DME Says Data Machine Not Acailable for CPAP
DME Says Data Machine Not Acailable for CPAP
I just talked to my DME, looking for a machine that has data on it, besides just time of use. They told me there is no CPAP machine available with that, you have to go to BPAP. You can only do this thru your doctor. Is this true?
-
sleepnationtv
- Posts: 283
- Joined: Mon Jul 19, 2010 7:23 pm
Re: DME Says Data Machine Not Acailable for CPAP
This is not true, every machine manufacturer has data capable machines. The DME just pays more for those. Here are a few videos you can watch all of these machines can be data capable.
http://www.sleepnation.tv/post/2337667545/resmeds9
http://www.sleepnation.tv/post/808897740/episode1
Just show them these as proof.
http://www.sleepnation.tv/post/2337667545/resmeds9
http://www.sleepnation.tv/post/808897740/episode1
Just show them these as proof.
Dane Schapper, Founder cpaplibrary.com
Check out the cpaplibrary video bloghttp://www.cpaplibrary.com
Check out the cpaplibrary video bloghttp://www.cpaplibrary.com
Re: DME Says Data Machine Not Acailable for CPAP
Sounds like you have one of those DME's that "If their lips are moving they are lying"! In your present machine the PR System One the Pro Model and the Auto both collect Data. Not all BiPaps collect data. Must be the BiPap Auto as I remember. In the Respirionics line nothing with Plus in the name will collect data. If it does not collect data you might as well go hook your mask up to the leaf blower in your garage, JMHO !!
Jerry
Jerry
_________________
| Humidifier: HC150 Heated Humidifier With Hose, 2 Chambers and Stand |
| Additional Comments: 11cm/H2O, Encore Pro 1.8i, Pro Analyzer, Encore Viewer1.0 - 3 Remstar Pro2's, 1 Remstar Auto |
Last edited by 6PtStar on Thu Dec 23, 2010 12:13 pm, edited 1 time in total.
Life's journey is not to arrive at the grave safely in a well preserved body, but rather to skid in sideways, totally worn out, shouting: "Wow what a ride!"
I still play Cowboys and Bad Guys but now I use real bullets. CAS
I still play Cowboys and Bad Guys but now I use real bullets. CAS
- BlackSpinner
- Posts: 9742
- Joined: Sat Apr 25, 2009 5:44 pm
- Location: Edmonton Alberta
- Contact:
Re: DME Says Data Machine Not Acailable for CPAP
If you believe that I have a bridge in Brooklyn for sale cheap. I have had a resmed s8Elite for over a year which gives me all that info.kentoboo wrote:I just talked to my DME, looking for a machine that has data on it, besides just time of use. They told me there is no CPAP machine available with that, you have to go to BPAP. You can only do this thru your doctor. Is this true?
Get a new DME with qualified people.
_________________
| Machine: PR System One REMStar 60 Series Auto CPAP Machine |
| Additional Comments: Quatro mask for colds & flus S8 elite for back up |
71. The lame can ride on horseback, the one-handed drive cattle. The deaf, fight and be useful. To be blind is better than to be burnt on the pyre. No one gets good from a corpse. The Havamal
Re: DME Says Data Machine Not Acailable for CPAP
kentoboo,
Talk to your insurance company and ask them for a comprehensive list of DMEs that are "in network" and start shopping around. You need to find another DME in my opinion.
You might also want to try to find a DME that specializes in CPAP equipment rather than a more generic DME. That's what I did. The CPAP-oriented DME that I found only carries full data APAP machines for their default CPAP/APAP machines. Most of the staff uses CPAPs. And the techs that set the machines up for customers always set the machines up so that the customers have access to the on-board data.
Talk to your insurance company and ask them for a comprehensive list of DMEs that are "in network" and start shopping around. You need to find another DME in my opinion.
You might also want to try to find a DME that specializes in CPAP equipment rather than a more generic DME. That's what I did. The CPAP-oriented DME that I found only carries full data APAP machines for their default CPAP/APAP machines. Most of the staff uses CPAPs. And the techs that set the machines up for customers always set the machines up so that the customers have access to the on-board data.
_________________
| Machine: DreamStation BiPAP® Auto Machine |
| Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear |
| Additional Comments: PR System DreamStation and Humidifier. Max IPAP = 9, Min EPAP=4, Rise time setting = 3, minPS = 3, maxPS=5 |
- chunkyfrog
- Posts: 34544
- Joined: Mon Jul 12, 2010 5:10 pm
- Location: Nowhere special--this year in particular.
Re: DME Says Data Machine Not Acailable for CPAP
what a bunch of dirt bags!(your current DME) --In my humble opinion, you need to switch DME's before your RX 'expires'.
The guy that read my sleep study used my request for a new Rx to require a totally unnecessary office visit--which puts me past the year-end deductible reset. You'll be lucky if insurance covers a full-data machine; if it doesn't, you can buy online; cpap.com, for example. You will find the price for what you want is probably far less than the current machine, (which you will probably use as a back-up.)
The guy that read my sleep study used my request for a new Rx to require a totally unnecessary office visit--which puts me past the year-end deductible reset. You'll be lucky if insurance covers a full-data machine; if it doesn't, you can buy online; cpap.com, for example. You will find the price for what you want is probably far less than the current machine, (which you will probably use as a back-up.)
_________________
| Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear |
| Additional Comments: Airsense 10 Autoset for Her |
Last edited by chunkyfrog on Thu Dec 23, 2010 12:23 pm, edited 1 time in total.
Re: DME Says Data Machine Not Acailable for CPAP
You need to go to another DME....if they would out and out lie to you, what else are they going to try. Run, don't walk, go, these guys need to be reported, there are lots of data capable machines that will run in straight cpap mode or auto. My spare machine is a Respironics M series auto, a flex, and I use it in straight cpap mode, I don't like it in auto, but I certainly do like the data I can get off of it. Just look up any manufacturer and you will see what liers these guys are.
Good luck in your quest you've come to right place for help.
Cheers
Nan
Good luck in your quest you've come to right place for help.
Cheers
Nan
_________________
| Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
| Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
Started cpap in 2010.. still at it with great results.
- Borgasmord
- Posts: 31
- Joined: Fri Oct 15, 2010 9:40 am
Re: DME Says Data Machine Not Acailable for CPAP
Yeh, but cpaptalk.com foiled them! Good move by coming here, kentoboo!chunkyfrog wrote:what a bunch of dirt bags! --In my humble opinion.
Re: DME Says Data Machine Not Acailable for CPAP
Just wanting to post accurate information concerning BiPAPs here: PR makes the System One REMstar BiPAP Pro (straight BiPAP) and the System One REMstart BiPAP Auto (auto BiPAP). Both of these machines record AHI, FL, RERA, snore, leak, CA, OA, HY, PB as well as compliance data. I don't believe that PR makes a BiPAP that only records compliance data.6PtStar wrote:Sounds like you have one of those DME's that "If their lips are moving they are lying"! In your present machine the PR System One the Pro Model and the Auto both collect Data. Not all BiPaps collect data. Must be the BiPap Auto as I remember. In the Respirionics line nothing with Plus in the name will collect data. If it does not collect data you might as well go hook your mask up to the leaf blower in your garage, JMHO !!
Jerry
But PR does indeed make a System One REMstar Plus CPAP that only records compliance data. *sigh*
_________________
| Machine: DreamStation BiPAP® Auto Machine |
| Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear |
| Additional Comments: PR System DreamStation and Humidifier. Max IPAP = 9, Min EPAP=4, Rise time setting = 3, minPS = 3, maxPS=5 |
- rested gal
- Posts: 12880
- Joined: Thu Sep 09, 2004 10:14 pm
- Location: Tennessee
Re: DME Says Data Machine Not Acailable for CPAP
My list of machines that record "full data" (AHI and leak info) and those that don't -- updated through March 2010:
viewtopic.php?p=307168#p307168
The ones in blue record full data. The ones in red do not.
In case the DME tries to convince you that a machine provides the kind of "data" you want to see, simply because it happens to have a card stuck in it:
viewtopic.php?p=344265#p344265
For some entertaining reading about that:
check out LinkC's post on page 2 of this thread:
viewtopic.php?p=342111#p342111
And roster's post on page 3 of that same thread.
viewtopic.php?p=307168#p307168
The ones in blue record full data. The ones in red do not.
In case the DME tries to convince you that a machine provides the kind of "data" you want to see, simply because it happens to have a card stuck in it:
viewtopic.php?p=344265#p344265
For some entertaining reading about that:
check out LinkC's post on page 2 of this thread:
viewtopic.php?p=342111#p342111
And roster's post on page 3 of that same thread.
ResMed S9 VPAP Auto (ASV)
Humidifier: Integrated + Climate Control hose
Mask: Aeiomed Headrest (deconstructed, with homemade straps
3M painters tape over mouth
ALL LINKS by rested gal:
viewtopic.php?t=17435
Humidifier: Integrated + Climate Control hose
Mask: Aeiomed Headrest (deconstructed, with homemade straps
3M painters tape over mouth
ALL LINKS by rested gal:
viewtopic.php?t=17435
Re: DME Says Data Machine Not Acailable for CPAP
Ken,
If you don't get anything else from this thread, I hope it helps you understand that your DME is a lying snake. I would hate to think that your DME is just ignorant--then they would have no business dealing in DME in the first place. But there are no rules against lying to people who aren't as smart as you to check their facts.
Your next question will be whether or not you want to do business with these creeps. If at all possible, you should find another DME, but be prepared to be lied to by the new one too (it seems to be the norm, not the exception). The next DME you will encounter will probably admit that yes, there are data capable machines in every line they carry, but be prepared for them to tell you either that 1) your insurance won't cover a data capable machine or 2) your prescription doesn't cover a data capable machine. Both of these are LIES also--unless you have an HMO or Medicare Advantage (Part C) Plan. Otherwise, in 99% of the cases, your insurance does not care if you get a data capable machine or not (because they are going to pay the same amount regardless of which machine you get) and your prescription does NOT have to require a data capable machine.
So, once you decide whether or not you are going to stick to your current lying DME or choose another, BEFORE you try to get the machine of your choice, have a look at my blog herehttp://maskarrayed.wordpress.com/what-y ... me-part-i/. This will help you understand the insurance coverage and how to deal with these lying SOB's--er, I meant "DMEs".
If you don't get anything else from this thread, I hope it helps you understand that your DME is a lying snake. I would hate to think that your DME is just ignorant--then they would have no business dealing in DME in the first place. But there are no rules against lying to people who aren't as smart as you to check their facts.
Your next question will be whether or not you want to do business with these creeps. If at all possible, you should find another DME, but be prepared to be lied to by the new one too (it seems to be the norm, not the exception). The next DME you will encounter will probably admit that yes, there are data capable machines in every line they carry, but be prepared for them to tell you either that 1) your insurance won't cover a data capable machine or 2) your prescription doesn't cover a data capable machine. Both of these are LIES also--unless you have an HMO or Medicare Advantage (Part C) Plan. Otherwise, in 99% of the cases, your insurance does not care if you get a data capable machine or not (because they are going to pay the same amount regardless of which machine you get) and your prescription does NOT have to require a data capable machine.
So, once you decide whether or not you are going to stick to your current lying DME or choose another, BEFORE you try to get the machine of your choice, have a look at my blog herehttp://maskarrayed.wordpress.com/what-y ... me-part-i/. This will help you understand the insurance coverage and how to deal with these lying SOB's--er, I meant "DMEs".
_________________
| Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
| Mask: DreamWear Nasal CPAP Mask with Headgear |
What you need to know before you meet your DME http://tinyurl.com/2arffqx
Taming the Mirage Quattro http://tinyurl.com/2ft3lh8
Swift FX Fitting Guide http://tinyurl.com/22ur9ts
Don't Pay that Upcharge! http://tinyurl.com/2ck48rm
Taming the Mirage Quattro http://tinyurl.com/2ft3lh8
Swift FX Fitting Guide http://tinyurl.com/22ur9ts
Don't Pay that Upcharge! http://tinyurl.com/2ck48rm
Re: DME Says Data Machine Not Acailable for CPAP
Is there such a thing?BlackSpinner wrote:kentoboo wrote:snip..........
Get a new DME with qualified people.
ResMed S9 Autoset
ResMed AirFit F20 mask with headgear
ResMed AirFit F20 mask with headgear
Re: DME Says Data Machine Not Acailable for CPAP
Unfortunately they've made a Remstar Bipap Plus with other models... our hosts don't seem to have that one that I can find... it's compliance only.robysue wrote:Just wanting to post accurate information concerning BiPAPs here: PR makes the System One REMstar BiPAP Pro (straight BiPAP) and the System One REMstart BiPAP Auto (auto BiPAP). Both of these machines record AHI, FL, RERA, snore, leak, CA, OA, HY, PB as well as compliance data. I don't believe that PR makes a BiPAP that only records compliance data.6PtStar wrote:Sounds like you have one of those DME's that "If their lips are moving they are lying"! In your present machine the PR System One the Pro Model and the Auto both collect Data. Not all BiPaps collect data. Must be the BiPap Auto as I remember. In the Respirionics line nothing with Plus in the name will collect data. If it does not collect data you might as well go hook your mask up to the leaf blower in your garage, JMHO !!
Jerry
But PR does indeed make a System One REMstar Plus CPAP that only records compliance data. *sigh*
Re: DME Says Data Machine Not Acailable for CPAP
Insurances pay by HCPCS (insurance) code NOT by brand and model. CPAPs and APAPs are the exact same HCPCS code: e0601. Bi-level PAPs (Bi-PAP is the proprietary name for Respironics' bi-levels, VPAP is Resmed's proprietary name for their bi-levels) are an entirely different HCPCS code. Bi-levels are considerably more expensive than CPAPs and APAPs.
Any Respironics or PR SystemOne PAP w/Plus in the model name are NOT fully data capable. Any Resmed w/Escape in the model name is NOT fully data capable.
Be sure to check out RestedGal's link to fully data capable CPAPs and APAPs at the link she posted.
Any Respironics or PR SystemOne PAP w/Plus in the model name are NOT fully data capable. Any Resmed w/Escape in the model name is NOT fully data capable.
Be sure to check out RestedGal's link to fully data capable CPAPs and APAPs at the link she posted.
_________________
| Mask: Quattro™ FX Full Face CPAP Mask with Headgear |
| Additional Comments: PR SystemOne BPAP Auto w/Bi-Flex & Humidifier - EncorePro 2.2 Software - Contec CMS-50D+ Oximeter - Respironics EverFlo Q Concentrator |
Women are Angels. And when someone breaks our wings, we simply continue to fly.....on a broomstick. We are flexible like that.
My computer says I need to upgrade my brain to be compatible with its new software.
My computer says I need to upgrade my brain to be compatible with its new software.
Re: DME Says Data Machine Not Acailable for CPAP
Every time I read one of these DME outrages my teeth start to hurt~~ To the OT, please take the good advice given to you and take charge of your own therapy! Knowledge is power and a data-capable machine and a properly fitted mask will give you that. And read everything you can on this forum and ask questions! Good luck.
_________________
| Mask: Mirage Quattro™ Full Face CPAP Mask with Headgear |
| Additional Comments: 14/8.4,PS=4, UMFF, 02@2L, |
"Do or Do Not-There Is No Try"-"Yoda"
"We are what we repeatedly do,so excellence
is not an act but a habit"-"Aristotle"
DEAR HUBBY BEGAN CPAP 9/2/08
"We are what we repeatedly do,so excellence
is not an act but a habit"-"Aristotle"
DEAR HUBBY BEGAN CPAP 9/2/08