likely useful, I don't know how "easy" they are. And you don't seem to give any firm information about this update. All you said is that it is possible. Of course it's possible. Does that mean there is an update available? Not necessarily.

_________________
Mask: Swift™ FX For Her Nasal Pillow CPAP Mask with Headgear
Additional Comments: titration 11
Sleep study (Aug 2010): AHI 16 (On mask AHI 0.2) <-- Now, if I could just attain that "0.2" again!
aPAP for 4 months, Switched to BiPap, 2nd sleep study Feb 2011 Possible PLMD
to quote Madalot..."I'm an enigma"

User avatar
Slinky
Posts: 11372
Joined: Wed Nov 01, 2006 3:43 pm
Location: Mid-Michigan

Re: Updating the firmware on the PR System One machines

Post by Slinky » Fri Jan 21, 2011 10:57 am

How does one check for the firmware version on the PR S1s?? I know how to do so on the Resmeds but have NO IDEA on the PR S1 BPAP Auto and now you have my curiosity up!!!

_________________
Mask: Quattro™ FX Full Face CPAP Mask with Headgear
Additional Comments: PR SystemOne BPAP Auto w/Bi-Flex & Humidifier - EncorePro 2.2 Software - Contec CMS-50D+ Oximeter - Respironics EverFlo Q Concentrator
Women are Angels. And when someone breaks our wings, we simply continue to fly.....on a broomstick. We are flexible like that.
My computer says I need to upgrade my brain to be compatible with its new software.

User avatar
GumbyCT
Posts: 5779
Joined: Fri Sep 14, 2007 6:22 pm
Location: CT
Contact:

Re: Updating the firmware on the PR System One machines

Post by GumbyCT » Fri Jan 21, 2011 11:04 am

Slinky wrote:How does one check for the firmware version on the PR S1s?? I know how to do so on the Resmeds but have NO IDEA on the PR S1 BPAP Auto and now you have my curiosity up!!!
You have to unplug the power then look at the LCD when you plug in the power again
Good Luck

_________________
Humidifier: HC150 Heated Humidifier With Hose, 2 Chambers and Stand
Additional Comments: New users can't remember they can't remember YET!
BeganCPAP31Jan2007;AHI<0.5
I have no doubt, how I sleep affects every waking moment.
I am making progress-NOW I remember that I can't remember
;)
If this isn’t rocket science why are there so many spaceshots?
Be your own healthcare advocate!

User avatar
Slinky
Posts: 11372
Joined: Wed Nov 01, 2006 3:43 pm
Location: Mid-Michigan

Re: Updating the firmware on the PR System One machines

Post by Slinky » Fri Jan 21, 2011 11:09 am

Thanks, Gumby. Youse a neat kid! (((hugs)))

Sheesh! Respironics and their d*mn unplug and plug in "hang up". I wonder when they are gonna "grow up" and run w/the "big boys".

This loaner from my DME is v.1.24 and it still dumps data every once in a while. PR really does need some new and better software engineers.

_________________
Mask: Quattro™ FX Full Face CPAP Mask with Headgear
Additional Comments: PR SystemOne BPAP Auto w/Bi-Flex & Humidifier - EncorePro 2.2 Software - Contec CMS-50D+ Oximeter - Respironics EverFlo Q Concentrator
Women are Angels. And when someone breaks our wings, we simply continue to fly.....on a broomstick. We are flexible like that.
My computer says I need to upgrade my brain to be compatible with its new software.

User avatar
blakepro
Posts: 507
Joined: Thu Oct 08, 2009 1:26 pm
Location: Utah

Re: Updating the firmware on the PR System One machines

Post by blakepro » Fri Jan 21, 2011 12:18 pm

Slinky wrote:Thanks, Gumby. Youse a neat kid! (((hugs)))

Sheesh! Respironics and their d*mn unplug and plug in "hang up". I wonder when they are gonna "grow up" and run w/the "big boys".

This loaner from my DME is v.1.24 and it still dumps data every once in a while. PR really does need some new and better software engineers.
Mine is version V1.07

User avatar
Slinky
Posts: 11372
Joined: Wed Nov 01, 2006 3:43 pm
Location: Mid-Michigan

Re: Updating the firmware on the PR System One machines

Post by Slinky » Fri Jan 21, 2011 12:25 pm

I just checked w/my friend I had borrowed his PR S1 BPAP Auto from earlier last year. The version on that one was V.1.22.

It sounds like patients are paying for equipment to be the "beta testers" for the Respironics devices. *sigh*