What kind of Dr would send you home with an APAP set at 4-18?!!! Call the Dr. now and tell him what you just told us. If you don't get immediate instructions on how to raise your lower pressure, let us know and someone will help you do that. I would not go to sleep tonight until I had that changed. Jeez!!!
Diagnosed 9/4/07
Sleep Study Titrated to 19 cm H2O
Rotating between Activa and Softgel
11/2/07 RemStar M Series Auto with AFlex 14-17
10/17/08 BiPAP Auto SV 13/13-23, BPM Auto, AHI avg <1
Sleep Study Titrated to 19 cm H2O
Rotating between Activa and Softgel
11/2/07 RemStar M Series Auto with AFlex 14-17
10/17/08 BiPAP Auto SV 13/13-23, BPM Auto, AHI avg <1
Re: Post your longest apnea events....
Stick w/the PAP therapy. Rome wasn't built in a day and this is an entirely new way to have to learn to sleep.
And since you can check your data, more important at this early stage is the Leak rate rather than the AHI.
Get your Leaks under control - 90%-95% leak rate preferably under 24 L/M but certainly under 40 L/M.
4-18 cms is too wide a gap. PAPs are too slow to get from that low a setting up to where the pressure is needed to stop the apneas. 6 cms would be a much wiser lower end pressure - unless - you also had an in-lab titration study in which case the best scenario is 1-2 cms below your titrated pressure and 2-3 cms or more above your titrated pressure.
And since you can check your data, more important at this early stage is the Leak rate rather than the AHI.
Get your Leaks under control - 90%-95% leak rate preferably under 24 L/M but certainly under 40 L/M.
4-18 cms is too wide a gap. PAPs are too slow to get from that low a setting up to where the pressure is needed to stop the apneas. 6 cms would be a much wiser lower end pressure - unless - you also had an in-lab titration study in which case the best scenario is 1-2 cms below your titrated pressure and 2-3 cms or more above your titrated pressure.
_________________
Mask: Quattro™ FX Full Face CPAP Mask with Headgear |
Additional Comments: PR SystemOne BPAP Auto w/Bi-Flex & Humidifier - EncorePro 2.2 Software - Contec CMS-50D+ Oximeter - Respironics EverFlo Q Concentrator |
Women are Angels. And when someone breaks our wings, we simply continue to fly.....on a broomstick. We are flexible like that.
My computer says I need to upgrade my brain to be compatible with its new software.
My computer says I need to upgrade my brain to be compatible with its new software.