EPR setting

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
trytryagain
Posts: 280
Joined: Fri Sep 03, 2010 6:28 am

EPR setting

Post by trytryagain » Tue Nov 02, 2010 11:11 am

When I started in September my EPR was set at 2, but I switched it to 3. Is there any theraputic reason to turn it back to 2 now that I am getting used to the machine?

_________________
Mask: Pilairo Q Nasal Pillow CPAP Mask with Headgear
Additional Comments: rescan 3.11

Janknitz
Posts: 8517
Joined: Sat Mar 20, 2010 1:05 pm
Location: Northern California

Re: EPR setting

Post by Janknitz » Tue Nov 02, 2010 12:46 pm

If it ain't broke, don't fix it!

If you are getting good numbers and feeling well, you can leave it just as it is. The manufacturers and sleep physicians claim that the EPR has no negative effect on AHI and if you are comfortable with it there's no need to change it.

Some people on this forum who have tried everything else when they are having problems find that turning the EPR down or off helps, but it doesn't seem to be necessary in most cases.
What you need to know before you meet your DME http://tinyurl.com/2arffqx
Taming the Mirage Quattro http://tinyurl.com/2ft3lh8
Swift FX Fitting Guide http://tinyurl.com/22ur9ts
Don't Pay that Upcharge! http://tinyurl.com/2ck48rm

dtsm
Posts: 1097
Joined: Sun Mar 08, 2009 8:34 am
Location: CT

Re: EPR setting

Post by dtsm » Tue Nov 02, 2010 5:17 pm

I actually did the reverse; started at 3 and reduced to 1. My understanding is EPR is a 'personal' comfort feature, to help exhalation. Keep in mind if you adjust EPR up or down, your cpap pressure needs to be adjusted by the same incremental setting.

User avatar
Breathe Jimbo
Posts: 954
Joined: Sun Sep 12, 2010 3:02 pm
Location: Pasadena, CA

Re: EPR setting

Post by Breathe Jimbo » Tue Nov 02, 2010 5:28 pm

dtsm, as I understand it, a change to EPR (ResMed machines) does NOT require any change to the pressure setting. The EPR is a temporary reduction in pressure at the moment the user begins to exhale. It operates only at the transition from inhalation to exhalation. It has no effect on the pressure during inhalation, and no effect for most of the time that one exhales.

_________________
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Since 9/9/10; 13 cm; ResScan 3.16; SleepyHead 0.9; PapCap

cwied
Posts: 52
Joined: Fri Oct 08, 2010 4:04 pm

Re: EPR setting

Post by cwied » Tue Nov 02, 2010 5:38 pm

Breathe Jimbo wrote:It has no effect on the pressure during inhalation, and no effect for most of the time that one exhales.
Are you sure about this? To me, it definitely feels like it's lowering the pressure for the whole time that I exhale, and then increasing it on the inhale. I actually changed my EPR speed to fast because I felt like it was ramping the pressure up too late on the inhale.

_________________
Machine: AirSense 11 Autoset
Mask: AirFit™ P30i Nasal Pillow CPAP Mask with Headgear Starter Pack
Additional Comments: Auto 6-9 (EPR 1)

User avatar
Breathe Jimbo
Posts: 954
Joined: Sun Sep 12, 2010 3:02 pm
Location: Pasadena, CA

Re: EPR setting

Post by Breathe Jimbo » Tue Nov 02, 2010 6:08 pm

Look at the graphs provided with the explanation of the Easy Breathe with EPR technology. The relief applies only at the beginning of an exhalation, not during the entire exhalation. The latter would be BIPAP.

_________________
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Since 9/9/10; 13 cm; ResScan 3.16; SleepyHead 0.9; PapCap

User avatar
sleepymama
Posts: 228
Joined: Mon Aug 02, 2010 6:46 pm
Location: Canada

Re: EPR setting

Post by sleepymama » Tue Nov 02, 2010 6:24 pm

not a dr here but i heard that sometimes if your EPR is too high (so say 3) you could be missing some events cause the machine is taking too much time to adjust the pressure to the pressure you need to stop the events. mine was at 3 and i turned it to 2. my next step would be a 1. not sure i would totally go to no EPR.

_________________
Mask: Opus 360 Nasal Pillow CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Using chinstrap plus taping mouth for mouth breathing. on my 5th mask...
~~SleepyMama~~
Diagnosed with OSA July 4/10
Titration Aug. 8/10
Treatment Started Aug. 25/10

Trying to find the right one...mask!

cwied
Posts: 52
Joined: Fri Oct 08, 2010 4:04 pm

Re: EPR setting

Post by cwied » Tue Nov 02, 2010 7:07 pm

Breathe Jimbo wrote:Look at the graphs provided with the explanation of the Easy Breathe with EPR technology. The relief applies only at the beginning of an exhalation, not during the entire exhalation. The latter would be BIPAP.
When I look at the graphs it looks to me like the pressure is low until you start to inhale. In fact the example that shows the difference between "fast" and "medium" inhale I think illustrates this pretty well.

As I understand it, Bi-PAP is defined to be at least 4cm pressure difference. I always assumed this is why EPR only goes to 3 cm.

I'm using an EPR of 1 because I find a higher setting makes me feel like the machine is forcing my breathing. My pressure is only 7cm, though, so I imagine if you have a higher pressure the higher EPR settings won't feel as overpowering.

I'm a little uncertain whether therapy pressure should be adjusted for EPR since I've seen a reference that with Bi-PAP the lower pressure is the one that should be set at the fixed CPAP titration value. This seems to be different than what is commonly assumed, so I don't know if it's reliable. If it is accurate, that would imply to me that you would do the same with EPR.

Edit: This is the fact sheet I'm looking at https://88559364175b176f8341-7518c0400f ... 0Sheet.pdf

_________________
Machine: AirSense 11 Autoset
Mask: AirFit™ P30i Nasal Pillow CPAP Mask with Headgear Starter Pack
Additional Comments: Auto 6-9 (EPR 1)
Last edited by cwied on Tue Nov 02, 2010 8:43 pm, edited 1 time in total.

User avatar
chunkyfrog
Posts: 34544
Joined: Mon Jul 12, 2010 5:10 pm
Location: Nowhere special--this year in particular.

Re: EPR setting

Post by chunkyfrog » Tue Nov 02, 2010 8:02 pm

It feels to me like EPR runs all the time I'm exhaling--if not, I'd be surprised.

_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear
Additional Comments: Airsense 10 Autoset for Her

User avatar
Breathe Jimbo
Posts: 954
Joined: Sun Sep 12, 2010 3:02 pm
Location: Pasadena, CA

Re: EPR setting

Post by Breathe Jimbo » Wed Nov 03, 2010 3:26 am

I could be wrong. I sent an email to ResMed in the hope of getting a definitive explanation whether the pressure reduction is transitional only, or during the entire exhalation. I'll let y'all know what ResMed says.

_________________
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Since 9/9/10; 13 cm; ResScan 3.16; SleepyHead 0.9; PapCap
Last edited by Breathe Jimbo on Wed Nov 03, 2010 8:21 pm, edited 1 time in total.

User avatar
jmelby
Posts: 463
Joined: Sat Nov 14, 2009 7:52 pm
Location: Minneapolis, MN

Re: EPR setting

Post by jmelby » Wed Nov 03, 2010 1:58 pm

The answer is "maybe"... I found that with any EPR setting, my AHI was way higher than without (even when adjusting my lower-end pressure setting by a corresponding amount as dtsm suggests). It depends on the individual on whether their breathing pattern is prone to this. My RT said they don't enable EPR at all for patients unless they are having discomfort exhaling. So, I have left the setting off after some experimenting with it early on, and have never felt the need to use it.

_________________
Mask: Mirage Activa™ LT Nasal CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: ResScan 3.12, APAP 9 - 13, no EPR, ClimateControl 75F
(yet another Jeff)

cwied
Posts: 52
Joined: Fri Oct 08, 2010 4:04 pm

Re: EPR setting

Post by cwied » Wed Nov 03, 2010 2:06 pm

jmelby wrote:So, I have left the setting off after some experimenting with it early on, and have never felt the need to use it.
I agree with you in general. The one thing I'll say is that the breathing assist that EPR gives me did help when I was congested following a cold last week. I turned up the EPR to 3 on the day after I stopped using the OTC nasal spray, and it definitely helped me breathe through the night. During my titration I had a 0 AHI even at 4cm, so I didn't adjust my pressure.

_________________
Machine: AirSense 11 Autoset
Mask: AirFit™ P30i Nasal Pillow CPAP Mask with Headgear Starter Pack
Additional Comments: Auto 6-9 (EPR 1)

dtsm
Posts: 1097
Joined: Sun Mar 08, 2009 8:34 am
Location: CT

Re: EPR setting

Post by dtsm » Wed Nov 03, 2010 8:05 pm

Breathe Jimbo wrote:dtsm, as I understand it, a change to EPR (ResMed machines) does NOT require any change to the pressure setting. The EPR is a temporary reduction in pressure at the moment the user begins to exhale. It operates only at the transition from inhalation to exhalation. It has no effect on the pressure during inhalation, and no effect for most of the time that one exhales.
This was something I learned from our Mother Superior [aka rested gal]. She went into a very detailed explanation for me in a private email back when I first started cpap.

All I can say is from personal experience, it made sense and was true in my case. I was able to validate by using apap rather than straight cpap. My 95th percentile pressure did change [drop} when I reduced my EPR.