Post by xenablue » Mon Nov 01, 2010 11:43 am

If someone would care to help me with this, I would be most grateful. Or possible point me to a site that will explain it to me.
http://i1226.photobucket.com/albums/ee4 ... pstudy.jpg

If you need more information, or data from the S9 I can provide that as well.

Thanks in advance,
Xenablue

_________________
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Dx 10/14/10. Also a T2 diabetic. High night/fasting numbers prompted a sleep study and here I am :-)

User avatar
Emilia
Posts: 1873
Joined: Mon Jul 12, 2010 8:56 am
Location: Florida, USA

Re: Need Help Interpreting Sleep Study Results

Post by Emilia » Mon Nov 01, 2010 12:09 pm

Ooops, you didn't embed it. Here it is for you....

Image
Yes, that blue eyed beauty is my cat! He is a seal point, bi-color Ragdoll. I adopted him in '08 from folks who could no longer care for him. He is a joy and makes me smile each and every day.

User avatar
xenablue
Posts: 1358
Joined: Tue Sep 07, 2010 12:52 pm
Location: Aussie living in balmy Wisconsin

Re: Need Help Interpreting Sleep Study Results

Post by xenablue » Mon Nov 01, 2010 12:24 pm

Thanks again Emilia - I must've missed one of the steps you gave me.

Xena

_________________
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Dx 10/14/10. Also a T2 diabetic. High night/fasting numbers prompted a sleep study and here I am :-)

User avatar
Emilia
Posts: 1873
Joined: Mon Jul 12, 2010 8:56 am
Location: Florida, USA

Re: Need Help Interpreting Sleep Study Results

Post by Emilia » Mon Nov 01, 2010 2:29 pm

You're welcome! This appears to be the results of a split study where you were titrated during the second half. In the first half, your total AHI was 21.3 events per hour--almost all were during REM sleep. This puts you into the moderate category of OSA. After being titrated, you fell to only 0.5 event per hour. Great results! Your oxygen levels were not great before titration....you spent 8 minutes below 89% and bottomed out at 78%. However, once using the mask, your oxygen was great with it being 90%+ for 99% of the time.

What pressure were you Rx'd? Are you using software to monitor you nightly results?
Yes, that blue eyed beauty is my cat! He is a seal point, bi-color Ragdoll. I adopted him in '08 from folks who could no longer care for him. He is a joy and makes me smile each and every day.

User avatar
xenablue
Posts: 1358
Joined: Tue Sep 07, 2010 12:52 pm
Location: Aussie living in balmy Wisconsin

Re: Need Help Interpreting Sleep Study Results

Post by xenablue » Fri Nov 05, 2010 12:48 pm

I came back to this thread to re-read and get a better grip on what the heck I'm working at.

Think I answered this in a PM, Emilia, but my pressure is set at 7, which seems to be working fine. I'm looking at my results almost every day and AHI is always well below 1, and usually .1 - .3.
The leaks seem inconsistent but I think that's due to me adjusting or taking off the mask while the pressure's still on (from what I read here).